Re: Poor Quality 50 Ohm Load - Where to get accurate ones?


W0LEV
 

I have N, SMA, and OSM series precision cal. kits from HP. I made a DC
measurement of the 50-ohm standard loads for these three connector series
from these kits using the following instruments (what I have).

1) Bench DMM: HP 3478A
2) Hand-Held DMM from Harbor Freight: P37772
3) Hand-Held DMM from Sperry: DM6400

*Results follow:*

*HP** Harbour Freight Sperry*

50.375 45.50 50.20 Type-N
(HP 909C)

50.179 45.50 50.10 SMA
(HP 909D)

50.330 47.00 50.10 OSM
Connector (the expensive ones from HP)

AVERAGE OF ALL READINGS: 48.801 Ω Clearly the Harbour Freight unit
is dragging down the average.

AVERAGE OF ALL READINGS (excluding the Harbour Freight measurements):
50.214 Ω (0.43 % error against 50.0000 Ω)

These are precision HP cal. standards. Sure, they are better than what
comes with the NANOVNA's. But consider the cost!!!!!! Is it worth arguing
and expecting something less than 2.26 % error when the load reads 51.13
ohms? Really, now?

This is a hobby, not a metrology lab!!!!!!

Dave - WØLEV

On Mon, Aug 17, 2020 at 4:20 PM Wolfgang Wilde via groups.io <wwilde69=
googlemail.com@groups.io> wrote:

Fully agree with Dave. Are you not awaiting a little bit to much from
the NanoVNA? What precision can be expected regarding impedance, levels,
noise figures and so on? The FR4 material may not be precise enough to
guarantee 100% precise 50 Ohms impedance for the input/ output lines of
the NanoVNA itself!

Can it really work that precise? Over what frequency range? And in what
impedance range will it show the right numbers?

Would you really expect similar figures and precision from it than you
would expect from HP/Agilent/Keysight/Tektronix/Rohde&Schwarz labor
grade measurement equipment that costs 100 times as much or more? Don't
get me wrong, it is really a brilliant DIY project that helps me a lot
and is very handy tool for me to get a rough idea of what is going on.
But I would never expect the NanoVNA being absolute precise as I think
it never was target on that and I doubt it could deliver that precise
results. It already does almost magic, so let's adjust expectations a
little bit for reality.

vy 73,
Wolfgang, (DG7NEF)




Am 17.08.20 um 17:33 schrieb Dr. David Kirkby, Kirkby Microwave Ltd:
On Fri, 14 Aug 2020 at 01:52, Glen Jenkins WB4KTF <wb4ktf@...>
wrote:

The 50 OHM load (SMA-Male) that came with my nanoVNA-H4 measures 51.13+
ohms. Not a good start for calibration. Where is a good source for
GOOD
loads that are accurate?

--

-----

Glen Jenkins, WB4KTF, Austin, TX
Do you have a specification in mind? DC resistance doesn’t tell you much,
other than it puts an *upper* limit on return loss. As soon as you go up
in frequency, the return loss will most probably decrease.

If you measure up to 3 GHz, it gives you no idea what it will be like at
6
GHz. Sometimes loads worsen dramatically with frequency, but others, from
the same batch will get better.

The last I looked, there was nothing in the NanoVNA firmware that
properly
accounted for the open or short calibration standards. So there are
limits
to what you accuracy you are going to achieve if you had a really
expensive
load, as the phase errors will kill you.

Dave




--
Mit freundlichen Gruessen,
------------------------------------------------------------
Wolfgang Wilde, -o) Email: wwilde@...
Nuernberg /\\ wwilde@...
_\_v

------------------------------------------------------------
Diskutiere nie mit einem Irren -
er zieht dich auf sein Niveau
und schlägt dich mit seiner Erfahrung.


Shit happens -
... manchmal bist du Taube, manchmal Denkmal.


Das Software Business ist binaer.
Du bist eine 1 oder eine 0,
lebendig oder tot.
Wer nicht innovativ sein will ist t o t.
Startup (von Peter Howitts)


Das Wissen der Menschheit gehört der Welt.
So wie Shakespeare oder Aspirin...
Startup (von Peter Howitts)


Es gibt keine Hindernisse, keine Grenzen.
ueberrascht mich, zeigt es mir, bertrefft mich,
bertrefft euch selbst.
Startup (von Peter Howitts)



Fürchtet euch nicht,
die Welt wird heute nicht untergehen.
Es ist nämlich schon morgen - in Australien!
Eckart von Hirschhausen




"Was ist das?"
"Das ist blaues Licht."
"Und was macht das?"
"Es leuchtet blau."
RAMBO III



"Gott muss lieben Verrückte Menschen"
"Warum?"
"Er macht so viele Davon!"
RAMBO III



FIAT:
F uer
I talianer
A usreichende
T echnik


mail is sent from F. U. C. K. U. P.
First Universal Cybernetic Kinetic Ultramicro Program


Man muss die Welt nicht verstehen,
man muss sich nur in ihr zurechtfinden!
Albert Einstein


Phantasie ist wichtiger als Wissen
denn Wissen ist begrenzt.
Albert Einstein


Die größten Wolhtäter der Welt sind Faulenzer.
Denken Sie mal daran
wieviel Unheil dadurch vermieden wurde
daß man einfach nichts getan hat!


Alle sagten: Das geht nicht.
Dann kam einer, der wußte das nicht und hat's gemacht.
(Unbekannt)


Stop wasting entropy - start using predictable tempfile names today!


RFC1925:
(3) With sufficient thrust, pigs fly just fine. However, this is
not necessarily a good idea. It is hard to be sure where they
are going to land, and it could be dangerous sitting under them
as they fly overhead.
(6a) (corollary). It is always possible to add another level of
indirection.
(8) It is more complicated than you think.


Mike Cox, I've got a new angle for ya. (BTW- Long-time fan!) Start doing
the total sarcasm posts like normal, but change them a bit so you sound
like the Iraqi Information Minister. (I miss him.)
"There IS no Linux within a hundred miles of any PC!"


I just found out that the brain is like a computer.
If that's true, then there really aren't any stupid people.
Just people running Windows.


Sie sind nicht berechtigt,
unrechtmäßige Kopien dieser Mail
anzufertigen oder zu vertreiben.


This software is an abomination in every sense of the word!
Run it! Hug it! Feel the power of the Dark Side!
[ MGM, (C) by Xiphophorus, see http://www.xiph.org]


-- 21 ist nur die halbe Wahrheit --


Aus einem Mailwechsel zum Thema Unterbrechungsfreie Stromversorgung:
Es gibt immer einen, der den Stecker ziehen kann.
Apple arbeitet schon an dem Problem... ;-)
Sie wollen die USV erfinden?
Nein, Stecker, die man zum Entfernen in den Papierkorb legen muß.


Muphy's Naturgesetze:
4.) Jedes Programm, das gut beginnt, endet schlecht.
Alles, was schlecht beginnt, endet furchtbar.


Muphy's Naturgesetze:
21.) Die Vorarbeit wurde immer von Personen ausgeführt, die noch
dabei sind, ihre unterste Stufe der Unfähigkeit zu erreichen.


Wer eine Personal Firewall verwendet, könnte auch eine Einkaufstüte
als Kondom benutzen: Den Sicherheits- und Lustgewinn beider Methoden
halten wir für vergleichbar. ;-)
[T-Online-Team in t-online.talk.internet]


Linux is like a wigwam:
no gates, no windows, and an apache inside.


"Bunt ist das Dasein und granatenstark."


COMMAND.COM kann nicht angehalten werden - System halted


-------------------------------------------------------
Mind your head and keep it simple !
-------------------------------------------------------

Höflichkeit ist wie ein Luftkissen.
Es mag nichts drin sein, aber es mildert die Stösse des Lebens.


The light at the end of the tunnel might be an oncoming dragon


Wenn Personal Firewalls die Lösung sind,
will ich das verdammte Problem zurück. [Robin S. Socha]


Good, Fast, Cheap: Pick any two (you can't have all three). RFC 1925


Stay the patient course.
Of little worth is your ire.
The network is down.


It's not denial! I'm just very selective about the reality I accept.


Erst jetzt nachdem du gesehen hast was du gehört hast
kannst du möglicherweise die Wahrheit akzeptieren....


Nichts
> Was hältst Du von Quotes am
Ende?


echo '@a=(A..Z,a..z," ");$t="a"x16;{for($x=0,$|=$n=30; $n--;){for($l=$t;
1<rand 7;) {substr$l,rand 16,1,$ a[rand@a]}$q=$_="zeluR xuniL
ESuS";for($a=$l;
$_;){s/.//;$q++ if chop$a eq$&}if($x<=$q){$e=$l;die$ e,$/if($x=$q)==16}}
print$t=$e,"\r";redo}'|perl


Thought is limitation. Free your mind.


--------------------------------------------------------------------
Verglichen mit jedem x-beliebigen Redmonder Betriebssystem-Clone
ist Linux geradezu eine leuchtende Perle der Datensicherheit.
------ Frank Rennemann (http://www.linux-knowledge-portal.org) -----


Trouble springs from idleness, and grievous toil from needless ease.
--Benjamin Franklin


-----------------------------------------------------------------
"Na schön, was sollen wir also tun, wenn wir von irgendwas
Lebensgefährlichem gebissen werden?" fragte ich.
Er sah mich an, als sei ich bescheuert.
"Na, was machen Sie dann wohl" sagte er. "Sie sterben.
Was denn sonst? Deshalb heißt es ja lebensgefährlich."
------------- Douglas Adams: Die Letzten ihrer Art --------------


"355/113 -- Not the famous irrational number PI, but an incredible
simulation!"


Was auch immer geschieht:
Nie dürft ihr so tief sinken,
von dem Kakao, durch den man euch zieht,
auch noch zu trinken!
- E. Kaestner


In a free world without walls and fences, who needs windows and gates?


"Reality is always controlled by the people who are most insane" Dogbert


Ich bin vorwiegend mit unterbrechungsintoleranten Arbeiten beschäftigt und
bevorzuge daher email gegenüber Telefonanrufen.


wonder why did i even answer to that user, i shouldn't do so, just flame
him to the hell as usual. Probably i was drunk then... -- [A'rpi]

ich denke, ich kann Dir für jede PHP-Dummheit eine Java-Dummheit
präsentieren. Scheiße wird nicht kleiner, wenn man ein UML-Diagramm
für den Darm malt, der sie produziert hat. -- [Kristian Köhntopp]

"Never underestimate the bandwidth of a station wagon full of
tapes hurtling down the highway" -- [Andrew S. Tanenbaum]

Wenn Personal Firewalls die Lösung sind,
will ich das verdammte Problem zurück. [Robin S. Socha]

While the IBM 650 was not a super-hot machine, it did have one feature that
made it sell [...] lots of blinking lights. [...] Some authors attribute
the
success of IBM to these blinking lights [...]
-- [
http://www.users.nwark.com/~rcmahq/jclark/ibm650.htm]

Whoever said, "Hire the handicapped; it's fun to watch 'em," never had to
work
in IT holding the hands of morons.
[
http://www.starbreaker.net/modules.php?op=modload&name=News&file=article&sid=3
]

7. Mitarbeiter (...) Jeder Mitarbeiter sollte stolz darauf sein, bei SuSE
zu
arbeiten und sollte sich mit den betrieblichen Grundsätzen
(Professionalität,
Kundenorientierung und eine offene und transparente Kommunikation gemäß den
Betriebsleitlinien) identifizieren. (...)
[
http://w3.suse.de/cw/company/suse_offiziell/strategy/strategie_de.html]

It's astounding, time is fleeting, madness takes its toll.

+--------------------+
| DAS IST KEIN HAUS! |
+--------------------+

"Multiple exclamation marks," he went on, shaking his head,
"are a sure sign of a diseased mind." -- [Terry Pratchett]

"The people who created MIME not only should be convicted, they should
be shot on the spot." [Linus Torvalds on linux-kernel 1996/05/15]
<Pine.LNX.3.91.960515095405.27927F-100000@...>

"Gerade gestern habe ich mich bei Slashdot wieder gefragt, ob das nicht
schlimmer als die Todesstrafe ist, wenn man Informatik Prof in Utah ist
und ein Gericht einen als Experten benennt, der sich den Windoze Quell-
code anschauen muß." [Felix von Leitner in dasr 2002/02/23]

"Everybody knows you've been discreet,
but there were so many people you just had to meet
without your clothes
And everybody knows." [Leonard Cohen]

"Mit Verlaub Herr Präsident, Sie sind ein Arschloch!" -- [Josef Fischer]

Heute schon Deine IntEDV umarmt? Already hugged your Internal IT today?

Good, Fast, Cheap: Pick any two (you can't have all three). RFC 1925

Heckler & Koch: The original point´n´click interface.

S.E.I.F.E.: Synthetic Electronic Individual Fabricated for Exploration

MCSE: Microsoft-Compatible Spongiforme Encephalitis

Wenn ich eine SuSE-CD an ein Schwein binde und dieses trete, laufen
KDE & Co. auch ohne RAM recht schnell.
--Robin S. Socha in de.comp.os.unix.linux.newusers--

"Was meinen wir mit offener und transparenter Kommunikation?
Offenheit: Wir halten nicht unnötigerweise Informationen zurück!
Transparenz: Wir legen unsere Entscheidungswege offen!
Ehrlichkeit und Verläßlichkeit: Wir sagen was wir meinen und stehen zu
dem
Gesagten!" <
http://w3.suse.de/ber/human_resources/unternehmensleitbild.html>

"Q: Wie sind Deine ersten Eindrücke hier?
A: Zunächst fühlt man sich etwas gehandicapped - Ich komme aus einer funk-
tionierenden und zuverlässigen Windows NT-Welt [...]"
<
http://w3/ber/human_resources/neue_mitarbeiter/archiv/neuemi08/index.html>

"Q: Wie kommst Du mit der Software hier klar?
A: Ich benutze Excel bei der Arbeit. [...]"
<
http://w3/ber/human_resources/neue_mitarbeiter/archiv/neuemi09/index.html>

If Java had true garbage collection, most programs would delete themselves
upon execution. [Robert Sewell]

Karneval ist die jährliche Überdosis Abführmittel für Spießer mit
chronischer Verstopfung. [Jens Grivolla]

Get back there in front of the computer NOW. Christmas can wait.
-- Linus "the Grinch" Torvalds, 24 Dec 2000 on linux-kernel

Yeah, yeah, it's 7PM Christmas Eve over there, and you're in the middle of
your Christmas dinner. You might feel that it's unreasonable of me to ask
you to test out my latest crazy idea. How selfish of you.
Get back there in front of the computer NOW. Christmas can wait.
-- Linus "the Grinch" Torvalds, 24 Dec 2000 on linux-kernel

Elektrosmog my ass. Wenn ich mal in's Keyboard beisse, dann wegen den
Unmengen an Kaffee, Zigaretten und dem ungesunden Essen, nicht wegen
diesem Esoterikschrott. [Jonas Luster]

"Du bist ein Dummer Haufen Luserscheisse. Deine Praesenz ekelt mich an,
ich muss mich uebergeben, wenn ich nur daran denken, dass Penner wie
Du auf das Netz, das ich liebe und heilige losgelassen werden."
[Jonas Luster zu Bernd Haug in dasr]

"Geh' fort. Geh' ganz weit fort. Paare Dich mit den anderen Wuersten in
microsoft.ALL und detebe, verschone mich, uns, diese Newsgruppe von
Deine Luserfresse, JETZT!" [Jonas Luster zu Bernd Haug in dasr]

"Du bist so scheisse, selbst das benutzerfreundlichste OS hasst Dich,
Mikrowellen schalten automatisch auf GRILL, wenn Du vorbeilaeufst,
denn selbst unanimierte Gegenstaende wuerden es hassen, wenn Du Dich
fortpflanzen koenntest." [Jonas Luster zu Bernd Haug in dasr]

BIND gehört genauso zum RL wie Fußballbundesliga.
Die einen haben Premiere abonniert, die anderen de.comp.security.firewall.
[Christoph von Nathusius in dasr]

"Ich muss weg!" [unbekannter Manager bei TV total]

"Es ist in Ordnung, wenn Du eine andere Meinung hast über das, was Du für
wichtig hältst. talk aber ist der einzige Platz dafür, die Meinung der
anderen darüber zu ignorieren." [Roman Drahtmüller am 10.4.2001 auf
talk]

"Das Neu starten eines Unix Rechners kann sehr schmerzhafte Folgen haben,
je nachdem wie schwer das naechste greifbare Buch ist, welches ihnen dann
an den Kopf fliegt." -- frei nach einer sehr alten c't

AOL-Kunden koennen keinen Videorecorder programmieren.
AOL-Kunden haben maximal 6 weitere Haushaltsangehoerige.
AOL-Kunden brauchen beide Haende, um 4 von 7 zu subtrahieren.
AOL-Kunden sind einfach drin. Hoffentlich bleiben sie da.

f u cn rd ths, u mst hv bn sng nx

"Da ich Pressemeldungen nur noch per Subject lese, l"auft mein Gehirn auch
nicht Gefahr, von meinem Dickdarm aus Verzweiflung erdrosselt zu werden"
-- [Daniel Bischof]

No matter how good she looks, someone else is sick and tired
of putting up with her shit.

I saw a woman wearing a sweatshirt with 'Guess' on it. I said, "Implants?"

Why is it that most nudists are people you don't want to see naked?


An expert is a person who avoids the small errors
as he sweeps on to the grand fallacy.
-- Benjamin Stolberg




Tech Support: "How may I help you?"
Customer: "I'm writing my first email."
Tech Support: "Ok, what seems to be the problem?"
Customer: "Well I can get the 'a'. But how do I put the circle around
it?"


All you have to decide
is what to do
with the time that is given to you
-- Gandalf at The Lord of the Rings


Yesterday,
All those backups seemed a waste of pay.
Now my database has gone away.
Oh I believe in yesterday.

Suddenly,
Tere's not half the files there used to be,
And there's a milestone hanging over me
The system crashed so suddenly.

I pushed something wrong
What it was I could not say.

Now all my data's gone
and I long for yesterday-ay-ay-ay.

Yesterday,
The need for backups seemed so far away.
I knew my data was all here to stay,
No I believe in yesterday.




Nuke Operator (16. Dezember 2003 23:22)

LE Big Macke schrieb am 16. Dezember 2003 20:12

> NCC-1701-A zum Bleistift?

Der Fileserver - unendliche Weiten...
Wir schreiben das Jahr 2004.
Dies sind die Abenteuer des Raumschiffs Nterprise,
Das mit seiner 400 Mann starken Besatzung fünf Jahre lang unterwegs
ist, um neue Software zu erforschen, neues Sourcen und neue Binaries.
Viele Kilobits vom User entfernt, dringt die Nterprise in
Verzeichnisse vor, die nie ein Mensch zuvor gesehen hat.




USB is for mice, FireWire is for men!


Boy, backups sure are a lot faster
since I linked /dev/st0 to /dev/null...


A computer is like an air conditioner.
It stops working when u open windows.


Was ist eine blondine mit gefärbten Haaren?
Künstliche Intelligenz


An der Kinokasse hängt ein Schild:
Sonderpreis! Soldaten nur ein Euro!
Geht die blonde Susi hin, legt zehn Euro auf den Tisch und säuselt
errötend:
"Bitte zehn von der Marine ...!?"




--
*Dave - WØLEV*
*Just Let Darwin Work*

Join {nanovna-users@groups.io to automatically receive all group messages.