VNA Calibration
Neil, G3RIR
Stupidly I have blown up the SMA load fro my NanoVNA. I put accidentally about 20W into it. It blew instantly.
Does any one have any spares I could purchase? It is just the load I need. Neil G3RIR |
|
Just checked my box - I have at least seven SMA-male loads. SOme with a little chain attached. Is that all you need? Andy Stupidly I have blown up the SMA load fro my NanoVNA. I put accidentally about 20W into it. It blew instantly. |
|
Neil, G3RIR
Andy, Brilliant, yes I just need an SMA male load. Can you email me direct at ackerley.neil@... please Neil G3RIR
On Monday, 20 March 2023 at 18:45:30 GMT, Andy G4JNT <andy.g4jnt@...> wrote:
Just checked my box - I have at least seven SMA-male loads. SOme with a little chain attached. Is that all you need? Andy Stupidly I have blown up the SMA load fro my NanoVNA. I put accidentally about 20W into it. It blew instantly. |
|
Mark GM4ISM
Hi Neil I can offer one calibrated) against an Anritsu 26GHz standard (long ago calibrated HI) Specced (35dB RL to 26.5GHz) I know I have several that had better return loss that the original supplied. (I don't use my original HI, it isn't good enough :) ) I can't offer traceability HI Mark GM4ISM On 20/03/2023 18:37, Neil, G3RIR via
groups.io wrote:
Stupidly I have blown up the SMA load fro my NanoVNA. I put accidentally about 20W into it. It blew instantly. -- Mark GM4ISM gm4ism@... dc2light.co.uk |
|
Neil, G3RIR
All sorted now thanks to G4JNT
Neil G3RIR |
|
Alan Bain
Hi,
I am curious about how you calibrate an SMA load respectably supposing access to another VNA (and its cal kit with coefficients!) My Anritsu came without the calibration kit (usual comments about people looking for test gear bargains apply) but I also tend to measure stuff with SMA connectors rather than N connectors so probably the N-type cal kit would just have sat on the shelf and gathered dust. If I'm honest most of the time I use Rosenberger SMA load/thru/open but have no idea what accuracy I'm losing doing this. There's a bit of a documentation shortage on how one provides cal coefficients to the Anritsu - I imagine it wants a capacitance at the end of the open and inductance for the short as a function of frequency, but less clear what is done about the load - which isn't exactly 50 ohm - I can measure that on a 4 terminal bridge at DC. The manual of course just says it wants a disk of cal kit coefficients which is a bit meaningless. Traceability not required here! Alan On Mon, 20 Mar 2023 at 18:54, Mark GM4ISM via groups.io <gm4ism@...> wrote:
|
|
Mark GM4ISM
Hi Alan There is quite a bit involved with getting confidence in the
measurements associated, thus this email is a little long
winded... I don't know the instrument or cal kit you are using but
typically the Open Short Load sets are 50 Ohm characterized.
When you calibrate most instruments with one, it measures and
calculates all the offsets and errors etc at the test point
(where you put the standard) and measures all subsequent WRT the
50 Ohm cal standard which the instrument assumes is 50 j0 ohms at
every frequency. Other cal kit coefficients that may be 'needed' are electrical length to the load / open / short. These help calibrate the phase components that a VNA will display. Assuming the instrument can be calibrated to the load, the measurement directivity will match that of the load RL irrespective of the cal coefficients not being loaded The ability to store and recall cal data is usually more to
enable a full test system (including cables, adapters etc) to be
set up quickly for a given application easily if nothing has
changed. Something I rarely did professionally, re-calibrating is
a little tedious but brings about a sense of certainty.
Its easy enough to damage a 50 Ohm standard load with over power but it usually stops measuring 50 ohms at DC at that point. In a 50 Ohm system, a 51 Ohm impedance is still 40dB RL A load with 34dB RL could be 52 Ohms at DC. The chances are your system is calibrated well enough. That said, the DC reading of the load should not be that far off 50 Ohms.
Formal calibration is about traceability in professional
standards (pun intended) as much as anything else. The equipment I have is all well passed its sell by date and overdue calibration in those circles. Having administered aspects of test gear standards professionally for years, I can say that I have never met an instrument (of this sort) that failed calibration and did not obviously have a fault. I have network analysers, a scalar analyser to 26GHz and
reference standards, one with specified return losses >49dB
(sub3 GHz). Having used them over the years and stored cal data,
they do not appear to have changed in the years since formal
calibration. The other standard (3.5mm), is specced at >40dB to 26.5GHz.
Using one as a reference for another will quickly show up any
issue. I have also had access for years to formally calibrated
standards and measurement comparisons were fine. So while I have no formal traceability, I have confidence that my reference standards are good. With any measurement of this type you need to reference a
'standard' and have confidence in that standard.
If you calibrate a VNA with a standard of 49dB return loss, it should read the calibration standard as having a return loss of infinity... Noise floors etc limit that but with a real network analyzer, at least 80dB. At these levels, just breathe on the test gear and the null into the reference load can be seen to twitch:) With a 'calibrated' network analyser showing >80dB return loss you can be assured that the directivity of the measurement system has become very close to the return loss of the standard (in this case 49dB worst case) Now substitute the load you want to measure (DUT) The DUT return loss indicated will have an accuracy dictated by
the directivity.. a reasonable gauge of the effect is here https://www.markimicrowave.com/assets/appnotes/directivity_and_vswr_measurements.pdf Measuring a DUT with a poor return loss with a good directivity system is really quite accurate. Measuring a good DUT with a mediocre directivity and you could get quite large numerical errors in the answer. Measuring a good DUT with a better directivity reduces the uncertainly. Typically you need a directivity 15dB better than the DUT return loss to have a reasonable levels of uncertainty. If you achieve an indicated >80dB RL, you have actually
achieved a RL close to that of the directivity of the instrument
(in this case 49dB) Thus with 49dB directivity you can measure a DUT to about 34dB
return loss and have a reasonable accuracy (within a couple of
dB) You are unlikely to get a flat line indicated RL >80dB! Having calibrated my test gear to the 49dB RL standard, I
measure the professional 3.5mm Marconi standard I have. It is specced at 40dB RL (up to 26.5GHz) and looks way better
than that up to 3GHz. It reads in the 60s dB RL up to 3GHz. I can
say is that it is very likely in spec ie >40dB This comparison gives a degree of confidence that the standard loads are in good condition and can be trusted.
I substitute other SMA loads (including one from my little nanoVNA) and pick the ones that closest match the professional standard. I was pleasantly surprised that the supplied standard for the nanoVNA was actually quite good, not quite as good as the Marconi standard but close. I measured a return loss of > 35dB +- 2dB I tested sundry SMA loads. Most were quite frankly not fit to be
used as a return loss standard, with measured RL at 25dB or worse
at higher frequencies. Not surprising, look at the specs for even
quality 50Ohm loads.. The spec required for a calibration standard
is way higher than a random good load If you used a 25dB RL load to calibrate a VNA, the VNA would have
a directivity of 25dB.. Nearly useless in my books, as an
indicated 20dB return loss could be anywhere from approx 16dB to
27dB . If you adjust the DUT for a 'perfect' match into such a
measurement system and achieve a very high measured RL, you will
actually have achieved a RL for the DUT that is equal to the
directivity. That may be good enough in many cases. Of course a 20dB ret loss is fine as a dummy load.
I hand picked from my extensive stock of SMA loads and ended up with 3 or 4 (about 10%) that were good enough (RL in the 35dB region) up to 3GHz Only one passed muster to >20GHz compared to the Marconi standard.
Above 3GHz, I am just comparing the Marconi standard to the DUT.
Any load that reads similar (>60dB) after the 26GHz analyser
has been 'calibrated' to the Marconi standard, will have a
similar return loss, ie about40dB Above 3GHz I don't have another good reference so there is a
degree of faith here on the instrumentation and that Marconi
standard load.
That faith is re-enforced by first principal measurements,
measuring standing waves with a swept frequency measurement using
a non-directional coupler using a mostly forgotten technique
called wobbulation (which can also indicate distance to fault so
long as you can do inverse FFTs in your head). Some on this list
will have also come across quaterwaving as a technique for
removing offsets and improving accuracy. I've got a waveguide
with a probe in a long slot for match measurements in WG16. Again
a 1st principal technique that doesn't rely on instrument
directivity as such.
All these measurement techniques have intrinsic uncertainly but the more you get the same answer by different techniques, the higher the confidence is. In short, I have reasonable confidence that up to 3GHz, I can
characterise a load to about 35dB RL with reasonable accuracy,
which is adequate for my purposes and I suspect adequate for most
amateurs as a standard against which to make measurements. A bog
standard SMA load may not cut it and introduce significant
uncertainly. That is what I consider 'calibration'. Without reference to what accuracy one expects to achieve or claims, the term calibration is open to interpretation. Its a far cry from what a professional calibration house can
access but the same principals are employed. I hope that helps Measuring return loss, frequency and power is all a bit of a
minefield in the amateur world. I am fortunate in having
instruments that let me have a fair go at all these. That said,
you will not loose many QSOs and are unlikely to blow up your PA
if the match of your antennas is only 17dB (so long as that isn't
because it has water in it) . Being 10 Hz of frequency is not
often an issue either.... Just don't believe your bird
wattmeter blindly .... Regards Mark GM4ISM
On 20/03/2023 20:24, Alan Bain wrote:
Hi, I am curious about how you calibrate an SMA load respectably supposing access to another VNA (and its cal kit with coefficients!) My Anritsu came without the calibration kit (usual comments about people looking for test gear bargains apply) but I also tend to measure stuff with SMA connectors rather than N connectors so probably the N-type cal kit would just have sat on the shelf and gathered dust. If I'm honest most of the time I use Rosenberger SMA load/thru/open but have no idea what accuracy I'm losing doing this. There's a bit of a documentation shortage on how one provides cal coefficients to the Anritsu - I imagine it wants a capacitance at the end of the open and inductance for the short as a function of frequency, but less clear what is done about the load - which isn't exactly 50 ohm - I can measure that on a 4 terminal bridge at DC. The manual of course just says it wants a disk of cal kit coefficients which is a bit meaningless. Traceability not required here! Alan On Mon, 20 Mar 2023 at 18:54, Mark GM4ISM via groups.io <gm4ism@...> wrote:Hi Neil I can offer one calibrated) against an Anritsu 26GHz standard (long ago calibrated HI) Specced (35dB RL to 26.5GHz) I know I have several that had better return loss that the original supplied. (I don't use my original HI, it isn't good enough :) ) I can't offer traceability HI Mark GM4ISM On 20/03/2023 18:37, Neil, G3RIR via groups.io wrote: Stupidly I have blown up the SMA load fro my NanoVNA. I put accidentally about 20W into it. It blew instantly. Does any one have any spares I could purchase? It is just the load I need. Neil G3RIR -- Mark GM4ISM gm4ism@... dc2light.co.uk -- Mark GM4ISM gm4ism@... dc2light.co.uk |
|