I did it once with boiling water and room temperature on what was supposed to be a low noise amp (144MHz) of the late 1980s era. I saw a genuine "fraction of a dB" but back then didn't have the means to measure mean noise power reliably or accurately, especially in a wide bandwidth. Simplistically: Assume a 1dB noise figure LNA with sufficient gain to overcome any following stages. 1dB is equivalent to a noise temp T of 75K ( since NF = 10.LOG(T / 290 + 1) So the delta noise power seen at the output of the system for input at 100C = 373K and Room Temp 290K = : 10.LOG[ (373 + 75) / (290 + 75) ] = 0.89dB Which ought to be measurable with modern wide bandwidth SDRs. It wasn't reliable using a diode on the output of an SSB Rx in 1988 ! Andy G4JNT