crushed walnut shells and Tek 500 series scopes 7 messages
Dennis – I hope you don’t consider this off topic. It is not on repairing a classic Tek scope – but an aspect on how they were made. After reading the posting from the Peter Keller on the Coketron, it
Dennis – I hope you don’t consider this off topic. It is not on repairing a classic Tek scope – but an aspect on how they were made. After reading the posting from the Peter Keller on the Coketron, it
|
By ditter2
·
|
|
In search of holdoff pot for 2235 7 messages
Due to a freak accident, the shaft of the Variable Holdoff potentiometer broke and I need to replace it. The details are: R9521 311-2146-00 Res, Var, Nonwir: 50 ohm, 20% 0.5W CM41773 It is a Clairstat
Due to a freak accident, the shaft of the Variable Holdoff potentiometer broke and I need to replace it. The details are: R9521 311-2146-00 Res, Var, Nonwir: 50 ohm, 20% 0.5W CM41773 It is a Clairstat
|
By John Hudak
·
|
|
Coketron - the "Real" Story from Peter Keller :) :) :) 13 messages
Peter Keller asked me to pass this along to the group. I'm sure he was smiling the entire time he wrote it. Dennis Tillman W7pF I guess it is time for the real story of the Coketron to be exposed. The
Peter Keller asked me to pass this along to the group. I'm sure he was smiling the entire time he wrote it. Dennis Tillman W7pF I guess it is time for the real story of the Coketron to be exposed. The
|
By Dennis Tillman W7pF
·
|
|
564 AC ripple in beam intensity 2 messages
Hi Folks - I'm seeing an intermittent intensity modulation at line frequency on my 564, and have been unable to measure the source. I've tried both a 3B3 timebase, and a 3A72 in the horizontal slot wi
Hi Folks - I'm seeing an intermittent intensity modulation at line frequency on my 564, and have been unable to measure the source. I've tried both a 3B3 timebase, and a 3A72 in the horizontal slot wi
|
By Paul
·
|
|
P6137 scope probe repair 7 messages
Hi all, I recently picked up a P6137 scope probe for a song and thought I'd try my luck. Predictably, the center conductor on the cable is open circuit..... Is there any way to repair or replace this
Hi all, I recently picked up a P6137 scope probe for a song and thought I'd try my luck. Predictably, the center conductor on the cable is open circuit..... Is there any way to repair or replace this
|
By Jared Cabot
·
|
|
2465 Ch1 attenuator repair question 8 messages
Hi, On my 2465, the Ch1 BNC is slightly "loose", and the signal is intermittent depending on which way the probe cable/connector tugs on it. The connector seems to be integral to the A10 attenuator as
Hi, On my 2465, the Ch1 BNC is slightly "loose", and the signal is intermittent depending on which way the probe cable/connector tugs on it. The connector seems to be integral to the A10 attenuator as
|
By saipan59 (Pete)
·
|
|
Missing knob for Tek 576 2 messages
Hi guys, im looking for a missing knob of the trace positioning (see pic below) https://i.gyazo.com/8e018f12acdef8d57bc2c949135d350f.jpg The original part number is 366-1027-00, unfortunately i cant f
Hi guys, im looking for a missing knob of the trace positioning (see pic below) https://i.gyazo.com/8e018f12acdef8d57bc2c949135d350f.jpg The original part number is 366-1027-00, unfortunately i cant f
|
By Dave S.
·
|
|
Tektronix 492BP 22 messages
Hi, I recently adquire this unit I had a diplexder and WG so to be able to increase the 21GHz frequency but this unit does not go over the 21GHz limit. It does not appear to have any options but it do
Hi, I recently adquire this unit I had a diplexder and WG so to be able to increase the 21GHz frequency but this unit does not go over the 21GHz limit. It does not appear to have any options but it do
|
By Mike Yepes
·
|
|
Tektronix 214 Storage Scope 3 messages
Hi Group, Some years ago I got 2 basket case scopes. The problems involved a leaky diode and bad trigger chip in one and for both I decided to remove the batteries, and not have them internal to the s
Hi Group, Some years ago I got 2 basket case scopes. The problems involved a leaky diode and bad trigger chip in one and for both I decided to remove the batteries, and not have them internal to the s
|
By Carl Hallberg
·
|
|
I built a TM500 mainframe tester, and updated the design. Someone might find this useful? 143 messages
Hi all, I have been accumulating TM500 modules and mainframes for a little while, so I recently finished building an 067-1201-99 TM500 mainframe tester from the Tek construction notes floating around.
Hi all, I have been accumulating TM500 modules and mainframes for a little while, so I recently finished building an 067-1201-99 TM500 mainframe tester from the Tek construction notes floating around.
|
By Jared Cabot
·
|
|
2247A Is this normal? 6 messages
Hello All I have a 2247a and when I put it in the voltmeter mode and pick Peak to Peak the cursers are not touching the peaks of the wave form. Is this normal or do i need to get the service manual ou
Hello All I have a 2247a and when I put it in the voltmeter mode and pick Peak to Peak the cursers are not touching the peaks of the wave form. Is this normal or do i need to get the service manual ou
|
By Tim / N8TIM
·
|
|
P6133 PROBE TIP NEEDED
All: Looking for a replacement Probe tip for P6133 - This could be 206-0265-12 or 206-0265-13. New or Good Used would be fine. Anyone who has and extra one and would be willing to sell it, please cont
All: Looking for a replacement Probe tip for P6133 - This could be 206-0265-12 or 206-0265-13. New or Good Used would be fine. Anyone who has and extra one and would be willing to sell it, please cont
|
By Michael W. Lynch
·
|
|
Grabber Slip-On -- P6105 Probe 3 messages
Is there any source for the slip-on "grabbers" for the P6105 probe. These are Tek P/N 013-0107-0. I seem to have lost one. This will teach me not to keep a trash bucket under the scope on the scope-mo
Is there any source for the slip-on "grabbers" for the P6105 probe. These are Tek P/N 013-0107-0. I seem to have lost one. This will teach me not to keep a trash bucket under the scope on the scope-mo
|
By ebrucehunter
·
|
|
Photo Tek_492_SwitchSettings.jpg uploaded
#photo-notice
The following photos have been uploaded to the Tek 492BP album of the TekScopes@groups.io group. Tek_492_SwitchSettings.jpg By: d.schilling.net@...
The following photos have been uploaded to the Tek 492BP album of the TekScopes@groups.io group. Tek_492_SwitchSettings.jpg By: d.schilling.net@...
|
By
TekScopes@groups.io Notification
·
|
|
[Resurrecting a Tek 485] Sweep not from left to right ? 6 messages
Hi guys ! The newbie is back ! Some news : PSU of the 485 delivers proper Low Voltage, I had to adjust the calibrator amplitude (the 5V was out of spec +/- 5mV). I've been checking (visually) several
Hi guys ! The newbie is back ! Some news : PSU of the 485 delivers proper Low Voltage, I had to adjust the calibrator amplitude (the 5V was out of spec +/- 5mV). I've been checking (visually) several
|
By marcosjl31@...
·
|
|
Groups.io Price Increases DO NOT Apply to TekScopes
I have read a copy of the email from Mark Fletcher very carefully line by line. (although I am annoyed I did not receive a copy sent to me directly). As I understand it these price increases DO NOT AP
I have read a copy of the email from Mark Fletcher very carefully line by line. (although I am annoyed I did not receive a copy sent to me directly). As I understand it these price increases DO NOT AP
|
By Dennis Tillman W7pF
·
|
|
Photo Notifications 2 messages
#photo-notice
Carl Hallberg added the album Tektronix 214 storage scope: Converted internal battery operation to external 12 Volts source #Tektronix214 The following photos have been uploaded to the Tektronix 214 s
Carl Hallberg added the album Tektronix 214 storage scope: Converted internal battery operation to external 12 Volts source #Tektronix214 The following photos have been uploaded to the Tektronix 214 s
|
By
TekScopes@groups.io Notification
·
|
|
Groups.io price increase? 8 messages
Is this relevant to this group? [HP-Agilent-Keysight-equipment] Massive groups.pricing increase - we need a plan B. However, it seems that groups.io are going to be changing their pricing https://beta
Is this relevant to this group? [HP-Agilent-Keysight-equipment] Massive groups.pricing increase - we need a plan B. However, it seems that groups.io are going to be changing their pricing https://beta
|
By Tom Gardner
·
|
|
WTB: SC-501 scope and other modules 3 messages
The ongoing discussion of the power supply tester has caused me to investigate the TM-500 series. I found it interesting, was able to get some items, and am interested in buying an SC501 scope and oth
The ongoing discussion of the power supply tester has caused me to investigate the TM-500 series. I found it interesting, was able to get some items, and am interested in buying an SC501 scope and oth
|
By greenboxmaven
·
|
|
T912 storage oscilloscope - No 2.7 KV high Voltage 14 messages
Greetings to all, I have an old T912 that no longer shows traces on the CRT screen. I checked and the 2.7 KV on the cathode of the CRT is missing. By disconnecting the diodes CR463 and CR465 from the
Greetings to all, I have an old T912 that no longer shows traces on the CRT screen. I checked and the 2.7 KV on the cathode of the CRT is missing. By disconnecting the diodes CR463 and CR465 from the
|
By Attilio
·
|