Topics

Tek Probe P6131 replacement tip ?


Graham
 

Probe P6131 replacement tip ?

Ive complet probe, missing the tip , as far as I can see , looking down the
tube, where the tip should be. there is a small off set, rectangular ''contact ??
push in fit ?

Are replacement tips available, if so from where ?

Tnx-Graham


Graham
 

Nb - Or Need replace the full assembly ?

This item >> tek replacement probe tip 206-0265-02 ??


Roy Thistle
 

Hi:
Did you look at the exploded drawing in the "P6131 Instruction Manual"?
I think you are referring to the "probe tip."
From:
"Figure 10: P6131 probe with standard accessories"
and the table:
"Replaceable parts: P6131 probe and standard accessories"
206–0265–10 "TIP,PROBE:10X,10.3PF,CLEAR/BLUE(STANDARD ONLY)" or
206–0265–12 "TIP,PROBE:10X,12.5PF,CLEAR YELLOW(OPTION 02 ONLY)"
I've seen this part(s) (probe tips) come up on Ebay... and it used to be sold by Tektronix; but, the P6131 is obsolete, so I don't think they sell the probe, or parts for it, anymore.


Phil Wong
 

What is a good replacement for a P6131 with broken tip?

-----Original Message-----
From: TekScopes@groups.io <TekScopes@groups.io> On Behalf Of Roy Thistle
Sent: Monday, November 02, 2020 9:40 AM
To: TekScopes@groups.io
Subject: Re: [TekScopes] Tek Probe P6131 replacement tip ?

Hi:
Did you look at the exploded drawing in the "P6131 Instruction Manual"?
I think you are referring to the "probe tip."
From:
"Figure 10: P6131 probe with standard accessories"
and the table:
"Replaceable parts: P6131 probe and standard accessories"
206–0265–10 "TIP,PROBE:10X,10.3PF,CLEAR/BLUE(STANDARD ONLY)" or
206–0265–12 "TIP,PROBE:10X,12.5PF,CLEAR YELLOW(OPTION 02 ONLY)"
I've seen this part(s) (probe tips) come up on Ebay... and it used to be sold by Tektronix; but, the P6131 is obsolete, so I don't think they sell the probe, or parts for it, anymore.


Roy Thistle
 

On Mon, Nov 2, 2020 at 07:15 AM, Phil Wong wrote:


What is a good replacement for a P6131 with broken tip?
A P6131 without a broken tip... or do you mean which different Tek probe... et. al., is a good substitution/replacement?
From TekWiki:
The Tektronix P6131 is a passive miniature 300 MHz ×10 scope probe for 1 MΩ // 14-18 pF scope inputs.It has a ×10 readout pin, and shares accessories with many other Tektronix subminiature modular probes.
Do you have a Tek 24XX scope, that needs the P6131 bandwidth? ... and you need a mini-probe?