Hi,
I've got a FG501 S/N B127250, ex-University of Louisville EE dept. Here's what
I did to get it up and running (interspersed with service manual calibration
steps).

0. The thing looks pretty beat up (ex EE student lab), had to get all the
knobs off, clean them with a toothbrush and windex, and clean the front panel
in the same fashion.

1. Power light out. +20V checks out OK, bulb out, so change R495 to 1.02k,
DS495 to a green LED.

2. Visually inspect: broken connecting wires to R27 (frequency vernier), also
loose mounting nut, resoldered and retightened.

3. Check out power supplies, stupidly enough missed DCPL -20V. Adjusted +20V
and current ref. 13.8V, all supplies (miss one) are fine.

4. No output, so check current sources and diodes with a multimeter,
everything is fine (current sources source/sink current, diodes are not
dead).

5. Remove U80, feed 4V/1kHz/450Ohm cal signal (from scope cal output) to
U80:13 to get diode bridge current switch to flip back and forth.

6. Set frequency to 10^1 x 10 to get a triangle on TP6.

7. TP6 shows 2Vpp triangle @ -6V, same at source of Q120 and emitter of Q125.

8. Noting is coming out of the following transistor, so check biases and DUH
-20V DCPL is dead. U420 (ua741) turns out to have the output pin (6) at 0V.
Had no 741s handy, so replaced U420 and its companion with OP177.

9. U195:3 <4> seems to be low, but methinks it's OK as it's below B130000 and
waveforms really apply to B130000 and up. Output span checks out OK at
+/-7.8V (compared to minimum +/-10V for B130000 & up).

10. Output [amplitude] and offset pots are very noisy, heating them up just a
bit with the heat gun and rotating them a few dozen times back and forth
seems to have gotten them back into shape.

11. Could not get 10^3x10 R20 in range, frequency is 4% high, so change R18 to
9.31k

12. After calibrating 10^3 and 10^5, check of various ranges shows 10^2 range
to be consistently 2% low in frequency.

13. Suspect 10^2 timing capacitor C74 0.1u/100V (bottom of board). Desoldered
it, and replaced with a mechanically larger capacitor from a junker 5B12
timebase plugin. The x1 and x10 frequencies are now within .5%.

Replacement BOM: two resistors (what a luck: they were adjacent!), one
precision cap, one LED, two opamps.

All in all, I think that I'm pretty lucky as at x1 and x10 all the frequencies
check out well within +/-1%. This is a nice, simple instrument and it was a
joy calibrating it.

Cheers, Kuba


um-gs@...
 

Hi,
just want to continue with my lessons learned on the same subject.
The device was operative, suplly voltages were adjustable to their nominal values, but no output.
1. No output
a) open R251 and inject signal voltage: Voltage appears at output, so amplifier is working
b) checking voltage at junction R32/R34 when turning the frequency dial: voltage varies
c) checking voltage at output of U15 pin 6 : does not vary - replace U15 which was defect - oscillator is working again, output signal is generated
2. Sine wave output distorted and adjustment R150 has no effect
a) replacing both R150 and R170 allows adjustment of distortion free sine wave output
3. Output signal not at symmetrical duty cycle
a) adjustment of R45 (step7) has no effect
b) output voltage at U40 pin6 does not vary when R45 is adjusted
c) swiper of R45 constantly at zero
d) opening up/replacing of C40 solved the problem (it was shorted) - output duty cycle can be adjusted

Regards, Gordian