Date
1 - 6 of 6
[Resurrecting a Tek 485] Sweep not from left to right ?
Tom Lee
Hi Leon,
toggle quoted messageShow quoted text
Your post takes me back in time! I feel like adjusting an ion trap magnet now. -- Cheers, Tom -- Prof. Thomas H. Lee Allen Ctr., Rm. 205 350 Jane Stanford Way Stanford University Stanford, CA 94305-4070 http://www-smirc.stanford.edu
On 1/15/2021 11:23, Leon Robinson wrote:
In the days of 60 degree deflection all it took was turning the yoke end for end.
|
|
Leon Robinson
In the days of 60 degree deflection all it took was turning the yoke end for end.
toggle quoted messageShow quoted text
Leon Robinson K5JLR
-------- Original message --------
From: Tom Lee <tomlee@ee.stanford.edu> Date: 01/15/2021 10:23 AM (GMT-06:00) To: TekScopes@groups.io Subject: Re: [TekScopes] [Resurrecting a Tek 485] Sweep not from left to right ? In the old days of crt TVs, it was fun to swap connections on the deflection yoke. Upside down, right-to-left TV? Nanoseconds of hilarity for kids. Simpler times. --Tom -- Prof. Thomas H. Lee Allen Ctr., Rm. 205 350 Jane Stanford Way Stanford University Stanford, CA 94305-4070 http://www-smirc.stanford.edu On 1/15/2021 08:18, marcosjl31@gmail.com wrote: that one was an easy fix.
|
|
Tom Lee
In the old days of crt TVs, it was fun to swap connections on the deflection yoke. Upside down, right-to-left TV? Nanoseconds of hilarity for kids.
toggle quoted messageShow quoted text
Simpler times. --Tom -- Prof. Thomas H. Lee Allen Ctr., Rm. 205 350 Jane Stanford Way Stanford University Stanford, CA 94305-4070 http://www-smirc.stanford.edu
On 1/15/2021 08:18, marcosjl31@gmail.com wrote:
that one was an easy fix.
|
|
marcosjl31@...
that one was an easy fix.
|
|
Roger Evans
The most likely explanation is that the two wires from the X amplifier board to the CRT X deflection plates .have been swapped over by mistake. Take care in correcting this, the pins on the CRT do like being bent and then straightened.
Regards, Roger
|
|
marcosjl31@...
Hi guys !
The newbie is back ! Some news : PSU of the 485 delivers proper Low Voltage, I had to adjust the calibrator amplitude (the 5V was out of spec +/- 5mV). I've been checking (visually) several things : => "Ready" indicator does not work. => on Chanel 1: leds (x1, x10 are not working), 1MOhm/50ohm selector has no light indication. => on chanel 2 : 1MOhm/50ohm selector has no light indication. Don't know if it's complicated/hard to fix ? Then, I was watching an interesting video on YouTube (https://www.youtube.com/watch?v=7Qg6ory3L8Y) and suddenly I realize that the sweep on the Tek 485 I've purchased seems to be inverted ! Here is what I get on Channel 2 (1MHz cal signal) : https://drive.google.com/drive/folders/1_9cXIKxCFtC60aFQ6zIm4ujq2paW4tSH Here is the same signal picture taken from video : https://drive.google.com/drive/folders/1_9cXIKxCFtC60aFQ6zIm4ujq2paW4tSH To check if I was right to suspect the sweep was not from left to right, I selected GND and turn TIME/DIV to 50ms : the sweep spot on the screen travels from right to left on this 485 !! Don't think this is correct... but I may be wrong. Could this be due to improper reassembly following a repair? Hope to here from you, Jose
|
|