TEK 7704A pre-death behaviour a clue to the fault?
Simon Owen
Hi all
I am embarking on the likely arduous journey to resurrect my 7704A from tick mode. It was actually in tick mode when I bought it so I got a good deal (I hope!). I got it home and pulled some plugins out and gave them a good shove home and presto, it hummed to life. I got a couple of months of tinkering with it but alas back in tick mode. I have put together a test load and I'll be measuring voltages and ripple shortly. It just occurred to me that before its most recent death there was potentially a warning sign: the scope would tick a few times before coming to life. Then I remember it would tick for longer and longer, but I found that if I pushed the beam finder button it would 'snap out of it' and get down to business. This clearly didn't last. Does anyone have any wisdom on this symptom so perhaps I will have an idea of the likely problem? When it was running during its 'pre-death' period I didn't notice anything else peculiar. Assuming I can rule out the power supply as the cause of the trouble, I wouldn't know where to begin! Thanks a million Simon Owen
|
|
Eric
First thing I would suspect is a leaky electrolytic or a tant that is
toggle quoted messageShow quoted text
shorted. I am leaning more to the electrolytic though. Seems like a "warm up" with a cap charging in time. Sounds like one of the rails is coming up slowly Eric
On Wed, Oct 7, 2020, 6:50 PM Simon Owen <Sowenbd@gmail.com> wrote:
Hi all
|
|
Simon Owen
Thanks Eric
toggle quoted messageShow quoted text
I forgot to mention I've replaced all the tantalum and electrolytic caps on the rect/filter board. Now I recall a tantalum cap on the board on top of the CRT burned out shortly after its first resurrection so I did a wholesale replacement of tantalum caps throughout. What you're saying makes sense to me though. Are you thinking it could be one or both of the huge beercan caps? Thanks, Simon
On Thu., 8 Oct. 2020, 09:56 Eric, <ericsp@gmail.com> wrote:
First thing I would suspect is a leaky electrolytic or a tant that is
|
|
I concur with Eric that most likely, it's a matter of one or more bad caps.
With the 7704A, the display unit may simply be separated from the base unit, at the "Main Interconnect". You just have to loosen the screws holding down the rails linking the top and bottom sections, to the left and right of the mainframe. After a quick look at the schematics, I can't find a reason why running the base unit powered on with the display unit disconnected, could do any harm. Checking the base unit under power on its own not only may allow to narrow down the problem area but measuring resistances across the power connections of the display unit may indicate the problem area(s). Raymond
|
|
On Thu, Oct 8, 2020 at 01:13 AM, Simon Owen wrote:
No, the beercan caps sit straight across the mains input, before the oscillator causing the ticking, so no tick mode. They fail not often but if they do, they usually cause the main fuse to blow and, with some bad luck, damage the input rectifier, Raymond
|
|
Simon Owen
Thanks a million, great advice both of you. I shall proceed as you suggest
toggle quoted messageShow quoted text
and report back!
On Thu., 8 Oct. 2020, 10:21 Raymond Domp Frank, <hewpatek@gmail.com> wrote:
I concur with Eric that most likely, it's a matter of one or more bad caps.
|
|
Simon Owen
Update: I just measured the +50V supply as being within tolerance but with
1.8 full volts of ripple. Further evidence for the case against an electrolytic capacitor? On Thu., 8 Oct. 2020, 10:26 Simon Owen via groups.io, <Sowenbd= gmail.com@groups.io> wrote: Thanks a million, great advice both of you. I shall proceed as you suggest
|
|
Simon Owen
Meant to say the AVERAGE voltage is within tolerance at 50.1V but with the
toggle quoted messageShow quoted text
measured ripple the peaks fall outside both bounds of tolerance (49.2V to 51.0V)
On Thu., 8 Oct. 2020, 10:37 Simon Owen, <sowenbd@gmail.com> wrote:
Update: I just measured the +50V supply as being within tolerance but with
|
|
On Thu, Oct 8, 2020 at 01:21 AM, Raymond Domp Frank wrote:
There's a failure mode of the big caps that may cause tick mode. I didn't realize earlier because I'm in a 230 VAC country and you're probably in a 115 VAC area. The big caps usually fail shorted where I live but in your area, before becoming shorted they may become (almost) open because of loss of capacity or detached internal links. It should be very easy for you to check that the Vpeak of your mains supply should be across the caps; something like 160 VDC if you're in a 115 VAC country. With low-capacity cans, a pulsating DC voltage may drive the inverter. That may possibly cause ticking. Far from sure, just keeping it in mind. *** Be aware that the big cans are directly connected to the mains, so use an isolation transformer if you can! *** Raymond
|
|
On Thu, Oct 8, 2020 at 01:37 AM, Simon Owen wrote:
You're not saying 60 or 120 Hz ripple, are you? If not, what approximate frequency? Raymond
|
|
Simon Owen
Thanks Raymond,
toggle quoted messageShow quoted text
240V 50Hz give or take (Sydney AU)
On Thu., 8 Oct. 2020, 10:42 Raymond Domp Frank, <hewpatek@gmail.com> wrote:
On Thu, Oct 8, 2020 at 01:21 AM, Raymond Domp Frank wrote:if
|
|
Simon Owen
I measure 1.8V peak-to-peak at approximately 7MHz. Yes that's Megahertz!
toggle quoted messageShow quoted text
On Thu., 8 Oct. 2020, 10:44 Raymond Domp Frank, <hewpatek@gmail.com> wrote:
On Thu, Oct 8, 2020 at 01:37 AM, Simon Owen wrote:with
|
|
On Thu, Oct 8, 2020 at 01:45 AM, Simon Owen wrote:
Aah, 240 VAC even! Do you have a Service Manual (SM)? If you do, it should be clear that no significant 50 Hz (i.e. mains) ripple should (can) come out of the low voltage power supply (LVPS), if the full-wave bridge rectifier works ok, because it produces 100 Hz! Do you also see 50 Hz ripple on the other outputs? If 50 Hz is indeed coming out of the LVPS, my idea of bad capacity value of one or both big cans becomes closer or, even much more likely, the bridge rectifier could be partly defective, because otherwise, you'd see the full-wave-rectified 100 Hz that the bridge produces. The idea would be that during the half-wave that the bridge functions ok, not enough voltage can be generated to produce a DC voltage that's flat enough. With a linear regulator, I'd bet 90% on that with a 50 Hz ripple! Raymond
|
|
On Thu, Oct 8, 2020 at 01:58 AM, Raymond Domp Frank wrote:
To be clear: Obviously, it isn't a linear regulator. If it isn't the bridge, which can be excluded with a good 340 VDC across the big caps and you do have 50 Hz across the LV outputs, the circuit around U3105 may be the culprit. That's on the lower part of DWG 8, "Inverter/rectifiers". OTOH: You say the power supply is in tick mode but you do have at least one voltage that is close (50 VDC)? Raymond
|
|
Simon Owen
I have put together a dummy load that demands about half of the total power
toggle quoted messageShow quoted text
the scope can consume, built from info I found in the forum. The PSU ticks a few times and then fires up with the dummy load connected. I thought: Great! The power supply is OK! Then: Oh no, if it's not the power supply... but now the measurements I am getting for the different supplies are baffling my novice brain. The +5V supply is putting out a waveform which is sinusoidal 1.2Vppk and centred on zero volts, the frequency of the wave is 7MHz. The -15V supply has 1.2V of ripple at the same 7MHz. +15V has 50Hz ripple of about 10mV. Probing continues.
On Thu., 8 Oct. 2020, 11:16 Raymond Domp Frank, <hewpatek@gmail.com> wrote:
On Thu, Oct 8, 2020 at 01:58 AM, Raymond Domp Frank wrote:To be clear: Obviously, it isn't a linear regulator.
|
|
On Thu, Oct 8, 2020 at 01:55 AM, Simon Owen wrote:
OK, timing in our messages probably caused my misunderstanding that you saw 50 Hz ripple on the 50 VDC. I now understand it's 7 MHz on the 50 VDC output. Tick mode applies to all low-voltage outputs together, so if 50 VDC is present, so should the other low voltages. Currently, I'm not sure I have a clear picture. Raymond
|
|
Simon Owen
Nor I. I will continue and put together my findings. Thanks for your help.
toggle quoted messageShow quoted text
S
On Thu., 8 Oct. 2020, 11:29 Raymond Domp Frank, <hewpatek@gmail.com> wrote:
On Thu, Oct 8, 2020 at 01:55 AM, Simon Owen wrote:OK, timing in our messages probably caused my misunderstanding that you
|
|
On Thu, Oct 8, 2020 at 02:25 AM, Simon Owen wrote:
Are you sure that your loads are all connected correctly between the output and ground each and that you're measuring between the outputs and GND? I'm not sure what to think of the 7 MHz. It's far too high for the inverter and looks more like a (parasitic) oscillation. That could happen around U3105 because of a missing or bad cap. Could it be there from when you recapped it? Did you forget a small cap or connected it wrongly? Can you try and look at the inputs of the linear regulators (dwg 9 and figure 6-16) and report on ripple, esp. if you're seeing the 7 MHz as well? I'm going to bed now (I'm in NL, GMT+2). Have fun! Raymond
|
|
On Thu, Oct 8, 2020 at 02:48 AM, Raymond Domp Frank wrote:
Quick afterburner: I understand now that the PSU is out of the 'scope and you're running with just a load connected? So it's still just the PSU after all? Raymond
|
|
Simon Owen
Thank you Raymond, good night, by the way I think I have found several
toggle quoted messageShow quoted text
power diodes not behaving properly with reverse voltages of 2 to 3 volts only. I think we might be zeroing in. Cheers, Simon
On Thu., 8 Oct. 2020, 11:51 Raymond Domp Frank, <hewpatek@gmail.com> wrote:
On Thu, Oct 8, 2020 at 02:25 AM, Simon Owen wrote:power
|
|