2465BDV slow reaction
Hi everybody,
I have a 2465BDV (early series, serial B013522) that has a weird problem. Everything works fine, display is crisp, readout/cursors/etc everything works. The problem is it is very slow to response to controls. E.g. when turning vertical position pot it doesn't react immediately so trace does not move. Then, after a second (may be half-second but very noticeable and long delay) it jumps to a new location. That is true for all controls e.g. it is very difficult to use trigger level. It is also slow to respond to switches (e.g. trigger mode) and when I turn off e.g. cursors they dissappear but the big character reading stays on the screen for a second or so after cursor turned off. If I turn a pot VERY slowly it follows it i.e. there is a very long lag on reading the control panel. Also all LEDs flash momentarily when it loses sync or any control changing LED state is activated (i.e. when there is a change in which LEDs are lit on the control panel.) Other than that everything works fine. All voltages are normal, no ripple, LVPS was recapped recently so there is no problems with the power. All caps were like new so no recap was actually required but once I've already pulled that PS off the scope I proceeded with the recap. It's been behaving exactly the same before that so that recap didn't change anything. Have anybody experienced anything like that? --- ****************************************************************** * KSI@home KOI8 Net < > The impossible we do immediately. * * Las Vegas NV, USA < > Miracles require 24-hour notice. * ******************************************************************
|
|
honolulusnowwhite
Hi Sergey,
toggle quoted messageShow quoted text
I haven't had this problem. I suggest that you follow the Preliminary Troubleshooting chart in the repair manual. Among other things, the chart has you check the clock speed of the microprocessor and the signals at selected other pins. The microprocessor is supposed to scan the front panel switches every 32 ms while the front panel pots are supposed to be scanned every 16 ms. Good luck, Patrick Wong AK6C
--- In TekScopes@..., Sergey Kubushyn <ksi@...> wrote:
|
|