Problem calibration TEK 495P
hh <ciulandro@...>
Hello,
i've this problem on my Tek 495P. When i start the calibration routine,with the internal generator 100Hhz -20 dB,the routine fails because the processor says taht there is any signal calibrator is present. I've checked the output signal at 100Mhz -20 Db and is ok. So i've connected to rf input an external RF generator at 100MHZ and the test start only when i give a signal @ 100Mhz 0(zero) dB. Seem to be a loss of gain. What can i verify in the instruments for repair this loss of gain? Thanks for Any help. MAX
|
|
|
|
Bob Koller <testtech@...>
--- On Sun, 7/3/11, hh wrote:
|
|
|
|
hh <ciulandro@...>
Thanks,
toggle quoted messageShow quoted text
i've made the test that you seggest and semm that the attenuator works. I can allign manually the level of the 495 with the front trimmer using the RF generator as reference. But the autocalibration routine always fail. If i give 0dBm the routine start and appear a square wave.The dispaly says to use the AMPL trimmer for made the square wawe as a straight line but i can't reach the straigth line.
--- In TekScopes@yahoogroups.com, Bob Koller <testtech@...> wrote:
|
|
|