Rom Images for Fluke 884xA/xx series DMM


Tom Miller
 

Thanks Stefan.
 
Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
the image for the 8842A/AF Ver 3.1.
 
Any others you can find will help.
 
 
Thanks,
Tom
 
 

----- Original Message -----
Sent: Wednesday, November 03, 2010 1:22 AM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I will be glad to provide the ROM image from my 8842A. Give me a few days to obtain it.
(Would be useful to start a new thread with this topic).

ST

On Wed, Nov 3, 2010 at 4:50 AM, Tom Miller <tmiller@...> wrote:


Hi Dave,
 
I am looking for a copy of the 8840A/AF EPROM. Jay needs the 8842A EPROM. If you have (1) a
working model of any of these models, a way to copy the binary data (prom programmer), and
willing to assist, it would be a great help. It seems we are seeing a number of EPROM failures in
these meters. I have two bad units alone and I see a large number of these showing up recently
as parts units.
 
The EPROM is a 2732A in a socket. It is marked U222 and is near the AC power line filter.
 
The four basic models we are seeking are:
 
1. 8840A
2. 8840A/AF
3. 8842A
4. 8842A/AF  (I have Ver 3.1) but there are others)
 
There may be several versions of the firmware running around so you may have more than one
version that is needed.
 
 
Thanks for any help you might provide.
 
Regards,
Tom
 
 
----- Original Message -----
From: Dave Casey
Sent: Tuesday, November 02, 2010 11:29 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I have two 8840As, both working after some effort (check the archives). What do you need?
 
Dave Casey
Sent from my PC
----- Original Message -----
From: Tom Miller
Sent: Tuesday, November 02, 2010 10:04 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

With out the battery, the + input to the op amp would float due to
the diode and possibly become unstable. The jumper terminates
the op amp with the 10 k resistor. I guess we would need to know
the serial number of the scope to use the correct manual.
 
Tom
 
OT PS to all, Do you have or know any one with a Fluke 8840A, 8840 A/AF, 8842 A, or 8842 A/AF?
 
Several of us are looking for the ROM binary images for these DMMs. They use a standard 2732A
EPROM that is socketed for the system memory. It seems these meters may be suffering the same
ROM issues the Tek scopes do.
 
 
----- Original Message -----
Sent: Tuesday, November 02, 2010 10:26 PM
Subject: [TekScopes] 2440 Battery Not Installed by Factory?

 

Hello Again:

I have a 2440 which appears to have a short in place of where the battery is supposed to go. This short looks to be installed by the factory, and the direct reason why It fails FFP. My question is why would the factory place a short across the battery, and how was the unit calibrated with a FFP failure?

I have looked at the technical and maintenance manual, and it appears as though this is not an option, but should have been standard in all models. Since the unit did come from USN surplus, it may have been requested that this unit be placed into this condition by the factory, as memory is a big security issue.  If that is the case, how would one clear a FFP FAIL on a product which come from the factory with no batteries for the memory?????

Thanks

Rich


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


Dave Casey
 

I believe I have two different versions of the 8840A ROM. Going back to what I dealt with two years ago:
 
DMM #1 had a blank display, and after being powered on would enter some sort of relay cycling tick mode. This turned out to be a watchdog reset signal. Upon powering off, the display would light briefly as it unblanked before the display lost power.
U202 is a mask ROM Zilog Z8, Fluke number 685438, marked V2.3 in silkscreen.
U222 (external PROM, Intel 2732) bears Fluke part number 728964 and V.2.3 on it's sticker.
 
DMM #2 has a frozen display when powered on, but seems audibly responsive to front panel button presses.
U202 is a OTP Z86E2112PSC 8k bearing Fluke part number 607256 and "V4.0" on it's sticker. I suspect it was replaced at some point.
U222 is Fluke number 855692 and labeled "REV 2.5"
 
Swapping the U202's between the two units produced a working DMM #1 and a failing DMM #2 with #1's old problem. DMM #1 also worked fine with either external PROM.
The part number for the 2.5 Zilog Z8 is 854067, and I found one on the internet for about $10.
The part that arrived was marked "Z0861108PSC R941 9311 BF JF 854067" and bore no mention of V2.5, though the parts listing from the manual says that's what it is. I tried it in DMM #1 with the V2.3 PROM and it wasn't happy. I then took the newer V4.0 U202 from DMM #2 and placed it in DMM #1 with the original V2.3 external PROM and had a working DMM #1.
I put the newly acquired V2.5 U202 into DMM #2 with its presumably original V2.5 PROM and had the same failure mode as when I first got it. (That failure is the display chip, U212, confirmed by swapping with from DMM #1. As soon as it's replacement (another $10) arrives, DMM #2 will be fully functional.)
 
Anyway, I don't know what's on the PROM, but it might be obsolete and not addressed with the V4.0 firmware anyway, and since the V4.0 chip is a OTP Z8, I presume one could duplicate it.
 
So to conclude, all the code is not on U222 and neither of my two DMMs suffered from a defective U222. However, I am happy to pull these images and will hopefully get that done tomorrow.
I don't have the immedate means to pull the images from the Z8s, but if someone can tell me how it's done, I'm willing to try.
 
Dave Casey
Sent from my PC
 

----- Original Message -----
From: Tom Miller
Sent: Wednesday, November 03, 2010 12:38 AM
Subject: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Thanks Stefan.
 
Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
the image for the 8842A/AF Ver 3.1.
 
Any others you can find will help.
 
 
Thanks,
Tom
 
 
----- Original Message -----
Sent: Wednesday, November 03, 2010 1:22 AM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I will be glad to provide the ROM image from my 8842A. Give me a few days to obtain it.
(Would be useful to start a new thread with this topic).

ST

On Wed, Nov 3, 2010 at 4:50 AM, Tom Miller <tmiller@...> wrote:


Hi Dave,
 
I am looking for a copy of the 8840A/AF EPROM. Jay needs the 8842A EPROM. If you have (1) a
working model of any of these models, a way to copy the binary data (prom programmer), and
willing to assist, it would be a great help. It seems we are seeing a number of EPROM failures in
these meters. I have two bad units alone and I see a large number of these showing up recently
as parts units.
 
The EPROM is a 2732A in a socket. It is marked U222 and is near the AC power line filter.
 
The four basic models we are seeking are:
 
1. 8840A
2. 8840A/AF
3. 8842A
4. 8842A/AF  (I have Ver 3.1) but there are others)
 
There may be several versions of the firmware running around so you may have more than one
version that is needed.
 
 
Thanks for any help you might provide.
 
Regards,
Tom
 
 
----- Original Message -----
From: Dave Casey
Sent: Tuesday, November 02, 2010 11:29 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I have two 8840As, both working after some effort (check the archives). What do you need?
 
Dave Casey
Sent from my PC
----- Original Message -----
From: Tom Miller
Sent: Tuesday, November 02, 2010 10:04 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

With out the battery, the + input to the op amp would float due to
the diode and possibly become unstable. The jumper terminates
the op amp with the 10 k resistor. I guess we would need to know
the serial number of the scope to use the correct manual.
 
Tom
 
OT PS to all, Do you have or know any one with a Fluke 8840A, 8840 A/AF, 8842 A, or 8842 A/AF?
 
Several of us are looking for the ROM binary images for these DMMs. They use a standard 2732A
EPROM that is socketed for the system memory. It seems these meters may be suffering the same
ROM issues the Tek scopes do.
 
 
----- Original Message -----
Sent: Tuesday, November 02, 2010 10:26 PM
Subject: [TekScopes] 2440 Battery Not Installed by Factory?

 

Hello Again:

I have a 2440 which appears to have a short in place of where the battery is supposed to go. This short looks to be installed by the factory, and the direct reason why It fails FFP. My question is why would the factory place a short across the battery, and how was the unit calibrated with a FFP failure?

I have looked at the technical and maintenance manual, and it appears as though this is not an option, but should have been standard in all models. Since the unit did come from USN surplus, it may have been requested that this unit be placed into this condition by the factory, as memory is a big security issue.  If that is the case, how would one clear a FFP FAIL on a product which come from the factory with no batteries for the memory?????

Thanks

Rich


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


 

Maybe I can read the Zilog witht the Beeprog?
<http://www.elnec.com/sw/dev_html/zilog_dev.htm>

ST


On Wed, Nov 3, 2010 at 6:46 AM, Dave Casey <dcasey@...> wrote:


I believe I have two different versions of the 8840A ROM. Going back to what I dealt with two years ago:
 
DMM #1 had a blank display, and after being powered on would enter some sort of relay cycling tick mode. This turned out to be a watchdog reset signal. Upon powering off, the display would light briefly as it unblanked before the display lost power.
U202 is a mask ROM Zilog Z8, Fluke number 685438, marked V2.3 in silkscreen.
U222 (external PROM, Intel 2732) bears Fluke part number 728964 and V.2.3 on it's sticker.
 
DMM #2 has a frozen display when powered on, but seems audibly responsive to front panel button presses.
U202 is a OTP Z86E2112PSC 8k bearing Fluke part number 607256 and "V4.0" on it's sticker. I suspect it was replaced at some point.
U222 is Fluke number 855692 and labeled "REV 2.5"
 
Swapping the U202's between the two units produced a working DMM #1 and a failing DMM #2 with #1's old problem. DMM #1 also worked fine with either external PROM.
The part number for the 2.5 Zilog Z8 is 854067, and I found one on the internet for about $10.
The part that arrived was marked "Z0861108PSC R941 9311 BF JF 854067" and bore no mention of V2.5, though the parts listing from the manual says that's what it is. I tried it in DMM #1 with the V2.3 PROM and it wasn't happy. I then took the newer V4.0 U202 from DMM #2 and placed it in DMM #1 with the original V2.3 external PROM and had a working DMM #1.
I put the newly acquired V2.5 U202 into DMM #2 with its presumably original V2.5 PROM and had the same failure mode as when I first got it. (That failure is the display chip, U212, confirmed by swapping with from DMM #1. As soon as it's replacement (another $10) arrives, DMM #2 will be fully functional.)
 
Anyway, I don't know what's on the PROM, but it might be obsolete and not addressed with the V4.0 firmware anyway, and since the V4.0 chip is a OTP Z8, I presume one could duplicate it.
 
So to conclude, all the code is not on U222 and neither of my two DMMs suffered from a defective U222. However, I am happy to pull these images and will hopefully get that done tomorrow.
I don't have the immedate means to pull the images from the Z8s, but if someone can tell me how it's done, I'm willing to try.
 
Dave Casey
Sent from my PC
 
----- Original Message -----
From: Tom Miller
Sent: Wednesday, November 03, 2010 12:38 AM
Subject: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Thanks Stefan.
 
Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
the image for the 8842A/AF Ver 3.1.
 
Any others you can find will help.
 
 
Thanks,
Tom
 
 
----- Original Message -----
Sent: Wednesday, November 03, 2010 1:22 AM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I will be glad to provide the ROM image from my 8842A. Give me a few days to obtain it.
(Would be useful to start a new thread with this topic).

ST

On Wed, Nov 3, 2010 at 4:50 AM, Tom Miller <tmiller@...> wrote:


Hi Dave,
 
I am looking for a copy of the 8840A/AF EPROM. Jay needs the 8842A EPROM. If you have (1) a
working model of any of these models, a way to copy the binary data (prom programmer), and
willing to assist, it would be a great help. It seems we are seeing a number of EPROM failures in
these meters. I have two bad units alone and I see a large number of these showing up recently
as parts units.
 
The EPROM is a 2732A in a socket. It is marked U222 and is near the AC power line filter.
 
The four basic models we are seeking are:
 
1. 8840A
2. 8840A/AF
3. 8842A
4. 8842A/AF  (I have Ver 3.1) but there are others)
 
There may be several versions of the firmware running around so you may have more than one
version that is needed.
 
 
Thanks for any help you might provide.
 
Regards,
Tom
 
 
----- Original Message -----
From: Dave Casey
Sent: Tuesday, November 02, 2010 11:29 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I have two 8840As, both working after some effort (check the archives). What do you need?
 
Dave Casey
Sent from my PC
----- Original Message -----
From: Tom Miller
Sent: Tuesday, November 02, 2010 10:04 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

With out the battery, the + input to the op amp would float due to
the diode and possibly become unstable. The jumper terminates
the op amp with the 10 k resistor. I guess we would need to know
the serial number of the scope to use the correct manual.
 
Tom
 
OT PS to all, Do you have or know any one with a Fluke 8840A, 8840 A/AF, 8842 A, or 8842 A/AF?
 
Several of us are looking for the ROM binary images for these DMMs. They use a standard 2732A
EPROM that is socketed for the system memory. It seems these meters may be suffering the same
ROM issues the Tek scopes do.
 
 
----- Original Message -----
Sent: Tuesday, November 02, 2010 10:26 PM
Subject: [TekScopes] 2440 Battery Not Installed by Factory?

 

Hello Again:

I have a 2440 which appears to have a short in place of where the battery is supposed to go. This short looks to be installed by the factory, and the direct reason why It fails FFP. My question is why would the factory place a short across the battery, and how was the unit calibrated with a FFP failure?

I have looked at the technical and maintenance manual, and it appears as though this is not an option, but should have been standard in all models. Since the unit did come from USN surplus, it may have been requested that this unit be placed into this condition by the factory, as memory is a big security issue.  If that is the case, how would one clear a FFP FAIL on a product which come from the factory with no batteries for the memory?????

Thanks

Rich


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





 

Hi Tom

FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the 8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume that the code would fit in the OTP device.

Still wondering whether a set of devices moved from a 8840A would work in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher than it should (1 volt = 100 volts), and resistance measurements are screwed up as well. I suppose there could be something else wrong with the meter, but I haven't taken a closer look thinking that something is wrong with the firmware revisions.

Jay

--- In TekScopes@..., "Tom Miller" <tmiller@...> wrote:

Thanks Stefan.

Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
the image for the 8842A/AF Ver 3.1.

Any others you can find will help.


Thanks,
Tom
[snip]


Tom Miller
 

Just looking at the Z8 datasheet, there is a way to read back the otp ROM. If the
protect function has not been activated.
 
Do you have one to read?
 
I did a bit of comparing the code in the image that jayW sent me. The last 14 bytes match but
there are  many differences.
 
 
Dave, where did you find the parts? Any more available?
 
Regards,
Tom
 

----- Original Message -----
Sent: Wednesday, November 03, 2010 2:06 AM
Subject: Re: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Maybe I can read the Zilog witht the Beeprog?
<http://www.elnec.com/sw/dev_html/zilog_dev.htm>

ST

On Wed, Nov 3, 2010 at 6:46 AM, Dave Casey <dcasey@...> wrote:


I believe I have two different versions of the 8840A ROM. Going back to what I dealt with two years ago:
 
DMM #1 had a blank display, and after being powered on would enter some sort of relay cycling tick mode. This turned out to be a watchdog reset signal. Upon powering off, the display would light briefly as it unblanked before the display lost power.
U202 is a mask ROM Zilog Z8, Fluke number 685438, marked V2.3 in silkscreen.
U222 (external PROM, Intel 2732) bears Fluke part number 728964 and V.2.3 on it's sticker.
 
DMM #2 has a frozen display when powered on, but seems audibly responsive to front panel button presses.
U202 is a OTP Z86E2112PSC 8k bearing Fluke part number 607256 and "V4.0" on it's sticker. I suspect it was replaced at some point.
U222 is Fluke number 855692 and labeled "REV 2.5"
 
Swapping the U202's between the two units produced a working DMM #1 and a failing DMM #2 with #1's old problem. DMM #1 also worked fine with either external PROM.
The part number for the 2.5 Zilog Z8 is 854067, and I found one on the internet for about $10.
The part that arrived was marked "Z0861108PSC R941 9311 BF JF 854067" and bore no mention of V2.5, though the parts listing from the manual says that's what it is. I tried it in DMM #1 with the V2.3 PROM and it wasn't happy. I then took the newer V4.0 U202 from DMM #2 and placed it in DMM #1 with the original V2.3 external PROM and had a working DMM #1.
I put the newly acquired V2.5 U202 into DMM #2 with its presumably original V2.5 PROM and had the same failure mode as when I first got it. (That failure is the display chip, U212, confirmed by swapping with from DMM #1. As soon as it's replacement (another $10) arrives, DMM #2 will be fully functional.)
 
Anyway, I don't know what's on the PROM, but it might be obsolete and not addressed with the V4.0 firmware anyway, and since the V4.0 chip is a OTP Z8, I presume one could duplicate it.
 
So to conclude, all the code is not on U222 and neither of my two DMMs suffered from a defective U222. However, I am happy to pull these images and will hopefully get that done tomorrow.
I don't have the immedate means to pull the images from the Z8s, but if someone can tell me how it's done, I'm willing to try.
 
Dave Casey
Sent from my PC
 
----- Original Message -----
From: Tom Miller
Sent: Wednesday, November 03, 2010 12:38 AM
Subject: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Thanks Stefan.
 
Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
the image for the 8842A/AF Ver 3.1.
 
Any others you can find will help.
 
 
Thanks,
Tom
 
 
----- Original Message -----
Sent: Wednesday, November 03, 2010 1:22 AM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I will be glad to provide the ROM image from my 8842A. Give me a few days to obtain it.
(Would be useful to start a new thread with this topic).

ST

On Wed, Nov 3, 2010 at 4:50 AM, Tom Miller <tmiller@...> wrote:


Hi Dave,
 
I am looking for a copy of the 8840A/AF EPROM. Jay needs the 8842A EPROM. If you have (1) a
working model of any of these models, a way to copy the binary data (prom programmer), and
willing to assist, it would be a great help. It seems we are seeing a number of EPROM failures in
these meters. I have two bad units alone and I see a large number of these showing up recently
as parts units.
 
The EPROM is a 2732A in a socket. It is marked U222 and is near the AC power line filter.
 
The four basic models we are seeking are:
 
1. 8840A
2. 8840A/AF
3. 8842A
4. 8842A/AF  (I have Ver 3.1) but there are others)
 
There may be several versions of the firmware running around so you may have more than one
version that is needed.
 
 
Thanks for any help you might provide.
 
Regards,
Tom
 
 
----- Original Message -----
From: Dave Casey
Sent: Tuesday, November 02, 2010 11:29 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I have two 8840As, both working after some effort (check the archives). What do you need?
 
Dave Casey
Sent from my PC
----- Original Message -----
From: Tom Miller
Sent: Tuesday, November 02, 2010 10:04 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

With out the battery, the + input to the op amp would float due to
the diode and possibly become unstable. The jumper terminates
the op amp with the 10 k resistor. I guess we would need to know
the serial number of the scope to use the correct manual.
 
Tom
 
OT PS to all, Do you have or know any one with a Fluke 8840A, 8840 A/AF, 8842 A, or 8842 A/AF?
 
Several of us are looking for the ROM binary images for these DMMs. They use a standard 2732A
EPROM that is socketed for the system memory. It seems these meters may be suffering the same
ROM issues the Tek scopes do.
 
 
----- Original Message -----
Sent: Tuesday, November 02, 2010 10:26 PM
Subject: [TekScopes] 2440 Battery Not Installed by Factory?

 

Hello Again:

I have a 2440 which appears to have a short in place of where the battery is supposed to go. This short looks to be installed by the factory, and the direct reason why It fails FFP My question is why would the factory place a short across the battery, and how was the unit calibrated with a FFP failure?

I have looked at the technical and maintenance manual, and it appears as though this is not an option, but should have been standard in all models. Since the unit did come from USN surplus, it may have been requested that this unit be placed into this condition by the factory, as memory is a big security issue  If that is the case, how would one clear a FFP FAIL on a product which come from the factory with no batteries for the memory?????

Thanks

Rich


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


Dave Casey
 

I pulled my two versions of U222 this evening, one being 2.3 and the other 2.5
I will send these to Bruce Lane and to Tom Miller and they can be put wherever.
While I had things apart, I did confirm that the version 4.0 OTP Z8 does not need U222 to run the DMM.
 
These chips do appear to be specific to the 8840A as they even have "8840" on the labels.
 
Dave Casey
Sent from my PC

----- Original Message -----
Sent: Wednesday, November 03, 2010 10:02 AM
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM

 

Hi Tom

FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the 8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume that the code would fit in the OTP device.

Still wondering whether a set of devices moved from a 8840A would work in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher than it should (1 volt = 100 volts), and resistance measurements are screwed up as well. I suppose there could be something else wrong with the meter, but I haven't taken a closer look thinking that something is wrong with the firmware revisions.

Jay

--- In TekScopes@..., "Tom Miller" wrote:
>
> Thanks Stefan.
>
> Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
> the image for the 8842A/AF Ver 3.1.
>
> Any others you can find will help.
>
>
> Thanks,
> Tom
>
[snip]


Dave Casey
 

I bought it used from Ucite two years ago. It appears to be the only one they had.
 
Dave Casey
Sent from my PC
 

----- Original Message -----
From: Tom Miller
Sent: Wednesday, November 03, 2010 10:29 PM
Subject: Re: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Just looking at the Z8 datasheet, there is a way to read back the otp ROM. If the
protect function has not been activated.
 
Do you have one to read?
 
I did a bit of comparing the code in the image that jayW sent me. The last 14 bytes match but
there are  many differences.
 
 
Dave, where did you find the parts? Any more available?
 
Regards,
Tom
 
----- Original Message -----
Sent: Wednesday, November 03, 2010 2:06 AM
Subject: Re: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Maybe I can read the Zilog witht the Beeprog?
<http://www.elnec.com/sw/dev_html/zilog_dev.htm>

ST

On Wed, Nov 3, 2010 at 6:46 AM, Dave Casey <dcasey@...> wrote:


I believe I have two different versions of the 8840A ROM. Going back to what I dealt with two years ago:
 
DMM #1 had a blank display, and after being powered on would enter some sort of relay cycling tick mode. This turned out to be a watchdog reset signal. Upon powering off, the display would light briefly as it unblanked before the display lost power.
U202 is a mask ROM Zilog Z8, Fluke number 685438, marked V2.3 in silkscreen.
U222 (external PROM, Intel 2732) bears Fluke part number 728964 and V.2.3 on it's sticker.
 
DMM #2 has a frozen display when powered on, but seems audibly responsive to front panel button presses.
U202 is a OTP Z86E2112PSC 8k bearing Fluke part number 607256 and "V4.0" on it's sticker. I suspect it was replaced at some point.
U222 is Fluke number 855692 and labeled "REV 2.5"
 
Swapping the U202's between the two units produced a working DMM #1 and a failing DMM #2 with #1's old problem. DMM #1 also worked fine with either external PROM.
The part number for the 2.5 Zilog Z8 is 854067, and I found one on the internet for about $10.
The part that arrived was marked "Z0861108PSC R941 9311 BF JF 854067" and bore no mention of V2.5, though the parts listing from the manual says that's what it is. I tried it in DMM #1 with the V2.3 PROM and it wasn't happy. I then took the newer V4.0 U202 from DMM #2 and placed it in DMM #1 with the original V2.3 external PROM and had a working DMM #1.
I put the newly acquired V2.5 U202 into DMM #2 with its presumably original V2.5 PROM and had the same failure mode as when I first got it. (That failure is the display chip, U212, confirmed by swapping with from DMM #1. As soon as it's replacement (another $10) arrives, DMM #2 will be fully functional.)
 
Anyway, I don't know what's on the PROM, but it might be obsolete and not addressed with the V4.0 firmware anyway, and since the V4.0 chip is a OTP Z8, I presume one could duplicate it.
 
So to conclude, all the code is not on U222 and neither of my two DMMs suffered from a defective U222. However, I am happy to pull these images and will hopefully get that done tomorrow.
I don't have the immedate means to pull the images from the Z8s, but if someone can tell me how it's done, I'm willing to try.
 
Dave Casey
Sent from my PC
 
----- Original Message -----
From: Tom Miller
Sent: Wednesday, November 03, 2010 12:38 AM
Subject: [TekScopes] Rom Images for Fluke 884xA/xx series DMM

 

Thanks Stefan.
 
Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
the image for the 8842A/AF Ver 3.1.
 
Any others you can find will help.
 
 
Thanks,
Tom
 
 
----- Original Message -----
Sent: Wednesday, November 03, 2010 1:22 AM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I will be glad to provide the ROM image from my 8842A. Give me a few days to obtain it.
(Would be useful to start a new thread with this topic).

ST

On Wed, Nov 3, 2010 at 4:50 AM, Tom Miller <tmiller@...> wrote:


Hi Dave,
 
I am looking for a copy of the 8840A/AF EPROM. Jay needs the 8842A EPROM. If you have (1) a
working model of any of these models, a way to copy the binary data (prom programmer), and
willing to assist, it would be a great help. It seems we are seeing a number of EPROM failures in
these meters. I have two bad units alone and I see a large number of these showing up recently
as parts units.
 
The EPROM is a 2732A in a socket. It is marked U222 and is near the AC power line filter.
 
The four basic models we are seeking are:
 
1. 8840A
2. 8840A/AF
3. 8842A
4. 8842A/AF  (I have Ver 3.1) but there are others)
 
There may be several versions of the firmware running around so you may have more than one
version that is needed.
 
 
Thanks for any help you might provide.
 
Regards,
Tom
 
 
----- Original Message -----
From: Dave Casey
Sent: Tuesday, November 02, 2010 11:29 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

I have two 8840As, both working after some effort (check the archives). What do you need?
 
Dave Casey
Sent from my PC
----- Original Message -----
From: Tom Miller
Sent: Tuesday, November 02, 2010 10:04 PM
Subject: Re: [TekScopes] 2440 Battery Not Installed by Factory?

 

With out the battery, the + input to the op amp would float due to
the diode and possibly become unstable. The jumper terminates
the op amp with the 10 k resistor. I guess we would need to know
the serial number of the scope to use the correct manual.
 
Tom
 
OT PS to all, Do you have or know any one with a Fluke 8840A, 8840 A/AF, 8842 A, or 8842 A/AF?
 
Several of us are looking for the ROM binary images for these DMMs. They use a standard 2732A
EPROM that is socketed for the system memory. It seems these meters may be suffering the same
ROM issues the Tek scopes do.
 
 
----- Original Message -----
Sent: Tuesday, November 02, 2010 10:26 PM
Subject: [TekScopes] 2440 Battery Not Installed by Factory?

 

Hello Again:

I have a 2440 which appears to have a short in place of where the battery is supposed to go. This short looks to be installed by the factory, and the direct reason why It fails FFP My question is why would the factory place a short across the battery, and how was the unit calibrated with a FFP failure?

I have looked at the technical and maintenance manual, and it appears as though this is not an option, but should have been standard in all models. Since the unit did come from USN surplus, it may have been requested that this unit be placed into this condition by the factory, as memory is a big security issue  If that is the case, how would one clear a FFP FAIL on a product which come from the factory with no batteries for the memory?????

Thanks

Rich


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10





No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1153 / Virus Database: 424/3233 - Release Date: 11/02/10


 

Hi Jay, i have a similar problems with my Fluke 8840, my z8 (Z0861108SPC) romles micro is dead, i have Data I/O, BP micro, ALL-11 programmers, and i tryed to tead one like 86E21, but without any success, from Ebay i bought Z8613RS microcontroller with piggyback eeprom, but i haven't binary dump for programming, please if you can read your, and upload somewhere.
 
p.s. i think that no way to read Z08611 with stadart programmer, but 86E2112 can be read.
 
Best regards.
NikiG
 
 

----- Original Message -----
Sent: Wednesday, November 03, 2010 5:02 PM
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM

 

Hi Tom

FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the 8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume that the code would fit in the OTP device.

Still wondering whether a set of devices moved from a 8840A would work in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher than it should (1 volt = 100 volts), and resistance measurements are screwed up as well. I suppose there could be something else wrong with the meter, but I haven't taken a closer look thinking that something is wrong with the firmware revisions.

Jay

--- In TekScopes@..., "Tom Miller" <tmiller@...> wrote:
>
> Thanks Stefan.
>
> Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
> the image for the 8842A/AF Ver 3.1.
>
> Any others you can find will help.
>
>
> Thanks,
> Tom
>
[snip]


 

Hi NikiG,

I posted it on KO4BB, but it is not yet available for download. I also sent a copy to Tom Miller, and he posted the file on the Yahoo group Fluke_DMM The file is called 8840A_ROMSET.zip

If you join the group, you should be able to download. Let me know how you make out.

Jay

--- In TekScopes@..., "Niki G." <trds.co@...> wrote:

Hi Jay, i have a similar problems with my Fluke 8840, my z8 (Z0861108SPC) romles micro is dead, i have Data I/O, BP micro, ALL-11 programmers, and i tryed to tead one like 86E21, but without any success, from Ebay i bought Z8613RS microcontroller with piggyback eeprom, but i haven't binary dump for programming, please if you can read your, and upload somewhere.

p.s. i think that no way to read Z08611 with stadart programmer, but 86E2112 can be read.

Best regards.
NikiG


----- Original Message -----
From: jayw_comark
To: TekScopes@...
Sent: Wednesday, November 03, 2010 5:02 PM
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM



Hi Tom

FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the 8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume that the code would fit in the OTP device.

Still wondering whether a set of devices moved from a 8840A would work in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher than it should (1 volt = 100 volts), and resistance measurements are screwed up as well. I suppose there could be something else wrong with the meter, but I haven't taken a closer look thinking that something is wrong with the firmware revisions.

Jay

--- In TekScopes@..., "Tom Miller" <tmiller@> wrote:
>
> Thanks Stefan.
>
> Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
> the image for the 8842A/AF Ver 3.1.
>
> Any others you can find will help.
>
>
> Thanks,
> Tom
>
[snip]


Tomas Larsson
 

You could always post it on ebaman as well

Tomas Larsson
www.ebaman.com

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of jayw_comark
Sent: Saturday, November 06, 2010 11:39 AM
To: TekScopes@...
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM

Hi NikiG,

I posted it on KO4BB, but it is not yet available for download. I also sent
a copy to Tom Miller, and he posted the file on the Yahoo group Fluke_DMM
The file is called 8840A_ROMSET.zip

If you join the group, you should be able to download. Let me know how you
make out.

Jay

--- In TekScopes@..., "Niki G." <trds.co@...> wrote:

Hi Jay, i have a similar problems with my Fluke 8840, my z8 (Z0861108SPC)
romles micro is dead, i have Data I/O, BP micro, ALL-11 programmers, and i
tryed to tead one like 86E21, but without any success, from Ebay i bought
Z8613RS microcontroller with piggyback eeprom, but i haven't binary dump for
programming, please if you can read your, and upload somewhere.

p.s. i think that no way to read Z08611 with stadart programmer, but
86E2112 can be read.

Best regards.
NikiG


----- Original Message -----
From: jayw_comark
To: TekScopes@...
Sent: Wednesday, November 03, 2010 5:02 PM
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM



Hi Tom

FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the
8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume
that the code would fit in the OTP device.

Still wondering whether a set of devices moved from a 8840A would work
in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher
than it should (1 volt = 100 volts), and resistance measurements are screwed
up as well. I suppose there could be something else wrong with the meter,
but I haven't taken a closer look thinking that something is wrong with the
firmware revisions.

Jay

--- In TekScopes@..., "Tom Miller" <tmiller@> wrote:
>
> Thanks Stefan.
>
> Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images
directory. I have and will add
> the image for the 8842A/AF Ver 3.1.
>
> Any others you can find will help.
>
>
> Thanks,
> Tom
>
[snip]



------------------------------------

Yahoo! Groups Links


 

Hi Jay, thanks a lot, i downloaded files, i'll program EPROM's and will give you result soon.
 
 
Best regards.
NikiG

----- Original Message -----
Sent: Saturday, November 06, 2010 12:39 PM
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM

 

Hi NikiG,

I posted it on KO4BB, but it is not yet available for download. I also sent a copy to Tom Miller, and he posted the file on the Yahoo group Fluke_DMM The file is called 8840A_ROMSET.zip

If you join the group, you should be able to download. Let me know how you make out.

Jay

--- In TekScopes@..., "Niki G." wrote:
>
> Hi Jay, i have a similar problems with my Fluke 8840, my z8 (Z0861108SPC) romles micro is dead, i have Data I/O, BP micro, ALL-11 programmers, and i tryed to tead one like 86E21, but without any success, from Ebay i bought Z8613RS microcontroller with piggyback eeprom, but i haven't binary dump for programming, please if you can read your, and upload somewhere.
>
> p.s. i think that no way to read Z08611 with stadart programmer, but 86E2112 can be read.
>
> Best regards.
> NikiG
>
>
> ----- Original Message -----
> From: jayw_comark
> To: TekScopes@...
> Sent: Wednesday, November 03, 2010 5:02 PM
> Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM
>
>
>
> Hi Tom
>
> FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the 8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume that the code would fit in the OTP device.
>
> Still wondering whether a set of devices moved from a 8840A would work in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher than it should (1 volt = 100 volts), and resistance measurements are screwed up as well. I suppose there could be something else wrong with the meter, but I haven't taken a closer look thinking that something is wrong with the firmware revisions.
>
> Jay
>
> --- In TekScopes@..., "Tom Miller" wrote:
> >
> > Thanks Stefan.
> >
> > Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
> > the image for the 8842A/AF Ver 3.1.
> >
> > Any others you can find will help.
> >
> >
> > Thanks,
> > Tom
> >
> [snip]
>


 

HI Jay, i'm sorry for big delay, but finaly i receive my Z8613 mcu and programmed eproms with your dumps V2.3, so all is ok and my 8840A work fine (thanks a lot again), also i tryed to read Z8 (now working mcu) but without success.
 
Besr regards.
NikiG
 

----- Original Message -----
Sent: Saturday, November 06, 2010 12:39 PM
Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM

 

Hi NikiG,

I posted it on KO4BB, but it is not yet available for download. I also sent a copy to Tom Miller, and he posted the file on the Yahoo group Fluke_DMM The file is called 8840A_ROMSET.zip

If you join the group, you should be able to download. Let me know how you make out.

Jay

--- In TekScopes@..., "Niki G." wrote:
>
> Hi Jay, i have a similar problems with my Fluke 8840, my z8 (Z0861108SPC) romles micro is dead, i have Data I/O, BP micro, ALL-11 programmers, and i tryed to tead one like 86E21, but without any success, from Ebay i bought Z8613RS microcontroller with piggyback eeprom, but i haven't binary dump for programming, please if you can read your, and upload somewhere.
>
> p.s. i think that no way to read Z08611 with stadart programmer, but 86E2112 can be read.
>
> Best regards.
> NikiG
>
>
> ----- Original Message -----
> From: jayw_comark
> To: TekScopes@...
> Sent: Wednesday, November 03, 2010 5:02 PM
> Subject: [TekScopes] Re: Rom Images for Fluke 884xA/xx series DMM
>
>
>
> Hi Tom
>
> FWIW, since I have a Z8 with the piggyback 2732 EPROM V2.3 for the 8840A, I'll get the binary and post it to KO4BB tomorrow. I would assume that the code would fit in the OTP device.
>
> Still wondering whether a set of devices moved from a 8840A would work in a 8842A... I suspect not, as when I do this my 8842A reads 100X higher than it should (1 volt = 100 volts), and resistance measurements are screwed up as well. I suppose there could be something else wrong with the meter, but I haven't taken a closer look thinking that something is wrong with the firmware revisions.
>
> Jay
>
> --- In TekScopes@..., "Tom Miller" wrote:
> >
> > Thanks Stefan.
> >
> > Jay put up the 8840A Ver 2.3 on the KO4BB site under the ROM Images directory. I have and will add
> > the image for the 8842A/AF Ver 3.1.
> >
> > Any others you can find will help.
> >
> >
> > Thanks,
> > Tom
> >
> [snip]
>


 

Hi Niki,

Glad to have been of some assistance, and congrats on your repair.

Jay

--- In TekScopes@..., "Niki G." <trds.co@...> wrote:

HI Jay, i'm sorry for big delay, but finaly i receive my Z8613 mcu and programmed eproms with your dumps V2.3, so all is ok and my 8840A work fine (thanks a lot again), also i tryed to read Z8 (now working mcu) but without success.

Besr regards.
NikiG
[snip]