intensity failure
Jerry Massengale <jmassen418@...>
|
|
|
|
Craig Sawyers <c.sawyers@...>
Ah – I had this precise problem while using a sampling set up. A tubular tant in the sampling head when short. What happened was that the scope went to max intensity and the power supply did not shut down.
Check the power lines for shorts before troubleshooting the z-axis.
Craig |
|
|
|
David DiGiacomo
On Tue, Aug 31, 2010 at 1:22 PM, Craig Sawyers
<c.sawyers@...> wrote: Ah – I had this precise problem while using a sampling set up. A tubular tant in the sampling head when short. What happened was that the scope went to max intensity and the power supply did not shut down.And then check the Z-axis DC restore circuit before looking at the logic board. |
|
|
|
Dave Baxter <dave@...>
Re: intensity failuredavid_digiacomo Tue Aug 31, 2010 3:28 pm (PDT)supply did not shut down. logic board.And then check the Z-axis DC restore circuit before looking at the Is it just the intensity flared right up, or did the display (incl' readout) shrink to about half size too? If so, I had that problem on my 7704, ended up replacing the chain of (I think) 30Meg resistors that load the Cathode HV, so the thing can regulate correctly. One had failed open, and as they run hot(!) I changed all of them for higher wattage types, that oddly are smaller than the originals.. Normal service was resumed... Regards. Dave B. |
|
|
|
Jerry Massengale <jmassen418@...>
--- On Wed, 9/1/10, Dave Baxter wrote:
|
|
|