Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs & compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
I can’t tell you about phase noise, but a key feature of rubidium standards is their operating life. The life is limited by depletion of the rubidium in the discharge lamp. Because studies by manufacturers of the performance of the standards they make, they can provide estimated lifetimes based on rubidium depletion. Because of the reliability requirements of frequency and time standards (for example in cell phone tower electronics) they are often swapped out for new oscillators as they get within the range of the expected end-of-life. The large number of rubidium standards that turn up on eBay is at least partly a result of this preventive maintenance program. Most of the used standards likely will have limited life. However, their low cost means you might be able to afford to buy more than one so you can swap in one that works for one that has failed. The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has a design life of 20 years. Just look up the model for a description of the long-term stability and low phase noise. The standard can monitor the lamp start voltage as it rises as the rubidium depletes. These standards sell for more on eBay because of the longer life and likelihood that they will work for you even as used units. Some honest and knowledgable sellers can tell you the lamp voltage. Disclosure: no financial connection with either Ball-Efratom or PRS. I do own standards from both and I bought used ones on eBay and they work (and are working after a couple of years) despite being used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that failed. I also have a couple of Frequency Electronics quartz standards - these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery packs that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter. Anyone interested in time and frequency should have a look at the Time Nuts site: http://leapsecond.com/time-nuts.htmSteveH
toggle quoted messageShow quoted text
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote: Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs & compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
I don't think Rb depletion is quite correct.
What I, and others, have found on units that just ran out of steam, is generally that the RF excited Rb vapor lamp has plated out its Rb metal on the envelope of the lamp, seriously reducing the intensity of the emitted light.
The cure is to re-vaporize the Rb metal off of the lamp's glass bulb.
To do this, take the Rb lamp assembly apart, and you will generally see the lamp's envelope is black in appearance. Heat it with a hot air gun and at a certain temperature, you will see the black coating simply disappear.
The Rb lamp is ready to go another decade or so.
-Chuck Harris
stevenhorii wrote:
toggle quoted messageShow quoted text
I can’t tell you about phase noise, but a key feature of rubidium standards is their operating life. The life is limited by depletion of the rubidium in the discharge lamp. Because studies by manufacturers of the performance of the standards they make, they can provide estimated lifetimes based on rubidium depletion. Because of the reliability requirements of frequency and time standards (for example in cell phone tower electronics) they are often swapped out for new oscillators as they get within the range of the expected end-of-life. The large number of rubidium standards that turn up on eBay is at least partly a result of this preventive maintenance program. Most of the used standards likely will have limited life. However, their low cost means you might be able to afford to buy more than one so you can swap in one that works for one that has failed.
The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has a design life of 20 years. Just look up the model for a description of the long-term stability and low phase noise. The standard can monitor the lamp start voltage as it rises as the rubidium depletes. These standards sell for more on eBay because of the longer life and likelihood that they will work for you even as used units. Some honest and knowledgable sellers can tell you the lamp voltage. Disclosure: no financial connection with either Ball-Efratom or PRS. I do own standards from both and I bought used ones on eBay and they work (and are working after a couple of years) despite being used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that failed. I also have a couple of Frequency Electronics quartz standards - these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery packs that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter.
Anyone interested in time and frequency should have a look at the Time Nuts site:
http://leapsecond.com/time-nuts.htm
SteveH
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote:
Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs & compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
Chuck, I got my information from this article: http://www.wriley.com/A%20History%20of%20the%20Rubidium%20Frequency%20Standard.pdfHowever, I think we are both right - see page 5 of the article. It notes the problem is a tradeoff between too much Rb (noise) and depletion from the Rb diffusing into the glass. There was a guy (Corby Dawson?) who was able to restore Rb frequency standards - maybe by doing what you suggested. Have you tried this? I've got a non-working Efratom standard I should try it on. Thanks for improving the answer and suggesting a fix. Steve H.
toggle quoted messageShow quoted text
On Mon, Jul 13, 2020 at 1:35 PM Chuck Harris <cfharris@erols.com> wrote: I don't think Rb depletion is quite correct.
What I, and others, have found on units that just ran out of steam, is generally that the RF excited Rb vapor lamp has plated out its Rb metal on the envelope of the lamp, seriously reducing the intensity of the emitted light.
The cure is to re-vaporize the Rb metal off of the lamp's glass bulb.
To do this, take the Rb lamp assembly apart, and you will generally see the lamp's envelope is black in appearance. Heat it with a hot air gun and at a certain temperature, you will see the black coating simply disappear.
The Rb lamp is ready to go another decade or so.
-Chuck Harris
stevenhorii wrote:
I can’t tell you about phase noise, but a key feature of rubidium standards
is their operating life. The life is limited by depletion of the rubidium in the discharge lamp. Because studies by manufacturers of the performance
of the standards they make, they can provide estimated lifetimes based on rubidium depletion. Because of the reliability requirements of frequency and time standards (for example in cell phone tower electronics) they are often swapped out for new oscillators as they get within the range of the expected end-of-life. The large number of rubidium standards that turn up on eBay is at least partly a result of this preventive maintenance program.
Most of the used standards likely will have limited life. However, their low cost means you might be able to afford to buy more than one so you can
swap in one that works for one that has failed.
The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has
a design life of 20 years. Just look up the model for a description of the
long-term stability and low phase noise. The standard can monitor the lamp
start voltage as it rises as the rubidium depletes. These standards sell for more on eBay because of the longer life and likelihood that they will work for you even as used units. Some honest and knowledgable sellers can tell you the lamp voltage. Disclosure: no financial connection with either
Ball-Efratom or PRS. I do own standards from both and I bought used ones on
eBay and they work (and are working after a couple of years) despite being
used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that
failed. I also have a couple of Frequency Electronics quartz standards - these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery packs that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter.
Anyone interested in time and frequency should have a look at the Time Nuts
site:
http://leapsecond.com/time-nuts.htm
SteveH
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote:
Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs &
compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
I did one in a HP5065A.
I had a bad oven, and a black Rb lamp, and fixed both. The oven had an internal short between its formvar covered nichrome wire wound heater, and the outer mumetal shield... This put the oven on full heat, and melted all of the solder on the lamp's board... parts fell out of their holes.
But, curiously didn't clean the glass on the bulb. I used a propane torch, as I recall. The glass is quartz, so heating it is not all that risky.
Corby Dawson is where I first got the idea.
Corby has made some optical modifications to the 5065A standards that eeek out another order of magnitude of S/N performance.
-Chuck Harris
stevenhorii wrote:
toggle quoted messageShow quoted text
Chuck,
I got my information from this article:
http://www.wriley.com/A%20History%20of%20the%20Rubidium%20Frequency%20Standard.pdf
However, I think we are both right - see page 5 of the article. It notes the problem is a tradeoff between too much Rb (noise) and depletion from the Rb diffusing into the glass.
There was a guy (Corby Dawson?) who was able to restore Rb frequency standards - maybe by doing what you suggested. Have you tried this? I've got a non-working Efratom standard I should try it on.
Thanks for improving the answer and suggesting a fix.
Steve H.
On Mon, Jul 13, 2020 at 1:35 PM Chuck Harris <cfharris@erols.com> wrote:
I don't think Rb depletion is quite correct.
What I, and others, have found on units that just ran out of steam, is generally that the RF excited Rb vapor lamp has plated out its Rb metal on the envelope of the lamp, seriously reducing the intensity of the emitted light.
The cure is to re-vaporize the Rb metal off of the lamp's glass bulb.
To do this, take the Rb lamp assembly apart, and you will generally see the lamp's envelope is black in appearance. Heat it with a hot air gun and at a certain temperature, you will see the black coating simply disappear.
The Rb lamp is ready to go another decade or so.
-Chuck Harris
stevenhorii wrote:
I can’t tell you about phase noise, but a key feature of rubidium standards
is their operating life. The life is limited by depletion of the rubidium in the discharge lamp. Because studies by manufacturers of the performance
of the standards they make, they can provide estimated lifetimes based on rubidium depletion. Because of the reliability requirements of frequency and time standards (for example in cell phone tower electronics) they are often swapped out for new oscillators as they get within the range of the expected end-of-life. The large number of rubidium standards that turn up on eBay is at least partly a result of this preventive maintenance program.
Most of the used standards likely will have limited life. However, their low cost means you might be able to afford to buy more than one so you can
swap in one that works for one that has failed.
The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has
a design life of 20 years. Just look up the model for a description of the
long-term stability and low phase noise. The standard can monitor the lamp
start voltage as it rises as the rubidium depletes. These standards sell for more on eBay because of the longer life and likelihood that they will work for you even as used units. Some honest and knowledgable sellers can tell you the lamp voltage. Disclosure: no financial connection with either
Ball-Efratom or PRS. I do own standards from both and I bought used ones on
eBay and they work (and are working after a couple of years) despite being
used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that
failed. I also have a couple of Frequency Electronics quartz standards - these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery packs that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter.
Anyone interested in time and frequency should have a look at the Time Nuts
site:
http://leapsecond.com/time-nuts.htm
SteveH
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote:
Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs &
compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
Chuck,
Quite a story! I am surprised that the oven did not have an upper limit temperature cutout.
That's like one of my Tracor/Sulzer quartz standards. Output went to zero on all outputs, but the power lamp was also out.
I surmised it was the power supply (in a separate cylinder). Changing power supplies fixed it.
However, I thought maybe it was just a bad capacitor or transistor in the regulator circuit, so I opened it up. The housing will mostly filled with black gook. It was all over everything - mostly anything dependent in the cylinder - the gook must have flowed "down". When I looked, I was surprised that it was not the capacitor in there (a cylindrical electrolytic) but the transformer. The transformer has in a sealed housing and the header had simply blown out and the electrolytic tar or whatever they used blew out. I eventually determined that the power supply had been running on "fast charge" mode continuously as one of the batteries in the backup pack had failed short. The supply must have sensed low output voltage from the batteries when I would turn off the supply when moving the standards rack and so switched to fast charge mode. It did not blow the fuse and the supply did not have an upper temp limit switch. I don't know if the oil boiled or just expanded enough to rupture the header, but it made a mess. I discarded all the stuff except the housing that was not covered by gook. After that, I took the battery packs out of the power supply cylinder (there were 21 of them in it) and made sure the supply ripple was OK and the output voltage was OK with no battery pack. The supply output was not affected by this. My application was not that critical that I needed battery backup - I was willing to live with having to watch the stability while the oscillator came back after a power down.
Isn't metal deposition on the inside of Nixie tubes the reason why old ones have dark glass?
SteveH
toggle quoted messageShow quoted text
On Mon, Jul 13, 2020 at 3:00 PM Chuck Harris <cfharris@erols.com> wrote: I did one in a HP5065A.
I had a bad oven, and a black Rb lamp, and fixed both. The oven had an internal short between its formvar covered nichrome wire wound heater, and the outer mumetal shield... This put the oven on full heat, and melted all of the solder on the lamp's board... parts fell out of their holes.
But, curiously didn't clean the glass on the bulb. I used a propane torch, as I recall. The glass is quartz, so heating it is not all that risky.
Corby Dawson is where I first got the idea.
Corby has made some optical modifications to the 5065A standards that eeek out another order of magnitude of S/N performance.
-Chuck Harris
stevenhorii wrote:
Chuck,
I got my information from this article:
http://www.wriley.com/A%20History%20of%20the%20Rubidium%20Frequency%20Standard.pdf
However, I think we are both right - see page 5 of the article. It notes the problem is a tradeoff between too much Rb (noise) and depletion from the Rb diffusing into the glass.
There was a guy (Corby Dawson?) who was able to restore Rb frequency standards - maybe by doing what you suggested. Have you tried this? I've got a non-working Efratom standard I should try it on.
Thanks for improving the answer and suggesting a fix.
Steve H.
On Mon, Jul 13, 2020 at 1:35 PM Chuck Harris <cfharris@erols.com> wrote:
I don't think Rb depletion is quite correct.
What I, and others, have found on units that just ran out of steam, is generally that the RF excited Rb vapor lamp has plated out its Rb metal on the envelope of the lamp, seriously reducing the intensity of the emitted light.
The cure is to re-vaporize the Rb metal off of the lamp's glass bulb.
To do this, take the Rb lamp assembly apart, and you will generally see the lamp's envelope is black in appearance. Heat it with a hot air gun and at a certain temperature, you will see the black coating simply disappear.
The Rb lamp is ready to go another decade or so.
-Chuck Harris
stevenhorii wrote:
I can’t tell you about phase noise, but a key feature of rubidium standards
is their operating life. The life is limited by depletion of the rubidium
in the discharge lamp. Because studies by manufacturers of the performance
of the standards they make, they can provide estimated lifetimes based
on
rubidium depletion. Because of the reliability requirements of
frequency
and time standards (for example in cell phone tower electronics) they
are
often swapped out for new oscillators as they get within the range of
the
expected end-of-life. The large number of rubidium standards that turn
up
on eBay is at least partly a result of this preventive maintenance program.
Most of the used standards likely will have limited life. However,
their
low cost means you might be able to afford to buy more than one so you can
swap in one that works for one that has failed.
The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has
a design life of 20 years. Just look up the model for a description of the
long-term stability and low phase noise. The standard can monitor the lamp
start voltage as it rises as the rubidium depletes. These standards
sell
for more on eBay because of the longer life and likelihood that they
will
work for you even as used units. Some honest and knowledgable sellers
can
tell you the lamp voltage. Disclosure: no financial connection with either
Ball-Efratom or PRS. I do own standards from both and I bought used
ones
on
eBay and they work (and are working after a couple of years) despite being
used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that
failed. I also have a couple of Frequency Electronics quartz standards -
these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery
packs
that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter.
Anyone interested in time and frequency should have a look at the Time Nuts
site:
http://leapsecond.com/time-nuts.htm
SteveH
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote:
Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs &
compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
As I recall, the short bypassed the thermal fuse.
HP has made a few bone-headed mistakes over the years... Allowing your Rb/Cs/xtal ovens to incinerate is very often one of them.
The battery inside of the Sulzer 2.5 and 5 references is an accident waiting for a place and time to happen.
I removed mine because it wouldn't hold a charge. The cells were all open circuit from the excess trickle charging.
Blackened bulbs are almost always metal that has been transported from where it belongs to the much cooler bulb envelope.
In the case of neons, and nixies... the same thing really, the black is metal that is ripped from the glowing elements. In the process of plating itself out on the inside of the glass, it will entrain some of the neon, reducing the pressure inside of the tube, and raising the striking voltage.
This is the same principle behind the ion pump.
-Chuck Harris
stevenhorii wrote:
toggle quoted messageShow quoted text
Chuck,
Quite a story! I am surprised that the oven did not have an upper limit temperature cutout.
That's like one of my Tracor/Sulzer quartz standards. Output went to zero on all outputs, but the power lamp was also out.
I surmised it was the power supply (in a separate cylinder). Changing power supplies fixed it.
However, I thought maybe it was just a bad capacitor or transistor in the regulator circuit, so I opened it up. The housing will mostly filled with black gook. It was all over everything - mostly anything dependent in the cylinder - the gook must have flowed "down". When I looked, I was surprised that it was not the capacitor in there (a cylindrical electrolytic) but the transformer. The transformer has in a sealed housing and the header had simply blown out and the electrolytic tar or whatever they used blew out. I eventually determined that the power supply had been running on "fast charge" mode continuously as one of the batteries in the backup pack had failed short. The supply must have sensed low output voltage from the batteries when I would turn off the supply when moving the standards rack and so switched to fast charge mode. It did not blow the fuse and the supply did not have an upper temp limit switch. I don't know if the oil boiled or just expanded enough to rupture the header, but it made a mess. I discarded all the stuff except the housing that was not covered by gook. After that, I took the battery packs out of the power supply cylinder (there were 21 of them in it) and made sure the supply ripple was OK and the output voltage was OK with no battery pack. The supply output was not affected by this. My application was not that critical that I needed battery backup - I was willing to live with having to watch the stability while the oscillator came back after a power down.
Isn't metal deposition on the inside of Nixie tubes the reason why old ones have dark glass?
SteveH
On Mon, Jul 13, 2020 at 3:00 PM Chuck Harris <cfharris@erols.com> wrote:
I did one in a HP5065A.
I had a bad oven, and a black Rb lamp, and fixed both. The oven had an internal short between its formvar covered nichrome wire wound heater, and the outer mumetal shield... This put the oven on full heat, and melted all of the solder on the lamp's board... parts fell out of their holes.
But, curiously didn't clean the glass on the bulb. I used a propane torch, as I recall. The glass is quartz, so heating it is not all that risky.
Corby Dawson is where I first got the idea.
Corby has made some optical modifications to the 5065A standards that eeek out another order of magnitude of S/N performance.
-Chuck Harris
stevenhorii wrote:
Chuck,
I got my information from this article:
http://www.wriley.com/A%20History%20of%20the%20Rubidium%20Frequency%20Standard.pdf
However, I think we are both right - see page 5 of the article. It notes the problem is a tradeoff between too much Rb (noise) and depletion from the Rb diffusing into the glass.
There was a guy (Corby Dawson?) who was able to restore Rb frequency standards - maybe by doing what you suggested. Have you tried this? I've got a non-working Efratom standard I should try it on.
Thanks for improving the answer and suggesting a fix.
Steve H.
On Mon, Jul 13, 2020 at 1:35 PM Chuck Harris <cfharris@erols.com> wrote:
I don't think Rb depletion is quite correct.
What I, and others, have found on units that just ran out of steam, is generally that the RF excited Rb vapor lamp has plated out its Rb metal on the envelope of the lamp, seriously reducing the intensity of the emitted light.
The cure is to re-vaporize the Rb metal off of the lamp's glass bulb.
To do this, take the Rb lamp assembly apart, and you will generally see the lamp's envelope is black in appearance. Heat it with a hot air gun and at a certain temperature, you will see the black coating simply disappear.
The Rb lamp is ready to go another decade or so.
-Chuck Harris
stevenhorii wrote:
I can’t tell you about phase noise, but a key feature of rubidium standards
is their operating life. The life is limited by depletion of the rubidium
in the discharge lamp. Because studies by manufacturers of the performance
of the standards they make, they can provide estimated lifetimes based
on
rubidium depletion. Because of the reliability requirements of
frequency
and time standards (for example in cell phone tower electronics) they
are
often swapped out for new oscillators as they get within the range of
the
expected end-of-life. The large number of rubidium standards that turn
up
on eBay is at least partly a result of this preventive maintenance program.
Most of the used standards likely will have limited life. However,
their
low cost means you might be able to afford to buy more than one so you can
swap in one that works for one that has failed.
The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has
a design life of 20 years. Just look up the model for a description of the
long-term stability and low phase noise. The standard can monitor the lamp
start voltage as it rises as the rubidium depletes. These standards
sell
for more on eBay because of the longer life and likelihood that they
will
work for you even as used units. Some honest and knowledgable sellers
can
tell you the lamp voltage. Disclosure: no financial connection with either
Ball-Efratom or PRS. I do own standards from both and I bought used
ones
on
eBay and they work (and are working after a couple of years) despite being
used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that
failed. I also have a couple of Frequency Electronics quartz standards -
these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery
packs
that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter.
Anyone interested in time and frequency should have a look at the Time Nuts
site:
http://leapsecond.com/time-nuts.htm
SteveH
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote:
Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs &
compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
Yeah, Chuck, I read that somewhere, that it's possible to rejuvenate a Rb lamp with heat. I'm hoping to get a spent Symetricomm (a more recent name of Ball Efratom, I understand) unit from Raytheon (Technologies it is now), my employer. Trouble is, it's buried inside a rack and not easy to get at. We cobbled up another Rb frequency standard of about the same age, so it may bite the dust soon, too. I'd take either or both and heat to drive the rubidium off the glass. I need a frequency standard with long term stability for holdover if my Leo Bodnar GPSDO drops out or something happens to the GPS. No, I'm not a time nut, not yet, anyway.
Jim Ford
toggle quoted messageShow quoted text
------ Original Message ------ From: "Chuck Harris" <cfharris@erols.com> To: TekScopes@groups.io Sent: 7/13/2020 10:35:43 AM Subject: Re: [TekScopes] PTB 100 Ball Efratom I don't think Rb depletion is quite correct.
What I, and others, have found on units that just ran out of steam, is generally that the RF excited Rb vapor lamp has plated out its Rb metal on the envelope of the lamp, seriously reducing the intensity of the emitted light.
The cure is to re-vaporize the Rb metal off of the lamp's glass bulb.
To do this, take the Rb lamp assembly apart, and you will generally see the lamp's envelope is black in appearance. Heat it with a hot air gun and at a certain temperature, you will see the black coating simply disappear.
The Rb lamp is ready to go another decade or so.
-Chuck Harris
stevenhorii wrote:
I can’t tell you about phase noise, but a key feature of rubidium standards is their operating life. The life is limited by depletion of the rubidium in the discharge lamp. Because studies by manufacturers of the performance of the standards they make, they can provide estimated lifetimes based on rubidium depletion. Because of the reliability requirements of frequency and time standards (for example in cell phone tower electronics) they are often swapped out for new oscillators as they get within the range of the expected end-of-life. The large number of rubidium standards that turn up on eBay is at least partly a result of this preventive maintenance program. Most of the used standards likely will have limited life. However, their low cost means you might be able to afford to buy more than one so you can swap in one that works for one that has failed.
The standard I know that was designed with longer life in mind is the Stanford Research Systems (SRS) line, Their PRS-10 model, for example, has a design life of 20 years. Just look up the model for a description of the long-term stability and low phase noise. The standard can monitor the lamp start voltage as it rises as the rubidium depletes. These standards sell for more on eBay because of the longer life and likelihood that they will work for you even as used units. Some honest and knowledgable sellers can tell you the lamp voltage. Disclosure: no financial connection with either Ball-Efratom or PRS. I do own standards from both and I bought used ones on eBay and they work (and are working after a couple of years) despite being used units. On the other hand, I have some HP and Tracor/Sulzer quartz frequency standards that are more than 20 (the HP) and 50 years old, respectively. I’ve had to repair a couple of the Tracor units, but the problem was almost always a transistor in the divider/amplifier stage that failed. I also have a couple of Frequency Electronics quartz standards - these are mil-spec units (URQ-10 and URQ-23) and work fine, but I’ve no idea how long they were in service. They all have built-in battery packs that are built up from “D” size nicad cells (I think - the packs are sealed) and all have failed. They run fine without the batteries - the power supply does not use the battery pack as a filter.
Anyone interested in time and frequency should have a look at the Time Nuts site:
http://leapsecond.com/time-nuts.htm
SteveH
On Mon, Jul 13, 2020 at 11:36 garp66 <hrgerson@gmail.com> wrote:
Does anyone know how the PTB-100 Ball Efratom Rubidium standard performs & compares technically (data & experience), with any of the other available frequency standards ?
(phase noise, etc...)
-- and how it ages ?
Is the PTB-100 still a useful, viable kit ?
thank you, rick
|
|
hi Chuck, Steven,
Where can one find Corby Dawson's detailed procedures and info ? Any .pdf's research or technical papers, or Blogs ?
For both the Rb cell Rejuvenation & the Optical mods to the HP 5065a ?
thank you, rick
|
|
join timenuts, and look in their archives.
-Chuck Harris
garp66 wrote:
toggle quoted messageShow quoted text
hi Chuck, Steven,
Where can one find Corby Dawson's detailed procedures and info ? Any .pdf's research or technical papers, or Blogs ?
For both the Rb cell Rejuvenation & the Optical mods to the HP 5065a ?
thank you, rick
|
|
toggle quoted messageShow quoted text
On Tue, Jul 14, 2020 at 11:08 AM Chuck Harris <cfharris@erols.com> wrote: join timenuts, and look in their archives.
-Chuck Harris
garp66 wrote:
hi Chuck, Steven,
Where can one find Corby Dawson's detailed procedures and info ? Any .pdf's research or technical papers, or Blogs ?
For both the Rb cell Rejuvenation & the Optical mods to the HP 5065a ?
thank you, rick
|
|