AM700 Resurrection
Hi,
I' am try to back to life an AM700. After changing capacitors, washing pcb, check for corroded track the unit boot up and enter diagnostic mode. After solve some issue I dropped in a FP RAM error, I have posted a a screenshot in photo album AM700 folder. Anyone could help me to decode this error? Best Regards Juri
|
|
Siggi
Hey Juri,
I've never seen one of those before, but TekWiki has a full service manual for it (http://w140.com/tekwiki/wiki/AM700). I find this pretty interesting, as this device is from a time when Tek (and others) had pretty much stopped issuing component-level service manuals for their devices. You have a copy, I hope? In the service manual, there's a table with the CPU's memory map, which claims this about the afflicted range: 0E00 0000 – 0EXX XXXX I/O Processor Shared Memory/Registers I took a quick look at the service manual, and my guess is that this'd be referring to the memory that's shared with the front-panel processor. In the Theory of Operation it says "The address bus to the front panel processor shared RAM is buffered by bidirectional data buffers on diagram 7", so maybe those could be at fault? The processor's data bus seems to be pulled high, so if you have a bad output driver, you'd get a stuck-hi bit. Siggi On Tue, Jun 9, 2020 at 3:11 PM Juri Ghisio via groups.io <yurighisio= iol.it@groups.io> wrote: Hi,
|
|
Hi,
After very complex suffered troubleshotting and firmware upgraded to latest version (many thanks to Giampy for his memory dump) finally my AM700 pass all low level diagnostics.... but hangup on splash screen. Powering up instrument in high level diagnostics mode and run it show this error "software abort, acqinit(), couldn' t download 96k DSP code "acqinfc.96k". Swap A7 dsp board with another (from working AM700) error disappear and instrument work fine. Any suggestion here?? best regards, Juri
|
|