I am new to scopes and am looking for a good analog scope for HF work. I like these scopes, but am torn between the three. I like the idea on auto set up, which would leave the 2465 out, but I don't know if the 2465B would be overkill. About me, I tend to over buy on purpose. I'm a believer in buy it once and save money. The notion of "beginner" or "starter" anything doesn't resonate with me - if you can learn on one, you can learn on any. I have read that the 2400 series tend to overheat, or get real hot, when operating. Also, that the power supplies are less than desirable. But then, those are a few opinions that may not be valid. The thing that concerns me would be having to replace older caps that may leak, blah, blah. Are there any thoughts about these, and are there any suggestions about where to look for refurbished units from good repair facilities.


Tom Gardner
 

I have had both 2465s and 2445Bs. I've sold my B and kept the 2465.

The first question I would ask is whether a scope is the right instrument for your HF work. You should understand how the sensitivity and linearity and the input capacitance (=> VSWR) will affect whatever you want to do. Frequently for RF work a 50ohm spectrum analyser is more appropriate.

Secondly, 24x5s don't have an internal battery for the calibration constants, which is an advantage to me. Replacing that battery isn't trivial.

All 24x5x should have their delayed action smoke generators (RIFA filter caps) replaced; I've had one remove part of the PCB prepreg when it erupted. Replacing other PSU caps is desirable.

The three SMD caps on the A5 board should be replaced before they leak.

Finally, your point about the autosetup. To me it is only a minor convenience; learning how to setup a scope is easy and an essential skill. Auto measurement is more convenient, but not essential since manually moving cursors isn't that difficult!

My opinion: if you want a scope, get a working 2465 which has had the caps replaced. YMMV.

On 01/04/20 04:31, Ross Hollinger via groups.io wrote:
I am new to scopes and am looking for a good analog scope for HF work. I like these scopes, but am torn between the three. I like the idea on auto set up, which would leave the 2465 out, but I don't know if the 2465B would be overkill. About me, I tend to over buy on purpose. I'm a believer in buy it once and save money. The notion of "beginner" or "starter" anything doesn't resonate with me - if you can learn on one, you can learn on any. I have read that the 2400 series tend to overheat, or get real hot, when operating. Also, that the power supplies are less than desirable. But then, those are a few opinions that may not be valid. The thing that concerns me would be having to replace older caps that may leak, blah, blah. Are there any thoughts about these, and are there any suggestions about where to look for refurbished units from good repair facilities.


GerryR <totalautomation1@...>
 

Hello Ross,
I have a 2565A that I am thinking of selling. It hasn't been re-capped (no issues there) but has had a new battery recently installed, and I have been through the recalibration procedure. It is in excellent condition. No issues that I know of. I am selling it complete with 3 probes (350 MHz) and with a Tektronix active probe (900 MHz) complete kit, and some extras. The power supply for the probe is a powerblock from Phoenix that I bought and added the proper connector for the probe. If you are interested, let me know; I was about to put it on EBay. I can send pictures and provide further details. I am retired and do not find as much use for it as I once needed.

Regards,
GerryR
KK4GER

----- Original Message -----
From: "Ross Hollinger via groups.io" <rhollinger=protonmail.com@groups.io>
To: <TekScopes@groups.io>
Sent: Tuesday, March 31, 2020 11:31 PM
Subject: [TekScopes] Tektronix 2465, 2465A, or 2465B


I am new to scopes and am looking for a good analog scope for HF work. I like these scopes, but am torn between the three. I like the idea on auto set up, which would leave the 2465 out, but I don't know if the 2465B would be overkill. About me, I tend to over buy on purpose. I'm a believer in buy it once and save money. The notion of "beginner" or "starter" anything doesn't resonate with me - if you can learn on one, you can learn on any. I have read that the 2400 series tend to overheat, or get real hot, when operating. Also, that the power supplies are less than desirable. But then, those are a few opinions that may not be valid. The thing that concerns me would be having to replace older caps that may leak, blah, blah. Are there any thoughts about these, and are there any suggestions about where to look for refurbished units from good repair facilities.


Chuck Harris <cfharris@...>
 

I don't know where you heard that they overheat. They do not.
They also do not get real hot when operating. The only time
they have a problem with heat is when run in the sun in a
desert... Or when kitty decides to fill the scope with leftover
fur. If house cleaning is not a completely foreign concept to
you, you should be fine.

The power supplies are intensely reliable, and very, very hard
to damage. But, they are also old. The capacitors used in the
switching sections of the supply do wear out, just like in any
other power supply.

Each of the scopes you mention was the top of its class at the
time it was offered. And they were not offered concurrently.

So, they are not starter scopes, or beginner scopes, but rather
just scopes. Very high quality, and popular scopes.

The original member of the series, the 2465 was a ground breaker
when it was introduced. It was a replacement for the venerable,
and highly popular 465, only it was microprocessor controlled,
had digital readout built on screen, measurement cursors, was
much more compact... and sported an honest 300MHz bandwidth.

In my experience, the 2465 operates the smoothest of the whole
family, because it had a CPU card that was designed to easily
support the needs of the scope, rather a scope that was designed
to fully utilize all of the capabilities of the CPU card.

The 2465A and B enhanced the operation to eliminate any controls
that were wired directly to the circuit they controlled... putting
the CPU card in between. But, they use exactly the same CPU and
the same clock frequency as the older 2465. This causes control
functions to stutter and balk from time to time as the CPU gets
around to servicing the control function. It also causes controls
that get moved while power is off, to have their rotation range,
become oddly biased... For instance, if you turn the scope off
while intensity is in mid range, and then turn the intensity knob
to the off position, on next power-up you will find that you can only
brighten the intensity from there, not dim it. You have to run
the intensity control its full range to get it to reset.

The Auto setup feature is nothing but marketing fluff. Nobody
needs such a feature after about a day's worth of scope use.

The only time I ever use the AUTO button is when I am testing a
customer's scope for full functionality. It is never necessary in
real life.

The same goes for all the other features in that control group.

I expect that you would be happy with any one of the scopes you
mentioned.

If you buy a "B" model, be warned that there are lots of franken
scopes modified from 2445B models, rebadged as 2465B. If you buy
any scope from someone whose ebay id is the same as that model,
be sure to get his assurances that he is selling you a genuine
scope that came from the tektronix factory as a 2465B. He will
know what you mean. You might want to look elsewhere.

-Chuck Harris

Ross Hollinger via groups.io wrote:

I am new to scopes and am looking for a good analog scope for HF work. I like these scopes, but am torn between the three. I like the idea on auto set up, which would leave the 2465 out, but I don't know if the 2465B would be overkill. About me, I tend to over buy on purpose. I'm a believer in buy it once and save money. The notion of "beginner" or "starter" anything doesn't resonate with me - if you can learn on one, you can learn on any. I have read that the 2400 series tend to overheat, or get real hot, when operating. Also, that the power supplies are less than desirable. But then, those are a few opinions that may not be valid. The thing that concerns me would be having to replace older caps that may leak, blah, blah. Are there any thoughts about these, and are there any suggestions about where to look for refurbished units from good repair facilities.




tekscopegroup@...
 

On Thu, Apr 2, 2020 at 03:52 AM, Tom Gardner wrote:

The three SMD caps on the A5 board should be replaced before they leak.
There are actually a total of 4 caps that need to be replaced. Three are in one group near the NVRAM chip, the other is located towards the right hand side top corner of the board.


victor.silva
 

Chuck,

Excellent! Couldn't have said it better.

--Victor


Siggi
 

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the 2465 is
as near silent as makes no difference - so long as the fan motor is in good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure the
airflow vents underneath the casework aren't blocked. You don't want to set
these down flat on a shelf, with the casework flat against the shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io <rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work. I
like these scopes, but am torn between the three. I like the idea on auto
set up, which would leave the 2465 out, but I don't know if the 2465B would
be overkill. About me, I tend to over buy on purpose. I'm a believer in
buy it once and save money. The notion of "beginner" or "starter" anything
doesn't resonate with me - if you can learn on one, you can learn on any.
I have read that the 2400 series tend to overheat, or get real hot, when
operating. Also, that the power supplies are less than desirable. But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any suggestions
about where to look for refurbished units from good repair facilities.




Stephen Hanselman
 

Look at what all the tech gurus are saying, they’re right. I have used all of them and find the features of our 2465B (for two of us we have four) to be the best.

Chuck is right on target with the franken2465, I have seen some of those come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC

On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the 2465 is
as near silent as makes no difference - so long as the fan motor is in good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure the
airflow vents underneath the casework aren't blocked. You don't want to set
these down flat on a shelf, with the casework flat against the shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io <rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work. I
like these scopes, but am torn between the three. I like the idea on auto
set up, which would leave the 2465 out, but I don't know if the 2465B would
be overkill. About me, I tend to over buy on purpose. I'm a believer in
buy it once and save money. The notion of "beginner" or "starter" anything
doesn't resonate with me - if you can learn on one, you can learn on any.
I have read that the 2400 series tend to overheat, or get real hot, when
operating. Also, that the power supplies are less than desirable. But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any suggestions
about where to look for refurbished units from good repair facilities.





stevenhorii
 

If you are looking for a 2465 of any vintage, I suggest checking out the
seller “2465b” on eBay. I have no financial connection with the seller, but
I have bought items from him. He supplies them fully checked out and
calibrated with a guarantee. The items I bought from him (not a 2465) were
very conservatively described and were in better condition than he
described.

He specializes in the 2465 and knows all the potential trouble points. For
example, if you buy one of the scopes he has worked on, one thing he does
is to put heat sinks on the chips that run hot.

His prices are not the lowest on eBay, but his knowledge and quality of
work are hard to beat. If you want a scope that will work for you likely
for years, look at his listings.

Steve

On Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com> wrote:

Look at what all the tech gurus are saying, they’re right. I have used
all of them and find the features of our 2465B (for two of us we have four)
to be the best.

Chuck is right on target with the franken2465, I have seen some of those
come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC
On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the 2465
is
as near silent as makes no difference - so long as the fan motor is in
good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a
re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only
use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure
the
airflow vents underneath the casework aren't blocked. You don't want to
set
these down flat on a shelf, with the casework flat against the shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io
<rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work.
I
like these scopes, but am torn between the three. I like the idea on
auto
set up, which would leave the 2465 out, but I don't know if the 2465B
would
be overkill. About me, I tend to over buy on purpose. I'm a believer
in
buy it once and save money. The notion of "beginner" or "starter"
anything
doesn't resonate with me - if you can learn on one, you can learn on
any.
I have read that the 2400 series tend to overheat, or get real hot, when
operating. Also, that the power supplies are less than desirable. But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any suggestions
about where to look for refurbished units from good repair facilities.







Jan Weber
 

Hi group,

I am sorry for asking, but curiosity has got the better part of me. How is
it done making a Frankenstein 2465 out of a 2445?

Not that I'm planning to do it, I already have a 2465CT and a 2467B.

Regards

Jan


stevenhorii <sonodocsch@gmail.com> schrieb am Do., 2. Apr. 2020, 19:09:

If you are looking for a 2465 of any vintage, I suggest checking out the
seller “2465b” on eBay. I have no financial connection with the seller, but
I have bought items from him. He supplies them fully checked out and
calibrated with a guarantee. The items I bought from him (not a 2465) were
very conservatively described and were in better condition than he
described.

He specializes in the 2465 and knows all the potential trouble points. For
example, if you buy one of the scopes he has worked on, one thing he does
is to put heat sinks on the chips that run hot.

His prices are not the lowest on eBay, but his knowledge and quality of
work are hard to beat. If you want a scope that will work for you likely
for years, look at his listings.

Steve

On Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com> wrote:

Look at what all the tech gurus are saying, they’re right. I have used
all of them and find the features of our 2465B (for two of us we have
four)
to be the best.

Chuck is right on target with the franken2465, I have seen some of those
come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC
On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There
are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the
2465
is
as near silent as makes no difference - so long as the fan motor is in
good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a
re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only
use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure
the
airflow vents underneath the casework aren't blocked. You don't want to
set
these down flat on a shelf, with the casework flat against the
shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io
<rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work.
I
like these scopes, but am torn between the three. I like the idea on
auto
set up, which would leave the 2465 out, but I don't know if the 2465B
would
be overkill. About me, I tend to over buy on purpose. I'm a believer
in
buy it once and save money. The notion of "beginner" or "starter"
anything
doesn't resonate with me - if you can learn on one, you can learn on
any.
I have read that the 2400 series tend to overheat, or get real hot,
when
operating. Also, that the power supplies are less than desirable.
But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any
suggestions
about where to look for refurbished units from good repair facilities.









Jim Ford
 

Hey, isn't that the guy who fakes 2445's as 2465's?  Just sayin'.Jim Ford Sent from my Verizon, Samsung Galaxy smartphone

-------- Original message --------From: stevenhorii <sonodocsch@gmail.com> Date: 4/2/20 10:08 AM (GMT-08:00) To: TekScopes@groups.io Subject: Re: [TekScopes] Tektronix 2465, 2465A, or 2465B If you are looking for a 2465 of any vintage, I suggest checking out theseller “2465b” on eBay. I have no financial connection with the seller, butI have bought items from him. He supplies them fully checked out andcalibrated with a guarantee. The items I bought from him (not a 2465) werevery conservatively described and were in better condition than hedescribed.He specializes in the 2465 and knows all the potential trouble points. Forexample, if you buy one of the scopes he has worked on, one thing he doesis to put heat sinks on the chips that run hot.His prices are not the lowest on eBay, but his knowledge and quality ofwork are hard to beat. If you want a scope that will work for you likelyfor years, look at his listings.SteveOn Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com> wrote:> Look at what all the tech gurus are saying, they’re right.  I have used> all of them and find the features of our 2465B (for two of us we have four)> to be the best.>> Chuck is right on target with the franken2465, I have seen some of those> come across my bench.>> So be careful, but get a 2465B.>> Regards,>> Stephen Hanselman> Datagate Systems, LLC> > On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:> >> > Hey Ross,> >> > I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There are> > minimal differences in features between the two scopes - see this link:> >> https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b> > .> > One feature difference that isn't mentioned there is the fan in the 2465> is> > as near silent as makes no difference - so long as the fan motor is in> good> > shape.> > The 2465 is obviously also older, and mine is due for a re-cap and a> re-cal> > by now.> >> > My 2467 has the CTT option, which is quite handy, though I really only> use> > it for the counter.> >> > I haven't noticed these scopes getting warm, but you have to make sure> the> > airflow vents underneath the casework aren't blocked. You don't want to> set> > these down flat on a shelf, with the casework flat against the shelving -> > they want to sit on the legs or raised up by the handle.> >> > Siggi> >> >> On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io> <rhollinger=> >> protonmail.com@groups.io> wrote:> >>> >> I am new to scopes and am looking for a good analog scope for HF work.> I> >> like these scopes, but am torn between the three.  I like the idea on> auto> >> set up, which would leave the 2465 out, but I don't know if the 2465B> would> >> be overkill.  About me, I tend to over buy on purpose.  I'm a believer> in> >> buy it once and save money.  The notion of "beginner" or "starter"> anything> >> doesn't resonate with me - if you can learn on one, you can learn on> any.> >> I have read that the 2400 series tend to overheat, or get real hot, when> >> operating.  Also, that the power supplies are less than desirable.  But> >> then, those are a few opinions that may not be valid.  The thing that> >> concerns me would be having to replace older caps that may leak, blah,> >> blah.  Are there any thoughts about these, and are there any suggestions> >> about where to look for refurbished units from good repair facilities.> >>> >>> >>> >>> >> >> >>> >>


johnhstrauch
 

Don't go near that guy. I am one of many who have been screwed over with one of those high dollar rebadged units. I tested the bandwidth and checked the innards and it was NOT a 2465B. Just a bit over 300MHz bandwidth. The A5 board failed within a couple of months.
When I confronted the good seller he was totally unremorseful. Said the unit was past his gauranty limit and his fee was $600 to fix it.

That was several years ago (early 2016) and it is still a bitter pill
for me.

Be aware of unscrupleous people and check the archives here for the plentiful prior messages concerning him.

JH

On 4/2/2020 12:08 PM, stevenhorii wrote:
If you are looking for a 2465 of any vintage, I suggest checking out the
seller “2465b” on eBay. I have no financial connection with the seller, but
I have bought items from him. He supplies them fully checked out and
calibrated with a guarantee. The items I bought from him (not a 2465) were
very conservatively described and were in better condition than he
described.
He specializes in the 2465 and knows all the potential trouble points. For
example, if you buy one of the scopes he has worked on, one thing he does
is to put heat sinks on the chips that run hot.
His prices are not the lowest on eBay, but his knowledge and quality of
work are hard to beat. If you want a scope that will work for you likely
for years, look at his listings.
Steve
On Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com> wrote:

Look at what all the tech gurus are saying, they’re right. I have used
all of them and find the features of our 2465B (for two of us we have four)
to be the best.

Chuck is right on target with the franken2465, I have seen some of those
come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC
On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the 2465
is
as near silent as makes no difference - so long as the fan motor is in
good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a
re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only
use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure
the
airflow vents underneath the casework aren't blocked. You don't want to
set
these down flat on a shelf, with the casework flat against the shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io
<rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work.
I
like these scopes, but am torn between the three. I like the idea on
auto
set up, which would leave the 2465 out, but I don't know if the 2465B
would
be overkill. About me, I tend to over buy on purpose. I'm a believer
in
buy it once and save money. The notion of "beginner" or "starter"
anything
doesn't resonate with me - if you can learn on one, you can learn on
any.
I have read that the 2400 series tend to overheat, or get real hot, when
operating. Also, that the power supplies are less than desirable. But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any suggestions
about where to look for refurbished units from good repair facilities.







Bill E
 

I personally like the A. I've used scopes of various flavors for 50 or so years (my first was an EICO kit I got for my 12th or so bday, 1 whole Mhz). I have a 2465 and a 2465A. Wouldn't mind trying a B, but there was a run of them that had some of the really flaky caps that were around, those need recapping for sure. But, any of these scopes are some of the last, great analog scopes. As for auto-setup, sure, anyone with some experience can set up a scope, I did that for decades. But, it is really convenient to just push a button if you're probing around a circuit with different voltages and currents, and you don't know exactly what the values are. Fluff? No. A crutch? Maybe. Useful? For sure. Both the A and B scopes add some useful features, mostly calculations related to the cursors.

Not to cause a flame-war, and yes, it's heresy, but I have to put a plug in for my HP 54542A scope, only HP scope I've ever used. But, for a digital scope, 4 chans, 2 gs/sec simultaneously on all 4, 500 Mhz bw, fantastic triggering,and actually usable out to 800+Mhz.. Sorry, I digress.

BTW, don't be afraid of the battery-backed calibration rams that have a separate battery, it's easy to replace the battery without losing cal data, you just have to do a little prep.OTOH, those awful Dallas ram-and-battery packages were one of the worst inventions of all time. But, even if you lose cal, you can actually recalibrate with some fairly common equipment and some improvisation, unlike some of the newer instruments that need an external PC running some proprietary and now non-existent special cal software. One reason I really dislike the later TEK scopes with embedded Windows. Facing that issue now with an HP 4195A spectrum analyzer. Not only does it need a PC and special sw, it expects specific equipment it diddles over the HP-IB chain. Essentially, boat anchor if you have to recal it.


satbeginner
 

I think he is also the person who sells or sold newly made 2465B badges for the handles...

Only, they show up not just on real 2465B's, but also on rebadged "upgraded" other 24xx models....


Chuck Harris <cfharris@...>
 

You might want to check the archives for this group to see all
of the kurfuffle that "2465b"'s "frankenscopes" (2445B scopes with
2465B labels) have caused.

-Chuck Harris

stevenhorii wrote:

If you are looking for a 2465 of any vintage, I suggest checking out the
seller “2465b” on eBay. I have no financial connection with the seller, but
I have bought items from him. He supplies them fully checked out and
calibrated with a guarantee. The items I bought from him (not a 2465) were
very conservatively described and were in better condition than he
described.

He specializes in the 2465 and knows all the potential trouble points. For
example, if you buy one of the scopes he has worked on, one thing he does
is to put heat sinks on the chips that run hot.

His prices are not the lowest on eBay, but his knowledge and quality of
work are hard to beat. If you want a scope that will work for you likely
for years, look at his listings.

Steve

On Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com> wrote:

Look at what all the tech gurus are saying, they’re right. I have used
all of them and find the features of our 2465B (for two of us we have four)
to be the best.

Chuck is right on target with the franken2465, I have seen some of those
come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC
On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the 2465
is
as near silent as makes no difference - so long as the fan motor is in
good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a
re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only
use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure
the
airflow vents underneath the casework aren't blocked. You don't want to
set
these down flat on a shelf, with the casework flat against the shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io
<rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work.
I
like these scopes, but am torn between the three. I like the idea on
auto
set up, which would leave the 2465 out, but I don't know if the 2465B
would
be overkill. About me, I tend to over buy on purpose. I'm a believer
in
buy it once and save money. The notion of "beginner" or "starter"
anything
doesn't resonate with me - if you can learn on one, you can learn on
any.
I have read that the 2400 series tend to overheat, or get real hot, when
operating. Also, that the power supplies are less than desirable. But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any suggestions
about where to look for refurbished units from good repair facilities.









 

Hi Ross,

From a somewhat different angle than the other comments, if you like to know what is inside your beloved instrument, the 2465s come with the excellent manuals in the Tektronix tradition. I downloaded the service manual for 2465A from http://bama.edebris.com/manuals/tek/ and browsed it with interest. Their "Theory of Operation" section and their circuit diagrams are informative and detailed. I learned here what a "Micro Channel Plate CRT" is (in the 2467). The circuit diagram of the high voltage supply is understandable, and even a "High Voltage Module" shows what is inside (!). In my judgment, this adds value to the instruments.

Ernesto


Chuck Harris <cfharris@...>
 

Briefly, because I have discussed this before:

1) Replace the labels that say 2445B with labels that say 2465B.
2) Add a jumper to the A1 motherboard connector that goes to the
A5 controller. This jumper misrepresents the scope type as a
2465B to the controller. The controller, thinking it has
a 2465B motherboard enables the 5ns/div step on the timing switch.
3) Isolate the two spiral inductor networks, by cutting traces, and
solder in a pair of jumper wires to bridge the signal path.
4) Do your level best to tweak the response as a 2465B, even though
you are missing 3 trimmer pots, 1 inductor, and a spiral delay
line matching network; have 3 wrong valued trimmer capacitors and
several wrong valued resistors; and your scope uses graded
vertical preamp hybrids that were culled out after production
because they could not meet the 2465B performance requirements.

Seems like a poor thing to do to your 2445B, but what do I know, I'm
just a dumb engineer.

-Chuck Harris

Jan Weber via groups.io wrote:

Hi group,

I am sorry for asking, but curiosity has got the better part of me. How is
it done making a Frankenstein 2465 out of a 2445?

Not that I'm planning to do it, I already have a 2465CT and a 2467B.

Regards

Jan


stevenhorii <sonodocsch@gmail.com> schrieb am Do., 2. Apr. 2020, 19:09:

If you are looking for a 2465 of any vintage, I suggest checking out the
seller “2465b” on eBay. I have no financial connection with the seller, but
I have bought items from him. He supplies them fully checked out and
calibrated with a guarantee. The items I bought from him (not a 2465) were
very conservatively described and were in better condition than he
described.

He specializes in the 2465 and knows all the potential trouble points. For
example, if you buy one of the scopes he has worked on, one thing he does
is to put heat sinks on the chips that run hot.

His prices are not the lowest on eBay, but his knowledge and quality of
work are hard to beat. If you want a scope that will work for you likely
for years, look at his listings.

Steve

On Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com> wrote:

Look at what all the tech gurus are saying, they’re right. I have used
all of them and find the features of our 2465B (for two of us we have
four)
to be the best.

Chuck is right on target with the franken2465, I have seen some of those
come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC
On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There
are
minimal differences in features between the two scopes - see this link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the
2465
is
as near silent as makes no difference - so long as the fan motor is in
good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a
re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only
use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure
the
airflow vents underneath the casework aren't blocked. You don't want to
set
these down flat on a shelf, with the casework flat against the
shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io
<rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF work.
I
like these scopes, but am torn between the three. I like the idea on
auto
set up, which would leave the 2465 out, but I don't know if the 2465B
would
be overkill. About me, I tend to over buy on purpose. I'm a believer
in
buy it once and save money. The notion of "beginner" or "starter"
anything
doesn't resonate with me - if you can learn on one, you can learn on
any.
I have read that the 2400 series tend to overheat, or get real hot,
when
operating. Also, that the power supplies are less than desirable.
But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak, blah,
blah. Are there any thoughts about these, and are there any
suggestions
about where to look for refurbished units from good repair facilities.











Tony Fleming
 

Ernesto, you are a wealth of information!!! Thank you very much again.

On Thu, Apr 2, 2020 at 1:07 PM Ernesto <ebordon@swbell.net> wrote:

Hi Ross,

From a somewhat different angle than the other comments, if you like to
know what is inside your beloved instrument, the 2465s come with the
excellent manuals in the Tektronix tradition. I downloaded the service
manual for 2465A from http://bama.edebris.com/manuals/tek/ and browsed it
with interest. Their "Theory of Operation" section and their circuit
diagrams are informative and detailed. I learned here what a "Micro
Channel Plate CRT" is (in the 2467). The circuit diagram of the high
voltage supply is understandable, and even a "High Voltage Module" shows
what is inside (!). In my judgment, this adds value to the instruments.

Ernesto




Tony Fleming
 

Chuck, my Tektronix is 2465 DMM - is this the 2465A version?
Can I do what you recommend in step # 2 and 3?
Thank you very much for your help.

On Thu, Apr 2, 2020 at 1:21 PM Chuck Harris <cfharris@erols.com> wrote:

Briefly, because I have discussed this before:

1) Replace the labels that say 2445B with labels that say 2465B.
2) Add a jumper to the A1 motherboard connector that goes to the
A5 controller. This jumper misrepresents the scope type as a
2465B to the controller. The controller, thinking it has
a 2465B motherboard enables the 5ns/div step on the timing switch.
3) Isolate the two spiral inductor networks, by cutting traces, and
solder in a pair of jumper wires to bridge the signal path.
4) Do your level best to tweak the response as a 2465B, even though
you are missing 3 trimmer pots, 1 inductor, and a spiral delay
line matching network; have 3 wrong valued trimmer capacitors and
several wrong valued resistors; and your scope uses graded
vertical preamp hybrids that were culled out after production
because they could not meet the 2465B performance requirements.

Seems like a poor thing to do to your 2445B, but what do I know, I'm
just a dumb engineer.

-Chuck Harris

Jan Weber via groups.io wrote:
Hi group,

I am sorry for asking, but curiosity has got the better part of me. How
is
it done making a Frankenstein 2465 out of a 2445?

Not that I'm planning to do it, I already have a 2465CT and a 2467B.

Regards

Jan


stevenhorii <sonodocsch@gmail.com> schrieb am Do., 2. Apr. 2020, 19:09:

If you are looking for a 2465 of any vintage, I suggest checking out the
seller “2465b” on eBay. I have no financial connection with the seller,
but
I have bought items from him. He supplies them fully checked out and
calibrated with a guarantee. The items I bought from him (not a 2465)
were
very conservatively described and were in better condition than he
described.

He specializes in the 2465 and knows all the potential trouble points.
For
example, if you buy one of the scopes he has worked on, one thing he
does
is to put heat sinks on the chips that run hot.

His prices are not the lowest on eBay, but his knowledge and quality of
work are hard to beat. If you want a scope that will work for you likely
for years, look at his listings.

Steve

On Thu, Apr 2, 2020 at 12:27 Stephen Hanselman <kc4sw.io@kc4sw.com>
wrote:

Look at what all the tech gurus are saying, they’re right. I have used
all of them and find the features of our 2465B (for two of us we have
four)
to be the best.

Chuck is right on target with the franken2465, I have seen some of
those
come across my bench.

So be careful, but get a 2465B.

Regards,

Stephen Hanselman
Datagate Systems, LLC
On Apr 2, 2020, at 07:39, Siggi <siggi@undo.com> wrote:

Hey Ross,

I have both a 2467 (350MHz, contemporary of 2465A) and a 2465. There
are
minimal differences in features between the two scopes - see this
link:
https://www.tek.com/support/faqs/what-are-differences-between-oscilloscope-models-2465-2465a-and-2465b
.
One feature difference that isn't mentioned there is the fan in the
2465
is
as near silent as makes no difference - so long as the fan motor is in
good
shape.
The 2465 is obviously also older, and mine is due for a re-cap and a
re-cal
by now.

My 2467 has the CTT option, which is quite handy, though I really only
use
it for the counter.

I haven't noticed these scopes getting warm, but you have to make sure
the
airflow vents underneath the casework aren't blocked. You don't want
to
set
these down flat on a shelf, with the casework flat against the
shelving -
they want to sit on the legs or raised up by the handle.

Siggi

On Thu, Apr 2, 2020 at 6:28 AM Ross Hollinger via groups.io
<rhollinger=
protonmail.com@groups.io> wrote:

I am new to scopes and am looking for a good analog scope for HF
work.
I
like these scopes, but am torn between the three. I like the idea on
auto
set up, which would leave the 2465 out, but I don't know if the 2465B
would
be overkill. About me, I tend to over buy on purpose. I'm a
believer
in
buy it once and save money. The notion of "beginner" or "starter"
anything
doesn't resonate with me - if you can learn on one, you can learn on
any.
I have read that the 2400 series tend to overheat, or get real hot,
when
operating. Also, that the power supplies are less than desirable.
But
then, those are a few opinions that may not be valid. The thing that
concerns me would be having to replace older caps that may leak,
blah,
blah. Are there any thoughts about these, and are there any
suggestions
about where to look for refurbished units from good repair
facilities.













Bill E
 

Wow, that seems like a lot of work to sell a fake, especially when you can pick up a 2465 for $250-350. But, the low priced ones need some careful due diligence. Look at the photos. If the seller doesn't have any that show an actual display, could be dead. I bought another 2465 as a parts unit, but turned out it actually worked fine after a few simple voltage checks and replacing a couple of caps. OTOH, could end up being exactly that, a parts unit.