Date
1 - 20 of 21
11801C and the T1331 error: lost factory TB calibration data.
freemail8877
Hi, first post to this forum. I have very much enjoyed browsing, and thought I could contribute with this information. Thanks guys for keeping this forum so interesting. The T1331 (loss of battery backed up information in the Time Base/Controller board NVRAMs) seems to be an error to avoid, by backing up/replacing the NVRAMs on that board. See e.g. the topic "Saving A5 Timing Board NVRAM calibration constants CSA803 (A/C) and 11801(B/C)" for how to. To answer some question raised there, here is some information from the Tektronix documentation.
In the 11801C User Manual (070-9971-01), on p. 3-112 (p.262/440 of the pdf), I found this: "Initializing and Erasing Nonvolatile RAM: ... The following information is not lost: ... - Factory calibration constants, which are established at the factory and cannot be changed by the user or by the 11801C - Enhanced Accuracy calibration constants, which are periodically calculated and updated ..." And in the 11801C Service Manual (070-9972-02) on p. 6-108 (p.244/286) it is stated: "Time Base Calibration Errors: ... Calibration data to enhance the time interpolator linearity is stored in NVRAM on the Time Base/Controller board. If this NVRAM fails, a T1331 error occurs during power-up diagnostics. The values are written at the factory with the GPIB command CALCORRECTION. CALCORRECTION<ui>:N where <ui> = 0..32 and N = -128..+127 CAUTION: The calibration enhancement values are set at time of manufacture. There is no query form of this command. Do NOT reset these values. If a T1331 failure occurs during extended power-up diagnostics or you suspect a time interpolator error, contact Tekronix factory service. A T1331 failure causes all N values in the 33 ui locations to be reset to zero. The oscilloscope will still run, but with decreased accuracy in the time interpolator linearity. Resetting NVRAM with the Teksecure feature does not affect the Time Base/Controller board NVRAM." So now the big question seems to be: Does anybody in the group know, or know who might know, how the factory calibration procedure is, how the 33 N-values are determined? Does anybody know how much the accuracy is affected? Monica A. PS: I have for many years been a Tektronix scopes addict, It started with the fascination in late 60's and early 70's, when I saw and used a 535 with Type CA plugin at my father's workplace. Being interested in electronics, this was heaven. I now have a 7904a with sampling and non-sampling plugins, a 7704a which rudely enough let the smoke out of the power supply, a newly acquired 7603, and a 468. The design and layout of the fronts are just so great. An 11801C is on the way in, so I have used search a lot lately. It is said to pass POST, and a picture shows the graticule on the screen, so hopefully it's OK. |
|
Reginald Beardsley
very interesting information. Thanks.
The 11801 series has what has to be the worst FW anyone has ever written for a scope as you will shortly experience first hand. But they really are *fantastic* despite that. Just watch out for the bugs. If I execute the wrong sequence of operations I get a 1-2 order of magnitude error in the times on my 11801. IIRC they will interpolate to 200 femtoseconds. Even at coax velocity of 2e8 m/s, that's so small a distance as to be rather "challenging". As in measuring 40 microns "challenging". Without a detailed theory of operation, schematics, etc, it is nearly impossible to divine what Tek was doing. I'd love to know. The circuitry is so fast, that even 30 years later, Keysight can just barely beat the rise time of the 11801 calibrator. I've got 2x SD-22, 4x SD-26 and a nice crisp 11801, with another 11801 and SD-26 for parts in CA and an SD-24 in transit. Honest, all I need now is just *one* SD-32 ;-) Prepare to be amazed. With an SD-24 or SD-26 & the calibrator, you can test RF connectors to 20 GHz as fast as you can connect and disconnect them. Have Fun! Reg |
|
Jim Ford
Hi, Reg.I am so envious of the 11000 series toys, er I mean tools, you have! For now, I will have to be content with my 7000 series sampling gear including two S-4 heads. Not that easy to even get a stable trace on the screen. But, at a few hundred dollars, the price is right these days. I agree with you that this is a golden age for test equipment acquisition. May it continue for posterity!Enjoy!Jim Ford Sent from my Verizon, Samsung Galaxy smartphone
toggle quoted message
Show quoted text
-------- Original message --------From: "Reginald Beardsley via Groups.Io" <pulaskite@...> Date: 8/17/19 5:40 PM (GMT-08:00) To: TekScopes@groups.io Subject: Re: [TekScopes] 11801C and the T1331 error: lost factory TB calibration data. very interesting information. Thanks.The 11801 series has what has to be the worst FW anyone has ever written for a scope as you will shortly experience first hand. But they really are *fantastic* despite that. Just watch out for the bugs. If I execute the wrong sequence of operations I get a 1-2 order of magnitude error in the times on my 11801.IIRC they will interpolate to 200 femtoseconds. Even at coax velocity of 2e8 m/s, that's so small a distance as to be rather "challenging". As in measuring 40 microns "challenging".Without a detailed theory of operation, schematics, etc, it is nearly impossible to divine what Tek was doing. I'd love to know. The circuitry is so fast, that even 30 years later, Keysight can just barely beat the rise time of the 11801 calibrator.I've got 2x SD-22, 4x SD-26 and a nice crisp 11801, with another 11801 and SD-26 for parts in CA and an SD-24 in transit. Honest, all I need now is just *one* SD-32 ;-)Prepare to be amazed. With an SD-24 or SD-26 & the calibrator, you can test RF connectors to 20 GHz as fast as you can connect and disconnect them.Have Fun!Reg
|
|
Albert Otten
[This topic continued from https://groups.io/g/TekScopes/message/158778 ]
toggle quoted message
Show quoted text
I know the discussion was about the later CSA803(A/B/C) and 11801(B/C). I have an 803 and an 803A. The U500/U511 batteries in my 803 were dead, so factory calibration constants would have been lost and I could freely experiment with the CCALCORRECTION command. To my surprise this command was unknown to the 803. This is also confirmed by the commands lists in U810/U910 EPROM images. The CSA803 Service Reference is now available at Tekwiki (thanks Håkan!). There is no reference at all in that manual to factory set calibration constants. The corresponding texts in the C version was obviously added later on. So I am convinced that in the CSA803 nothing is lost when an A5 time base board Dallas chip dies. [[BTW in the pdf manuals provided by Tek you won't find words in the added text with an ordinary search. Likely Tek used a custom character coding in the added text. If you copy-paste for instance T1331 from the added text into the search field then you will find the other instances of T1331 in the added text.] Albert On Thu, Aug 15, 2019 at 04:21 PM, freemail8877 wrote:
|
|
Everyone:
toggle quoted message
Show quoted text
The following subject threads with their email responses are "very" or "some what" relevant to this discussion of fixing lost cal data in the 11K series. These are the TekScopes archive email numbers and the dates they were posted: VERY RELEVANT Programming a new NVRAM (Dallas DS1220Y) with a EPROM programmer By Dennis Tillman W7PF · #119113 · 10/01/15 By Dennis Tillman W7PF · #119111 · 10/01/15 Repairing an expired NVRAM batteryin11A32,11A33,11A34, 11A72 plugins By Vince Vielhaber · #118126 · 08/19/15 By Dennis Tillman W7PF · #118121 · 08/19/15 By Dennis Tillman W7PF · #118115 · 08/19/15 By Dennis Tillman W7PF · #118063 · 08/16/15 Addendum: Repairing an expired NVRAM battery in 11A32, 11A33, 11A34, 11A72 plugins By Dennis Tillman W7PF · #118098 · 08/17/15 11801 Sampling Head By Dennis Tillman W7PF · #119056 · 09/29/15 · By Dennis Tillman W7PF · #119043 · 09/28/15 · SOME-WHAT RELEVANT Tektronix 11000 series vs. modern scopes. Sampling rate and aliasing questions. By ditter2 · #96382 · 07/26/13 11A33 - what are possible problems, gotchas, and revisions? By cheater cheater · #96393 · 07/27/13 By fjh001 · #96406 · 07/27/13 I hope this helps. Often, although it may not be the exact answer to your specific problem, it will guide you to a solution. If that is the case please share it with the forum so we can add it to our "bag of tricks". Dennis Tillman W7PF -----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Albert Otten Sent: Friday, August 23, 2019 9:22 AM To: TekScopes@groups.io Subject: Re: [TekScopes] 11801C and the T1331 error: lost factory TB calibration data. [This topic continued from https://groups.io/g/TekScopes/message/158778 ] I know the discussion was about the later CSA803(A/B/C) and 11801(B/C). I have an 803 and an 803A. The U500/U511 batteries in my 803 were dead, so factory calibration constants would have been lost and I could freely experiment with the CCALCORRECTION command. To my surprise this command was unknown to the 803. This is also confirmed by the commands lists in U810/U910 EPROM images. The CSA803 Service Reference is now available at Tekwiki (thanks Håkan!). There is no reference at all in that manual to factory set calibration constants. The corresponding texts in the C version was obviously added later on. So I am convinced that in the CSA803 nothing is lost when an A5 time base board Dallas chip dies. [[BTW in the pdf manuals provided by Tek you won't find words in the added text with an ordinary search. Likely Tek used a custom character coding in the added text. If you copy-paste for instance T1331 from the added text into the search field then you will find the other instances of T1331 in the added text.] Albert On Thu, Aug 15, 2019 at 04:21 PM, freemail8877 wrote:
-- Dennis Tillman W7PF TekScopes Moderator |
|
For some unexpected reason the formatting of my previous post was lost so here it is again in what I hope will be a more readable version:
toggle quoted message
Show quoted text
Everyone: The following subject threads with their email responses are "very" or "some what" relevant to this discussion of fixing lost cal data in the 11K series. These are the TekScopes archive email numbers and the dates they were posted: VERY RELEVANT Programming a new NVRAM (Dallas DS1220Y) with a EPROM programmer By Dennis Tillman W7PF · #119113 · 10/01/15 By Dennis Tillman W7PF · #119111 · 10/01/15 Repairing an expired NVRAM battery in 11A32, 11A33, 11A34, 11A72 plugins By Vince Vielhaber · #118126 · 08/19/15 By Dennis Tillman W7PF · #118121 · 08/19/15 By Dennis Tillman W7PF · #118115 · 08/19/15 By Dennis Tillman W7PF · #118063 · 08/16/15 Addendum: Repairing an expired NVRAM battery in 11A32, 11A33, 11A34, 11A72 plugins By Dennis Tillman W7PF · #118098 · 08/17/15 SOME-WHAT RELEVANT 11801 Sampling Head By Dennis Tillman W7PF · #119056 · 09/29/15 · By Dennis Tillman W7PF · #119043 · 09/28/15 · Tektronix 11000 series vs. modern scopes. Sampling rate and aliasing questions. By ditter2 · #96382 · 07/26/13 11A33 - what are possible problems, gotchas, and revisions? By cheater cheater · #96393 · 07/27/13 By fjh001 · #96406 · 07/27/13 I hope this helps. Often, although it may not be the exact answer to your specific problem, it will guide you to a solution. If that is the case please share it with the forum so we can add it to our "bag of tricks". Dennis Tillman W7PF -----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Albert Otten Sent: Friday, August 23, 2019 9:22 AM To: TekScopes@groups.io Subject: Re: [TekScopes] 11801C and the T1331 error: lost factory TB calibration data. [This topic continued from https://groups.io/g/TekScopes/message/158778 ] I know the discussion was about the later CSA803(A/B/C) and 11801(B/C). I have an 803 and an 803A. The U500/U511 batteries in my 803 were dead, so factory calibration constants would have been lost and I could freely experiment with the CCALCORRECTION command. To my surprise this command was unknown to the 803. This is also confirmed by the commands lists in U810/U910 EPROM images. The CSA803 Service Reference is now available at Tekwiki (thanks Håkan!). There is no reference at all in that manual to factory set calibration constants. The corresponding texts in the C version was obviously added later on. So I am convinced that in the CSA803 nothing is lost when an A5 time base board Dallas chip dies. [[BTW in the pdf manuals provided by Tek you won't find words in the added text with an ordinary search. Likely Tek used a custom character coding in the added text. If you copy-paste for instance T1331 from the added text into the search field then you will find the other instances of T1331 in the added text.] Albert On Thu, Aug 15, 2019 at 04:21 PM, freemail8877 wrote:
-- Dennis Tillman W7PF TekScopes Moderator -- Dennis Tillman W7PF TekScopes Moderator |
|
Reginald Beardsley
I'm beginning to think I need to make a femtosecond adjustable trombone line. Though the horizontal display jitter is more important as it says failing part in the video display.
BTW If you'll forgive a gloat, I got an SD-24 yesterday via ebay for $250 + shipping. Return address on the label said "Tektronix, Inc" :-) It is *so* sweet. Now if I can just get an SD-32 for $500. |
|
John Griessen
On 8/23/19 3:37 PM, Reginald Beardsley via Groups.Io wrote:
I got an SD-24 yesterday via ebay for $250 + shipping. Return address on the label said "Tektronix, Inc":-) It is*so* sweet.So it must have been Tek ebay-selling old stock that had been lurking. New old stock! |
|
Bob Koller <testtech@...>
Good deal on the SD-24!
CHIPXS on ebay, is the Tektronix "Country Store", as it used to be called, RAMS now. They have a physical location, and are open to the public on the 1st and 3rd Thursdays of the month. I visit almost every time they are open. The items listed on eBay are not available in the store, but there are tons of other stuff! |
|
Reginald Beardsley
I plan to browse chipxs regularly. This is hardly an NOS SD-24. Last cal was 2011 due next in 2013. Measured rise times are ~14 ps with uncertainties of 4 ps. I am extremely pleased with it.
I did some calculations this morning. With an adjustable shorted stub on a 0.0001" reading micrometer head, the reflection would shift ~25 femtoseconds for each 0.0001" movement. At 5 ps/div and 5120 samples we've got samples at ~10 femtosecond intervals. So measuring timebase nonlinearity is rather challenging. I don't think any piece of gear I have has been as just plain amazing. The idea that it is 30 year old technology blows me away. I just wish I had better service data for it. I'm quite alarmed that after being idle for a few months I now have display jitter even after warming up for several hours. I'd love to know who designed it and whether they are still alive. |
|
Bob Koller <testtech@...>
Reg,
The display jitter; Is this timing/trigger related, or a problem with the display, ie. CRT circuitry proper? |
|
Leo Bodnar
On Sat, Aug 24, 2019 at 05:06 PM, Reginald Beardsley wrote:
I'd love to know who designed it and whether they are still alive.I think most heads were designed by Agoston Agoston. Other than that: ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ This product was made possible by the dedicated efforts of a group of inspired, tireless, and often times ingenious individuals of the design team. The following is a list of a few of the cast of many that contributed to the development of this oscilloscope. Design Team: Mark Anderson Brian Colony John DeLacy Rob DeVoe Jim Edwards Larry Hattery Tim Holte Roy Kaufman Jim Lamb Les Larson Chan Lee Jim Long Ben McCarroll Clark Morgan Jim Peterson Jim Schlegel Bob Simpson Ken Smith David White Wayne Wilburn Contributing Engineers: Ray Blohm Thomas Dye Scott Halsted Ron Henricksen Steve Herring Bev Kramlich Mehrab Sedeh Jim Stanley Gil Stephens Engineering Support: Bob Chandler Hilde Cochran John Eskeldson Pam Fodrea Lorna Goedel Dick Griffin John Hazard Diane Kemper Beverly McClenathan Jerry Sternes Wally Sutton Pat Wiley Manuals: Brian Diehm Amy Farrell Susan Grace Julie Leonetti Krystn McCaleb Chuck Melikian Roland Parenteau David Powe Mark Stade Carolyn Strong Manufacturing: Tim Bennington-Davis Donna Brown Jerry Brucken Kevin Cosgrove Lew Cummings Jeanne Eick Jerry Feickert Berdine Garner Clark Jarvis Paulette Jesse Murlan Kaufman Phil Lloyd Woody Ngincharoen Tina Noll Howard Nutt Carol Parks Al Phillips Steve Ratner Doug Rowe Bill Schell Colleen Swanson Mark Swenson Ron Tegner Rod Van Loon Walt Ventgen Marketing/Service: Denny Chamberlin Roger Ensrud Tom Freeman Theresa Graf Ivan Jackson Dennis Kucera Paul Kristof Ray Blohm John Boatwright Betty Bonham Bob Bousquet Jim Carter Dennis Chamberlin Jo Chi John Cooper Jerry Coulter Connie DeClerck Laszlo Dobos Tom Dye Brad Figg Dean Gehnert Ted Gerlinger Larry Hershiser Ken Holland Will Hott Shirley Humphreys David Irwin Stan Kaveckis Ray Kazlauskas Hedy Leidelmeyer Roy Lewallen Ken Longgrear Tina Newkirk Reba Norris Oris Nussbaum Mike O'Shanecy Janet Peters Barbara Ports V. Prasannan Mary Rehse John Rettig Glen Rollins Ava Stupek Dan Taylor Erik Teose Mark Tilden Jan Todd Harold Vandecoevering Mona Webber Vaughn Weidel Bob Windham |
|
Reginald Beardsley
Wow!!! Leo, where did you find that information? Does anyone reading this know if any of them might have some schematics in a box some where? I'd really like to prevent these from all dying off for lack of repair data. And if you've worked around a packrat, you know who they are. So if someone recognizes the name of a packrat they worked with who might still be around, please drop me a private email.
I have been having a blast with my SD-24. Being able to see the SMA-F to SMA-M connection separately from the N-F to N-M connection on an adapter which is ~32 mm long is incredible. I could sort of manage it using the calibrator output and a tee, but the mismatch produced by parallel 50 ohm loads at the tee obscured things. Lots of reflections in the SMA tee to sort out. It's especially fun for me coming from reflection seismology because I'm so used to interpreting time domain data in the frequency domain by inspection. I'm still getting used to having complex reflection coefficients instead of the pure real coefficients of elasticity, but the wave equation is still the same and transmission lines are *much* simpler than 3D. All I need now is an SD-32 at a sensible price. The jitter I'm referring to is video system jitter. It makes me very nervous not having any data for it. Can it be repaired without data? Sure, but that's a long slog even with very simple gear. And these are anything but simple. BTW To return to the thread topic, has anyone investigated substituting a GPSDO for the timebase clock? It would seem to me that one of Leo's units would completely obviate the T1331 issue. But the errors might be gate and divider delays in which case it would be device specific. However, in that case I'd question whether they would be the same after 30 years. Have Fun! Reg |
|
Hi Leo,
toggle quoted message
Show quoted text
The S-4 and S-6 were designed by George Fry. They were a radical departure from the previous, much slower sampling heads. I once asked George why they were faster and he said it was because he saw the problem in terms of a wave front moving through the sampling gates. George is alive and well. He sells custom hearing aids in the Portland area. A few years ago I visited Agoston Agoston in Beaverton. I'm not sure when he left Tek but I don't believe he designed the newer heads for the 11K series. Right after he left Tek he formed HyperLabs and produced a series of sampling plugins that are the same form factor as the S series plugins. That's why I don't think he ever designed anything for the 11K series. His S series heads had better specs than the Tek heads. Here are the plugins HyperLabs made: HL-11 100 pS 50Ω BNC SAMPLING HEAD HL-12 100 pS 10 KΩ BNC SAMPLING HEAD HL-13 20 pS 50Ω, K (SMA), SAMPLING HEAD HL-14 <35 pS 50Ω, K (SMA), TDR HEAD HL-15 25 pS PHOTONICS HEAD HL-16 25 pS, 50Ω, K (SMA), PULSE GENERATOR HEAD Hyperlabs is still in business. You can read about what they are up to at http://www.hyperlabsinc.com/about-us.aspx I have datasheets somewhere for each of these sampling heads but I never saw a price list so I have no idea how much they sold for. In 20 years searching for them on eBay I have never come across any of them. When I was at his house we did not discuss the sampling heads. His company is now making TDR based systems that test and characterize the performance of extremely fast cables. His wife worked at Tek at the time of my visit. Dennis Tillman W7PF -----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Leo Bodnar Sent: Saturday, August 24, 2019 3:47 PM On Sat, Aug 24, 2019 at 05:06 PM, Reginald Beardsley wrote: I'd love to know who designed it and whether they are still alive.I think most heads were designed by Agoston Agoston. Other than that: ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ This product was made possible by the dedicated efforts of a group of inspired, tireless, and often times ingenious individuals of the design team. The following is a list of a few of the cast of many that contributed to the development of this oscilloscope. Design Team: Mark Anderson Brian Colony John DeLacy Rob DeVoe Jim Edwards Larry Hattery Tim Holte Roy Kaufman Jim Lamb Les Larson Chan Lee Jim Long Ben McCarroll Clark Morgan Jim Peterson Jim Schlegel Bob Simpson Ken Smith David White Wayne Wilburn Contributing Engineers: Ray Blohm Thomas Dye Scott Halsted Ron Henricksen Steve Herring Bev Kramlich Mehrab Sedeh Jim Stanley Gil Stephens Engineering Support: Bob Chandler Hilde Cochran John Eskeldson Pam Fodrea Lorna Goedel Dick Griffin John Hazard Diane Kemper Beverly McClenathan Jerry Sternes Wally Sutton Pat Wiley Manuals: Brian Diehm Amy Farrell Susan Grace Julie Leonetti Krystn McCaleb Chuck Melikian Roland Parenteau David Powe Mark Stade Carolyn Strong Manufacturing: Tim Bennington-Davis Donna Brown Jerry Brucken Kevin Cosgrove Lew Cummings Jeanne Eick Jerry Feickert Berdine Garner Clark Jarvis Paulette Jesse Murlan Kaufman Phil Lloyd Woody Ngincharoen Tina Noll Howard Nutt Carol Parks Al Phillips Steve Ratner Doug Rowe Bill Schell Colleen Swanson Mark Swenson Ron Tegner Rod Van Loon Walt Ventgen Marketing/Service: Denny Chamberlin Roger Ensrud Tom Freeman Theresa Graf Ivan Jackson Dennis Kucera Paul Kristof Ray Blohm John Boatwright Betty Bonham Bob Bousquet Jim Carter Dennis Chamberlin Jo Chi John Cooper Jerry Coulter Connie DeClerck Laszlo Dobos Tom Dye Brad Figg Dean Gehnert Ted Gerlinger Larry Hershiser Ken Holland Will Hott Shirley Humphreys David Irwin Stan Kaveckis Ray Kazlauskas Hedy Leidelmeyer Roy Lewallen Ken Longgrear Tina Newkirk Reba Norris Oris Nussbaum Mike O'Shanecy Janet Peters Barbara Ports V. Prasannan Mary Rehse John Rettig Glen Rollins Ava Stupek Dan Taylor Erik Teose Mark Tilden Jan Todd Harold Vandecoevering Mona Webber Vaughn Weidel Bob Windham -- Dennis Tillman W7PF TekScopes Moderator |
|
Leo Bodnar
On Sun, Aug 25, 2019 at 12:54 AM, Dennis Tillman W7PF wrote:
A few years ago I visited Agoston Agoston in Beaverton. I'm not sure when heI have probably concluded it rather than established it. In 1987 Aki was with Tek, in 1992 he was already with Hyperlabs. 11800 came out in late 1980's? Agoston Agoston is listed as an inventor in dual TDR head design https://patentimages.storage.googleapis.com/6e/80/e3/84f58712244c9d/US4755742.pdf which is, essentially what SD-24 is http://w140.com/tekwiki/images/f/f0/SD24_4.jpg Leo |
|
Egge Siert
Hi To All,
Around April this year I uploaded the most relevant SD-Patent Papers in a special Folder on TekScopes. named: Patent Papers SD_heads e.a. Enjoy. Greetings, Egge Siert |
|
Leo Bodnar
On Sun, Aug 25, 2019 at 12:48 AM, Reginald Beardsley wrote:
All I need now is an SD-32 at a sensible price.What are you going to use it for? I have two SD-30 that I use interchangeably that I have bought broken and fixed: one had blown termination resistors, another - faulty trigger pick-off hardline. I don't trust *any* SD heads, even sold as guaranteed working. My approach with them: if it works - it's a bonus. SD series had really bad manufacturing quality track record. From delaminating PCB gold plating to cracked solder to tin whiskers to missing bond wires to dying EEPROMs. I reckon S-series will still be alive and kicking while SD-series will perish in time. Cheers Leo |
|
Reginald Beardsley
These are just toys for an old man. I bought my 11801 because of the CSA803A plots that you send out with your pulsers. I thought it was just *so* cool. I got lucky and picked up an 11801 with a bad NVRAM error for $145 delivered and a pair of SD-22s for $150 delivered. With some help from a UK member of the list I got the NVRAM replaced. To my delight it all works fine Except for the video display jitter I mentioned earlier.
I bought 4x SD-26 heads for $200 each, one of which would not calibrate correctly. The seller replaced it. Thursday I got my SD-24 which eliminates the issues of impedance mismatch using a tee to pick off reflections. I bought the 4x SD-26s to measure timing skew for various DSP filter topologies implemented on a Zynq for use in a DSO. After spending some time with an MSOX3104T and RTM3104 I am quite underwhelmed by what you get for $20K. So rather than complain, I'm going to fix the issue by writing proper FW. I have a 2nd 11801 sitting in CA at my sister's with an SD-22 or 26 head. I'm told one channel is wonky. but won't know for sure until I get it. I know it has an NVRAM error. I primarily got it as a parts mule. About all I can think of right now to do with an SD-32 is check my Leo Bodnar pulsers ;-) If I do get an SD-32 I'll try to build a faster rise time pulser than yours. I've read a bunch on non-linear pulse generators. I have had more sheer fun fooling around doing completely insignificant experiments with this beast than any other piece of electronics I have ever owned. Partly this is because a transmission line is the EM analog to the 1D plane layered medium problem in reflection seismology. So it's *very* familiar. Being able to see on the screen things I have calculated in countless numerical simulations is just a blast. Have Fun! Reg |
|
Leo Bodnar
Reg,
If you are out for exotic toys, try finding 1180x or CSA803x with fast calibrator. You can tell it apart visually by the calibrator jack being gold 3.5mm type instead of Nickel plated SMA. It's basically a TDR part of SD-24 in a separate enclosure used instead of regular calibrator. They are rare but not unique. Here's random example https://www.artisantg.com/itemimages/Tektronix_CSA_803_View1.jpg I have never found who and why did this and why this is not documented anywhere. Leo |
|
Reginald Beardsley
Mine has it and it's documented in the 11801 user and service manuals. The user reference manual on p 42 says:
"The signal from the calibrator output is a 250 mV square wave with a rise time of approximately 20 ps and a period of approximately 10 us." I did a lot of work with it and my SD-26s until I found the SD-24. It was what go me hooked. |
|