2465B U950 Hybrid replacement ...
Yep, this sounds exactly like a failed Z Axis hybrid. I have replaced many of them. They are indeed in the top 5 failures. No soldering required, just remove the 4 nuts. When replacing the part, you m
Yep, this sounds exactly like a failed Z Axis hybrid. I have replaced many of them. They are indeed in the top 5 failures. No soldering required, just remove the 4 nuts. When replacing the part, you m
|
By
...
· #80182
·
|
|
Tek vs Hickok and Fairchild - Boeing vs Tupalov
I think that Tek made instruments like your CA for IBM under contract. If I recall correctly the 454 ( or one of the original 45X instruments) was designed in response to an IBM need. Some of these we
I think that Tek made instruments like your CA for IBM under contract. If I recall correctly the 454 ( or one of the original 45X instruments) was designed in response to an IBM need. Some of these we
|
By
...
· #80002
·
|
|
SI5020 MicroWave Switches
If it's any help, any Tek part number ending in -99 is a non-salable in house test fixture.
If it's any help, any Tek part number ending in -99 is a non-salable in house test fixture.
|
By
...
· #79393
·
|
|
2432A Digital Oscilloscope
Can you post a picture of the CRT showing the problem?
Can you post a picture of the CRT showing the problem?
|
By
...
· #78816
·
|
|
TG501 is missing 1 ns output
Well, you can see the 1ns markers on lesser scopes. I just did see them on the 2465A I have on my bench. The amplitude and triggering aren't great, but certainly viewable. I have adjusted the TG501 ou
Well, you can see the 1ns markers on lesser scopes. I just did see them on the 2465A I have on my bench. The amplitude and triggering aren't great, but certainly viewable. I have adjusted the TG501 ou
|
By
...
· #78509
·
|
|
TG501 is missing 1 ns output
The output circuit may be misadjusted. It requires a good output cable and termination, a good scope, and cre to adjust. Also clean the push button switches.
The output circuit may be misadjusted. It requires a good output cable and termination, a good scope, and cre to adjust. Also clean the push button switches.
|
By
...
· #78501
·
|
|
244B internal temperature
The fan control circuit in all these model variations have a thermistor to vary the fan speed as the internal temperature increases. The outgoing airflow is mainly directed out the right rear corner,
The fan control circuit in all these model variations have a thermistor to vary the fan speed as the internal temperature increases. The outgoing airflow is mainly directed out the right rear corner,
|
By
...
· #78435
·
|
|
I need CRT for 305DMM
Martin, I looked today, and I do not have a CRT for your 305. The tube I had in mind is for the 221. Bob
Martin, I looked today, and I do not have a CRT for your 305. The tube I had in mind is for the 221. Bob
|
By
...
· #78172
·
|
|
I need CRT for 305DMM
Martin, I think I have one. I will look within the next few days.
Martin, I think I have one. I will look within the next few days.
|
By
...
· #78150
·
|
|
Just got a tek 555 for free, and in need of some advice!
Why? Evidence please. The archives are sketchy at best.. Or, just power it up, and wait for a cap to short.
Why? Evidence please. The archives are sketchy at best.. Or, just power it up, and wait for a cap to short.
|
By
...
· #77830
·
|
|
Just got a tek 555 for free, and in need of some advice!
If you want to be really careful, slowly raise the mains voltage with a Variac if you have one. This will allow all those old electrolytic caps to reform slowly. this is about all you can do. You can
If you want to be really careful, slowly raise the mains voltage with a Variac if you have one. This will allow all those old electrolytic caps to reform slowly. this is about all you can do. You can
|
By
...
· #77816
·
|
|
2430A CCD
I have brand new, factory packed 165-2074-04 for $65.00 shipped CONUS. If interested contact me off forum.
I have brand new, factory packed 165-2074-04 for $65.00 shipped CONUS. If interested contact me off forum.
|
By
...
· #77606
·
|
|
2465B Jitter and non-linear display
Jim, This seems to be the dreaded bad Channel Switch hybrid. Unfortunately fairly common. Try one from another, any, 24xx analog scope. A replacement will require a recal. Bob
Jim, This seems to be the dreaded bad Channel Switch hybrid. Unfortunately fairly common. Try one from another, any, 24xx analog scope. A replacement will require a recal. Bob
|
By
...
· #77508
·
|
|
576 +/- 12.5 Volt Supply Issues
Replace all the PS output caps. I bet the oscillations stop...
Replace all the PS output caps. I bet the oscillations stop...
|
By
...
· #76719
·
|
|
2465B DC Balance/Parametric measurement problem
The Parametric measurements use the U500. You will need to recal the entire scope if it is to work completely correctly. Jim, contact me off list regarding the cal if you want it done. I have all the
The Parametric measurements use the U500. You will need to recal the entire scope if it is to work completely correctly. Jim, contact me off list regarding the cal if you want it done. I have all the
|
By
...
· #76677
·
|
|
2465B DC Balance/Parametric measurement problem
Well, you may not need to change the caps. How many hours on the scope? I know it can be changed, but if they are low-ish, I would just use it.
Well, you may not need to change the caps. How many hours on the scope? I know it can be changed, but if they are low-ish, I would just use it.
|
By
...
· #76666
·
|
|
2465B DC Balance/Parametric measurement problem
Ah, OK, the U500 sounds like it is good then. You really need a calibration generator to run the cal's. It needs to be a 1Khz square wave of the correct voltage. Most function generators can't do it.
Ah, OK, the U500 sounds like it is good then. You really need a calibration generator to run the cal's. It needs to be a 1Khz square wave of the correct voltage. Most function generators can't do it.
|
By
...
· #76663
·
|
|
2465B DC Balance/Parametric measurement problem
Jim, You need to run CAL 2 and 3 first. The different Trig hybrid will need different cal constants, plus the bad one probably caused some too-far-afield constants to be generated. If the CAL's don't
Jim, You need to run CAL 2 and 3 first. The different Trig hybrid will need different cal constants, plus the bad one probably caused some too-far-afield constants to be generated. If the CAL's don't
|
By
...
· #76660
·
|
|
2465B DC Balance/Parametric measurement problem
I really think it is the trigger hybrid. It is used to make the parametric measurements, so if it is bad, you will get the LIMIT message.
I really think it is the trigger hybrid. It is used to make the parametric measurements, so if it is bad, you will get the LIMIT message.
|
By
...
· #76598
·
|
|
2465B DC Balance/Parametric measurement problem
You probably have a bad trigger hybrid. A pretty common failure. You can try running CAL 2, but I don't think that will fix it..
You probably have a bad trigger hybrid. A pretty common failure. You can try running CAL 2, but I don't think that will fix it..
|
By
...
· #76592
·
|