Tek 500 series valved scopes. Any available?
I spotted a 547 on Ebay last night. Might fit the bill? Item 200618872943. Kevin
I spotted a 547 on Ebay last night. Might fit the bill? Item 200618872943. Kevin
|
By
Kevin Wood G7BCS
· #62275
·
|
|
1240 Logic Analyser display problem
Ahh, good point. I should try that. It feeds the next stage through a 10k resistor so unlikely it's being dragged down but you never know... Yep, the thought crossed my mind too. Actually, the H&V cou
Ahh, good point. I should try that. It feeds the next stage through a 10k resistor so unlikely it's being dragged down but you never know... Yep, the thought crossed my mind too. Actually, the H&V cou
|
By
Kevin Wood G7BCS
· #62107
·
|
|
1240 Logic Analyser display problem
Hi Guys, I wonder if anyone has ever troubleshooted display issues with a 1240 Logic Analyser? Symptoms are that I get a stable raster on the screen if brightness is turned up but no display. All powe
Hi Guys, I wonder if anyone has ever troubleshooted display issues with a 1240 Logic Analyser? Symptoms are that I get a stable raster on the screen if brightness is turned up but no display. All powe
|
By
Kevin Wood G7BCS
· #62105
·
|
|
Pease help diagnosing my 2235
I would fix the intensity problem first if I were you.Running the tube as bright as it looks in those pictures for any length of time will do it no favours at all. Check over all the components in the
I would fix the intensity problem first if I were you.Running the tube as bright as it looks in those pictures for any length of time will do it no favours at all. Check over all the components in the
|
By
Kevin Wood G7BCS
· #61242
·
|
|
Pease help diagnosing my 2235
Yep, that's it. You can regard it as a network that turns the Z axis signal (let's say it varies from 0-100V for the sake of argument. I'm not sure on the exact range) into an AC signal with an amplit
Yep, that's it. You can regard it as a network that turns the Z axis signal (let's say it varies from 0-100V for the sake of argument. I'm not sure on the exact range) into an AC signal with an amplit
|
By
Kevin Wood G7BCS
· #61232
·
|
|
Pease help diagnosing my 2235
Hi Bryan, In addition to the excellent suggestions already made, if the cathode is OK at about -2KV I would have a careful look at the DC restorer circuit that drives the control grid of the tube. I.E
Hi Bryan, In addition to the excellent suggestions already made, if the cathode is OK at about -2KV I would have a careful look at the DC restorer circuit that drives the control grid of the tube. I.E
|
By
Kevin Wood G7BCS
· #61207
·
|
|
2236 DMM issue - Now fixed :-)
Well, I found the problem. U1909 had what appeared to be an internal short from the +7.5V supply rail to some of the output pins. Everything was fine until the range changes started to switch in the A
Well, I found the problem. U1909 had what appeared to be an internal short from the +7.5V supply rail to some of the output pins. Everything was fine until the range changes started to switch in the A
|
By
Kevin Wood G7BCS
· #61205
·
|
|
2236 DMM issue - anyone able to help please?
Yep, I've come across the dreaded increasing resistors before! I think I measured those two correctly but will double check. I will do, although I got the scope on the collectors of Q1910 and Q1911 an
Yep, I've come across the dreaded increasing resistors before! I think I measured those two correctly but will double check. I will do, although I got the scope on the collectors of Q1910 and Q1911 an
|
By
Kevin Wood G7BCS
· #61014
·
|
|
2236 CTM self diagnostic
The CTM runs through a series of checks at power-up and reports an error code if anything fails. Otherwise, it just enters normal operation. There is a good description of the self test and further di
The CTM runs through a series of checks at power-up and reports an error code if anything fails. Otherwise, it just enters normal operation. There is a good description of the self test and further di
|
By
Kevin Wood G7BCS
· #60974
·
|
|
2236 DMM issue - anyone able to help please?
I think I might be getting somewhere. Wind up the voltage slowly (so the Hi-Z buffer output only reaches 0.5v or a little more) and on some of the ranges the +7.5v is getting pulled down by a volt or
I think I might be getting somewhere. Wind up the voltage slowly (so the Hi-Z buffer output only reaches 0.5v or a little more) and on some of the ranges the +7.5v is getting pulled down by a volt or
|
By
Kevin Wood G7BCS
· #60970
·
|
|
2236 DMM issue - anyone able to help please?
Hi Jim, Many thanks for the help. The switches checked out OK. I tried it with c1811 removed just for good measure in case the input amp. was overshooting and there was no difference. Incidentally, ap
Hi Jim, Many thanks for the help. The switches checked out OK. I tried it with c1811 removed just for good measure in case the input amp. was overshooting and there was no difference. Incidentally, ap
|
By
Kevin Wood G7BCS
· #60907
·
|
|
2236 DMM issue - anyone able to help please?
Hi Guys, Does anyone have any experience with the DMM on the 2236? I have currently got my 2236 apart for another repair which has been successful. While I had the lid off I thought I'd fix a niggle I
Hi Guys, Does anyone have any experience with the DMM on the 2236? I have currently got my 2236 apart for another repair which has been successful. While I had the lid off I thought I'd fix a niggle I
|
By
Kevin Wood G7BCS
· #60866
·
|
|
How to disassemble Tek 2236 ? ... for A1 mainboard HV section
Mine exhibited a similar problem once and the culprit was a group of 4 series connected 510k resistors in the dropper chain that fed the focus control. These run a little hot and one had failed. Kevin
Mine exhibited a similar problem once and the culprit was a group of 4 series connected 510k resistors in the dropper chain that fed the focus control. These run a little hot and one had failed. Kevin
|
By
Kevin Wood G7BCS
· #51915
·
|
|
Tektronix 2235 preregulator
I had a similar problem with my 2236 a while back. The switching transistor (from memory, a MOSFET on a heatsink just behind the cooling fan) had died. The fan had also seized up so I suspect it proba
I had a similar problem with my 2236 a while back. The switching transistor (from memory, a MOSFET on a heatsink just behind the cooling fan) had died. The fan had also seized up so I suspect it proba
|
By
Kevin Wood G7BCS
· #45241
·
|
|
Request - information on how to diagnose a dying 547/545B HV ...
I appreciate that, Chuck. Since I have a spare core sitting around anyway it is probably a no-brainer to turn it into a known good spare while the opportunity exists. Kevin
I appreciate that, Chuck. Since I have a spare core sitting around anyway it is probably a no-brainer to turn it into a known good spare while the opportunity exists. Kevin
|
By
Kevin Wood G7BCS
· #44865
·
|
|
Request - information on how to diagnose a dying 547/545B HV ...
I'd be interested to know too. As it happens I came across a replacement HV transformer for my 546 on Ebay last week and it appears to be a good one (was dated 1972 IIRC). Chuck's post is quite timely
I'd be interested to know too. As it happens I came across a replacement HV transformer for my 546 on Ebay last week and it appears to be a good one (was dated 1972 IIRC). Chuck's post is quite timely
|
By
Kevin Wood G7BCS
· #44854
·
|