maxim.vlasov@...
 

Chuck,

I've been using all sorts of LAs since the beginning of 90s in the reverse engineering for the systems with scarse or no technical documentation. Simply at the start I had my own home built machine initially based on Intel FlexLogic ICs (they had the fastest clock rate in 1993) then switching to Altera FLEX/APEX families. Then at the beginning of 2000s HP/Agilent LAs with a deep memory became available at the bargain prices. I talk mainly about the arcade machine HW (as this role goes I analyze the chipset architecture in details and re-produce its functional analogues for the FPGA/semicustom IP implementation) which are not the simple uP based HW (Tek is a such example). So, I know something about the reverse engineering bringing the concrete examples ;))

Everyone in the embedded "professional" hacking knows IDA. This is not a panacea as a cross-platform disassembler, but almost and certainly the best of all. But since there is a lot's of the dynamic code in the FW (especially when ppl translating from C++), which can't be analyzed statically (by the disassembler), this runtime can be only analyzed on the working maching by using the LA even without any built-in inverse assembler. So the synergy of both are required to understand the code architecture. Getting the architecture, control and data flow and symbolizing the source is the most important steps. Also we shouldn't forget that the source can be actually compiled by not a very optimal compiler generating the spagetty code.
From all the above, the current example (Tek 6802 base platform) is quite simple, HW is all documented and it can be analyzed statically. But due to the code volume superimposing the calls will be necessary and there the LA will be undispensable.

Back to the oscilloscope, I've got 2445a CAL01 re-calibrated. This time I got it right (digital scopes unfortunately are quite useless). Still CAL02 gives the LIMIT error. But what I also noticed is that when I put the time cursors to the 1st and 11st vertical line the displayed frequency/period gives the precise value in seconds. However, it seems that when I just put them one division apart or a few, then I do have the error in the displayed delta T. Negative if both cursors are to the left from the center and positive when both to the right. It seems like there is a problem with the DAC linearity or something in its path.
IMHO this kind of error then hits back in CAL02.

What do you think?

Thank you again and all the Best,

Maxim

Re: vintageTEK museum releases Replaceable Parts Registry (RPR)

 

Thanks so much for another great RPR scan, the 311 set is very handy to have!
I can hardly wait for that 366 scan, it will answer a lot of questions.
the page link is here: http://w140.com/tekwiki/wiki/RPR

I notice it doesn't seem to tie to the general main page at all.

thank you for all the hard work!
-walter

--
Walter Shawlee 2
Sphere Research Corp. 3394 Sunnyside Rd.
West Kelowna, BC, V1Z 2V4 CANADA
Phone: +1 (250-769-1834 -:- http://www.sphere.bc.ca
We're all in one boat, no matter how it looks to you. (WS2)
All you need is love. (John Lennon)
But, that doesn't mean other things don't come in handy. (WS2)

Re: 2445A calibration

Harvey White
 

On Sat, 13 Oct 2018 09:55:50 -0400, you wrote:

I have been designing and debuging embedded systems for more than
40 years. I started with the 8008, and have designed and programmed
hardware for the 8080, Z80, 8085, 8086, 80188, 80186, ..., 68000, 68010,
PIC, ARM, ..... Believe it or not, but they are all substantially similar,
some a little more similar than others.

However, whenever anyone tells me that a Logic Analyzer, with an internal
disassembler probe, will ease the job of understanding anything, I know
immediately that they don't know what they are talking about.

The state of LA disassembler probes never reached a point where useful
work could be done, and are in fact a tantalizing way of wasting money.
It seems like it should be possible, was sold as the software engineers
panacea, but it never is.
There are some situations where the LA can be of use, but mostly for
tracking specific events under specific conditions, IMHO.




The largest problem is that the disassembler probes for given microprocessors
always lags the deployment of those microprocessors, and the fast evolution of
new processors means that the art is always running away from your investment
in probes.
Now for the 6800 and such, you could find the most up-to-date probe,
so that becomes less of a factor for that generation of processors,
I'd think.



A far better way of understanding an assembly program is to use a
good disassembler, that allows you to use your own labels for data and
jump addresses, and it is generally free. Plus, the software interface to
the hardware processors is surprisingly slow to change, and simple and
cheap to adapt.
Hmmm, got one for the 6800 series? I have core images...



Software emulators, debuggers, and disassembler/decompilers, have largely
supplanted the disassembling Logic Analyzer, and they did so back in the days
of the Data General Eagle, and DEC's VAX..

I don't think anyone was recommending using 6502 tools on a 6802. I think
it was simply either someone showing a parallel, or mistaking my mention of
the 2465's 6802, for the more famous "02" processor the 6502.
IIRC, the 6802 did more integration of the RAM into the processor. I
think that the DM5010 uses one as well.

Harvey


-Chuck Harris

maxim.vlasov@... wrote:
6802 shares the 6800 instruction set, however for the inverse assembler for HP/Agilent 165xx, 166xx and 167xx still a different adapter is required due to the slightly different bus timing (or one can be made by using a simple 5V tolerant CPLD). Mostek/WDC 650x & Motorola 680x are completely different animals architecture wise.
The TASM PC (for DOS) cross assembler can be used to compile and link the code natively for 6802/6800. IDA disassembler will re-create the disassmbled code to the level of the original source (or sometimes even better documented). HP/Agilent INVASM will disassemble in real-time all the event driven routines, for example interrupts, insitu on the running oscilloscope.
By disassembling and recompiling the original source (disassembly done automatically by IDA) the generated ELF (embedded linker file with symbol data) can be loaded into the HP/Agilent 16505A or Agilent 167xx software analyzer suite to display the statistical distribution and recurrence of the service routine calls. This sometimes very handy to analyze the real-time FW. I've done a few projects in the past for 68040/PPC in that way.

IMHO, there is no need to complicate the life with 6502 tools since they'll provide no help with 6800.

Re: 2445A calibration

Harvey White
 

On Sat, 13 Oct 2018 01:47:20 -0400, you wrote:

Or, "Square Wave".
To be a bit pedantic, to me, square wave doesn't talk about the
levels, so where's 0 and where's Vmax...?

bet there's lots of opinions on that one.

Harvey



-Chuck Harris

Harvey White wrote:
On Fri, 12 Oct 2018 16:22:55 -0700, you wrote:
A useful term is "pulsating DC", or perhaps unipolar pulses, if you
wish.

Harvey

Re: 468 horizontal jitter

 

In case you haven't, it is important to check if it is a centering problem or a triggering problem. Even in auto or free run it is still triggered, so if it is the problem could be all the way back at the ramp generator.

If the trace itself is moving and not the sweep (I hope you got that...) then I would more look for hash on a PS line somewhere. Of course that applies to the ramp generator as well but different PS legs of course.

Re: vintageTEK museum releases Replaceable Parts Registry (RPR)

Vincent Trouilliez
 

On Sat, Oct 13, 2018 at 09:05 AM, walter shawlee wrote:
I notice it doesn't seem to tie to the general main page at all.
Yes I was wondering about thaat too ! Such a nice resource... that's not even visible/accessible, kinda defeats the purpose of making it public ^^

I guess the link to that RPR page might fit at the bottom of the main page, under the section labelled " Manuals, Catalogs and Other Publications " ?!

Thanks for all the scanned sections so far, and for all those to come :-)


Vincent Trouilliez

Re: 468 horizontal jitter

Brendan
 

On Sat, Oct 13, 2018 at 09:51 AM, Jeff Urban wrote:


In case you haven't, it is important to check if it is a centering problem or
a triggering problem. Even in auto or free run it is still triggered, so if it
is the problem could be all the way back at the ramp generator.

If the trace itself is moving and not the sweep (I hope you got that...) then
I would more look for hash on a PS line somewhere. Of course that applies to
the ramp generator as well but different PS legs of course.
Thanks Jeff.
There is jitter on the A gate out with no input just a free running trace. I had to put my 465 in 10x to see it but it does match the jitter on the 468. So... This looks like its going to be a fun one to figure out.

Re: vintageTEK museum releases Replaceable Parts Registry (RPR)

Dave Brown
 

It is supposed to be on the main page Manuals, Catalogs, and Other Publications > Reference Materials but I don't see it listed there. I'll contact Kurt.

Dave

Re: 2445A calibration

Mark Litwack
 

On Sat, Oct 13, 2018 at 08:40 AM, <maxim.vlasov@...> wrote:

6802 shares the 6800 instruction set, however for the inverse assembler for
HP/Agilent 165xx, 166xx and 167xx still a different adapter is required due to
the slightly different bus timing (or one can be made by using a simple 5V
tolerant CPLD). Mostek/WDC 650x & Motorola 680x are completely different
animals architecture wise.
The TASM PC (for DOS) cross assembler can be used to compile and link the code
natively for 6802/6800. IDA disassembler will re-create the disassmbled code
to the level of the original source (or sometimes even better documented).
HP/Agilent INVASM will disassemble in real-time all the event driven routines,
for example interrupts, insitu on the running oscilloscope.
By disassembling and recompiling the original source (disassembly done
automatically by IDA) the generated ELF (embedded linker file with symbol
data) can be loaded into the HP/Agilent 16505A or Agilent 167xx software
analyzer suite to display the statistical distribution and recurrence of the
service routine calls. This sometimes very handy to analyze the real-time FW.
I've done a few projects in the past for 68040/PPC in that way.
...

Actually, the 167xx doesn't always need a special adapter. There's a good thread on HP/Agilent disassemblers here:

http://www.eevblog.com/forum/testgear/hp-logic-analyzer-inverse-assemblers

In post #3 in that thread there's an example of using a 16702B logic analyzer to probe a 6802 processor using flying lead probes directly to the processor pins (no adapter). The thing being probed was a Tek 2465.

On the 2445A/2465A and 2465B/2465B, Tek uses bank-switched memory, making it even more fun to untangle the code. The 2465 does not have bank switching. (Maybe IDA handles bank switching?) If someone wants to head down the disassembly path, here's some info on how the banks are handled, specifically for a 2465A, but I think the B series is the same:

http://www.eevblog.com/forum/testgear/just-got-a-2465a-couple-questions/msg1469675/#msg1469675

I've done some disassembly of the code along with real-time tracing to figure out how some things work. It's time consuming and some of the code seems a bit hack-y as it was ported from one platform to the next.

-mark

PG 506 IS IT NECESSARY FOR CALIBRATING?

Jim Olson <v_12eng@...>
 

---------- Original Message ----------
From: Jim Olson <v_12eng@...>
To: TekScopes@groups.io
Date: October 11, 2018 at 3:54 PM
Subject: [TekScopes] PG 506 IS IT NECESSARY FOR CALIBRATING?

This is for the folks here that are a lot more knowledgeable and experienced than me with these 4xx series scopes with the testing and calibrating of them. I have a number of the TM series plugins plus case for them that I got really lucky and got for a great deal. Here's a list of the modules: DM501, DM502A, PG501, PG502, PS503A, FG504, Type 187 Time mark Generator, Type 191 Signal Generator and a CT3, need to get the BNC adapters for it I see there two types one straight through and one with a 50ohm terminator should i get both? I still need the 50ohm terminators and coax cables. I have 100' of RG58U just need to pick up some cable connectors for it.
I also have a good 2215 scope with a P6208 probe.
So my question is do I need the PG506 to really calibrate the scopes and also should i get a SG503 to get the higher Mhz output?
Once i get the 466 together I will really need to calibrate it so do i have enough cal instruments? and I will be asking for much help too!

Jim O


Re: vintageTEK museum releases Replaceable Parts Registry (RPR)

Dave Brown
 

The link is now on Manuals, Catalogs, and Other Publications > Reference Materials.
Dave

Re: 2445A calibration

Chuck Harris
 

Harvey White wrote:
The state of LA disassembler probes never reached a point where useful
work could be done, and are in fact a tantalizing way of wasting money.
It seems like it should be possible, was sold as the software engineers
panacea, but it never is.
There are some situations where the LA can be of use, but mostly for
tracking specific events under specific conditions, IMHO.
Indeed, but I find that the LA disassemblers are problematic. It is difficult
to tell with a piplined processor whether the memory reads you are processing
are reading data, or reading instructions. Without a probe specific to the
processor that can make those determinations, you are just looking at
data translated to assembler mnemonics... not useful. The 8086 was the first
commonly used microprocessor to suffer this problem... and solution. Intel
included extra pads that brought out pipeline tracking information for emulators
and logic analyzers.

Tracking specific events under specific conditions is where a LA shines.

The largest problem is that the disassembler probes for given microprocessors
always lags the deployment of those microprocessors, and the fast evolution of
new processors means that the art is always running away from your investment
in probes.
Now for the 6800 and such, you could find the most up-to-date probe,
so that becomes less of a factor for that generation of processors,
I'd think.
Possibly, most of the older processor's pods have been long since scrapped,
but some are certain to still exist.

I just ditched some 6502, Z80, and 8085 emulators... They probably are still
available from the recycler if anyone is interested.



A far better way of understanding an assembly program is to use a
good disassembler, that allows you to use your own labels for data and
jump addresses, and it is generally free. Plus, the software interface to
the hardware processors is surprisingly slow to change, and simple and
cheap to adapt.
Hmmm, got one for the 6800 series? I have core images...
A couple, actually. Sourceforge, and github are littered with them,
particularly the 6809, which is an upfeatured 6802 used in the automotive
industry.

PHK, Poul-Henning Kamp wrote a nice one that actually analyzes the code
by executing it. He used it to disassemble the code in some HP counters.

...

IIRC, the 6802 did more integration of the RAM into the processor. I
think that the DM5010 uses one as well
Tek tended to use the 68B02, which is a 6802 without the internal RAM.

-Chuck Harris

Re: 2445A calibration

Harvey White
 

On Sat, 13 Oct 2018 16:05:14 -0400, you wrote:

Harvey White wrote:
<snip>
The largest problem is that the disassembler probes for given microprocessors
always lags the deployment of those microprocessors, and the fast evolution of
new processors means that the art is always running away from your investment
in probes.
Now for the 6800 and such, you could find the most up-to-date probe,
so that becomes less of a factor for that generation of processors,
I'd think.
Possibly, most of the older processor's pods have been long since scrapped,
but some are certain to still exist.

I just ditched some 6502, Z80, and 8085 emulators... They probably are still
available from the recycler if anyone is interested.
I probably would have liked the ones for the 8085 and the 6502, just
because. Every once and while, I need to look at some of those old
ones. Right now it's the 6802 that I need to look at.



A far better way of understanding an assembly program is to use a
good disassembler, that allows you to use your own labels for data and
jump addresses, and it is generally free. Plus, the software interface to
the hardware processors is surprisingly slow to change, and simple and
cheap to adapt.
Hmmm, got one for the 6800 series? I have core images...
A couple, actually. Sourceforge, and github are littered with them,
particularly the 6809, which is an upfeatured 6802 used in the automotive
industry.
I found a few, they want DOS.

I have one that I wrote that runs under windows and is gui driven.


PHK, Poul-Henning Kamp wrote a nice one that actually analyzes the code
by executing it. He used it to disassemble the code in some HP counters.
Hmmm, some people want money for those things, not that I begrudge him
the income, but I tend to go to open source.

That's what mine does, although I get tired of writing tools to make
tools....



...

IIRC, the 6802 did more integration of the RAM into the processor. I
think that the DM5010 uses one as well
Tek tended to use the 68B02, which is a 6802 without the internal RAM.
The one in the DM5010 isn't, I think, and actually has internal RAM.
For analysis cases, all it does (of course) is to change the location
of some of the RAM.

Harvey


-Chuck Harris


Re: 7000-series power-hungry plug-ins

tinkera123
 

Okay, understand that now. I have played with/built many linear supplies, but none with much protection circuitry.
Thank you, Chuck.
Cheers,
Ian

In regulated supplies, there are often over current protection
circuits that are used in lieu of fuses. These circuits reduce
the voltage from the power supply in an attempt to reduce the
current, and the power dissipation, to something that the supply
can safely survive. This is called current foldback.

The Tektronix power supplies for the most part all have this
capability.

-Chuck Harris

Re: 2445A calibration

Chuck Harris
 

There are many disassemblers that actually emulate the target
code to determine the exact way it operates... all in software.

CAL01 doesn't have much to do with CAL02.

Which attenuator settings give you the RANGE error when running
CAL02?

-Chuck Harris

maxim.vlasov@... wrote:

Chuck,

I've been using all sorts of LAs since the beginning of 90s in the reverse engineering for the systems with scarse or no technical documentation. Simply at the start I had my own home built machine initially based on Intel FlexLogic ICs (they had the fastest clock rate in 1993) then switching to Altera FLEX/APEX families. Then at the beginning of 2000s HP/Agilent LAs with a deep memory became available at the bargain prices. I talk mainly about the arcade machine HW (as this role goes I analyze the chipset architecture in details and re-produce its functional analogues for the FPGA/semicustom IP implementation) which are not the simple uP based HW (Tek is a such example). So, I know something about the reverse engineering bringing the concrete examples ;))

Everyone in the embedded "professional" hacking knows IDA. This is not a panacea as a cross-platform disassembler, but almost and certainly the best of all. But since there is a lot's of the dynamic code in the FW (especially when ppl translating from C++), which can't be analyzed statically (by the disassembler), this runtime can be only analyzed on the working maching by using the LA even without any built-in inverse assembler. So the synergy of both are required to understand the code architecture. Getting the architecture, control and data flow and symbolizing the source is the most important steps. Also we shouldn't forget that the source can be actually compiled by not a very optimal compiler generating the spagetty code.
From all the above, the current example (Tek 6802 base platform) is quite simple, HW is all documented and it can be analyzed statically. But due to the code volume superimposing the calls will be necessary and there the LA will be undispensable.
Back to the oscilloscope, I've got 2445a CAL01 re-calibrated. This time I got it right (digital scopes unfortunately are quite useless). Still CAL02 gives the LIMIT error. But what I also noticed is that when I put the time cursors to the 1st and 11st vertical line the displayed frequency/period gives the precise value in seconds. However, it seems that when I just put them one division apart or a few, then I do have the error in the displayed delta T. Negative if both cursors are to the left from the center and positive when both to the right. It seems like there is a problem with the DAC linearity or something in its path.
IMHO this kind of error then hits back in CAL02.

What do you think?

Thank you again and all the Best,

Maxim

Re: 2445A calibration

Chuck Harris
 

PHK is an open source legend. His disassembler is available
for free on github.

PHK wrote the NTP routines for openbsd.

-Chuck Harris

Harvey White wrote:
On Sat, 13 Oct 2018 16:05:14 -0400, you wrote:
...

PHK, Poul-Henning Kamp wrote a nice one that actually analyzes the code
by executing it. He used it to disassemble the code in some HP counters.
Hmmm, some people want money for those things, not that I begrudge him
the income, but I tend to go to open source.

That's what mine does, although I get tired of writing tools to make
tools....



...

IIRC, the 6802 did more integration of the RAM into the processor. I
think that the DM5010 uses one as well
Tek tended to use the 68B02, which is a 6802 without the internal RAM.
The one in the DM5010 isn't, I think, and actually has internal RAM.
For analysis cases, all it does (of course) is to change the location
of some of the RAM.

Harvey


-Chuck Harris





Re: 7000-series power-hungry plug-ins

Chuck Harris
 

If you have ever used a 7805, or LM309, LM319, or LM340, ... you
have already used current foldback.

-Chuck Harris

tinkera123 wrote:

Okay, understand that now. I have played with/built many linear supplies, but none with much protection circuitry.
Thank you, Chuck.
Cheers,
Ian

In regulated supplies, there are often over current protection
circuits that are used in lieu of fuses. These circuits reduce
the voltage from the power supply in an attempt to reduce the
current, and the power dissipation, to something that the supply
can safely survive. This is called current foldback.

The Tektronix power supplies for the most part all have this
capability.

-Chuck Harris

Re: 2445A calibration

Chuck Harris
 

Hi Maxim,

The normalizer is just a resistor and a trimmer capacitor
in parallel, but in series with the center conductors of
a male and a female BNC connector. It is built in a little
Pomona box.

You can do the same thing with a scope probe, using the
probe's trimmer capacitor to square up the waveform on
one channel, and the C205 to square it up on the other.

I doubt it needs adjustment, assuming that you haven't moved
it already.

That CAL03 throws a LIMIT on CH3 and 4 makes me even more
certain that your problem is your generator.

-Chuck Harris

maxim.vlasov@... wrote:

Chuck,

I read again the input capacitance adjustment procedure for CH1/CH2. I have no 15pF normalizer and the high voltage calibration generator.

I can observe that when 1 MOhm input on the scope is selected, the square waveform signal delivered by the BNC cable has a visible overshoot on the rising edge. So, likely the C105 & C205 have to be adjusted too. I wonder if this overshoot could potentially contribute to the CAL02 LIMIT problem on the channel 1 & 2.
However, when calibration CAL03 routing with other channels 3 & 4 the scope also throws the LIMIT error.

Thank you,

Maxim

Re: 2465B Blue Screen Filter - 378-0270-00

Miklos Koncz
 

Hi Victor,

Please, reserve me also 2 pieces of blue filter!

You can contact me at kmiklos at vnet.hu.

Thanks

Miklos

Re: 2445A calibration

maxim.vlasov@...
 

To be more concrete I've thrown to IDA the 2465a ROM files. 160-3302-09 & 160-3303-09 are the part number downloaded from here:
http://w140.com/tekwiki/wiki/ROM_images

The code structure is quite interesting. The bank switching routines are in the first 2K RAM working area. I've quickly found the voltage reference setting routine:
https://groups.io/g/TekScopes/photo/74577/22?p=Name,,,20,1,20,0

See the comments in the code. The 0xAC=172d code thrown in the DAC sets the reference voltage of 1.36V:
1ma*(4095-172)*4/4096=3831uA. Solving for x=TP2421 voltage:
x*14.2KOhm/13.0KOhm + 681Ohm*(4095-172)*0.004A/4096 = -x
x=(4095-172)*0.004A/4096*681Ohm/(1Ohm+14200Ohm/13000Ohm) = -1.247V

The above assumes that the DAC is calibrated for the 1mA reference current.

Here are some other examples of the DAC and the counter access:
https://groups.io/g/TekScopes/photo/74577/21?p=Name,,,20,1,20,0
https://groups.io/g/TekScopes/photo/74577/20?p=Name,,,20,1,20,0

The code is complex in structure due to the ugly banking implementation. There are similar sections in all four 32K ROM pages. DAC is accessed all over the place in byte fashion and also by direct 16 bit X or even S register store. The DAC work area in the first 2K of RAM is also used heavily and many locations are duplicated.

If completely stuck, then I'll try tracing the LIMIT error, but looking just at the ADC routine complexity (uses almost 200 bytes of various runtime data to do the conditional branching) likely it will consume all the time for several months...