Date   
telonics sweep generator

battyhugh
 

Hi I have a Telonics 2003 sweep generator - has anyone still got one? - or a manual. I have had it for years mouldering away in the store room. As it is dual voltage - I can run it here (but the original 3 pin plug seems unobtainable) - it's a beautiful piece of late 1960's US technology - and I'd like to get it going.
(that's what spectrum analysers are for!) - thanks and Cheers
Hugh

Re: TM5xx/TMxxx Pass Transistors. SUBJECT CHANGE, WAS: DC-508 repair advice

 

Hi Michael,
There a lots of possible replacements provided the voltage, current, and power ratings are equivalent. The other thing to watch for is be sure the replacement pass transistor leads are in the same order as the originals. Tek changed the parts they used over the years and you must make sure your replacements have the same order for the E, B, and C pins. Beyond these 4 things the package outline is the only other detail that you need to match.

The rightmost slot of the TM504, TM506, and TM5006 are high power compartments so they probably have beefier pass transistors.

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Mlynch001
Sent: Sunday, December 15, 2019 1:49 PM
Subject: Re: [TekScopes] TM5xx/TMxxx Pass Transistors. SUBJECT CHANGE, WAS: DC-508 repair advice

Dennis,

I was able to find the original specs for my particular unit and then found some replacements that I thought might work. It has 5 bays with the same NPN and PNP parts and the high power compartment used something different. I found some replacements and then used the curve tracer to match the old parts to the new parts as close as possible. I’m going to look at what you have written, to see how close I got by dead reckoning and using my Curve Tracer. I appreciate your guidance, I always learn something from the members of this group.

--
Michael Lynch
Dardanelle, AR




--
Dennis Tillman W7PF
TekScopes Moderator

Re: Paying it forward with David's Part's Exchange

nonIonizing EMF
 

I recently ordered some components with the same thought and intent, needed about 8 and had to order 50, though I noted on Hackaday. I'll email you to update the list with the components I have once I get a more accurate inventory. I have a bunch of 0603 SMD components also. Thanks David.

Re: Paying it forward with David's Part's Exchange

David Berlind
 

Just noticing that the web version masked my email address. So, just in case you didn't get it... it is my first name at my last name (berlind) dot "org".

Re: Paying it forward with David's Part's Exchange

David Berlind
 

Thanks Dennis...

Anyone who is interested in something that I've listed should contact me
off-group via email at david@....

The process will basically be:

1. Double check that I have the part (please only request parts that I've
listed)
2. Give you an address to send your SAS-package to.
3. I'll return the package to you with the part

Thanks.

David

On Sun, Dec 15, 2019 at 2:37 PM Dennis Tillman W7PF <@Dennis_Tillman_W7pF>
wrote:

David,

Please provide YOUR STREET ADDRESS for people to send to you for your list
or include YOUR EMAIL ADDRESS.

Interested members PLEASE contact David Berlind directly if you are
interested in his generous offer.
All inquiries should be handled OFF-LIST to keep unnecessary personal
correspondence off of TekScopes.

Dennis Tillman W7PF


-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of David
Berlind
Sent: Saturday, December 14, 2019 9:18 AM
Subject: [TekScopes] Paying it forward with David's Part's Exchange

Hi Tekkies...

I have launched a Google Spreadsheet that I am very slowly updating with
the inventory of all my parts. In an effort to pay it forward to you and
others that have been so generous in helping me to get my fingernails
dirty, I'm making 99% of the parts available to anyone for free.
Personally, it drives me nuts I need one part and I have to pay more to buy
it in bulk, or the shipping costs more than the part itself. I will never
use most of these parts and so I'd rather give them away (not in bulk). All
I ask is that you send me a self-addressed stamped envelope (or other
packaging) and I will put the part in that packaging and return it to you.
It's going to take me months to get this spreadsheet fully up to date.
Maybe years. I don't know. But each week, I'm chipping away at the iceberg,
adding new parts. So, let me know if you see something you need or like.
I'm calling this David's Part Exchange. Let me know if you have any
questions.
https://docs.google.com/spreadsheets/d/1Q4M7dnS--TXb5PW1VqEBmM2Sxonknz0vVFVk_rfho_o/edit#gid=437389817




--
Dennis Tillman W7PF
TekScopes Moderator



Re: Paying it forward with David's Part's Exchange

David Berlind
 

Thanks Michael. If you want me to add tabs to my spreadsheet that list your
inventory, let me know.

On Sat, Dec 14, 2019 at 1:41 PM Mlynch001 <@mlynch001> wrote:

David,

Very Generous of you. You have inspired me to consider doing something
similar for those on this forum.

--
Michael Lynch
Dardanelle, AR



Re: DC-508 repair advice

Mlynch001
 

This sounds interesting! I need to look into this further.

--
Michael Lynch
Dardanelle, AR

Re: DC-508 repair advice

 

On Sun, Dec 15, 2019 at 09:16 AM, Mlynch001 wrote:


Check your fuses on your plugins. They are prone to open when the pass
transistors fail. Also check your voltages as well. Plugin crowbars are also
prone to fail. Pay special attention to the complementary transistor in the
plugin I.e, the one that’s mated to the failed pass transistor. For
example, in a TG501 on the 15v Supply, this would be Q610.

--
Michael Lynch
Dardanelle, AR
There ares several manuals for the 067-1201-99 TM500 power supply tester (I don't think it was ever commercially available) page on the Tekwiki web site at http://w140.com/tekwiki/wiki/067-1201-99. The manuals contain construction details for the tester, to be built into a 040-0652-02 blank plug-in chassis. I have one of the blank chassis kits, and have been wondering what kind of project I could build into it. These questions reminded me of the tester and I think it would be worthwhile, since I have several of the TM500 power supply units.

After revisiting the parts list for the tester, I think the 4pole 6position meter switch would be the most hard-to-find, but I found several of them available on Ebay. Not cheap, though. I think a 1pole 6 position switch could be used to switch in relays without much trouble. Small relays could be used, since none of the switched contacts seem to be in high current circuits.

Has anyone in the group built one of the testers? If so, did you make a PCB for it?

Cheers,
Dave M

Tek 2465 Problem

Panos
 

Hello to all.
Taking some measurements on my TEK 2465 using another oscilloscope, I see something strange.

I read for CHN5B signal -1,2 Volt versus -3,6 Volt. This only with BW LIMIT off.
Also for CHN 5A I read signal 600mV p-p versus -6 Volt as look at waveforms on service manual.

Am I doing something wrong, or I have problem somewhere in my TEK scope?

Waveforms 28 and 46 in pages 238 and 239 in service manual. These appears in diagram No 5 in page 244.

Looking for part

 

Does anyone have a 2221A, 2216, 2224 or 2232 parts donor ?
I'm looking for the flex circuit board that goes from the bezel buttons to the main board.
P/N 259-0065-0x. Contact me off list if you have one to sell.
/Håkan

Re: Great video for 500 series Tek restorers

John Williams
 

Thanks Kurt. These are very interesting videos. Even though I have restored a dozen 555’s and still have three of the beasts, I still can learn a lot from his information. Tektronix made so many great and interesting instruments during the tube era. I really like them all. It is nice to see a video tour around the instruments.

On an off topic note, I am just finishing a book called “Dec is dead, long live Dec. it is the story of the rise and fall of Digital Equipment Corporation and the legacy they left behind. Perhaps something similar was a force at Tektronix but was overcome better by that company. Anyway it is a good and interesting read. Another favourite - Bill and Dave

Re: TM5xx/TMxxx Pass Transistors. SUBJECT CHANGE, WAS: DC-508 repair advice

Mlynch001
 

Dennis,

I was able to find the original specs for my particular unit and then found some replacements that I thought might work. It has 5 bays with the same NPN and PNP parts and the high power compartment used something different. I found some replacements and then used the curve tracer to match the old parts to the new parts as close as possible. I’m going to look at what you have written, to see how close I got by dead reckoning and using my Curve Tracer. I appreciate your guidance, I always learn something from the members of this group.

--
Michael Lynch
Dardanelle, AR

TM5xx/TMxxx Pass Transistors. SUBJECT CHANGE, WAS: DC-508 repair advice

 

Hi Michael and all TekScopes members,

As my late friend, Jerry Scheltgen (AKA Kilowatt on eBay), and I once documented, knowing which pass transistors are used in which TM500 / TM5000 mainframe, is not easy to figure out because Tek changed them often and some mainframes use different ones for no apparent reason. To make matters worse on some occasions Tek changed to a different transistor with ENTIRELY DIFFERENT lead order so they require contortions to reorient the leads to make them reach the holes the old leads went into.

To solve this mess in 8/2014, Jerry and I documented which pass transistors were used where and when you had to be careful because the lead orientation mattered.

I uploaded the list we created just now. It can be found at
https://groups.io/g/TekScopes/files/TM5xxx%20Pass%20Transistor%20List.pdf

If you search our archives of messages regarding any topic with "pass transistors" you will find several recommendations our members made for "better" or "more powerful" or "more sensitive" or "less expensive" or "more available" substitutes for what Tek used. The only consideration I would caution that you observe it is desirable (to prevent oscillation) that the pass transistor have a low cutoff frequency (Ft). Less than 1 MHz would be ideal.

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Mlynch001
Sent: Sunday, December 15, 2019 11:51 AM
Subject: Re: [TekScopes] DC-508 repair advice

The hard part is finding replacement transistors for the pass transistors. I blew one up by plugging one of my plugins onto my extension cable upside down. The plugin extension did not have an indexing pin, so it would fit either way.
--
Michael Lynch
Dardanelle, AR




--
Dennis Tillman W7PF
TekScopes Moderator

Re: DC-508 repair advice

Mlynch001
 

The hard part is finding replacement transistors for the pass transistors. I blew one up by plugging one of my plugins onto my extension cable upside down. The plugin extension did not have an indexing pin, so it would fit either way.
--
Michael Lynch
Dardanelle, AR

Re: Paying it forward with David's Part's Exchange

 

David,

Please provide YOUR STREET ADDRESS for people to send to you for your list or include YOUR EMAIL ADDRESS.

Interested members PLEASE contact David Berlind directly if you are interested in his generous offer.
All inquiries should be handled OFF-LIST to keep unnecessary personal correspondence off of TekScopes.

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of David Berlind
Sent: Saturday, December 14, 2019 9:18 AM
Subject: [TekScopes] Paying it forward with David's Part's Exchange

Hi Tekkies...

I have launched a Google Spreadsheet that I am very slowly updating with the inventory of all my parts. In an effort to pay it forward to you and others that have been so generous in helping me to get my fingernails dirty, I'm making 99% of the parts available to anyone for free. Personally, it drives me nuts I need one part and I have to pay more to buy it in bulk, or the shipping costs more than the part itself. I will never use most of these parts and so I'd rather give them away (not in bulk). All I ask is that you send me a self-addressed stamped envelope (or other packaging) and I will put the part in that packaging and return it to you. It's going to take me months to get this spreadsheet fully up to date. Maybe years. I don't know. But each week, I'm chipping away at the iceberg, adding new parts. So, let me know if you see something you need or like. I'm calling this David's Part Exchange. Let me know if you have any questions. https://docs.google.com/spreadsheets/d/1Q4M7dnS--TXb5PW1VqEBmM2Sxonknz0vVFVk_rfho_o/edit#gid=437389817




--
Dennis Tillman W7PF
TekScopes Moderator

Re: Paying it forward with David's Part's Exchange

 

PLEASE take requests like this OFF-LINE.

Thank you
Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of DaveH52
Sent: Sunday, December 15, 2019 11:05 AM
To: TekScopes@groups.io
Subject: Re: [TekScopes] Paying it forward with David's Part's Exchange

Where would I send the SASE and wish list?
Dave





--
Dennis Tillman W7PF
TekScopes Moderator

Re: TDS684C "Cal Fails during initialization".

Siggi
 

Hey Reg,

I can't speak to these errors, but wonder if you know what you're getting
into with the TDS 6XX series. These are CCD-based scopes, meaning they
capture in real time at up to 5GS/sec per channel into a CCD FIFO, which is
then digitized at a pretty leisurely rate after a trigger hits.
The upshot of this is that they have a fairly limited memory depth, and
some other acquisition quirks. The selling point for those scopes
originally is that you get the full 5GS/sec for all the channels
simultaneously.
Another thing to watch out for is that the 6XX*A*s don't have a peak detect
mode, which might be a problem - though only if you're looking at the As.

If you contrast the 7XX series, those scopes capture at up to 4GS/sec
across all channels. This is however multiplexed, so the capture rate per
channel drops off as you turn on more channels. With all four channels
turned on, you're down to 1GS/sec each.
You do however have equivalent time capture for repetitive signals, hi-res
for slow signals, peak detect across the board, and (I'd guess) much less
dead time between acquisitions.

I don't know if any of this matters to you - and apologies if I'm preaching
to the choir...

Siggi

On Sun, Dec 15, 2019 at 1:23 PM Reginald Beardsley via Groups.Io <pulaskite=
yahoo.com@groups.io> wrote:

I've seen this error in a number of ebay listings. What does it mean? I
could find nothing in the service manual, 071062702 on Tek's web site. I"d
bought a LeCroy DDA-120, but FedEx dropped it, so I'm looking for another
cheap 1 GHz DSO.

Is this an NVRAM failure with error messages even worse than the 11801?
What is required to repair the fault? So far no search has turned up
useful information.

Reg



Re: Paying it forward with David's Part's Exchange

DaveH52
 

Where would I send the SASE and wish list?
Dave

TDS684C "Cal Fails during initialization".

Reginald Beardsley
 

I've seen this error in a number of ebay listings. What does it mean? I could find nothing in the service manual, 071062702 on Tek's web site. I"d bought a LeCroy DDA-120, but FedEx dropped it, so I'm looking for another cheap 1 GHz DSO.

Is this an NVRAM failure with error messages even worse than the 11801? What is required to repair the fault? So far no search has turned up useful information.

Reg

Re: DC-508 repair advice

radioconnection@...
 

Thanks... I have manuals on order for most of them. Will check to make sure.

I managed to blow up the DM-501 at 2AM playing around (installed it upside down, since I had
the slides removed for servicing.) Another project. I need to go to bed instead of doing bench work at 2AM half asleep.

The only plug in I care about is the tracking generator which is mated with the 497P. The others were free for the taking.
I ended up with the 504 since the 503 was getting filled up rather quickly. But, I am anal and will get them all working
eventually.