Date   

Re: 114xx bezzel buttons

Pete Lancashire
 

I didn't look at text website for service information do you have the URL
or can I just type in 11403 ?

On Fri, Jul 6, 2018, 3:39 PM <jbau@...> wrote:

Not sure exactly what your question is, but I've removed the panel/trim
from several 1140x's to gain access to the crt, using the procedure in the
service manual on Tek's website, and I don't recall encountering anything
unusual.




Re: Questions about TM mainframes with Option 2

Dave Casey
 

You have some conflicting terms.

The 5Axx modules are 5000 series oscilloscope amplifier plugins. They are
not compatible with the TM500/TM5000 series mainframes.

The AM502 is a TM500 plugin. It is not compatible with the 5000 series
oscilloscopes. Some TM500 plugins are not compatible with the TM5000 series
because of power supply issues. There are only a couple of these and they
are listed in the manuals for the TM5000 series mainframes. Most of the
TM5000 series plugins are not compatible with the TM500 series mainframes,
but there are some exceptions to this as well.

For a given TM500/TM5000 plug-in, the manual will have a section on the
rear interface connections. Any connection listed there can be wired to a
rear panel BNC or Dsub connector on the Option 02/Option 12 mainframes.
Sometimes you must move an internal jumper in the plug-in to route the
signal to the rear interface instead of the front panel. It's all in the
manuals.

Dave Casey

On Thu, Jul 5, 2018 at 5:49 AM, PHB via Groups.Io <
mole_org=yahoo.com@groups.io> wrote:

Hi all, first post here.

I'm looking to set up a TM 5000 series mainframe and trying to sort out
the compatibility gotchas before I "pull the trigger" on one of these.

My main interest is in the differential amplifier plugins. I do mostly
audio work and these are the ones I am thinking about:

5A13
5A19
5A20
5A21
5A22
5A26
AM502

Question #1) Will the AM502 work in a 5000 series mainframe? (there seems
to be conflicting information about whether or not 500 series plugins work
in 5000 series mainframes)

Question #2) If I get a 5000 series mainframe with Option 2 (w/ the BNC
connector panel on the back of the unit), would I be able to wire from an
A5xx plugin's single ended output to a BNC connector on the back panel, and
then connect this to a standard o-scope input?

My understanding is that Tek 5000 series scopes' baseline sensitivity is
10 mV/division, so I'm assuming I can set my scope to 10 mV/division, hook
up the BNC out (wired to the A5xx plugin's output) from the back of the TM
5000 series mainframe and just plug it into my scope. Does this sound right?

I really appreciate any advice you have about this.




Questions about TM mainframes with Option 2

PHB
 

Hi all, first post here.

I'm looking to set up a TM 5000 series mainframe and trying to sort out the compatibility gotchas before I "pull the trigger" on one of these.

My main interest is in the differential amplifier plugins. I do mostly audio work and these are the ones I am thinking about:

5A13
5A19
5A20
5A21
5A22
5A26
AM502

Question #1) Will the AM502 work in a 5000 series mainframe? (there seems to be conflicting information about whether or not 500 series plugins work in 5000 series mainframes)

Question #2) If I get a 5000 series mainframe with Option 2 (w/ the BNC connector panel on the back of the unit), would I be able to wire from an A5xx plugin's single ended output to a BNC connector on the back panel, and then connect this to a standard o-scope input?

My understanding is that Tek 5000 series scopes' baseline sensitivity is 10 mV/division, so I'm assuming I can set my scope to 10 mV/division, hook up the BNC out (wired to the A5xx plugin's output) from the back of the TM 5000 series mainframe and just plug it into my scope. Does this sound right?

I really appreciate any advice you have about this.


Re: 114xx bezzel buttons

Precaud
 

Not sure exactly what your question is, but I've removed the panel/trim from several 1140x's to gain access to the crt, using the procedure in the service manual on Tek's website, and I don't recall encountering anything unusual.


Re: MEDIA MAIL (Re: [TekScopes] I have two Tek manuals I don't think I need)

Vince A
 

David
You got to be vigilant at the PO, know before you go! Google and asking
questions here really helps you out. If it sounds fishy go back another day
or see another PO. Use a different guy/girl on another day at the PO.
Vince :)

---
-vince :")
Thank you for your reply
Got computer troubles, We have the answers!
Visit Starlite Systems
Starlitesystems Dot Com
My other computer cost's $35.00. Ask me about it

On Tue, Jul 3, 2018 at 9:13 AM, David Berlind <david@...> wrote:

So, just fyi.. I went back to the PO and asked "What about the media mail
option?" ... Bascially, it was "Oh, that? $3.17" (for what was originally
$11.80). Pretty shocked that, like Dave says, they neglected to offer that
option in the first place when I requested "least expensive shipping
option."

On Tue, Jun 26, 2018 at 2:38 PM, Artekmedia <manuals@...>
wrote:

They are trained NOT to tell you. You have to ASK for it or as the
previous poster said it is available through PayPal shipping includes
tracking. Tell them it just a "manuscript" and DO NOT INCLUDE any
correspondence. You put a one page letter in there and suddenly it is a
1st
class mail

For a description google is your friend

-DC
ArtekManuals.com

On 6/26/2018 1:40 PM, John Griessen wrote:

On 06/26/2018 11:26 AM, David Berlind wrote:

what is "media mail?" when I go to the post office, they offer me a
slew
of options.. but "media mail" is not one of them.
use paypal shipnow to get a label for media mail with a little extra
discount.




--
Dave
Manuals@...
www.ArtekManuals.com








Re: DC 508

Ed (SCSKITS)
 

The option 7 on the DC508 increases its value as it allows it to work with the TM500 series tracking generators. If it has option 1, the oven oscillator, that also increases the value. The oven oscillator is a silver box at the rear of the PCB, easy to see if it is there.

Also, you do not need to be a HAM operator to play with the old test equipment. You do need a HAM license if you want to build transmitters for the HAM bands, no license needed for receivers and other gear..


Owon SDS7102 Deep Memory Digital Storage Oscilloscope

Vince A
 

Morning Gentleman

Well I finally bit the bullet and bought a 100mghz Scope. Now I need to
learn how to use it. :) Is there a oscilloscope for dummies book out there
or just some quick pointers you guys would like share with me? I am sure
a'll have a LOT of questions but I don't have a clue what to do next and
what all the fancy controls could or should do for me? I am mostly into
vacuum tube audio and some solid state electronics. The package did come
with some really nice probes and almost like a very small jewelers type of
screwdriver driver? And Manuel

What I would like to do first with this is to measure the Pre-amp signal of
the 12AT7 to see if i am getting a clean signal to the output of my first
pair of 6L6 tube amp I built using Russian output tubes. And if not, how
could I clean that up by doing other things to get the signal better though
the path? I don't learn very well by reading books guys, I am more of a
hands on guy. When I read books, the pages melt into each other after about
2 pages. So reading online is much better for me. That is why I like the
combined knowledge here.

How did you guys learn how to use a Scope? How did you go about getting
familiar with a Scope? What was your very first Scope? This is my VERY
first Scope of anything guys.
Thanks loads
Vince :)
---
-vince :")
Thank you for your reply
Got computer troubles, We have the answers!
Visit Starlite Systems
Starlitesystems Dot Com
My other computer cost's $35.00. Ask me about it


TDS744

Vintage Test
 

Hi All,

I've just laid my hands on one of these for a decent price, knowing it didn't work. The owner said it worked one day and jus did tick, tick etc the next - a familiar story! When I stripped it, I found a load of duff diodes/transistors and two-way devices in the standby circuit, which I replaced. I powered it up - still the same, except that the front-panel LEDs all flashed too now. On a whim, I disconnected everything from the processor board, and lo and behold -it powered up and worked properly into a VGA monitor. Now, to the problem; the CRT driver assembly is clearly faulty, but there's no schematic for it anywhere. I have the manual for TDS540, which has a lot of similarities, but is monochrome and not colour, so the CRT driver assembly is way different. Do any of you happen to have any experience with these 'scopes?

Cheers,
--
you can never have enough oscilloscopes, DMMs, valve testers or soldering irons . . .


Re: Dead 7603

Vintage Test
 

Hi Jim,

On a 7603 - and I've sorted a lot of them - the main culprits of no power-up, when you've checked all the usual suspects, is the power supply filter capacitors. To check these, you will have to remove the raw voltage power unit, situated at the bottom rear of the 'scope. There are three large cross-head screws on each side of the chassis, which allows you to carefully draw out the mains transformer and attendant rectifier/capacitor PCB. If the capacitors are all metal cased, it is likely that these are faulty now and need replaced. Looking at the rear of the 'scope, there is a rectifier attached to the bottom left corner , underneath the PCB. This is the +/- 15V rectifier and the two caps nearest the left edge are the reservoirs. If these don't have about 24V DC on them, then nothing else will work. Have a look at this and get back to us.
--
you can never have enough oscilloscopes, DMMs, valve testers or soldering irons . . .


Re: Kludging together a TDR or similar?

 

hi guys, thanks for the replies.

I don't have the parts necessary to build a pulser. Can I use the
mainframe's pulse generator?

The cable goes from a local distribution box, part underground, part
overground, and then goes to a wall socket used only by my modem. There is
no splitting and no neighbors using this. The run in my yard that is
overground is maybe 30m at most.

There was no thunder when this stopped working, just heavy rain.

On Fri, 29 Jun 2018, 06:53 Michael A. Terrell, <mike.terrell@...>
wrote:


If lightning hit that drop, there is likely more than one damaged area, or
even long stretches of the shield is gone. I worked in CATV repairing the
electronics and I have seen damaged coax where the entire outer conductor
had vaporized. The worst damage is usually at sharp turns, like the corner
of a building. We often had to run temporary drops to replace buried lines,
until or Ditchwitch was available to bury it. This was in the Cincinnati
area, so lightning was a big problem. It wiped out thousands of power
transformers in the set top boxes, as well.


There should be a tap to reduce the signal level to the modem which may
also be damaged. It is usually at the house end of the drop, and it is
easily damaged by lightning. Mine is 17dB.

Michael A. Terrell
--


-----Original Message-----
From: Kevin Oconnor <@KO3Y>
Sent: Jun 28, 2018 11:04 PM
To: TekScopes@groups.io
Subject: Re: [TekScopes] Kludging together a TDR or similar?

TV coax problem....
As noted, u don’t indicate the run length u have. If it goes from ur
house terminal to the typical green post in the yard and it is 100ft or
less, it would be more efficient to just run a new coax over the ground to
verify operation.
If you have buried coax with a fault, I wouldn’t trust any of it.

Kjo



Double messages

Jim Ford
 

To whom it may concern, now that I  can post to TekScopes, I get double copies of every message.   Very annoying!  How do I get it to stop?
Thanks!
Jim


Sent from my Verizon, Samsung Galaxy smartphone


Re: 2215 avoiding future CRT damage

tom jobe <tomjobe@...>
 

Thank you Hakan!
That is an excellent test that will probably work on other scopes to get an idea if the CRT is good.
Your contributions to Tekscopes are always good, and usually you are the only one on the planet that has the information you post.
tom jobe...

On 7/6/2018 11:08 AM, zenith5106 wrote:
So my CRT / filament is cooked then, by now, after nearly 40 years of use...
oh well... not much I can do about it, but at least if I find the trace too
dim at fast sweep speeds (haven't checked specifically for that yet), I will
know why and won't waste trying to "fix"/troubleshoot something that just
can't be cured ! :-(
At the time we had a simple procedure at our Field Office to determine
if the CRT should be replaced or not. I don't recall if it was an official procedure
or something we came up locally. We set the T/B to the fastest sweep speed with
X10 Magnifier, Hold Off to max, Auto Trig and no input signal. If the trace could
be seen with Intensity set to max , even just quite dim, the CRT was considered OK.

/Håkan


Re: 7A13 Comparison Voltage coarse knob

 

The Tek part number for the knob is 366-1084-00 and it just may be used on
some other item of equipment.
This knob was also used on AM502, 26A2, 5A19N, 5A22N, 7A22, 7S11 & 7S12.

/Håkan


Re: 2215 avoiding future CRT damage

Vincent Trouilliez
 

On Fri, Jul 6, 2018 at 11:08 am, zenith5106 wrote:
We set the T/B to the fastest sweep speed with
X10 Magnifier, Hold Off to max, Auto Trig and no input signal. If the trace
could be seen with Intensity set to max , even just quite dim, the CRT was
considered OK.
OK... just perform said test, here is the result :

https://groups.io/g/TekScopes/photo/62323/0?p=Name,,,20,1,0,0

I adjusted the ISO setting of the camera so that the brightness on the pic actually matches the brightness level that my eyes perceive.
That's under normal conditions : a couple fluorescent tubes above the bench for lightning, and me sitting right in front of the scope, at a close but reasonable distance, I mean I did glue my face to the screen...

Looks like an "acceptable" tube then ? I mean it's very difficult to see, and any dimmer than this, would make it invisible but... as you say, you can still see the trace.

Plus, if I turn the var hold-off to the min, brightness comes back to normal, the trace becomes perfectly visible, no effort needed whatsoever.

I will admit I don't really master the var hold-off feature, I am not sure I get why it impacts the brightness level.. but obviously it does. Will need to do some reading and experimenting ! ^^

Anyway, looks like my CRT is not worn out in the end, what do you say ? That's good ! :-)
I mean I have sooo many problems left to fix on this scope, I really didn't need yet another (costly ! ) problem on my list ! LOL


Regards,


Vincent Trouilliez


SC504 Gain Knob Removal

Larry McDavid
 

I have a SC504 scope module with a damaged Channel 1 knob skirt and I have a Tek replacement skirted knob. I can remove the small red central knob by loosening a set screw.

How is the skirted knob removed? I've tried gently pulling on it without success. The knob has a long plastic shaft with some internal ribs to fit over another shaft but I don't know how the knob is attached.

Anyone done this?

--
Best wishes,

Larry McDavid W6FUB
Anaheim, California (SE of Los Angeles, near Disneyland)


7A13 Comparison Voltage coarse knob

Colin Herbert
 

I have a later model 7A13 with the LED volts readout. It has the serial # 104748 and was made in Guernsey. It works fine, but the Comparison Voltage coarse knob is not the correct one. Would anyone have one of these to spare? Being where they are located, they are pretty vulnerable to damage and so may be a bit rare (hence the silly prices sometimes quoted). The Tek part number for the knob is 366-1084-00 and it just may be used on some other item of equipment. Reasonable costs understood.
TIA, Colin.


Re: 2215 avoiding future CRT damage

 


So my CRT / filament is cooked then, by now, after nearly 40 years of use...
oh well... not much I can do about it, but at least if I find the trace too
dim at fast sweep speeds (haven't checked specifically for that yet), I will
know why and won't waste trying to "fix"/troubleshoot something that just
can't be cured ! :-(
At the time we had a simple procedure at our Field Office to determine
if the CRT should be replaced or not. I don't recall if it was an official procedure
or something we came up locally. We set the T/B to the fastest sweep speed with
X10 Magnifier, Hold Off to max, Auto Trig and no input signal. If the trace could
be seen with Intensity set to max , even just quite dim, the CRT was considered OK.

/Håkan


Re: 468 DSO Calibration/troubleshooting

Colin Herbert
 

Hi Fabio,
I have now completed the "STORAGE HORIZONTAL" section of the Calibration procedure. Absolutely everything checked out as well within tolerance, until I got to the item 4 "Check Jitter Correction". This all checked out good, too, except that while there was no jitter generally on the right-hand side of the screen, it could be seen for the first two or so cycles. This was most obvious when using 50 kHz sinewave, 20 us/DIV, 2 mV/DIV, and X10 MAG. Again the centre part of the display was within specification, but the first two or so full cycles showed noticeable jitter.
Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Fabio Trevisan
Sent: 05 July 2018 23:35
To: TekScopes@groups.io
Subject: Re: [TekScopes] 468 DSO Calibration/troubleshooting

OK Collin,
Now it makes some sense that there's no instruction to set the "NON STORE" to Off.
It's set to off at Section #2, step i, when it asks to set storage mode to "NORM".
Fine then... it means that during this step the scope is in full "digital" mode, and the waveform we're looking at is the digitized one.
regarding your triggering problem... well it does seem that it's being caused by something else, because if we think of the entire STORAGE triggering check / calibration instructions, it's rather minimalist.
It basically covers checks of the storage window mode (post trig and pre trig) and the A and B gates, and this one and only adjustment (for Storage NORM DC Bal).
I don't think the jitter is caused anywhere here, and from your description of when you try to align the rising edge of the 50KHz signal to the 2nd graticule line. that it comes out of trigger, that doesn't really make sense.
This fact alone suggests that the triggering point of the digital section is, level wise, completely unmatched to the analog section... Because, if you display this very same 50KHz in analog mode in this time base of 0.1ms, and set the trigger point of the 1st cycle at the 0 crossing... the 6th cycle will be dead on the correct spot (and so every one of the displayed cycles will be crossing the center line exactly at each of the HOR minor divisions).

And talking about your jitter / alias... Well, at 50 points per division (as the storage mode is expected to acquire/display), this 50KHz signal at 0.1ms/div would still have a healthy 10 points / cycle so, a jitter of +- 1 sample would be 1/10th of a minor division and, at this acquisition speed, much lower than the limit of the scope, I wouldn't expect nothing worse than +-1 sample.
Definitely there shouldn't be aliasing at play here either (with this healthy 10 sample / cycle). If much, a jitter of +- 0.05 div.

I will try to dig a little bit on the circuitry to see if I can come with any suggestion... but basically, from this point on, I don't think you're getting the calibration instructions wrong...or that you're missing something. It really seems that something is not working well on the digital triggering section.

KRgrds,

Fabio


On Thu, Jul 5, 2018 at 12:03 pm, Colin Herbert wrote:


Hi Fabio,
The difficulty with looking at a 50-KHZ at 0.1 ms/div is that the triggering
is really not that great and some aliasing is evident. I do see your point,
though and I have given it a go, but the jitter doesn't help and I can't get
any of the rising edges to align with the second vertical graticule line using
the A trigger level. What I have just noticed, though, is that the stable
triggering only takes place between the zero crossing and just after the next
crest and, oddly, it seems to be the wrong slope! That is, with the slope
switch in the + position, triggering is on the falling edge and vice versa.
The non-storage triggering is exactly the way it should be. I think there may
be something wrong, here.

With regard to the storage/non-storage buttons. They are all latched. Pushing
any of the storage buttons (Norm, Envelope, Avg and Save) releases any of the
other three and the Non-stope button. Pushing Non-store releases the storage
buttons. I think that's what you might expect.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Fabio
Trevisan
Sent: 05 July 2018 19:29
To: TekScopes@groups.io
Subject: Re: [TekScopes] 468 DSO Calibration/troubleshooting

Hi Collin,
You're welcome... It's more challenging to solve puzzles when you're not
actually on the driver's seat.

Right out of the box I can't see a flaw in your reasoning... I confess I
didn't make the math that 50KHz at 0.1ms/div would give something like 50
cycles...
I agree it seems really silly to try to adjust the crossing of the waveform at
that particular cross-hair, with such dense waveform. 20us/div would really
make more sense in that respect.
Or maybe... they really meant it to be densely packed with cycles, so that we
could "disregard" the rising edge slope (at that density, the rising edge is
almost a vertical line, that we would only need to make sure is aligned to the
vertical graticule).
If there's no mistake in the instructions, it should have - at least -
mentioned that we should be aligning the leading edge of the 6th displayed
cycle, and not just "Align the leading edge"... that's too vague to say the
least.

Now, regarding my doubt whether the procedure is meant to be carried out in
storage mode or not, I have a doubt: How does the "NON STORE" button mode
works (mechanically)?
Is that when we press some of the STORAGE MODE buttons (say, NORM, ENVELOPE,
AVG, or SAVE), does it make the "NON STORE" button to pull out?
That would explain why there's no instruction to explicitly set the "NON
STORE" button to OFF (out).

KRgrds,

Fabio





On Thu, Jul 5, 2018 at 10:07 am, Colin Herbert wrote:


Hi Fabio,

Firstly, thank you for giving this some thought. Some of my observations:

1. Check STORAGE WINDOW Operation. This all works fine, except that when
going
to "d. set STORAGE WINDOW PRE TRIG" the triggering jitter becomes quite
noticeable; it is less so in POST TRIG.
2. Check A and B+ GATES.
Set NON-STORE On; A TIME/DIV 0.1 ms; B TIME/DIV 20 us. All works as
expected,
except for the jitter when in NORM storage mode - Non-storage there is a
rock-solid trace with no jitter at all.
3. Check/Adjust Storage NORM Trigger DC Balance (R126)
Set CH1 VOLTS/DIV 5mV; HORIZ DISPLAY A. Remember that the A sweep is still
at 0.1 ms/DIV and so there are some 50 full cycles of a 50-kHz sine-wave.
Trying to set anything with the A TRIGGER LEVEL is nigh on impossible. I
think
the sweep should be A at 20 us/DIV, showing ten full cycles. The problem now
is that trying to align the leading edge with the second vertical graticule
line at the horizontal centre graticule line results in loss of triggering.
This alignment can be done using the horizontal position control, however.
When aligned in this way, switching between AC and DC Trigger coupling of
the
A sweep shows no movement.

Have I achieved what is required, or am I again missing something?

It is a pity that Reed Dickinson hasn't seen this thread, as I think he is a
bit of an expert on the 468 and he certainly holds them in some respect.
However, I repeat my thanks to you, Fabio.

Regards, Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Fabio
Trevisan
Sent: 05 July 2018 17:09
To: TekScopes@groups.io
Subject: Re: [TekScopes] 468 DSO Calibration/troubleshooting

Hello Collin,

Regarding my last statement (the P.S.), that this check should be made in
"non-store" mode, I'm giving second thoughts at it.
I`m not sure yet, but I just found out that there are two "NORM triggering
DC
balance" adjustments... one on the analog side and another one on the
storage
side...
So, it seems weird that we could even be able to adjust the STORAGE NORM
triggering DC balance in any mode that isn't a storage mode.

On the other hand, I may still be correct on my initial assumption, because
the flow of the instructions and the explicit instruction to set "NON STORE"
to ON, at Section #2, step g.
I`m only not sure if that adjustment (R126) would be even doable while in
NON
STORE mode...

I`ll get back to it as soon as I can dig a little bit more on the manual.

Rgrds,

Fabio



On Thu, Jul 5, 2018 at 08:23 am, Fabio Trevisan wrote:


Hello Colin,

I don't own a 468 and have no particular experience with it, but I have
the
manuals (in PDF form) and since your doubt is related to understaing of
the
instructions, I decided to give it a try and see if I could make sense out
of
this part of the instruction that you're doubtful at.

At this step, the only thing that doesn't seem really clear to me is why
it
asks specifically to align the leading edge of the sine-wave to the 2nd
graticule line (why the 2nd.???)..
But, regardless of not understanding that, I think the purpose of this
calibration step is clear... Is to make a "differential" assessment of the
triggering level shift, between using AC triggering coupling and DC
triggering
coupling...
So, basically what it is asking for, is for you to make sure that the
triggering point doesn't move between AC and DC coupling, when you're
actually
feeding a signal that already swings about 0V...
I mean, if the input signal is swinging about 0V (i.e. if it doesn't
contain
a
DC level), changing the triggering coupling from AC to DC should not make
any
difference, and therefore, the triggering point, whatever it was set to
before
switching to DC, the wave form must remain in the same position.
If the waveform shifts horizontally, than it means that the triggering
point
"seen" by the triggering circuit have changed when switching over from AC
to
DC, which it shouldn't.

Giving second thoughts to the instruction, it's not much different than
adjusting the Norm DC triggering balance of a 464 (which I own), and
probably
the same as the 465 (more common).
the only difference is that, for the 464, we're asked for looking at the
very
beginning of the sweep, and not at the 2nd vertical graticule line (i.e. 1
division after the sweep actually started).
This difference in procedure (from the non store scope), doesn't actually
matter to what the DC balance adjustment is doing (which is nothing more
than
shifting the DC level of the actual signal being fed to the triggering
input
buffer, with the intent to making it match with the level of the same
signal
when AC coupled).
The only reason I think they're doing that (adjusting at the 2nd vertical
graticule line iso at the beginning of the sweep), is because this point
is
about the actual point on the display where the triggering point of the
digital stored signal is supposed to be displayed.

P.S. Also note that, although this adjustment step is called "Check /
Adjust
STORAGE NORM Trigger DC balance", by following the instructions coming
from
the previous steps, I understand (it's my conclusion though) that this
adjustment is not meant to be performed in storage mode.
The previous section, "#2 Check A and B+ gates", at step g. it asks you
to
set the "NON STORE" button to ON position... and then it doesn't ever ask
you
to change it to OFF, so, by the flow of the procedure, the #3 check is to
be
performed in NON STORE mode.
Therefore, this jitter that you mentioned you're getting on storage mode
should not interfere with you performing this adjustment (because it's
done
in
NON STORE mode).

Now, talking about this jitter... I don't think it has any relation to
this
adjustment, and it's probably not going to be fixed by setting this
adjustment
right.

I hope I was able to help you somehow.

KRgrds,

Fabio




On Tue, Jul 3, 2018 at 10:08 am, Colin Herbert wrote:


I have got through the calibration/adjustment procedure up to "Storage
Triggering". My problems now are twofold. I can't understand Section #3
-
"Check/Adjust Storage NORM Trigger DC Balance" (P4-89 in the manual).
Apart
from the fact that the storage triggering has a deal of jitter, I can't
get
step (e) to make sense. I am supposed to "use the A TRIGGER LEVEL to
align
the
point where the sine-wave's leading edge intersects the center
horizontal
graticule line with the 2nd vertical graticule line". Because of the the
jitter and the fact that the triggering fails when I adjust the A
TRIGGER
LEVEL, this is impossible. Am I doing something wrong (quite possible),
or
is
the manual giving me the wrong information? Help!
If anyone has an idea of how I can eliminate the jitter, that would be
nice,
too. There is no jitter in "NON STORE" operation. I was hoping it would
just
be something simple like a switch needing cleaning, or an IC needing
reseating, but nothing has given me any clues so far...
Colin.







Re: PG506 trig out jitter?

Siggi
 

On Fri, 6 Jul 2018 at 04:17 Albert Otten <aodiversen@...> wrote:

This time I didn't study the manual but just looked dual-trace at the
waveforms of trigger and Hi Ampl. Trigger always properly terminated. But I
confused the traces when saying that the down-going edge of the trigger was
slow; that was the terminated Hi Ampl. I checked again. It's nice to see
how the terminated Hi Ampl negative going edge "climbs" along the same path
to its maximum, when you vary the amplitude knob.
When Hi Amp is unterminated and loaded with coax cable then the rise and
fall are almost symmetrically, both governed by the same RC time as
mentioned already. I found about 200 ns rise time when using about 1 m coax
which measured 128 pF including the scope input. This seems not far off at
600 Ohm.
I took a look at the (un-terminated) hi-rise output of the PG506 and it was
pretty slow - I forget how slow. It doesn't help when the TD pulser is the
"load" on there, and
when I look at the time from trigger out to TD pulser edge, it's on the
order of 600ns. I guess it's a bit of hack to use the hi-rise output of the
PG506 as a power supply for a pulser :). Clearly this arrangement won't
work for looking at the TD edge with a sampler. Either I need a delay line,
or else a different TD pulser, one with a stable pre-trigger.


Re: Another Soviet plugin: 7CT1N curve tracer equivalent Я4С-92

cmjones01
 

Yes, I do have the manual which includes schematics (and even the winding details of the power converter transformer!). I definitely want to scan the schematics when I get chance. My old office had a really good scanner, and sadly the new one doesn't so I'll have to go to the copy shop down the road.

The manual runs to 78 pages plus schematics which are all odd sizes.

Incidentally, browsing the manual just now, it becomes clear how the X-Y mode is implemented. There's an X output on pins 15A/B of the plugin interface, and a Y output on pins 25A/B. The vertical amplifier plugin also has its signal output on 25A/B, which differs from the Tek allocation (11A/B). Trigger output is on 13A/B the same as Tek, though.

The mainframe has an X-axis switch for each plugin as well as a Y-axis switch, it turns out, so plugins are allowed to generate both X and Y signals. Interesting.

I wonder how the
Chris

On Fri, Jul 6, 2018 at 06:32 am, Artekmedia wrote:
I think you said you had the manual as well when you get a chance
perhaps you can scan the schematics and post them here or on KO4BB. It
would be interesting to see how they accomplished some of these features

Cheers
Dave
manuals@...

On 7/6/2018 8:50 AM, cmjones01 wrote:
Good question, I wondered that too. I've just tried out the plugin in a
mainframe for the first time and noticed two things. First, it has readout!
The current and voltage per division are displayed, together with the mode -
'n' for N-type FETs and 'npn' for NPN bipolars, and so on.

It doesn't need help from any other plugins to do X-Y mode, so the plugin
interface must have been changed from the Tek one (though I've already
observed that a lot of the signals are in the same places). The plugin has a
'measure' button which seems to enable and disable it. I've no idea how it
interacts with having a timebase in the horiztontal slot. Perhaps it's handled
gracefully, perhaps it all becomes a terrible mess, I'll have to try it at
some point.

There's a picture of the tracer in action on a pair of IRLB3034 MOSFETs I
found lying around on the bench. They're switching MOSFETs so they don't make
very nice curves, but it's possible to see that the two start to switch on at
different gate voltages. I presume the looping visible is because the slope of
drain current against gate voltage is so steep, so even the tiniest variation
in gate voltage results in visible loop.

https://groups.io/g/TekScopes/photo/62315/3?p=Name,,,20,1,0,0

A neat feature is that device 'A' is displayed with a solid trace and device
'B' gets a dotted one so you can tell them apart. The buttons on the adapter
each *disable* one of the devices momentarily, otherwise they're both
displayed all the time.

Chris

On Fri, Jul 6, 2018 at 05:22 am, Dave Casey wrote:
Interesting....but how does it do X-Y mode?

Dave Casey

On Fri, Jul 6, 2018 at 6:01 AM, cmjones01 <chris@...> wrote:

It was with some excitement that I unpacked today's new acquisition. From
a surplus dealer in Ukraine, via some helpful friends in that country,
I've
got my hands on the Я4С-92, the Soviet version of the 7CT1N curve tracer
plugin. in its original transit case with an almost complete set of
adapters and its original manual and calibration certificate dated 1988!
It's not new old stock and has definitely been used, but the seals on the
plugin are intact.

There are five different adapters which fit on the front of the plugin,
one for two-lead devices and the others for various transistors and FETs.
They all have two buttons for comparing devices, and little plastic lids
with a microswitch to enable the tracer when they're safely closed. The
only bits which seem to be missing are a couple of connector blocks which
seem to have something to do with the calibration process, but I'm not too
worried about those.

I have to say that the Soviet makers have the edge over Tektronix when it
comes to the sheer robustness of the packaging. It's a plywood box with
steel corner protectors, lined with polystyrene and foam to fit all the
parts. The whole thing weighs about 10.5kg (23lb).

Pictures here:
https://groups.io/g/TekScopes/album?id=62315

It'll be fun learning to drive it, and it will be really useful in the
lab, too.

Chris




--
Dave
Manuals@...
www.ArtekManuals.com