Re: 497P owners-- bad RADIAL lead through-hole tantalum caps!
https://nepp.nasa.gov/files/29192/NEPP-TR-2018-Teverovsky-T22-Capacitors-TN52048.pdf
-----Mensaje original-----
De: TekScopes@groups.io [mailto:TekScopes@groups.io] En nombre de Chuck Harris
Enviado
https://nepp.nasa.gov/files/29192/NEPP-TR-2018-Teverovsky-T22-Capacitors-TN52048.pdf
-----Mensaje original-----
De: TekScopes@groups.io [mailto:TekScopes@groups.io] En nombre de Chuck Harris
Enviado
|
By
Miguel Work
·
#157275
·
|
|
Re: Isolation???
Using the two-probe method that Mike suggested is the usual way of making a
differential measurement (when expensive diff probes are not available, or
when the voltages are too high for them).
One
Using the two-probe method that Mike suggested is the usual way of making a
differential measurement (when expensive diff probes are not available, or
when the voltages are too high for them).
One
|
By
dnmeeks
·
#157274
·
|
|
Re: 7B92A time-base plug-in odd behaviour
Not totally the case, Albert. As I said, the instability can be present after warm-up and can be stopped or induced by "wiggling" the Time/Div knob. Specifically, if the delaying or delayed time-base
Not totally the case, Albert. As I said, the instability can be present after warm-up and can be stopped or induced by "wiggling" the Time/Div knob. Specifically, if the delaying or delayed time-base
|
By
Colin Herbert
·
#157273
·
|
|
Re: 497P owners-- bad RADIAL lead through-hole tantalum caps!
That type of capacitor is a wet slug tantalum capacitor, and
it uses sulfuric acid as its electrolyte. Over time, the
positive lead gets etched away, and the acid leaks out. Lots
are starting to
That type of capacitor is a wet slug tantalum capacitor, and
it uses sulfuric acid as its electrolyte. Over time, the
positive lead gets etched away, and the acid leaks out. Lots
are starting to
|
By
Chuck Harris
·
#157272
·
|
|
497P owners-- bad RADIAL lead through-hole tantalum caps!
During trouble shooting, I've found several bad 50 uF at 25 VDC radial lead (not SMD!) on the first few boards I pulled for inspection.
All of those caps showed severe leakage and many had open leads
During trouble shooting, I've found several bad 50 uF at 25 VDC radial lead (not SMD!) on the first few boards I pulled for inspection.
All of those caps showed severe leakage and many had open leads
|
By
radioconnection@...
·
#157271
·
|
|
Re: Latest firmware for TDS210 and TDS2CM?
Hmm, good info. I think I probably have a late firmware then, it had the earthing recall fix already done too, so it was post recall as well.
I guess I'll leave this as it is unless something turns
Hmm, good info. I think I probably have a late firmware then, it had the earthing recall fix already done too, so it was post recall as well.
I guess I'll leave this as it is unless something turns
|
By
Jared Cabot
·
#157270
·
|
|
Re: 7B92A time-base plug-in odd behaviour
Hi Colin,
Since most of the problem disappears after warming up I guess it's more like a semiconductor issue.
Do you have the model with "Alt Off" at the ccw position of the TRACE SEP knob, S/N
Hi Colin,
Since most of the problem disappears after warming up I guess it's more like a semiconductor issue.
Do you have the model with "Alt Off" at the ccw position of the TRACE SEP knob, S/N
|
By
Albert Otten
·
#157269
·
|
|
Re: 7B92A time-base plug-in odd behaviour
Thanks for the info, guys. I have now located where these switches go, but I am unclear as to what their functions are. Perhaps I don't really need to know, after all.
I am still perplexed by the odd
Thanks for the info, guys. I have now located where these switches go, but I am unclear as to what their functions are. Perhaps I don't really need to know, after all.
I am still perplexed by the odd
|
By
Colin Herbert
·
#157268
·
|
|
Re: 7904 excessive shadow/flare
O.K.! Thanks to everyone for the comprehensive answers, very instructive to me!
Undoubtedly the driving circuits of the 7904 and 7804 CRTs are completely different and this could partly explain the
O.K.! Thanks to everyone for the comprehensive answers, very instructive to me!
Undoubtedly the driving circuits of the 7904 and 7804 CRTs are completely different and this could partly explain the
|
By
unclebanjoman
·
#157267
·
|
|
Re: Tektronix 564 WW resistor for -100V supply
I would use a low Tempco metal film, say 15 to 50ppm. The only reason I can think of using WW in that location is that carbon comp and WW were the primary choices when that scope was designed, and
I would use a low Tempco metal film, say 15 to 50ppm. The only reason I can think of using WW in that location is that carbon comp and WW were the primary choices when that scope was designed, and
|
By
Dave Hills
·
#157266
·
|
|
Re: 465B Attenuator servicing
Easy and cheap enough to get a pump sprayer bottle
from the store, and fill it with IPA. That is what
I do when I need to flood a circuit/switch/keyboard
with IPA.
-Chuck Harris
Colin Herbert via
Easy and cheap enough to get a pump sprayer bottle
from the store, and fill it with IPA. That is what
I do when I need to flood a circuit/switch/keyboard
with IPA.
-Chuck Harris
Colin Herbert via
|
By
Chuck Harris
·
#157265
·
|
|
Re: 465B Attenuator servicing
OK, so the existence of spray-can IPA is new to me. Perhaps I should see if
I can get it in the UK, but it is cheap enough anyway and I apply it with
cotton-buds, tissues or slivers of paper,
OK, so the existence of spray-can IPA is new to me. Perhaps I should see if
I can get it in the UK, but it is cheap enough anyway and I apply it with
cotton-buds, tissues or slivers of paper,
|
By
Colin Herbert
·
#157264
·
|
|
Re: Tektronix 564 WW resistor for -100V supply
The choice of wire-wound was probably for temperature stability of the -100 volt power supply. Since R617 forms a voltage divider with pot R616 and R618, both also specified as wire-wound, the
The choice of wire-wound was probably for temperature stability of the -100 volt power supply. Since R617 forms a voltage divider with pot R616 and R618, both also specified as wire-wound, the
|
By
Trevor
·
#157263
·
|
|
Re: Craigslist- Northern Virginia Tek 545B for sale
The description says:
"Tektronix Oscilloscope from the 1960s or early 1970s complete with rolling stand and leads. When I plugged it in the lights came on and the screen works. As far as it working
The description says:
"Tektronix Oscilloscope from the 1960s or early 1970s complete with rolling stand and leads. When I plugged it in the lights came on and the screen works. As far as it working
|
By
Colin Herbert
·
#157262
·
|
|
Re: 465B Attenuator servicing
Thanks for your feedback.
Whilst I admit the specific tin I've just used does only have "Ambersil Electronic Safety Solvent" written on it I have nevertheless used it extensively for years. It has a
Thanks for your feedback.
Whilst I admit the specific tin I've just used does only have "Ambersil Electronic Safety Solvent" written on it I have nevertheless used it extensively for years. It has a
|
By
NigelP
·
#157261
·
|
|
Re: 465B Attenuator servicing
Hi, Colin,
IPA IS available in spray cans e.g.
Hi, Colin,
IPA IS available in spray cans e.g.
|
By
Redguuz
·
#157260
·
|
|
Re: Isolation???
What's immediately jumping out at me is that you have a 7904A with a 7A26. You can put a scope probe on channel 1 to the positive DC output, and the scope probe from channel 2 on the 41v test
What's immediately jumping out at me is that you have a 7904A with a 7A26. You can put a scope probe on channel 1 to the positive DC output, and the scope probe from channel 2 on the 41v test
|
By
Harvey White
·
#157259
·
|
|
Re: Isolation???
The rule of thumb when isolation is required is "Don't isolate the instrument - isolate the DUT." Plug the Motorola supply into an isolation transformer - a true isolation transformer like my RCA
The rule of thumb when isolation is required is "Don't isolate the instrument - isolate the DUT." Plug the Motorola supply into an isolation transformer - a true isolation transformer like my RCA
|
By
Dale H. Cook
·
#157258
·
|
|
Re: Isolation???
Hi Mike,
This is where it pays to have a 7A13 Differential Comparator and a P6055 differential probe set if you are going to see some high frequency signals (DC to 110MHz) when you do
Hi Mike,
This is where it pays to have a 7A13 Differential Comparator and a P6055 differential probe set if you are going to see some high frequency signals (DC to 110MHz) when you do
|
By
Dennis Tillman W7pF
·
#157257
·
|
|
Re: Isolation???
Mike:
As a suggestion: Use the 7a26 with two x10 probes. Set the 7a26 with
"DISPLAY MODE" set to "ADD" and with "CH 2 POLARITY" set to "INVERT".
Connect the two probe ground leads (or
Mike:
As a suggestion: Use the 7a26 with two x10 probes. Set the 7a26 with
"DISPLAY MODE" set to "ADD" and with "CH 2 POLARITY" set to "INVERT".
Connect the two probe ground leads (or
|
By
Mike Dinolfo
·
#157256
·
|