Re: 2440 CCD Problem - HELP!
Rich Miller
Hi Peter:
toggle quoted message
Show quoted text
To answer your question, over by the gain cell board's; two SD5000's Switching IC's, one of which is common to channel one. I investigated this IC, and could find nothing wrong. I then brushed my hand against the CCD hybrid heat sink, and was surprised how hot it was. I did have a small muffin fan blowing over the circuit board, but I don't think it was enough. As a result, I now have large blower in place, and things seemed to have stabilized. I am going to attempt CCD calibration one more time, and hopefully this is the answer to the problem.
-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf Of Peter Gottlieb Sent: Saturday, September 25, 2010 4:35 PM To: TekScopes@... Cc: Rich Miller Subject: Re: [TekScopes] 2440 CCD Problem - HELP! Do you think the fault is temperature related or time running related? I don't have my 2440 manual here but what is common to all 4 "sides"? Could an input start getting a DC drift? Peter On 9/25/2010 1:35 PM, Rich Miller wrote:
------------------------------------ Yahoo! Groups Links |
|
|
|
Re: 7704A Saga continued.
John Griessen
On 09/25/2010 04:07 PM, Jerry Massengale wrote:
I have an attitude problem at the moment. I like the idea of connecting the clamps to ground rather than the + and - 15V.Sounds like a reasonable mod to do. Or you could clamp with two diode drops each side. If you're talking about the signal going into an op amp one diode drop would be a fine protection circuit and not bother the signal. John |
|
|
|
Re: 7704A Saga continued.
Jerry Massengale <jmassen418@...>
--- On Sat, 9/25/10, Göran Krusell wrote:
|
|
|
|
Re: 7603
Tom Miller <tmiller@...>
Have you verified all the power supply voltages are
good?
Tom
|
|
|
|
Re: 7603
jvanderwall1941
Just to narrow it down, can you use a different vertical plug-in? Have you tried both channels of the 7A26?
toggle quoted message
Show quoted text
What do you see when you either free-run the sweep or trigger it from the line and activate the beam finder? Regards. Jonathan
--- In TekScopes@..., "milesdj8" <bwex@...> wrote:
|
|
|
|
7704A Saga continued.
Göran Krusell <goran.krusell@...>
Jerry,
I had a very similar problem with my 7904. The IC was broken in some way, the input offset was 1V, and my guess was that the circuit was destroyed by an input pulse being too high. I exchanged the IC and moved the two protecting diodes to clip around ground instead of +/-15V. There after no problem. Göran |
|
|
|
Re: 2440 CCD Problem - HELP!
Peter Gottlieb <hpnpilot@...>
Do you think the fault is temperature related or time running related?
toggle quoted message
Show quoted text
I don't have my 2440 manual here but what is common to all 4 "sides"? Could an input start getting a DC drift? Peter
On 9/25/2010 1:35 PM, Rich Miller wrote:
|
|
|
|
7603
Göran Krusell <goran.krusell@...>
David,
what do you get in the other vertical slot? Or with no vertical plug-in at all, only the time base in auto mode? Göran |
|
|
|
7603
milesdj8 <bwex@...>
Hello .
I have a 7603 with a 7A26 vert amp, now with no signal trace. If I connect a sig gen with 2Khz sine wave signal, all I see on the screen is the screen information numbers jumping up and down for a while before they settle down, but no signal trace. I have noticed for some months, that it was difficult to get the trace in the centre of the screen on some ranges, but I didn't think anything of it. Has anyone any idea what could be wrong? David Miles. |
|
|
|
F/S: Redout encoding BNC connectors
Dave Casey <dcasey@...>
Howdy all -
I have a limited quantity of crimp style male BNC
connectors available. These have the plastic outer housing with the spring
loaded pin and resistor for 10x readout encoding. These look exactly like
the connectors used on the probe compensation boxes except they are
to be crimped on to a cable.
These appear to be sized for RG-58. I DO NOT have
the center pins or crimping ferrules (this design already has the teflon
insulator built in), so you'll have to source those elsewhere.
These are brand new and the plastic is pliable, so
you could remove the readout bits and put them on another connector, however the
connector has to be notched in order for things to fit properly.
You could also probably use the parts to repair a
broken probe compensation box if the readout pin was damaged or
missing.
Photos (including an exploded view of one I took
apart) are available upon request.
$3 each, plus $5 S/H within the US, regardless of
quantity. International shipping will be extra.
First come, first serve until all are spoken for.
If you'd like more than 5, please consider your fellow TekScopers and defer
yourself to the back of the line for any additional pieces.
Please reply off list.
Dave Casey
Sent from my PC
|
|
|
|
Re: HV supply tektronix 535A
majefepe
Hi, sorry for my
toggle quoted message
Show quoted text
--- In TekScopes@..., guy allo <tek508@...> wrote:
hi, actually the high voltage is +8650 v. The comparator is a 12AU7A valve and 6au6 for 100v. I downloaded the cathode voltage up to-1200v and then went up to -1400V. but in the grid I have always comparator-161v. I checked the circuit, but always I have this voltage. so that any variation in load does not affect this tension, is always fixed. Any idea? Thank you. |
|
|
|
Re: HV supply tektronix 535A
guy allo
Hi,
I have one one my bench.
Which high voltage supply ? CRT high voltage +8650Volts?
What is the comparator valve number? V664 for the +100V supply?
Tek508
From: majefepe <majefepe@...> To: TekScopes@... Sent: Sat, September 25, 2010 5:57:29 PM Subject: [TekScopes] HV supply tektronix 535A I need help, I have a tektronix 535A and I have a problem at the source of high voltage, the comparator output is always 160 volts, the output is independent of the cathode voltage, which does not stabilize this tension, which can be the problem? changed and high voltage capacitors and comparator valve to no avail. The resistance of the comparator are good, what can be the problem? Thank you all for your help. ------------------------------------ Yahoo! Groups Links <*> To visit your group on the web, go to: http://groups.yahoo.com/group/TekScopes/ <*> Your email settings: Individual Email | Traditional <*> To change settings online go to: http://groups.yahoo.com/group/TekScopes/join (Yahoo! ID required) <*> To change settings via email: TekScopes-digest@... TekScopes-fullfeatured@... <*> To unsubscribe from this group, send an email to: TekScopes-unsubscribe@... <*> Your use of Yahoo! Groups is subject to: http://docs.yahoo.com/info/terms/ |
|
|
|
2440 CCD Problem - HELP!
Rich Miller
Hello All:
I had hoped fixing the +5V power supply issue would have solved my CCD issue. Unfortunately, this was not the case. Last night I attempted to complete the CCD Adjustment Procedure, as outlined in section 5. When I completed Section 8E - Do a SEL CAL and verify initial CCD gain, I am getting a FAIL status on 7230 for Sides 1, 2, 3, and 4 of Channel 1. At that point I decided to swap hybrid's U450, and U350 to see if the fault followed the hybrid to Channel 2, but to my surprise the faults remained on Channel 1.
Another symptom which makes this fault incredibly difficult to diagnose, is the fact that is only shows up after the unit has been running for a period of time (i.e. 15 - 20 minutes). If I restart the scope, all test pass without issue, and it's not until after some time has passed before FAULT 7230 shows up. For this reason it makes it very difficult to test while the fault condition exists.
I am going to proceed with the calibration procedure (Beyond Step 8E), in the hope that a parameter is so far out of whack, it is triggering the fault. Has anyone else seen this before, and if so what where your findings?
Rich
|
|
|
|
7704A Saga continued.
Jerry Massengale <jmassen418@...>
|
|
|
|
Re: Tek 7904 vertical drift.
Jerry Massengale <jmassen418@...>
--- On Sat, 9/25/10, Peter Ricardo wrote:
|
|
|
|
HV supply tektronix 535A
majefepe
I need help, I have a tektronix 535A and I have a problem at the source of high voltage, the comparator output is always 160 volts, the output is independent of the cathode voltage, which does not stabilize this tension, which can be the problem? changed and high voltage capacitors and comparator valve to no avail. The resistance of the comparator are good, what can be the problem? Thank you all for your help.
|
|
|
|
Tek 7904 vertical drift.
peter19a
Hi forum.
My 7904 is basically working fine, but... sometimes after switching the unit on, the horizontal trace and the readout are not where they should be, meaning both are about 2/10 of an inch (5mm) below there proper position. After a while both slowly move to there correct position. The same problem occurs sometimes while the scope was running for hours. I found out that (scope still on power) by pushing one or both of the Y units the trace(s) and readout are moving to the correct position. So what was is more obious as looking for a bad contact? I removed both Y plug in units and switched on. The X trace as well as the readout are shifted by about the same amount as described above. The vertical Amp is located between the two Y channel sockets. The ser# of the scope is 260000 up. So when I pushed slightly on the printed circuit, especially one of the two big IC's, the trace on the CRT jumps to the correct position and does not move again. I removed the Y amp from the frame but could not find anything to be suspected. If only one vertical plug in is inserted, the problem occurs also, then by sliding my hand in to the empty Y location and pushing at the Y PCB the trace and readout jump to there correct position. Is there anything known about bad contacts I could not see or find? Alle the little coax connectors are ok. The little stud connectors look ok. It's not really a big thing, but no problem at all would be better. Thanks a lot for any help. Peter |
|
|
|
Re: 7704a crt repair
Jerry Massengale <jmassen418@...>
--- On Sat, 9/25/10, Greg_A wrote:
|
|
|
|
Re: 7704a crt repair
Greg_A <greg.a@...>
Jerry hi,
toggle quoted message
Show quoted text
was good to talk to you on phone already.... 7704acrt 004 - those only from Sept 24th? Show the anode area RTV high voltage...good pictures. Question - how to upload like saved photo's to the Yahoo site - like you did yours? Is some other password needed and maybe some software? be in touch, Greg At 08:07 AM 9/24/2010, you wrote:
|
|
|
|
THS 700 series portable scope
coolhand_21_21 <coolhand21_21@...>
Hi does anyone have experience with these portable scope, how do they compare with the fluke 190 series?
They seen to have limited memory, would aliasing be a issue? thank cool |
|
|