Re: 7904 readout on the wrong place - vertical signal squeezed
Hi,
Sorry,
I have none in my stock. I have 7603s and a 7704. No 7904 stuff.
jerry
Hi,
Sorry,
I have none in my stock. I have 7603s and a 7704. No 7904 stuff.
jerry
|
By
Jerry Massengale <jmassen418@...>
·
#44636
·
|
|
Re: A Tale Of Two Teks (454A CRT Swapout)
Hi:
It was a bit of a chore because I had to tear it down a good bit further than I otherwise would have. This is because I needed to remove the rear casting on #2 to replace the damaged one on #1.
Hi:
It was a bit of a chore because I had to tear it down a good bit further than I otherwise would have. This is because I needed to remove the rear casting on #2 to replace the damaged one on #1.
|
By
TomS <bullet@...>
·
#44635
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
Hi Jerry,
sorry, I can't answer your question at the moment, because I destroyed the hybrid U685 on the vertical ampliefier board by a short circuit during checking the output signal.
Do you know,
Hi Jerry,
sorry, I can't answer your question at the moment, because I destroyed the hybrid U685 on the vertical ampliefier board by a short circuit during checking the output signal.
Do you know,
|
By
kupkagerhard <ghm1@...>
·
#44634
·
|
|
Re: NV ram in 2465B
Hey guys,
I wonder if I could use an ECL one shot triggeredfrom a HP 8656B signal generator to get the three pulses needed to finish thecal?
It could have a 5, 2, and 1 nS pulse and triggeredfrom
Hey guys,
I wonder if I could use an ECL one shot triggeredfrom a HP 8656B signal generator to get the three pulses needed to finish thecal?
It could have a 5, 2, and 1 nS pulse and triggeredfrom
|
By
Tom Miller <tmiller@...>
·
#44633
·
|
|
Re: Crazy question regarding 2445A...
Thanks for the reply Victor.
Ok then, any way to center the focus knob? Seems both my 2430 and 2445A come into focus right about 2 o'clock on the potentiometer.
Just curious if this was
Thanks for the reply Victor.
Ok then, any way to center the focus knob? Seems both my 2430 and 2445A come into focus right about 2 o'clock on the potentiometer.
Just curious if this was
|
By
amstel78 <amstel78@...>
·
#44632
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
Hi,
I forgot to ask if the signal on the vertical plates swing positive above zero.
jerry
Hi,
I forgot to ask if the signal on the vertical plates swing positive above zero.
jerry
|
By
Jerry Massengale <jmassen418@...>
·
#44631
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
Hi,
You might check continuity of the lead. If the lead is good, I would suspect the CRT itself. You can often inspect the internal leads through the glass. Sometimes the internal glass rods that
Hi,
You might check continuity of the lead. If the lead is good, I would suspect the CRT itself. You can often inspect the internal leads through the glass. Sometimes the internal glass rods that
|
By
Jerry Massengale <jmassen418@...>
·
#44630
·
|
|
Re: Tek 2432 failing self test
Tek has the 2440 SM on its site:
http://www2.tek.com/cmswpt/madetails.lotr?ct=MA&cs=msv&ci=15047&lc=EN
You really need the SM for the 2432, it's very different from the 2440 and even the 2432A.
If
Tek has the 2440 SM on its site:
http://www2.tek.com/cmswpt/madetails.lotr?ct=MA&cs=msv&ci=15047&lc=EN
You really need the SM for the 2432, it's very different from the 2440 and even the 2432A.
If
|
By
victor.silva
·
#44629
·
|
|
Re: Tek 2432 failing self test
The CCDs are the crucial Fast In Slow Out portion of the main sampling
circuit and yes CCD does stand for charge coupled device.
The PA failure points to the Pre-Amp, but is often a secondary result
The CCDs are the crucial Fast In Slow Out portion of the main sampling
circuit and yes CCD does stand for charge coupled device.
The PA failure points to the Pre-Amp, but is often a secondary result
|
By
David C. Partridge <david.partridge@...>
·
#44628
·
|
|
Re: Tek 2432 failing self test
Hi Jayw,
Thanks for your response.. I know tek used hybrids in alot of scopes, what part of the ckt are these CCD's used in, and what does the CCD stand for, charge coupled device?
Thanks,
Ryan
Hi Jayw,
Thanks for your response.. I know tek used hybrids in alot of scopes, what part of the ckt are these CCD's used in, and what does the CCD stand for, charge coupled device?
Thanks,
Ryan
|
By
Ryan Scott
·
#44627
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
Checking the vertical plates are connected inside the CRT is easy - check
ohms from vertical plate input pin on CRT to the corresponding pin for the
terminator. Also of course check that the
Checking the vertical plates are connected inside the CRT is easy - check
ohms from vertical plate input pin on CRT to the corresponding pin for the
terminator. Also of course check that the
|
By
David C. Partridge <david.partridge@...>
·
#44626
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
This sounds like the same issue I have with my 7704A. Just curious; can you move the trace above the center line when you activate the beam finder button?
Glenn
This sounds like the same issue I have with my 7704A. Just curious; can you move the trace above the center line when you activate the beam finder button?
Glenn
|
By
Glenn Butzlaff <gbutzlaff@...>
·
#44625
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
Before you condemn the CRT (or the vertical output amp) check the values on
the large emitter resistors R724, R725. They should be 500 ohms (early
7904) or 332R in the later (post B260000) version.
Before you condemn the CRT (or the vertical output amp) check the values on
the large emitter resistors R724, R725. They should be 500 ohms (early
7904) or 332R in the later (post B260000) version.
|
By
David C. Partridge <david.partridge@...>
·
#44624
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
But the + wire from the horizontal main amplifier is plugged on the contact on the CRT neck, or do you think inside the crt?
But the + wire from the horizontal main amplifier is plugged on the contact on the CRT neck, or do you think inside the crt?
|
By
kupkagerhard <ghm1@...>
·
#44623
·
|
|
Re: test 05 fail 24
Hi Gerald,
Whenever problems of this sort appear on Tekscopes, I search the archives
for additional information about the problems if they are of interest to me.
A search of the archives for "test 05
Hi Gerald,
Whenever problems of this sort appear on Tekscopes, I search the archives
for additional information about the problems if they are of interest to me.
A search of the archives for "test 05
|
By
tom jobe <tomjobe@...>
·
#44622
·
|
|
Re: 7904 readout on the wrong place - vertical signal squeezed
Hi,
Sounds like the horizontal + pin on the neck of the crt is open.
jerry
Hi,
Sounds like the horizontal + pin on the neck of the crt is open.
jerry
|
By
Jerry Massengale <jmassen418@...>
·
#44621
·
|
|
Re: test 05 fail 24
Hi Gerald
I've done a couple of 2465B rebuilds.
The best capacitors (IMHO) for this application are Panasonic EB's. They are my first choice for about 90% of the replacement caps on the 2465B.
Hi Gerald
I've done a couple of 2465B rebuilds.
The best capacitors (IMHO) for this application are Panasonic EB's. They are my first choice for about 90% of the replacement caps on the 2465B.
|
By
M Yachad
·
#44620
·
|
|
7904 readout on the wrong place - vertical signal squeezed
In my 7904 the readout is placed 1 division below the middle of the screen. Additionally the vertical signal can't be moved to upper half of the screen. When I move the vertical signal from bottom of
In my 7904 the readout is placed 1 division below the middle of the screen. Additionally the vertical signal can't be moved to upper half of the screen. When I move the vertical signal from bottom of
|
By
kupkagerhard <ghm1@...>
·
#44619
·
|
|
Re: NV ram in 2465B
Hello Stan,
You said
"Theslewed edge consists of a series of very accurately
> spaced "edges",space accurately "in time" from the CG551AP Trigger Output
> pulse. Ascope display of it will consist
Hello Stan,
You said
"Theslewed edge consists of a series of very accurately
> spaced "edges",space accurately "in time" from the CG551AP Trigger Output
> pulse. Ascope display of it will consist
|
By
Tom Miller <tmiller@...>
·
#44618
·
|
|
Re: Tek 2432 failing self test
Hi Ryan,
These scopes had a high failure rate in the CCD hybrids. It is likely that on or both are defective. You can determine which using the extended service diagnostics. (Or looking at their
Hi Ryan,
These scopes had a high failure rate in the CCD hybrids. It is likely that on or both are defective. You can determine which using the extended service diagnostics. (Or looking at their
|
By
Jay Walling
·
#44617
·
|