Re: New here, got some Tek scopes (let's start with the 2236)
Thanks for mentioning the thread.. it just confirms my suspicions that there is a possible hazard and I already had one of these things (not Schaffner branded, but still) stink up my apartment, so I'm
Thanks for mentioning the thread.. it just confirms my suspicions that there is a possible hazard and I already had one of these things (not Schaffner branded, but still) stink up my apartment, so I'm
|
By
Luca
·
#195776
·
|
|
Introduction
Thank you for accepting me into this group. I'm a retired electronics design engineer; that is, "retired" to the extent I no longer go to a job. I do however stay very active with my own small
Thank you for accepting me into this group. I'm a retired electronics design engineer; that is, "retired" to the extent I no longer go to a job. I do however stay very active with my own small
|
By
gkricorissian
·
#195775
·
|
|
Re: New here, got some Tek scopes (let's start with the 2236)
I just happened to notice this cross over between groups. There's a conversation going on in the HP/Agilent/Keysight group regarding RIFA caps, and the question Schaffner filters is broached there as
I just happened to notice this cross over between groups. There's a conversation going on in the HP/Agilent/Keysight group regarding RIFA caps, and the question Schaffner filters is broached there as
|
By
Dave Peterson
·
#195774
·
|
|
Re: PG506 PSU Problem
OK current update is that VR45 and Q45 are in good shape even if VR45 could be a shade higher (only 6.05V versus 6.2V @ 10mA). This was verified on a 577 CT in both swept and DC modes and also with
OK current update is that VR45 and Q45 are in good shape even if VR45 could be a shade higher (only 6.05V versus 6.2V @ 10mA). This was verified on a 577 CT in both swept and DC modes and also with
|
By
NigelP
·
#195773
·
|
|
Re: PG506 PSU Problem
Nigel,
In my previous message I mixed up "crowbar" and "current foldback". I intended to say that there is no current foldback, so the current stays at the maximum value when a short occurs.
Albert
Nigel,
In my previous message I mixed up "crowbar" and "current foldback". I intended to say that there is no current foldback, so the current stays at the maximum value when a short occurs.
Albert
|
By
Albert Otten
·
#195772
·
|
|
Re: PG506 PSU Problem
Albert thanks for your inputs but I would beg to differ, there is for sure a crowbar and it's comprised of Q45, an SCR clearly designed to clamp the 5V directly to ground, and it's triggering device
Albert thanks for your inputs but I would beg to differ, there is for sure a crowbar and it's comprised of Q45, an SCR clearly designed to clamp the 5V directly to ground, and it's triggering device
|
By
NigelP
·
#195771
·
|
|
Re: New here, got some Tek scopes (let's start with the 2236)
Hi, I've had good luck with the probes I received (1x P6109 and 2x P6122). They're in used but not abused condition and seem to work with no issues. Unfortunately, the probe leads of the P6122 have
Hi, I've had good luck with the probes I received (1x P6109 and 2x P6122). They're in used but not abused condition and seem to work with no issues. Unfortunately, the probe leads of the P6122 have
|
By
Luca
·
#195770
·
|
|
Re: Easy DIY Build - AC Volts / Amps, Current Limiter Test Bench
Point taken, and previously considered.
Yes, those sockets are exposed to live mains voltage, WHEN and ONLY WHEN the Leftmost switch is ON.
Those are my choice, and I am well aware of the dangers,
Point taken, and previously considered.
Yes, those sockets are exposed to live mains voltage, WHEN and ONLY WHEN the Leftmost switch is ON.
Those are my choice, and I am well aware of the dangers,
|
By
M Yachad
·
#195769
·
|
|
Re: 465 With no beam
Issac,
Good, then you are ahead of me on this. The HV Transformers on the 465 are also known to fail. Interested to learn what you finally find as the problem.
--
Michael Lynch
Dardanelle, AR
Issac,
Good, then you are ahead of me on this. The HV Transformers on the 465 are also known to fail. Interested to learn what you finally find as the problem.
--
Michael Lynch
Dardanelle, AR
|
By
Mlynch001
·
#195768
·
|
|
Re: 465 With no beam
Yes i desoldered it, In-fact i expected originally it to be faulty having seen the HVM in my 475 fail, so i jumped the gun and outright removed it from the scope before testing to see if the -2kv
Yes i desoldered it, In-fact i expected originally it to be faulty having seen the HVM in my 475 fail, so i jumped the gun and outright removed it from the scope before testing to see if the -2kv
|
By
Isaac Walsh
·
#195767
·
|
|
Re: PG506 PSU Problem
Nigel,
Current is limited to a maximum by conduction of VR55 but there is no crowbar function. VR45 fires in case of overvoltage.
I would put a voltmeter on +5 V and an ammeter in series with F65 (or
Nigel,
Current is limited to a maximum by conduction of VR55 but there is no crowbar function. VR45 fires in case of overvoltage.
I would put a voltmeter on +5 V and an ammeter in series with F65 (or
|
By
Albert Otten
·
#195766
·
|
|
Re: Tek TDS744a scope Options 2C, 3c, 4C hardware ?
I don't think there is any additional hardware required for 2C, but for 3C and 4C, optic fiber to BNC optical probes are required for each option, they cover different light frequency ranges.
The
I don't think there is any additional hardware required for 2C, but for 3C and 4C, optic fiber to BNC optical probes are required for each option, they cover different light frequency ranges.
The
|
By
Jared Cabot
·
#195765
·
|
|
Re: Tek TDS744a scope Options 2C, 3c, 4C hardware ?
Those options are not for 744A or any other A-scope.
http://www.hakanh.com/dl/docs/kitinstructions/tdsupgr.pdf
/Håkan
Those options are not for 744A or any other A-scope.
http://www.hakanh.com/dl/docs/kitinstructions/tdsupgr.pdf
/Håkan
|
By
zenith5106
·
#195764
·
|
|
Re: Which TDS7000 (Non-B) Serial Numbers are to be avoided?
The TDS (and DPO) 7000 series seem to be very few and far between on the open market, I think it may be a case of taking what's avaialble rather than being able to pic and choose.
The TDS (and DPO) 7000 series seem to be very few and far between on the open market, I think it may be a case of taking what's avaialble rather than being able to pic and choose.
|
By
David
·
#195763
·
|
|
Re: Easy DIY Build - AC Volts / Amps, Current Limiter Test Bench
At the very least some kind of shrouded plug/socket arrangement. Something like the fluke DMM leads mean you cant get your fingers in there.
At the very least some kind of shrouded plug/socket arrangement. Something like the fluke DMM leads mean you cant get your fingers in there.
|
By
David
·
#195762
·
|
|
Tek TDS744a scope Options 2C, 3c, 4C hardware ?
Does anyone know what hardware is required for the TDS744a Options:
2C (comm analyzer) ,
3C (Optical short Lambda)
&
4C (Optical, long Lambda) ?
& where to get this hardware ?
thank
Does anyone know what hardware is required for the TDS744a Options:
2C (comm analyzer) ,
3C (Optical short Lambda)
&
4C (Optical, long Lambda) ?
& where to get this hardware ?
thank
|
By
garp66
·
#195761
·
|
|
Re: Tek TDS 744a "Unstable Histogram" Error
I had an occasional image "shear" or jitter on my TDS784D. Recapping didn't
fix it, and eventually I discovered that the problem was simply that the
connector to the horizontal deflection yoke had a
I had an occasional image "shear" or jitter on my TDS784D. Recapping didn't
fix it, and eventually I discovered that the problem was simply that the
connector to the horizontal deflection yoke had a
|
By
Siggi
·
#195760
·
|
|
Re: TDS684A NV RAM Replacement\Issues
Hi Jared,
I've watched your video on YouTube, it is interesting you have re-designed probably a variant of the DS1486 and DS1250Y replicas. As you might probably know, there is an american person
Hi Jared,
I've watched your video on YouTube, it is interesting you have re-designed probably a variant of the DS1486 and DS1250Y replicas. As you might probably know, there is an american person
|
By
tantratron
·
#195759
·
|
|
Re: Easy DIY Build - AC Volts / Amps, Current Limiter Test Bench
Are those banana jacks exposing bare metal that is at mains voltage?......
I really think you should rethink that design...
Are those banana jacks exposing bare metal that is at mains voltage?......
I really think you should rethink that design...
|
By
Jared Cabot
·
#195758
·
|
|
Which TDS7000 (Non-B) Serial Numbers are to be avoided?
Hi to All,
Rather straightforward question. Anybody? Early TDS7000 Series NVRAM's have sometimes a too high current drain. Later ones (like the B-Series) use the DS1245Y-120 Devices.
These seem more
Hi to All,
Rather straightforward question. Anybody? Early TDS7000 Series NVRAM's have sometimes a too high current drain. Later ones (like the B-Series) use the DS1245Y-120 Devices.
These seem more
|
By
Egge Siert
·
#195757
·
|