Re: 3B3 Time Base Interesting Problem
Check that the blanking (unblanking) is working correctly. Your backwards trace motion could be retrace.
Check that the blanking (unblanking) is working correctly. Your backwards trace motion could be retrace.
|
By
Bruce Atwood
·
#193229
·
|
|
CM360 Oscilloscope checker manuals or literature??
Hi all,
I have come across a Tektronix/Sony CM360 Oscilloscope Checker and was trying to find any manuals or information on this equipment.
Unfortunately, the internet seems a bit lacking in any real
Hi all,
I have come across a Tektronix/Sony CM360 Oscilloscope Checker and was trying to find any manuals or information on this equipment.
Unfortunately, the internet seems a bit lacking in any real
|
By
Jared Cabot
·
#193228
·
|
|
Re: 3B3 Time Base Interesting Problem
Let me look at the 3A6 Jim, that is one plugin I am not familiar with. I am certainly interested though. Send me an email at jeromeleach17 at gmail.
Let me look at the 3A6 Jim, that is one plugin I am not familiar with. I am certainly interested though. Send me an email at jeromeleach17 at gmail.
|
By
Jerome D Leach
·
#193227
·
|
|
Re: 465M with probe for sale $60
The owner does not seem to know much about it. He is listing it as 50 MHz but the 465M is 100MHz!
The owner does not seem to know much about it. He is listing it as 50 MHz but the 465M is 100MHz!
|
By
Ananda
·
#193226
·
|
|
Are Dallas DS1250 and DS1650 interchangeable in Tektronix Scopes
Hi,
I have just seen a video on YouTube where the person says that it is possible to interchange Dallas DS1250 and DS1650 in all Tektronix scopes, since it seems that the protected area is not used by
Hi,
I have just seen a video on YouTube where the person says that it is possible to interchange Dallas DS1250 and DS1650 in all Tektronix scopes, since it seems that the protected area is not used by
|
By
Rogerio O
·
#193225
·
|
|
Re: Analog storage scope howto
Analog or digital storage scope, the use is the same.
1) single shot events. Use the single sweep mode to digitize/store/display a one time event. Photographic film was once used to do this as
Analog or digital storage scope, the use is the same.
1) single shot events. Use the single sweep mode to digitize/store/display a one time event. Photographic film was once used to do this as
|
By
Harvey White
·
#193224
·
|
|
Re: TDS540 repair; drifting vertical position at start-up
Hi Ralph,
My comments are below.
This may be another open trace or control voltage issue. Interpolator makes heavy use of calibration voltages.
I had similar errors on TDS520’s that cleared up
Hi Ralph,
My comments are below.
This may be another open trace or control voltage issue. Interpolator makes heavy use of calibration voltages.
I had similar errors on TDS520’s that cleared up
|
By
Ozan
·
#193223
·
|
|
Re: Analog storage scope howto
The "STORAGE" section in the operators manual (starting at page 1-20) tells you what the modes and controls are (named), what they do and how to operate them. It doesn't tell you what you'd use a
The "STORAGE" section in the operators manual (starting at page 1-20) tells you what the modes and controls are (named), what they do and how to operate them. It doesn't tell you what you'd use a
|
By
Raymond Domp Frank
·
#193222
·
|
|
465M with probe for sale $60
https://sacramento.craigslist.org/ele/d/folsom-oscilloscope/7474698836.html
No connection although Folsom is where i live.
Bob
https://sacramento.craigslist.org/ele/d/folsom-oscilloscope/7474698836.html
No connection although Folsom is where i live.
Bob
|
By
Robert Simpson
·
#193221
·
|
|
Re: Analog storage scope howto
A previous user pre-burned my 7514 for me. :(
It isn't too bad but it is noticeable.
Thanks,
Barry - N4BUQ
A previous user pre-burned my 7514 for me. :(
It isn't too bad but it is noticeable.
Thanks,
Barry - N4BUQ
|
By
n4buq
·
#193220
·
|
|
Re: Analog storage scope howto
Hi Tom,
Can you please upload the screenshots to the TekScope's photo section?
This way everyone can benefit from it.
I personally would be interested to know about the capabilities of storage
Hi Tom,
Can you please upload the screenshots to the TekScope's photo section?
This way everyone can benefit from it.
I personally would be interested to know about the capabilities of storage
|
By
Ke-Fong Lin
·
#193219
·
|
|
Re: Analog storage scope howto
Like I said, I did not find the operators manual section useful in terms of
teaching me to use storage mode to solve problems, it might be somewhat
useful in demonstrating correct behaviour, but IMO
Like I said, I did not find the operators manual section useful in terms of
teaching me to use storage mode to solve problems, it might be somewhat
useful in demonstrating correct behaviour, but IMO
|
By
Andy Warner
·
#193218
·
|
|
Re: Analog storage scope howto
Hi Andy-
I hope you wind up liking your 7633 as much as I do mine..... It's the one that I keep on the bench as a "first call" for anything that doesn't require something that it can't do.
I find
Hi Andy-
I hope you wind up liking your 7633 as much as I do mine..... It's the one that I keep on the bench as a "first call" for anything that doesn't require something that it can't do.
I find
|
By
fiftythreebuick
·
#193217
·
|
|
Re: Analog storage scope howto
Andy,
Congratulations on your new 7633. In its time, it was Tek's most powerful 100MHz analog storage mainframe. It is capable of all useful and not-so-useful storage modes that analog 'scopes
Andy,
Congratulations on your new 7633. In its time, it was Tek's most powerful 100MHz analog storage mainframe. It is capable of all useful and not-so-useful storage modes that analog 'scopes
|
By
Raymond Domp Frank
·
#193216
·
|
|
Re: Analog storage scope howto
Tek's Concept Series are a good way to go.
Since you've got a 7633, use its Operator's Manual, which is available for free online. It contains some pretty good introductory information on what you're
Tek's Concept Series are a good way to go.
Since you've got a 7633, use its Operator's Manual, which is available for free online. It contains some pretty good introductory information on what you're
|
By
Raymond Domp Frank
·
#193215
·
|
|
Re: Analog storage scope howto
I need to take a look at that. I recently acquired a 7514 and I'm quite puzzled about how the anode lead makes it to the CRT. The HV out of the multiplier goes to a typical cylindrical anode-style
I need to take a look at that. I recently acquired a 7514 and I'm quite puzzled about how the anode lead makes it to the CRT. The HV out of the multiplier goes to a typical cylindrical anode-style
|
By
n4buq
·
#193214
·
|
|
Re: Analog storage scope howto
You might find useful information in the Tek Concepts Series volume on storage crts and circuits (see Tekwiki for "Concepts Series"). Even though it isn't written as a how-to-use manual, knowing how
You might find useful information in the Tek Concepts Series volume on storage crts and circuits (see Tekwiki for "Concepts Series"). Even though it isn't written as a how-to-use manual, knowing how
|
By
Tom Lee
·
#193213
·
|
|
Re: 3B3 Time Base Interesting Problem
The better choice is the 3A6, which is just a 3A1 with a delay line, so the scope can display the event that triggers the sweep. I've only got one 3A1, but I have several 3A6s, all in good cosmetic
The better choice is the 3A6, which is just a 3A1 with a delay line, so the scope can display the event that triggers the sweep. I've only got one 3A1, but I have several 3A6s, all in good cosmetic
|
By
Jim Adney
·
#193212
·
|
|
Re: 3B3 Time Base Interesting Problem
I've got several 3B3s, all in good working and cosmetic condition. Make me an offer.
I always thought the 3B4 was the better timebase plugin. Should I take a better look at the 3B3?
I've got several 3B3s, all in good working and cosmetic condition. Make me an offer.
I always thought the 3B4 was the better timebase plugin. Should I take a better look at the 3B3?
|
By
Jim Adney
·
#193211
·
|
|
Re: 7A16A high frequency compensation.
A simultaneous failure of all trimmers is highly improbable. Your problem is elsewhere.
Possibilities include: Bad connection at input BNC, poor contacts at attenuator socket pins, etc. So, your plan
A simultaneous failure of all trimmers is highly improbable. Your problem is elsewhere.
Possibilities include: Bad connection at input BNC, poor contacts at attenuator socket pins, etc. So, your plan
|
By
Tom Lee
·