Re: Getting calibration equipment calibrated
For fun last night I characterized my bench DMM with my PG506 in DC mode. I was able to use the deflection error to determine that the entirety of the DMMs range held to within 0.2%. Up to the 100v
For fun last night I characterized my bench DMM with my PG506 in DC mode. I was able to use the deflection error to determine that the entirety of the DMMs range held to within 0.2%. Up to the 100v
|
By
Dave Peterson
·
#183002
·
|
|
Re: Getting calibration equipment calibrated
I use one of these. And one can send it back for recalibration.
https://dmmcheckplus.com/
DaveD
I use one of these. And one can send it back for recalibration.
https://dmmcheckplus.com/
DaveD
|
By
Dave Daniel
·
#183001
·
|
|
Re: Getting calibration equipment calibrated
Dave,
for a voltage reference you can get something quite economically (less than $25) on Amazon or eBay that is based off an
Dave,
for a voltage reference you can get something quite economically (less than $25) on Amazon or eBay that is based off an
|
By
Jeff Dutky
·
#183000
·
|
|
Re: Getting calibration equipment calibrated
Walter,
I second your comments. We had a “cal-light” lab. Two GPSDO’s and a HP3458 and a set of Julie labs precision resistors. The 3458 got the stds lab cal at keysight ($1800) and then
Walter,
I second your comments. We had a “cal-light” lab. Two GPSDO’s and a HP3458 and a set of Julie labs precision resistors. The 3458 got the stds lab cal at keysight ($1800) and then
|
By
Stephen Hanselman
·
#182999
·
|
|
Tektronix 2246A faint readout
Hi all,
Having repaired the power supply I've now powered the 'scope up. I've measured the calibration signal on all 4 channels, the buttons and switches all seem to work and the self tests
Hi all,
Having repaired the power supply I've now powered the 'scope up. I've measured the calibration signal on all 4 channels, the buttons and switches all seem to work and the self tests
|
By
Jon Harrison-Hughes
·
#182998
·
|
|
Re: P6302 / AM503B 330MHz Oscillation Where is this from?
(continued) No pronounced resonance with SG503 either. Albert
(continued) No pronounced resonance with SG503 either. Albert
|
By
Albert Otten
·
#182997
·
|
|
Re: Tektronix 2246A power supply repair and safe bench testing
So having repaired the power supply I've powered the 'scope up and tested the LV outputs on J1204, supplies are as
So having repaired the power supply I've powered the 'scope up and tested the LV outputs on J1204, supplies are as
|
By
Jon Harrison-Hughes
·
#182996
·
|
|
Re: Getting calibration equipment calibrated
There's a lot of stuff you don't need to do unless you want to.
Remember that the analog scopes you're checking, unless they have digital counters (7D15, et al), you're deciding whether or not you
There's a lot of stuff you don't need to do unless you want to.
Remember that the analog scopes you're checking, unless they have digital counters (7D15, et al), you're deciding whether or not you
|
By
Harvey White
·
#182995
·
|
|
Re: Getting calibration equipment calibrated
Outside calibration is complex situation, especially since it is understood differently by the
providers of the service, and their customers. this disconnect can be severe, and often leads to
some
Outside calibration is complex situation, especially since it is understood differently by the
providers of the service, and their customers. this disconnect can be severe, and often leads to
some
|
By
walter shawlee
·
#182994
·
|
|
Re: P6302 / AM503B 330MHz Oscillation Where is this from?
Dennis, in your pictures I see about 3 periods of oscillation in 20 ns. Isn't that corresponding to 150 MHz in stead of about 300 MHz? (I hesitate to bring this up...).
I tried this with similar
Dennis, in your pictures I see about 3 periods of oscillation in 20 ns. Isn't that corresponding to 150 MHz in stead of about 300 MHz? (I hesitate to bring this up...).
I tried this with similar
|
By
Albert Otten
·
#182993
·
|
|
Re: 2465B on a 200C scope cart?
Sorry I mixed up 200, all of my carts are K212
Jon
Sorry I mixed up 200, all of my carts are K212
Jon
|
By
Jean-Paul
·
#182991
·
|
|
Re: Getting calibration equipment calibrated
Hi Dave,
Ok that changes a LOT of things. If you have access to that lecroy you should be set to do the calibrations on all the plugins. A 6.5 Digit DMM is plenty for the PG 506 did mine on a keithly
Hi Dave,
Ok that changes a LOT of things. If you have access to that lecroy you should be set to do the calibrations on all the plugins. A 6.5 Digit DMM is plenty for the PG 506 did mine on a keithly
|
By
Eric
·
#182990
·
|
|
Re: 547 Restoration conundrum
#photo-notice
Thank you for the recommendations all!
I took a look at the neck as best I could without fully removing it and I don't see any spare magnets or damage inside the tube neck itself. That said, I will
Thank you for the recommendations all!
I took a look at the neck as best I could without fully removing it and I don't see any spare magnets or damage inside the tube neck itself. That said, I will
|
By
Jason A.
·
#182989
·
|
|
P6042 Probe
Hi all, is it possible to install and adapt a P6302 probe tip into a P6042 tip? The transformer is bad on my P6042 probe.
Thanks
Hi all, is it possible to install and adapt a P6302 probe tip into a P6042 tip? The transformer is bad on my P6042 probe.
Thanks
|
By
Rick
·
#182988
·
|
|
Re: PG506 Repair - I had a Great Day today/
Hi Albert, interesting stuff and info but I have the coming period almost no time for my hobby. An urgent repair in the garage and my car, so when there is more time I dive in again and post the
Hi Albert, interesting stuff and info but I have the coming period almost no time for my hobby. An urgent repair in the garage and my car, so when there is more time I dive in again and post the
|
By
tek_547
·
#182987
·
|
|
Re: 2465B on a 200C scope cart?
I tried doing that when I received my 2465B a while back. The combination of the 2465B and the 200C seemed very unstable. I bought a K212 cart for a very reasonable price and that works just
I tried doing that when I received my 2465B a while back. The combination of the 2465B and the 200C seemed very unstable. I bought a K212 cart for a very reasonable price and that works just
|
By
Dave Daniel
·
#182986
·
|
|
Re: My eye sensitivity to blue-violet higher than normal
ISTR that there are a number of people who have four types of colour sensors which extend vision into what most call the low-UV.
David
ISTR that there are a number of people who have four types of colour sensors which extend vision into what most call the low-UV.
David
|
By
David C. Partridge
·
#182985
·
|
|
Re: 2465B on a 200C scope cart?
Until a few weeks ago, I had a 7704A on one of the carts meant for portables (forget the model, but it's the one with larger rear wheels). I needed extra straps to anchor it down and careful
Until a few weeks ago, I had a 7704A on one of the carts meant for portables (forget the model, but it's the one with larger rear wheels). I needed extra straps to anchor it down and careful
|
By
Dave Seiter
·
#182984
·
|
|
Re: 2465B on a 200C scope cart?
Hello, good question:
I have several 2465/7/B and carts.
Not a problem to mount the scopes, feet and straps need careful fiddling to position.
I recall t scope rests about 1" forward of
Hello, good question:
I have several 2465/7/B and carts.
Not a problem to mount the scopes, feet and straps need careful fiddling to position.
I recall t scope rests about 1" forward of
|
By
Jean-Paul
·
#182983
·
|
|
Re: Getting calibration equipment calibrated
I might be a bit of an odd bird: I quite enjoy twiddling the variety of pots vs. pure automation. I work in the IC industry, so I get my fill of high level integration, and have spent many hours in
I might be a bit of an odd bird: I quite enjoy twiddling the variety of pots vs. pure automation. I work in the IC industry, so I get my fill of high level integration, and have spent many hours in
|
By
Dave Peterson
·
#182982
·
|