source for blue filters for 2465 family scopes ??
anyone know of a source for the blue crt filters that slid in front of the actual CRT ?
-bob
anyone know of a source for the blue crt filters that slid in front of the actual CRT ?
-bob
|
By
bob
·
#172757
·
|
|
Re: What are 012-0057 and 012-057 Coax Cable used for?
I know they are specified in most of the scope calibration documents for scopes 400 series and 7000 series. The cal documents call for 2. I think they are just a high quality 42 inch cable. With a
I know they are specified in most of the scope calibration documents for scopes 400 series and 7000 series. The cal documents call for 2. I think they are just a high quality 42 inch cable. With a
|
By
Eric
·
#172756
·
|
|
What are 012-0057 and 012-057 Coax Cable used for?
I was going through the coax cables I have looking for one I could cut the
BNC off one end when I noticed two of the cables have Tek part numbers on
them.
Both cables are 42" long.
One cable has a
I was going through the coax cables I have looking for one I could cut the
BNC off one end when I noticed two of the cables have Tek part numbers on
them.
Both cables are 42" long.
One cable has a
|
By
Dennis Tillman W7pF
·
#172755
·
|
|
Re: 468 Storage Mode Failure
My analog section works in the "Bypass" mode. When I go back to the standard configuration, analog does not show a trace in "Non-Store", but the digital portion shows signs of life in the form of a
My analog section works in the "Bypass" mode. When I go back to the standard configuration, analog does not show a trace in "Non-Store", but the digital portion shows signs of life in the form of a
|
By
Michael W. Lynch
·
#172754
·
|
|
Re: 468 Storage Mode Failure
I am following this conversation and working on my 468 as well. It may be beneficial to mention here that when isolating the analog section for testing, as outlined in the service manual; there is an
I am following this conversation and working on my 468 as well. It may be beneficial to mention here that when isolating the analog section for testing, as outlined in the service manual; there is an
|
By
Michael W. Lynch
·
#172753
·
|
|
Re: Scope grounding and generators
You have to ask yourself what you want to see.
If you want to know the distortion of the "sine wave", you only
need at most DC to a few kilohertz of bandwidth.
If you want to know the RF on the
You have to ask yourself what you want to see.
If you want to know the distortion of the "sine wave", you only
need at most DC to a few kilohertz of bandwidth.
If you want to know the RF on the
|
By
Chuck Harris
·
#172752
·
|
|
Re: Scope grounding and generators
They should be ok, as long as you are using only the probe
tips, and not the grounding clips to connect to the line and
neutral leads of the generator. The probe grounding clips
should connect to the
They should be ok, as long as you are using only the probe
tips, and not the grounding clips to connect to the line and
neutral leads of the generator. The probe grounding clips
should connect to the
|
By
Chuck Harris
·
#172751
·
|
|
Re: Carry-on adventures (was: Re: [TekScopes] Tek R453 - a rack version of a portable scope!)
Things have gotten more challenging for air travelers who need to carry
expensive equipment rather than risk it to baggage handling or loss. I once
had a computer security cable that TSA would not let
Things have gotten more challenging for air travelers who need to carry
expensive equipment rather than risk it to baggage handling or loss. I once
had a computer security cable that TSA would not let
|
By
stevenhorii
·
#172750
·
|
|
Re: Scope grounding and generators
I wonder whether optical isolation would work. The photodiode types have
very fast response, but can be nonlinear. Though usually used for logic
signals, there are optocouplers designed for analog
I wonder whether optical isolation would work. The photodiode types have
very fast response, but can be nonlinear. Though usually used for logic
signals, there are optocouplers designed for analog
|
By
stevenhorii
·
#172749
·
|
|
Re: Carry-on adventures (was: Re: [TekScopes] Tek R453 - a rack version of a portable scope!)
Not Tek instrument related, but still relevant. During the mid-late 80s I just to often have to travel to remote sites via air travel, and to avoid airport delays used to take everything aboard as
Not Tek instrument related, but still relevant. During the mid-late 80s I just to often have to travel to remote sites via air travel, and to avoid airport delays used to take everything aboard as
|
By
tekscopegroup@...
·
#172748
·
|
|
Re: Scope grounding and generators
P6009 is rated to 1.5 KV. And for 60 Hz nominal frequency doesn't need to be derated. (It's flat to around 100 KHz.) The can be compensated to the 465's input capacitance.
So yes and no.
Your
P6009 is rated to 1.5 KV. And for 60 Hz nominal frequency doesn't need to be derated. (It's flat to around 100 KHz.) The can be compensated to the 465's input capacitance.
So yes and no.
Your
|
By
Roy Thistle
·
#172747
·
|
|
Re: Scope grounding and generators
Getting back to a Tektronix question, I have P6009 probes. Wouldn't using those be safe from over voltage outputs (If there are any) of a generator to the 465M oscilloscope?
Bob
Getting back to a Tektronix question, I have P6009 probes. Wouldn't using those be safe from over voltage outputs (If there are any) of a generator to the 465M oscilloscope?
Bob
|
By
Robert Simpson
·
#172746
·
|
|
Re: Desoldering Iron, was Re: [TekScopes] 466-464 stray wire
One quick comment about air compressors.
The connector that you use can be a major limiting factor.
I had a box of odds & sods that I had obtained over the years & went through them recently when
One quick comment about air compressors.
The connector that you use can be a major limiting factor.
I had a box of odds & sods that I had obtained over the years & went through them recently when
|
By
Brian Symons
·
#172745
·
|
|
Switched Mode Power Supplies (was: Scope Grounding and Generators)
As requested by the moderator.
As requested by the moderator.
|
By
Roy Thistle
·
#172744
·
|
|
Re: Scope grounding and generators
And yet, I do.
If anyone chooses to ignore, then I'll have less to read, less to ponder, and less to write about.
Go ahead make your day, if that's what makes it a better day for you.
My day has
And yet, I do.
If anyone chooses to ignore, then I'll have less to read, less to ponder, and less to write about.
Go ahead make your day, if that's what makes it a better day for you.
My day has
|
By
Roy Thistle
·
#172743
·
|
|
Re: refurbishing TM500 and TM5000 hardware
Chuck,
In actuality the TM5000 has both a Linear supply and a Switching supply.
GOD Bless and Thanks,
rich!
Chuck,
In actuality the TM5000 has both a Linear supply and a Switching supply.
GOD Bless and Thanks,
rich!
|
By
Richard R. Pope
·
#172742
·
|
|
Re: refurbishing TM500 and TM5000 hardware
TM500 has a mains transformer powered linear supply.
TM5000 has an inverter powered switching power supply.
-Chuck Harris
Richard Erlacher wrote:
TM500 has a mains transformer powered linear supply.
TM5000 has an inverter powered switching power supply.
-Chuck Harris
Richard Erlacher wrote:
|
By
Chuck Harris
·
#172741
·
|
|
Re: refurbishing TM500 and TM5000 hardware
I’ve not looked inside one of these frames, of which I have several, to see what sort of power supply they have, but I don’t remember anything that looked like a switch-mode supply. They have
I’ve not looked inside one of these frames, of which I have several, to see what sort of power supply they have, but I don’t remember anything that looked like a switch-mode supply. They have
|
By
Richard Erlacher
·
#172740
·
|
|
Re: Scope grounding and generators
May I suggest that the subject of many of these posts should be changed to Switching Mode Power Supplies
May I suggest that the subject of many of these posts should be changed to Switching Mode Power Supplies
|
By
Dennis Tillman W7pF
·
#172739
·
|
|
Re: refurbishing TM500 and TM5000 hardware
Regarding the OP, in the case of TM500/5000 stuff, I'd tend to go with Chuck's recommendation and just plug it in, and turn it on. It would be good to first fire up the empty mainframe, then check the
Regarding the OP, in the case of TM500/5000 stuff, I'd tend to go with Chuck's recommendation and just plug it in, and turn it on. It would be good to first fire up the empty mainframe, then check the
|
By
Ed Breya
·
#172738
·
|