Re: TDS544A major surgery - CRT issues
To the best of my knowledge. the displays are much the same, the difference being the LCD shutter feed from the MB. From what I know, the LCD shutter is "fastened" to the display with an
To the best of my knowledge. the displays are much the same, the difference being the LCD shutter feed from the MB. From what I know, the LCD shutter is "fastened" to the display with an
|
By
Harvey White
·
#160900
·
|
|
TDS544A major surgery - CRT issues
A few years back, I reported on restoring a TDS544A that had the capacitor plague on the acquisition board. That part worked out OK, but a couple months or so later, the HV transformer in the CRT
A few years back, I reported on restoring a TDS544A that had the capacitor plague on the acquisition board. That part worked out OK, but a couple months or so later, the HV transformer in the CRT
|
By
Ed Breya
·
#160899
·
|
|
Re: 7834 HT transformer rewind?
Chuck,
You are a cornucopia of knowledge!
Thanks!
--
Michael Lynch
Dardanelle, AR
Chuck,
You are a cornucopia of knowledge!
Thanks!
--
Michael Lynch
Dardanelle, AR
|
By
Mlynch001
·
#160898
·
|
|
Re: 130 LC meter
I have spent lots of time and lost lots of sleep over this. I got up in the middle of the night last night and discovered that the problem appears to be a defective tube socket. Not the best news
I have spent lots of time and lost lots of sleep over this. I got up in the middle of the night last night and discovered that the problem appears to be a defective tube socket. Not the best news
|
By
Bob Albert
·
#160897
·
|
|
Re: 7834 HT transformer rewind?
Has anyone investigated the compatability of transformers or cores from Tektronix televison studio equipment? I really don't care for the idea of trashing that gear, but I can't imagine any of the
Has anyone investigated the compatability of transformers or cores from Tektronix televison studio equipment? I really don't care for the idea of trashing that gear, but I can't imagine any of the
|
By
greenboxmaven
·
#160896
·
|
|
Re: 7834 HT transformer rewind?
The 7834 transformer is an all-in-one transformer, and as
such it has all of the power supply windings, and the EHT
on a single core.
It is a cornucopia of wire sizes, taps, tapes, and wedges.
The
The 7834 transformer is an all-in-one transformer, and as
such it has all of the power supply windings, and the EHT
on a single core.
It is a cornucopia of wire sizes, taps, tapes, and wedges.
The
|
By
Chuck Harris <cfharris@...>
·
#160895
·
|
|
Re: 7834 HT transformer rewind?
David,
I would be interested in a quote for the HV transformer of the Type 576 as well. The winding specs are readily available, unfortunately, the ferrite cores are not. There are very likely some
David,
I would be interested in a quote for the HV transformer of the Type 576 as well. The winding specs are readily available, unfortunately, the ferrite cores are not. There are very likely some
|
By
Mlynch001
·
#160894
·
|
|
Re: 130 LC meter
My Tek 130 had several Sprague Black Beauty capacitors in the oscillator
circuit
All were bad. I've had enough trouble (and been shocked enough) with these
caps
that I replace them on
My Tek 130 had several Sprague Black Beauty capacitors in the oscillator
circuit
All were bad. I've had enough trouble (and been shocked enough) with these
caps
that I replace them on
|
By
Jack Wills
·
#160893
·
|
|
Re: 7834 HT transformer rewind?
That is good news to hear , some years ago I went to a company that said they could wind a 549 transformer , in todays money it cost about the same -- but it didnt work when I got it ?? . I was never
That is good news to hear , some years ago I went to a company that said they could wind a 549 transformer , in todays money it cost about the same -- but it didnt work when I got it ?? . I was never
|
By
Brian
·
#160892
·
|
|
7834 HT transformer rewind?
I may have located a company in UK that can handle the rewinding of these
transformers.
It won't be dagger cheap - they gave me a ball-park figure of about GBP100
plus tax and delivery, but that was
I may have located a company in UK that can handle the rewinding of these
transformers.
It won't be dagger cheap - they gave me a ball-park figure of about GBP100
plus tax and delivery, but that was
|
By
David C. Partridge
·
#160891
·
|
|
Re: 130 LC meter
Have you checked the smoothing capacitor on the power supply? Mine was open circuit.
Craig
Have you checked the smoothing capacitor on the power supply? Mine was open circuit.
Craig
|
By
Craig Sawyers <c.sawyers@...>
·
#160890
·
|
|
130 LC meter
Is there anyone familiar with this unit? I have one that isn't working and have traced the problem to the variable oscillator not oscillating.
It's plate voltage is too low but everything seems good
Is there anyone familiar with this unit? I have one that isn't working and have traced the problem to the variable oscillator not oscillating.
It's plate voltage is too low but everything seems good
|
By
Bob Albert
·
#160889
·
|
|
260-2500-00 Encoder
There was some announcement a year or so ago that somebody has several of
those in stock. Don't know who it was and if there are still some available
bue I need one for my recently acquired CG5011
There was some announcement a year or so ago that somebody has several of
those in stock. Don't know who it was and if there are still some available
bue I need one for my recently acquired CG5011
|
By
Sergey Kubushyn
·
#160888
·
|
|
Re: Chart Recorder and Its Replacement
I did some testing on the terminals that supposedly provide the START signal. It turns out to be just a relay that closes when the recording starts and opens again when the scan is finished. The
I did some testing on the terminals that supposedly provide the START signal. It turns out to be just a relay that closes when the recording starts and opens again when the scan is finished. The
|
By
Pax Electronica
·
#160887
·
|
|
Re: 465B fan question
Received and installed a new Thermister and 15Ohm resistor. The fan does start up better but there is still a problem. Watching it carefully, I see that the fan shaft has about 1/32 inch forward and
Received and installed a new Thermister and 15Ohm resistor. The fan does start up better but there is still a problem. Watching it carefully, I see that the fan shaft has about 1/32 inch forward and
|
By
Robert Simpson
·
#160886
·
|
|
Re: 175 Curve Tracer Rear 9-Pin Connector
Hi :
It seem the data for ebay item were missing after I post them .
Add the ebay ID for easier find them .
RegardTony CheungOCT 19 2019
Hi :
It seem the data for ebay item were missing after I post them .
Add the ebay ID for easier find them .
RegardTony CheungOCT 19 2019
|
By
Yiu On Tony C
·
#160885
·
|
|
Re: 175 Curve Tracer Rear 9-Pin Connector
Hi :
Cable connector for the rear of 575 connection is a Amphenol 165-13 ( male ) , for the rear of 175 connection is a Amphenol 165-14 ( female ) , all those connection pin are occupied ( 9
Hi :
Cable connector for the rear of 575 connection is a Amphenol 165-13 ( male ) , for the rear of 175 connection is a Amphenol 165-14 ( female ) , all those connection pin are occupied ( 9
|
By
Yiu On Tony C
·
#160884
·
|
|
CSA803A cal constants addresses in U500/U511
As mentioned in other topics the 33 factory calibration constants in a CSA803A/B/C (and corresponding 11800 models) can be set, but not read, in U500/U511 using the GPIB command calcorrection. These
As mentioned in other topics the 33 factory calibration constants in a CSA803A/B/C (and corresponding 11800 models) can be set, but not read, in U500/U511 using the GPIB command calcorrection. These
|
By
Albert Otten
·
#160883
·
|
|
Yahoo Groups are deleting all files in December
Hi Richard,
Thank you for the notice. I never received any notice from Yahoo about this (that was typical for Yahoo!) so your heads up was appreciated.
We moved everything over to Groups.io a long
Hi Richard,
Thank you for the notice. I never received any notice from Yahoo about this (that was typical for Yahoo!) so your heads up was appreciated.
We moved everything over to Groups.io a long
|
By
Dennis Tillman W7pF
·
#160882
·
|
|
Re: 175 Curve Tracer Rear 9-Pin Connector
I have a 175. It is pins
Craig
I have a 175. It is pins
Craig
|
By
Craig Sawyers <c.sawyers@...>
·
#160881
·
|