Re: 11801 NVRAM
/
/ The challenge would be how to iteratively adjust the NVram constants without removing the NVram chips a thousand times. I don't know of any GPIB way / to SET the value of each NVram cal constant,
/
/ The challenge would be how to iteratively adjust the NVram constants without removing the NVram chips a thousand times. I don't know of any GPIB way / to SET the value of each NVram cal constant,
|
By
Albert Otten
·
#160106
·
|
|
Re: From 2014: Tektronix Announces Winner of Europe’s Oldest Working Oscilloscope Contest
Did the contest apply to only Tek scopes? Sadly, my 511A and 514AD don't work (The HV assemblies (I think) on both are dead- both had puddles of melted stuff around the assembly just above the main
Did the contest apply to only Tek scopes? Sadly, my 511A and 514AD don't work (The HV assemblies (I think) on both are dead- both had puddles of melted stuff around the assembly just above the main
|
By
Dave Seiter
·
#160107
·
|
|
Re: Thermal Insulator question
Thermal Conductivity
--------------------
Mica = 0.75W/mK
Silicone/fiberglass = 1.3-3.0W/mK
The unit of thermal conductivity is Watts per milliKelvin temperature rise.
So, higher W/mK is better
Thermal Conductivity
--------------------
Mica = 0.75W/mK
Silicone/fiberglass = 1.3-3.0W/mK
The unit of thermal conductivity is Watts per milliKelvin temperature rise.
So, higher W/mK is better
|
By
Chuck Harris
·
#160108
·
|
|
Re: 2465 DVS Repair - was; Re: [TekScopes] Wanted; Person to repair Tektronix 2465 DVS
The power inside of the 2465 supplies is pretty dirty from
an RF point of view. It is pretty quiet from a 2x60Hz point
of view.
How can you tell the difference?
The RF noise is composed of sliver
The power inside of the 2465 supplies is pretty dirty from
an RF point of view. It is pretty quiet from a 2x60Hz point
of view.
How can you tell the difference?
The RF noise is composed of sliver
|
By
Chuck Harris
·
#160109
·
|
|
Re: From 2014: Tektronix Announces Winner of Europe’s Oldest Working Oscilloscope Contest
The 310 is close, but no cigar.
The 511 was the original Tek scope, the 511AD was a
later refined model, with a distributed vertical amplifier.
My working (?) 513D is probably older than Alan's
The 310 is close, but no cigar.
The 511 was the original Tek scope, the 511AD was a
later refined model, with a distributed vertical amplifier.
My working (?) 513D is probably older than Alan's
|
By
Chuck Harris
·
#160110
·
|
|
Re: Thermal Insulator question
Be careful here!
Sometimes, the screw passes through the heatsink to a
socket, and the screw is the electrical connection to
the tab. Other times the screw screws into the heatsink,
and must be
Be careful here!
Sometimes, the screw passes through the heatsink to a
socket, and the screw is the electrical connection to
the tab. Other times the screw screws into the heatsink,
and must be
|
By
Chuck Harris
·
#160111
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
Michael,
The cores are universal among tekscopes.
I have never had one stick like you are describing. Epoxy
just isn't that good.
Send it to me, and I will remove it. If, in my zeal I
Michael,
The cores are universal among tekscopes.
I have never had one stick like you are describing. Epoxy
just isn't that good.
Send it to me, and I will remove it. If, in my zeal I
|
By
Chuck Harris
·
#160112
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
Chuck,
Leave it to me to have the “odd ball” of the group. I have seen many others that look the same or very similar. I may take you up on your offer and contact you off list for an
Chuck,
Leave it to me to have the “odd ball” of the group. I have seen many others that look the same or very similar. I may take you up on your offer and contact you off list for an
|
By
Michael W. Lynch
·
#160113
·
|
|
Re: 2465 DVS Repair - was; Re: [TekScopes] Wanted; Person to repair Tektronix 2465 DVS
Very well and detail explanation!
Thank you Chuck.
Very well and detail explanation!
Thank you Chuck.
|
By
Tony Fleming
·
#160114
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
I have split a number of transformers, A few thoughts
1, You should take Chuck up on the offer
2, But how are you getting the number 260F? oven dials can be way off and
the temp inside an oven can
I have split a number of transformers, A few thoughts
1, You should take Chuck up on the offer
2, But how are you getting the number 260F? oven dials can be way off and
the temp inside an oven can
|
By
ykochcal
·
#160115
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
John,
I will try this again at a higher temp. You are very certainly right about this. I have not done any “practice” since I don’t have anything to practice on. I guess this is my
John,
I will try this again at a higher temp. You are very certainly right about this. I have not done any “practice” since I don’t have anything to practice on. I guess this is my
|
By
Mlynch001
·
#160116
·
|
|
Re: 11801 NVRAM
I'm very interested in how to derive and store the correction terms. There has been discussion of a command to write them, but it appeared not to work when attempted IIRC.
It would be very helpful
I'm very interested in how to derive and store the correction terms. There has been discussion of a command to write them, but it appeared not to work when attempted IIRC.
It would be very helpful
|
By
Reginald Beardsley
·
#160117
·
|
|
Re: AM700
Hi,
Wow! I waiting for your flash image very happy. Do you know if this flash simm are commercial type?
In this days I continued capacitors check, I find some through hole capacitors that just start
Hi,
Wow! I waiting for your flash image very happy. Do you know if this flash simm are commercial type?
In this days I continued capacitors check, I find some through hole capacitors that just start
|
By
Juri Ghisio
·
#160118
·
|
|
Re: From 2014: Tektronix Announces Winner of Europe’s Oldest Working Oscilloscope Contest
Yes, after perusing my copy of "Oscilloscope Development, 1943-57" by Peter D. Hiscocks, I realize that one cannot beat a 511 as the earliest 'Tektronix 'scope.
When I was much younger, I had a
Yes, after perusing my copy of "Oscilloscope Development, 1943-57" by Peter D. Hiscocks, I realize that one cannot beat a 511 as the earliest 'Tektronix 'scope.
When I was much younger, I had a
|
By
Dave Daniel
·
#160119
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
There are practice cores in most old electronics, which are everywhere
around here if you look.
In particular Old PC power supplies typically have one of similar size but
the wrong orientation.
I am
There are practice cores in most old electronics, which are everywhere
around here if you look.
In particular Old PC power supplies typically have one of similar size but
the wrong orientation.
I am
|
By
ykochcal
·
#160120
·
|
|
Re: 2465 DVS Repair - was; Re: [TekScopes] Wanted; Person to repair Tektronix 2465 DVS
Assuming TP46 is the PCB edge side of R567 It has about 1VAC 60hz maybe a 20 ms dc offset. It doesn't change as I exit the error message or _fiddle_ with the triggering adjustments.
- Jeff
Assuming TP46 is the PCB edge side of R567 It has about 1VAC 60hz maybe a 20 ms dc offset. It doesn't change as I exit the error message or _fiddle_ with the triggering adjustments.
- Jeff
|
By
jhalbrecht
·
#160121
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
Just remember the ferrite is a ceramic mixture,
so it was once heated much hotter than your oven
will ever go.
Good gloves so you don't burn yourself, don't worry
about the winding. Soak it in the
Just remember the ferrite is a ceramic mixture,
so it was once heated much hotter than your oven
will ever go.
Good gloves so you don't burn yourself, don't worry
about the winding. Soak it in the
|
By
Chuck Harris
·
#160122
·
|
|
Re: 2465 DVS Repair - was; Re: [TekScopes] Wanted; Person to repair Tektronix 2465 DVS
It should be exactly 0.8Vp-p, with no offset. See the waveform
picture in the manual.
The scope's power on test routing attempts to measure the peaks
of that sine'ish waveform as a verification of
It should be exactly 0.8Vp-p, with no offset. See the waveform
picture in the manual.
The scope's power on test routing attempts to measure the peaks
of that sine'ish waveform as a verification of
|
By
Chuck Harris
·
#160123
·
|
|
Re: 11801 NVRAM
FWIW I set up my 11801 using the internal trigger and feeding a ~100 kHz ramp from my 33622A. By making slight adjustments to the ramp frequency I get a straight line which intersects the graticule
FWIW I set up my 11801 using the internal trigger and feeding a ~100 kHz ramp from my 33622A. By making slight adjustments to the ramp frequency I get a straight line which intersects the graticule
|
By
Reginald Beardsley
·
#160124
·
|
|
Re: Tek 576 Curve Tracer HV Transformer winding
Ha. This reminds me of Jeff Potter's book "Cooking for Geeks", where he intentionally disables the self-cleaning interlock on his range to see how fast he can cook a pizza. ISTR that he got the oven
Ha. This reminds me of Jeff Potter's book "Cooking for Geeks", where he intentionally disables the self-cleaning interlock on his range to see how fast he can cook a pizza. ISTR that he got the oven
|
By
Dave Daniel
·
#160125
·
|