Re: Tek 575 tube adapter
magnustoelle
Hi Julian,
I am glad to see you are making good (further use) of your 575. We just had that topic some months ago. I'd say give the search function a try and it shall come up with this: http://glydeck.blogspot.de/2012/02/testing-miniature-pentodes-with-tek-575.html and George's postings on subject such as http://tech.groups.yahoo.com/group/TekScopes/message/84878 He did some quite marvelous things... Cheers, Magnus
|
|
Re: TEK 453 Vert Amp - not 453A
Mark Wendt <mark.wendt@...>
On 05/20/2013 12:46 PM, David wrote:
The 453A manual uses a thick light blue line for the demarcation,Okay, mea culpa. In my defense, I did say I was a bit dense. I just realized what I thought were circuit traces were the board edges... Doh! Mark
|
|
Re: FIXED: TDS544A error "acqdataconf"
Hi Andy,
toggle quoted messageShow quoted text
Nice job. For future reference, I don't think the memory controllers and SRAM chips are laid out in order 4, 3, 2, 1 from left to right of the ACQ board (with the scope upside down and the BNCs facing you) like one would think lining up with the input channels. I believe that in this orientation it is 3, 4, 1, 2. I will check later on when I get home - I have an old parts board that I marked up. I'm not even sure that the memory map I posted earlier has the channel references correct; I may have done that before I realized that the channels were not in order. In fact it could also be 4, 2, 3, 1. Jay
--- In TekScopes@yahoogroups.com, "baltimora86" <acuffe@...> wrote:
|
|
Re: 434 Storage Scope triggering problem
fiftythreebuick <ae5i@...>
It was indeed. I have an early one and a late one and they are both very nice scopes. The later one has a much more elaborate power supply control system than the early one. Both work quite well though....
toggle quoted messageShow quoted text
Tom
--- In TekScopes@yahoogroups.com, larrys@... wrote:
very nice bistable storage scope.text removed<< Properly adjusted, the 434 was a
|
|
TDS 784D CRT Adjustment Questions
W1WF <hwashcraft@...>
I am working on a TDS 784D, Serial # B010510
I have removed and installed a new CRT/LCD Shutter assembly and now need to adjust the display. The manual I have tells me to set the dip switches closed except 3, 6 & 7 that should be open, start the scope, and then enter the utility menu and display the composite test screen. (So far, so good). I am then supposed to adjust the trace rotation using R401 (described as second adjustment from the fan), brightness using R403 and contrast using R404. The problem is that the main board doesn't have R401, R403 and R404. I have looked all over the unit and can't find any adjustable resistors where the manual says they should be. So...where are they? Or do they not apply to this serial number? And if so, how do you make the adjustments? Thanks, Howard
|
|
Re: 434 Storage Scope triggering problem
VR635 Zener, fed through R636 from the +15 volt supply. The Zener is shown as 5.1 V on the schematic. (Trigger Pickoff and Generator) P 181
toggle quoted messageShow quoted text
of the manual at w140.com/tektronix_434_service.pdf near the top and towards the right side of the page.
-----Original Message-----
From: cybertheque_museum <msg.together@gmail.com>
Michael A. Terrell
|
|
Re: Tek 7000 Series deflection plate precision?
Don Black <donald_black@...>
I'm no expert on electron microscopes but you should be aware of two
types, TEM (Transmission Electron Microscopes) and SEM (Scanning
Electron Microscopes). A TEM shines a beam of electrons through the
specimen as if though a photographic slide and then expands the beam
onto a fluorescent screen to obtain the image (often viewed with an
optical microscope to see small details). A SEM scans the target
with a raster of s fine electron beam. The beam in a SEM is
relatively large since it's like an illuminating light source but
the SEM needs a very fine beam, smaller in diameter than the
smallest detail you are trying to see. I believe the high school
microscope is a TEM and so doesn't need as fine a beam as a SEM.
toggle quoted messageShow quoted text
Others who really know what they are talking about please jump in and correct any mistakes I've made here please. Don Black.
On 21-May-13 2:36 PM, iglesia_cristiana_arpas_eternas wrote:
|
|
Re: Tek 7000 Series deflection plate precision?
iglesia_cristiana_arpas_eternas
Hi Damian,Here my two cent about:
toggle quoted messageShow quoted text
You need to put an extra objective lens "magnetical prefereble" in order to obtain a very small dot at least half what minimum size yo like to see.and ..or.. increase the tube length, for the same reason. Most people use a Cristal plate with phosphor where the second emission electron emitted from the target impact and release a photon ,which in turn will be detected by the photomultiplier itself, I believe that such optical conversion is not necessary , just remove the cap on photomultiplier and allow the electron hit the dinodes directly ,If this approach work, surely will be give you a enormous gain and increase in S/N ratio, which in turn allow to use less beam intensity with all benefits.obviously all in high vacuum, inside the chamber. End of my two cent. Gabriel.
--- In TekScopes@yahoogroups.com, "cheater00 ." <cheater00@...> wrote:
|
|
Re: Tek 7000 Series deflection plate precision?
Don Black <donald_black@...>
Maybe an image orthicon rather than a vidicon? They have an electron
multiplier as part of the tube. I think you might find one to try
but I hope there aren't too many IOs destroyed for the project.
toggle quoted messageShow quoted text
Don Black.
On 21-May-13 12:13 AM, cleyson@... wrote:
|
|
Re: 434 Storage Scope triggering problem
teamlarryohio
"Ed Breya" <edbreya@yahoo.com> wrote:
BTW, I think the 434 is one of the nicest scopes ever made. I don'tOne thing the 432 / 434 brought was an early SMPS predating the custom Tek-made control IC. Properly adjusted, the 434 was a very nice bistable storage scope. -ls-
|
|
Re: Tek 7000 Series deflection plate precision?
Carlos
Brad: I'm not familiar with electron microscopes, but I remembered having read many years ago an article about an electron microscope built by students. It was featured in the 'Amateur Scientist' section of Scientific American magazine (when it was an excellent publication). I just found the article here: http://jesseenterprises.net/amsci/1973/09/1973-09-fs.html Maybe this could be of some interest for you. (By the way, I found also a source for the book 'The Scientific American book for the Amateur Scientist' by Clair L. Stong, published in 1960. 605 pages of very interesting material, in my opinion: http://www.sciencemadness.org/library/books/projects_for_the_amateur_scientist.pdf Just in case someone is interested) Regards, Carlos
|
|
Re: 434 Storage Scope triggering problem
Ed Breya
I'd guess that the guts are OK, but you have a switch or pot (including tweaks) that has contact problems. If you exercise and sufficiently jiggle all of the controls a few dozen times, it may start acting more normal - or not.
toggle quoted messageShow quoted text
You seem to know what you're doing, so you know you can figure it out if you have the right info - instruction manual, etc, and you can access the points to make the right measurements. I think your results with going around the normal TD bias are key to understanding what's going on. Yes, the TDs and associated passive components can drift out of spec, but I'd look at the electromechanical stuff first. BTW, I think the 434 is one of the nicest scopes ever made. I don't have one, nor have I seen or used one since over thirty years ago - it made a great and lasting first impression on me. Ed
--- In TekScopes@yahoogroups.com, "cybertheque_museum" <msg.together@...> wrote:
|
|
434 Storage Scope triggering problem
cybertheque_museum
Advice sought regarding the following failure on a 434 storage scope and also a question about the power supply:
1. Scope would not sweep regardless of trigger source, level, polarity; single sweep inoperative (and lamp would not light) and auto mode also would not produce a trace. Waveform tracing up to Q645 base was ok (as per manual) -- tested tunnel diodes CR644 and CR675 on curve tracer(both good) -- tested Q645 on curve tracer - displayed parasitic oscillations but IsubC - VsubC curve family was in spec, however base lead broke off so replaced with a BC557B -- adjusted +15 volt supply (was +15.11 V) -- verified that ripple on all supply rails is in spec -- verified high voltages (above +15) are in spec At this point, in AUTO or NORM mode could get a single or few sweeps by switching the slope switch back and forth when the level control was adjusted to produce correct waveform on TP640 (as per manual). Verified that AUTO mode monostable was working properly (AUTO mode same as NORM however - no AUTO). Discovered that by injecting a current of about 730 uA into TP678 a stable display with proper triggering can be obtained (used a 100 ohm resistor to a bench supply) - slope switch works as intended. -- with kludge, verified all sweep rates work and are linear -- verified triggering works at 1/3 division (has some instability) -- single sweep runs continuously (no difference to NORM except lamp lights or flickers with sweep rates) -- AUTO mode same as NORM (no auto). Interior of scope is very clean, no burnt smells or visible discolorations, no brittle insulation, etc. Injecting current into the TD moves its operating point enough to permit trigger to begin and end the sweep but the auto mode and single sweep depend on a reset current which could be still out of range. There are a number of connections to '+5 V' in these circuits but the voltage measures at +5.14 V (I have been unable to locate the source of +5 V in the diagrams). How likely is component aging out of specs in this scope (especially resistors)? This scope was working well, including all of the storage controls, until it abruptly had this failure when powered up two years ago. So far I have not found any 'dead' semiconductors. Within the tolerances specified, the measured voltages at test points and marked points on the schematic are in range for the specified operating conditions, as are the waveforms. 2. Where is the source of +5 VDC? (manual does not say and I looked carefully over each schematic but failed to find it). Thanks, Michael
|
|
Re: 1N3718 tunnel diode testing (475 scope trigger TD)
cybertheque_museum
Using a B&K 501 curve tracer instead of a breadboard kludge, I can verify that all four TDs display good characteristic curves and the funky shift of the negative resistance region I had previously observed is absent.
I am taking a break from work on the 475 in order to repair a triggering problem on a 434 since I need a better scope on the bench than the Ballentine 1032A for working the the 475.
|
|
Tektronix 502 available in Ottawa
r_corriveau
Someone contacted me to see if I was interested in a Tektronix 502. It is located in Ottawa, Canada. If anyone local is interested please contact me and I will foward the message to him.
Robert
|
|
7854 parts 'scope on eBay
Dave Daniel
I found this while perusing eBay - Looks like a(n) (currently) affordable parts mule.
I know there were several people looking for 7854 parts a while back ... DaveD
|
|
FIXED: TDS544A error "acqdataconf"
Andy
It was U217, one of the SRAMs for channel 3. I found the problem by replacing one SRAM, and then reinstalling the one I just replaced in the next location, and repeating until the error went away. I hit it on the third try, and only needed one replacement chip.
toggle quoted messageShow quoted text
Andy
--- In TekScopes@yahoogroups.com, "baltimora86" <acuffe@...> wrote:
|
|
Re: Tek 575 tube adapter
lindberg.adam
I usually use a 1kohm resistor connected between the base and emitter connection, doing so converts the current markings to voltage markings in the step generator.
so 1mA = 1V and 0,5mA = 0,5V and so on this is also handy when measuring mosfets I built a small rigg with different tube bases and banana jacks to patch the heater and external g2 source if needed then you can just connect the rigg to the curve tracer.
|
|
Re: Repair question for Type D High-Gain Differential plug-in
I once had the joy of having to find 16 matched 6146 out of 150 tubes. This was for a video modulator, where they were run in parallel to get the required power and a low output impedance. I managed two almost matched sets, but they aged rapidly in use. Wasn't '50's brute force technology interesting? ;-)
toggle quoted messageShow quoted text
-----Original Message-----
From: "tubesnthings@aol.com" <tubesnthings@aol.com>Michael A. Terrell
|
|
U400 (155-0236-00) Channel Switch replacement
thomas.lafay
Greetings. I have a replacement for the U400 in development, revision B, that is near finished in layout amd am wondering if there's any interest in this project. It would be nice to have others review the schematic before layout is completed. Send me an email at my yahoo address and I'll forward a copy of the schematic to you. Tom
|
|