Barry,
Goran is right about that area. VR3147 being open will throw the circuit off. C3127 and C3129 are marked ,1mfd. Some models use 1mfd in this place. I have replaced mine with the 1mfd 63V film types because the ESR of the originals was too high causing too much ripple and low B+. After replacing the parts, the B+ voltages at the pins on the 0067 IC were correct. These parts can be checked while the parts you have on order are being shipped. I also replaced the 1 and 2mfd electrolytics on the inverter board with film types of a higher voltage. The originals were bad in mine. Mark
|
|
Re: 465M
Bert Haskins
On 9/13/2021 12:16 PM, Michael W. Lynch via groups.io wrote:
On Mon, Sep 13, 2021 at 10:57 AM, Michael wrote:I have had a total of six.I have owned one for many years. It still works perfectly. I mainly use it forOh, Yes. Even though they are not as rugged as the other 4XX series, they are lighter and just as competent in performance. TEK certainly did not cheap out in the performance department. I like my 455 for many tasks and find it to be a fine scope as well. Now it's one plus a parts mule which could be restored if I could find a handle with all hardware. The one that's complete serves mainly as a display for any of my many curve tracers. A huge advantage of the 465M is that if it's the lower scope in a stack, you don't have to worry about blocking any heat vents. Every one of these had bad 2.0 and 2.2 caps in the horizontal power supply section! - Bert
|
|
Re: Tektronix 067-1338-0x calibration step generator for SD-series sampling heads
On Mon, 13 Sep 2021, Ken Eckert wrote:
Thought this might be of interest................That "Add to basket" on UK site is a charm :) --- ****************************************************************** * KSI@home KOI8 Net < > The impossible we do immediately. * * Las Vegas NV, USA < > Miracles require 24-hour notice. * ******************************************************************
|
|
Re: 7704A (et. al.) Inverter Regulator Question
n4buq
Until I get that bad transistor replaced, then I don't plan on trying it connected to the display unit as a load. I tried that before I was aware of the bad transistor. Once I get that, I do plan to test it connected to the display unit. If it's still ticking, then it's either a problem with the display unit or the PS. In that case, the external load resistors will at least eliminate the display unit from the equation. I'll report back once I get the transistor replaced.
toggle quoted messageShow quoted text
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Göran Krusell" <mc1648pp@gmail.com>
|
|
Re: 7704A (et. al.) Inverter Regulator Question
n4buq
Very good! I appreciate the link.
toggle quoted messageShow quoted text
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Ozan" <ozan_g@erdogan.us>
|
|
Re: 7704A (et. al.) Inverter Regulator Question
Hi, I have not followed this discussion very closely but my guess is that you are looking in the wrong direction.
You need to check U3105 pin 2, BAL SENSE, where the voltage should be 0V in normal operation. If it deviates more than 200mV (+ or -) then the PS should turn off. Also check U3105 pin 13, I SENSE, where the voltage shall be positive, perhaps +0.23V, in normal operation. In an over-current situation the voltage drops below 0V and the PS should turn off. If you hear a hissing and sparking sound you may have a problem with your voltage multiplier. Carry on and good luck Göran
|
|
Re: 7704A (et. al.) Inverter Regulator Question
Ozan
Following link also has debug information on 7904 PSU which may contain useful information for 7704A PSU:
https://w140.com/Tek_7904_Power_Supply.pdf Ozan
|
|
Re: FG502 No wave
Ozan
Hi Chris,
There could be multiple issues. You can jump to “Collector voltage at Q315 = -0.6V at Q325 = -3.5V.” if you are looking for a quick answer. Turn the dial so that output (pin 6) of U135 is about +10V, pick triangle waveform, pick 10^3 range.Earlier you measured 7.28 to 19.7V which is about what is expected. Although 9.8V is fine for analysis something seems to be wrong with R110, is the pot securely mounted with no loose nuts? This is not the root cause of the “no oscillation” problem but something that needs to be debugged separately later. Voltage across R140 = +2.66V, across R175 = -2.52V, measured with + of DVM probe on collector of transistors.Thank you for giving the + reference, it eliminates any ambiguity. This says up and down currents are about 5mA. There is slight mismatch in currents, and the value is higher than expected but this should still make the circuit oscillate. This step eliminates the current source circuit as the suspect. We can re-visit the currents when you fix R110. Collector voltage of Q292 = +0.15V, R155 is quite hot and voltage at supply side is + 16.23V. No sign of any wave.Collector of Q292 should be either +5V or -5V (comparator output) in normal operation. +0.15V is not right but see below. R155 burns 16.23^2/820=0.3W. Should be warm but not very hot. If it is getting really hot you may want to measure its resistance, we are looking for much smaller than 820-ohm so in circuit testing can give you an idea. See the next item before checking it though. Collector voltage at Q315 = -0.6V at Q325 = -3.5V.This portion is the “gate function”. This is (one of?) the reasons for stopped oscillations. If collector voltage of Q325 is -3.5V it will turn on Q315 and it will clamp collector of Q292 to stop the oscillations. You can short collector of Q325 to its emitter with a piece of wire, if you get waveforms out then this is it. As for debug: I assume nothing is connected to the “GATE IN” BNC. Base of Q332 should be about two diode voltages above ground ~ 1.2-1.4V. This should turn off Q332 and turn on Q330 which supplies base bias to Q325. Base of Q325 should be one diode drop above is emitter (0.7V above -17V). This should turn on Q325 and make its collector voltage ~ -17V. You are measuring -3.5V so something is not working. In normal operation Q315 should be off but in your case I suspect Q315 is on. You can confirm by measuring voltage drop across R318 which should be 0V in normal operation. I recommend measuring base, emitter, collector voltages of Q325, Q330, and Q332. Rest can wait until this is debugged. Your “GATE” circuit is gating off the oscillations. Base voltage of Q230A = +0.14V (for reference base voltage at Q230B = +.05V) Both A and B were checked out of circuit and appear to be OK.Base of Q230A is OK given other side of the bridge is at +0.15V. Base of Q230B depends on emitter voltage of Q210, which is OK for your measurement below. Gate voltage of Q200 = +7.9V fixed i.e.. not going up / down. Please confirm that the gate connection is what would be the base if it were a transistor.Yes, you can think gate of a JFET similar to base of a transistor. Just to make sure: gate terminal is the one connected to cathode of CR145 and anode of CR175. I suspect this measurement is not at the gate of Q200 because of the following point. Emitter voltage of Q210 = +0.16V This transistor gets very hot (not unexpectedly, as it is fitted with a heat sink). Q204 also gets very hot.Emitter of Q210 should try to follow gate of Q200. We expect gate of Q200 to be ~ 0.15V and the emitter is at 0.16V which would be correct if we ignore the 7.9V measurement. Could you confirm gate voltage of Q200 again, which is the same node as cathode of CR145 and anode of CR175. Ozan
|
|
Re: 465M
On Mon, Sep 13, 2021 at 10:57 AM, Michael wrote:
Oh, Yes. Even though they are not as rugged as the other 4XX series, they are lighter and just as competent in performance. TEK certainly did not cheap out in the performance department. I like my 455 for many tasks and find it to be a fine scope as well. -- Michael Lynch Dardanelle, AR
|
|
Re: "mottled effect" on anodized front panel
Greg Muir
I’ve seen similar effects from contaminated anodizing solutions. If the rinse is inadequate after the item leaves the electrolyte bath some of the acid can remain in the porous aluminum oxide layer during sealing. Over time the remaining acid can slowly affect the finish.
Greg
|
|
Re: 465M
Michael
I have owned one for many years. It still works perfectly. I mainly use it for stereo and hf transceiver repair work.
|
|
Re: "mottled effect" on anodized front panel
Chris Wilkson
On Sun, Sep 12, 2021 at 09:34 AM, Tom Norman wrote:
Slight correction...Novus #3 is the MOST aggressive. Beware! Novus #1 is for cleaning and polishing. It's non-abrasive and pretty safe. Novus #2 is for fine scratch removal and heavy cleaning. It is mildly abrasive and will damage fine finishes and artwork. Novus #3 is for heavy scratch removal. VERY ABRASIVE!! It will destroy the finish and remove artwork completely. Novus is great for restoration and maintenance. Used extensively in pinball and video arcade hobbies and some automotive applications. Always start with #1. Use #2 gently to remove fine scratches and blemishes. #3 is rarely used...only in extreme restoration cases.
|
|
Re: 7704A (et. al.) Inverter Regulator Question
n4buq
Hi Albert,
toggle quoted messageShow quoted text
No, I hadn't seen that. I've downloaded that manual and am reviewing those sections. That should be a big help. Thanks! Barry - N4BUQ
----- Original Message -----
From: "Albert Otten" <aodiversen@concepts.nl>
|
|
Tektronix 067-1338-0x calibration step generator for SD-series sampling heads
Ken Eckert
|
|
Re: help with identity of 7000 series PSU part no 620-0276-00
Hi Hakan , thank you for that info , I'll have to keep looking or try harder to fix the supply in my 7834 which just sits there and ticks .
toggle quoted messageShow quoted text
Brian
On Monday, 13 September 2021, 16:42:35 BST, zenith5106 via groups.io <hahi=telia.com@groups.io> wrote:
On Mon, Sep 13, 2021 at 04:21 PM, Brian wrote: 620-0276-00 was used in 7104 only and in case of replacement 620-0283-02 should be used. 620-0283-02 was used in 7103, 7104, 7704A, 7854, 7904A and 7934. No replacement info on 620-0250-00 which was used in 7834 possibly because 7834 went out of support in 1995. /Håkan
|
|
Re: help with identity of 7000 series PSU part no 620-0276-00
On Mon, Sep 13, 2021 at 04:21 PM, Brian wrote:
620-0276-00 was used in 7104 only and in case of replacement 620-0283-02 should be used. 620-0283-02 was used in 7103, 7104, 7704A, 7854, 7904A and 7934. No replacement info on 620-0250-00 which was used in 7834 possibly because 7834 went out of support in 1995. /Håkan
|
|
Re: TCP202 with 15mA of ~150KHz noise.
Andy Warner
I tried adding some ferrites with decent impedance in the ~100KHz range to
the +/- 15V rails: https://photos.app.goo.gl/RaxHFTVdvF6T7LAJA No significant change, next step repeat for the +/- 5V rails, and if that doesn’t help, move on to augmenting the decoupling… On Fri, Sep 10, 2021 at 16:17 Andy Warner <andyw@pobox.com> wrote: Thank you for the datapoint, I wish I were so lucky.Andy
|
|
Re: 7704A (et. al.) Inverter Regulator Question
Albert Otten
Hi Barry, just in case it escaped your attention: for instance the 7904A manual contains a more extensive fault finding procedure for the similar SMPS.
Albert
|
|
502A Dual Beam Oscilloscope on Shopgoodwill
Bert Haskins
No affiliation I just watch this hoping to see a good 7603 ( not a rackmount ) come up.
|
|
Re: Suggestion for our group Administrator
FWIW I've seen a spam of microedits just this month on HPAK - and I
toggle quoted messageShow quoted text
barely even read that list.
On Sun, Sep 12, 2021 at 6:43 PM Jim Adney <jadney@vwtype3.org> wrote:
|
|