Re: Type 519 -- Why 125 ohms?

Harvey White
 

IIRC, that's the impedance, or close to it, of twisted pair non shielded cable.

Harvey

On 5/10/2021 9:54 PM, Sean Turner wrote:
Pondering the beast that is sitting in my lab got me thinking, why 125 ohm? Does anyone know of the history of why this was chosen? I haven't been able to find much about this, so insights are appreciated!

Sean





Re: Type 519 -- Why 125 ohms? Zo formula

Jeff Kruth
 

Hi Sean!A reasonable swag is that it is about the top end of coax impedance, hard to make a higher Z as OD of inner conductor gets vanishingly small. 73Jeff Kruth In a message dated 5/10/2021 9:54:24 PM Eastern Standard Time, sdturne@q.com writes: 
Pondering the beast that is sitting in my lab got me thinking, why 125 ohm? Does anyone know of the history of why this was chosen? I haven't been able to find much about this, so insights are appreciated! Sean


Re: TM-500 Test Module Question

Jared Cabot
 

Ok, I just updated the photo album with a description of the fix for future reference. This should make it clear how to perform the fix.

Let's call it a 'service update', just like you find with all good test equipment... :)


https://groups.io/g/TekScopes/album?id=263748


Re: Chaining power supplies together.

Torch
 

At the risk of being branded a philistine, not all Chinese power supplies are completely useless or disposable. I bought one of the early Korad KA3305P units and have been quite satisfied with it. Effectively 3 power supplies in 1, and the two variable 31V, 5.1A outlets can be chained in series or parallel with the push of a button. IE: it can deliver 62V @ 5.1A or 31V @10.2A.

I did a tear down at the time, which can be seen here: https://www.eevblog.com/forum/testgear/inside-the-new-korad-ka3305p-linear-psu/


Type 519 -- Why 125 ohms?

Sean Turner
 

Pondering the beast that is sitting in my lab got me thinking, why 125 ohm? Does anyone know of the history of why this was chosen? I haven't been able to find much about this, so insights are appreciated!

Sean


Re: Looking for sensitive audio RMS meter

Liam Perkins
 

I'm going thru a PAR CR-4A LNA for a guy in Germany who wants to use it as
an effects box with his synth'. There are a couple of YTs by another guy
over there who loves his and has it make the God Awfullest noises you ever
heard, and he loves it. I think those guy get a kick out torturing
germanium . . .

On Mon, May 10, 2021 at 7:05 PM George Kerber <gk5220@gmail.com> wrote:

Sorry about that. Best bet is to check out eBay listings for 323. It's
probably really pricey when purchased directly from Ballantine (I was about
$700 new in the 1970's). I was surprised to see it on their website.

FFT spectrum analyzer is perfect for noise measurements. You get the
spectrum and total noise level number if needed. I use a, not so portable,
HP 35665A. A good low noise preamp such as PAR 113 is a good complement to
any RMS meter or FFT analyzer.

George

On Mon, May 10, 2021 at 5:41 PM cheater cheater <cheater00social@gmail.com
wrote:

Hi George,

On Tue, May 11, 2021 at 1:58 AM George Kerber <gk5220@gmail.com> wrote:

Bob,
Actually I'm the one looking for the volt meter :-)

You might consider a HP distortion analyzer such as this one:
http://www.barrytech.com/hewlett-packard/analyzers/hp339a.html
https://www.ebay.com/itm/233587661597?hash=item3662e8db1d:g:KrMAAOSwiGBevSVB

I love the fact that it's got a dBV display which is what I'd like to
be doing my measurements in anyways (it's for audio equipment).

Also, a Ballantine 323 True RMS analog AC voltmeter. Only measure to
300
uV.
https://www.ballantinelabs.com/323/
https://www.ebay.com/itm/384146432011?hash=item5970e9600b:g:j8MAAOSwcXBgk0HN

Thanks, any clue how much the Ballantine 323 costs? They don't have
any prices on their website. However, ebay shows units going for sub
$100 which is pretty great.

Any chance of being able to convert its scale to dBV? I would love to
not have to pull out a calculator every time I measure something.

Both are somewhat portable.

An HP 3561A 100kHz FFT spectrum analyzer would probably meet your spec,
it's portable, but really hard to lug around.
I'm afraid it might be way too large for me. Always liked how it
looks, but it's just not practical in my circumstances, and also I'm
not looking for a spectrum analyzer.

On Tue, May 11, 2021 at 2:35 AM George Kerber <gk5220@gmail.com> wrote:

You might also consider soundcard-based FFT for rms and spectrum
measurements to 100kHz and a laptop. There are a lot of choices and
price
ranges. One professional package is by Pioneer Hill Software:
https://www.spectraplus.com/ Very nice, but $$$.

Another is Spectrum Lab, by DL4YHF Spectrum Lab (Audio Signal
Analyser).
I
have played with it a little, but it has a very steep learning curve.

George
I'm looking for a meter to measure total noise level, rather than its
spectrum. I agree that FFT measurements are useful in their own right.

I'm on a short budget. If I had the money I'd just buy Audio Precision.










Re: Looking for sensitive audio RMS meter

Liam Perkins
 

I just lit up the 3457s with inputs shorted and they pass self-test. One of
them has Opt 007 which I -think- might be the 7th digit option but it's a
slog thru about 5lbs of manual to sort that for certain.
You started here saying tens of uV now you're into nVs which is a place
that ought to have a gate saying, "Abandon Hope All Ye Who Enter". Read a
slew of Jim Williams app notes from, I think, Analog Devices and start
buying coffee in 2 and 5lb tins because you're gonna need 'em nested like
Russian dolls.
What -are- you wanting to do ?

On Mon, May 10, 2021 at 7:05 PM Bill Perkins <hifi@telus.net> wrote:

The 465A is an about 1960 design and isn't even in the ballpark

On Mon, May 10, 2021 at 6:46 PM Göran Krusell <mc1648pp@gmail.com> wrote:

hp 465 A amplifier!






Re: Looking for sensitive audio RMS meter

Liam Perkins
 

The 465A is an about 1960 design and isn't even in the ballpark

On Mon, May 10, 2021 at 6:46 PM Göran Krusell <mc1648pp@gmail.com> wrote:

hp 465 A amplifier!






Re: Looking for sensitive audio RMS meter

George Kerber
 

Sorry about that. Best bet is to check out eBay listings for 323. It's
probably really pricey when purchased directly from Ballantine (I was about
$700 new in the 1970's). I was surprised to see it on their website.

FFT spectrum analyzer is perfect for noise measurements. You get the
spectrum and total noise level number if needed. I use a, not so portable,
HP 35665A. A good low noise preamp such as PAR 113 is a good complement to
any RMS meter or FFT analyzer.

George

On Mon, May 10, 2021 at 5:41 PM cheater cheater <cheater00social@gmail.com>
wrote:

Hi George,

On Tue, May 11, 2021 at 1:58 AM George Kerber <gk5220@gmail.com> wrote:

Bob,
Actually I'm the one looking for the volt meter :-)

You might consider a HP distortion analyzer such as this one:
http://www.barrytech.com/hewlett-packard/analyzers/hp339a.html
https://www.ebay.com/itm/233587661597?hash=item3662e8db1d:g:KrMAAOSwiGBevSVB

I love the fact that it's got a dBV display which is what I'd like to
be doing my measurements in anyways (it's for audio equipment).

Also, a Ballantine 323 True RMS analog AC voltmeter. Only measure to 300
uV.
https://www.ballantinelabs.com/323/
https://www.ebay.com/itm/384146432011?hash=item5970e9600b:g:j8MAAOSwcXBgk0HN

Thanks, any clue how much the Ballantine 323 costs? They don't have
any prices on their website. However, ebay shows units going for sub
$100 which is pretty great.

Any chance of being able to convert its scale to dBV? I would love to
not have to pull out a calculator every time I measure something.

Both are somewhat portable.

An HP 3561A 100kHz FFT spectrum analyzer would probably meet your spec,
it's portable, but really hard to lug around.
I'm afraid it might be way too large for me. Always liked how it
looks, but it's just not practical in my circumstances, and also I'm
not looking for a spectrum analyzer.

On Tue, May 11, 2021 at 2:35 AM George Kerber <gk5220@gmail.com> wrote:

You might also consider soundcard-based FFT for rms and spectrum
measurements to 100kHz and a laptop. There are a lot of choices and price
ranges. One professional package is by Pioneer Hill Software:
https://www.spectraplus.com/ Very nice, but $$$.

Another is Spectrum Lab, by DL4YHF Spectrum Lab (Audio Signal Analyser).
I
have played with it a little, but it has a very steep learning curve.

George
I'm looking for a meter to measure total noise level, rather than its
spectrum. I agree that FFT measurements are useful in their own right.

I'm on a short budget. If I had the money I'd just buy Audio Precision.






Re: Looking for sensitive audio RMS meter

Göran Krusell
 

hp 465 A amplifier!


Re: Looking for sensitive audio RMS meter

 

Hi George,

On Tue, May 11, 2021 at 1:58 AM George Kerber <gk5220@gmail.com> wrote:

Bob,
Actually I'm the one looking for the volt meter :-)

You might consider a HP distortion analyzer such as this one:
http://www.barrytech.com/hewlett-packard/analyzers/hp339a.html
https://www.ebay.com/itm/233587661597?hash=item3662e8db1d:g:KrMAAOSwiGBevSVB
I love the fact that it's got a dBV display which is what I'd like to
be doing my measurements in anyways (it's for audio equipment).

Also, a Ballantine 323 True RMS analog AC voltmeter. Only measure to 300 uV.
https://www.ballantinelabs.com/323/
https://www.ebay.com/itm/384146432011?hash=item5970e9600b:g:j8MAAOSwcXBgk0HN
Thanks, any clue how much the Ballantine 323 costs? They don't have
any prices on their website. However, ebay shows units going for sub
$100 which is pretty great.

Any chance of being able to convert its scale to dBV? I would love to
not have to pull out a calculator every time I measure something.

Both are somewhat portable.

An HP 3561A 100kHz FFT spectrum analyzer would probably meet your spec,
it's portable, but really hard to lug around.
I'm afraid it might be way too large for me. Always liked how it
looks, but it's just not practical in my circumstances, and also I'm
not looking for a spectrum analyzer.

On Tue, May 11, 2021 at 2:35 AM George Kerber <gk5220@gmail.com> wrote:

You might also consider soundcard-based FFT for rms and spectrum
measurements to 100kHz and a laptop. There are a lot of choices and price
ranges. One professional package is by Pioneer Hill Software:
https://www.spectraplus.com/ Very nice, but $$$.

Another is Spectrum Lab, by DL4YHF Spectrum Lab (Audio Signal Analyser). I
have played with it a little, but it has a very steep learning curve.

George
I'm looking for a meter to measure total noise level, rather than its
spectrum. I agree that FFT measurements are useful in their own right.

I'm on a short budget. If I had the money I'd just buy Audio Precision.


Re: Looking for sensitive audio RMS meter

George Kerber
 

You might also consider soundcard-based FFT for rms and spectrum
measurements to 100kHz and a laptop. There are a lot of choices and price
ranges. One professional package is by Pioneer Hill Software:
https://www.spectraplus.com/ Very nice, but $$$.

Another is Spectrum Lab, by DL4YHF Spectrum Lab (Audio Signal Analyser). I
have played with it a little, but it has a very steep learning curve.

George

On Mon, May 10, 2021 at 4:58 PM George Kerber via groups.io <gk5220=
gmail.com@groups.io> wrote:

Bob,
You might consider a HP distortion analyzer such as this one:
http://www.barrytech.com/hewlett-packard/analyzers/hp339a.html

https://www.ebay.com/itm/233587661597?hash=item3662e8db1d:g:KrMAAOSwiGBevSVB

Also, a Ballantine 323 True RMS analog AC voltmeter. Only measure to 300
uV.
https://www.ballantinelabs.com/323/

https://www.ebay.com/itm/384146432011?hash=item5970e9600b:g:j8MAAOSwcXBgk0HN

Both are somewhat portable.

An HP 3561A 100kHz FFT spectrum analyzer would probably meet your spec,
it's portable, but really hard to lug around.

George

On Mon, May 10, 2021 at 4:21 PM Bob Albert via groups.io <bob91343=
yahoo.com@groups.io> wrote:

I would be interested in one or both of those 3457s.
Bob K6DDX
On Monday, May 10, 2021, 04:13:47 PM PDT, Liam Perkins <
hifi@telus.net>
wrote:

I have a couple of HP 3457s in good nick and looking for a home.

We just parted out a pair of 3456As for gold scrap (broke my heart)
and
I can tell you the '57s are smaller and much lighter, same width give or
take but about 4" shallower front to back.

Liam

On Mon, May 10, 2021 at 3:18 PM cheater cheater <
cheater00social@gmail.com
wrote:

Hi all, looking for a true RMS meter that can do roughly ~DC to ~100
kHz measurement down to 1 uV and up to say tens of volts. Hoping to
find something inexpensive and can be calibrated. Can someone suggest
anything? Either Tek or HP/A/K or something else. Bench or portable is
fine. Battery operation a big plus either way. I'd like 5 digits or
more, but fewer is fine around 1 uV. My budget is a few hundred $.

Thanks



















Wanted - Tektronix 305 Oscilloscope Knobs and Caps

Sampler_one
 

I need 3 Tektronix 305 scope Parts ;
366-1057-00 Gray knob <<<<<<<<<<<<<<<<<<<<<<<<<<<<<Mine cracked in two and the scope is hard to use now.
366-1405-06 Red Knob , screw and cap set <<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<Just in case the old one breaks.
200-1342-00 Handel cover <<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<<Missing one to complete the scope.

Happy to pay a reasonable price / postage or trade.
I have a ton of Vacuum Tubes

Best regards
Please contact me.


Re: Looking for sensitive audio RMS meter

 

On Tue, May 11, 2021 at 1:11 AM Liam Perkins <hifi@telus.net> wrote:

I have a couple of HP 3457s in good nick and looking for a home.
Thanks Liam. It seems like the 3457's lowest range for AC voltage is
30 mV, and I need to be able to measure tenths of a microvolt. Correct
me if I'm wrong please. The specs I've found weren't very conclusive.

On Tue, May 11, 2021 at 1:35 AM Liam Perkins <hifi@telus.net> wrote:

Also, have a look at the erstwhile Geller Labs site on the WayBack here:
https://web.archive.org/web/20131018055721/http://www.gellerlabs.com/index.html
and dig for the J-Can article, you'll find it highly informative. The
ultra-low noise SMD JFET he specs is history but I have a bunch and they're
also available on eBay although it's anyone's guess what they might
actually be. Rather than looking for a uV capable meter you're probably
farther ahead to build a little +40dB gain box, battery power it and put
panel mount BNCs on it for in-s and out-s.
A Google for J-Can will pull his PDF on the J-Can, a worthwhile read in any
case.
Liam
Thanks, that's a super useful pointer. I've found an article that goes
deep into the theory and build, but doesn't actually tell me much
about what the thing does. So it's a +40 dB amplifier? What is its
self-noise?

Would you be willing to sell me a few of those jfets? I would probably
like to set up a balanced measurement.

Best regards


Re: Looking for sensitive audio RMS meter

George Kerber
 

Bob,
You might consider a HP distortion analyzer such as this one:
http://www.barrytech.com/hewlett-packard/analyzers/hp339a.html
https://www.ebay.com/itm/233587661597?hash=item3662e8db1d:g:KrMAAOSwiGBevSVB

Also, a Ballantine 323 True RMS analog AC voltmeter. Only measure to 300 uV.
https://www.ballantinelabs.com/323/
https://www.ebay.com/itm/384146432011?hash=item5970e9600b:g:j8MAAOSwcXBgk0HN

Both are somewhat portable.

An HP 3561A 100kHz FFT spectrum analyzer would probably meet your spec,
it's portable, but really hard to lug around.

George

On Mon, May 10, 2021 at 4:21 PM Bob Albert via groups.io <bob91343=
yahoo.com@groups.io> wrote:

I would be interested in one or both of those 3457s.
Bob K6DDX
On Monday, May 10, 2021, 04:13:47 PM PDT, Liam Perkins <hifi@telus.net>
wrote:

I have a couple of HP 3457s in good nick and looking for a home.

We just parted out a pair of 3456As for gold scrap (broke my heart) and
I can tell you the '57s are smaller and much lighter, same width give or
take but about 4" shallower front to back.

Liam

On Mon, May 10, 2021 at 3:18 PM cheater cheater <cheater00social@gmail.com
wrote:

Hi all, looking for a true RMS meter that can do roughly ~DC to ~100
kHz measurement down to 1 uV and up to say tens of volts. Hoping to
find something inexpensive and can be calibrated. Can someone suggest
anything? Either Tek or HP/A/K or something else. Bench or portable is
fine. Battery operation a big plus either way. I'd like 5 digits or
more, but fewer is fine around 1 uV. My budget is a few hundred $.

Thanks















Re: Manual for a Telequipment S61?

Tom Lee
 

Wow, very nice (and fast) work. Thanks, Håkan!

--Cheers,
Tom

--
Prof. Thomas H. Lee
Allen Ctr., Rm. 205
350 Jane Stanford Way
Stanford University
Stanford, CA 94305-4070
http://www-smirc.stanford.edu

On 5/10/2021 07:22, zenith5106 wrote:
On Sun, May 9, 2021 at 09:43 PM, Liam Perkins wrote:

that way I can make a proper PDF of it
The manual has been scanned and sent to Kurt to add on TekWiki.
In the meantime until it shows up there it can be found here: www.hakanh.com/dl/temp/S61_Instruction.pdf
OCR'ed and bookmarked, hope it is proper enough for you.
/Håkan




Re: Looking for sensitive audio RMS meter

Liam Perkins
 

Also, have a look at the erstwhile Geller Labs site on the WayBack here:
https://web.archive.org/web/20131018055721/http://www.gellerlabs.com/index.html
and dig for the J-Can article, you'll find it highly informative. The
ultra-low noise SMD JFET he specs is history but I have a bunch and they're
also available on eBay although it's anyone's guess what they might
actually be. Rather than looking for a uV capable meter you're probably
farther ahead to build a little +40dB gain box, battery power it and put
panel mount BNCs on it for in-s and out-s.
A Google for J-Can will pull his PDF on the J-Can, a worthwhile read in any
case.
Liam

On Mon, May 10, 2021 at 5:13 PM Liam Perkins via groups.io <hifi=
telus.net@groups.io> wrote:

I have a couple of HP 3457s in good nick and looking for a home.

We just parted out a pair of 3456As for gold scrap (broke my heart) and
I can tell you the '57s are smaller and much lighter, same width give or
take but about 4" shallower front to back.

Liam

On Mon, May 10, 2021 at 3:18 PM cheater cheater <cheater00social@gmail.com
wrote:

Hi all, looking for a true RMS meter that can do roughly ~DC to ~100
kHz measurement down to 1 uV and up to say tens of volts. Hoping to
find something inexpensive and can be calibrated. Can someone suggest
anything? Either Tek or HP/A/K or something else. Bench or portable is
fine. Battery operation a big plus either way. I'd like 5 digits or
more, but fewer is fine around 1 uV. My budget is a few hundred $.

Thanks










Re: Looking for sensitive audio RMS meter

Bob Albert
 

I would be interested in one or both of those 3457s.
Bob K6DDX

On Monday, May 10, 2021, 04:13:47 PM PDT, Liam Perkins <hifi@telus.net> wrote:

    I have a couple of HP 3457s in good nick and looking for a home.

    We just parted out a pair of 3456As for gold scrap (broke my heart) and
I can tell you the '57s are smaller and much lighter, same width give or
take but about 4" shallower front to back.

    Liam

On Mon, May 10, 2021 at 3:18 PM cheater cheater <cheater00social@gmail.com>
wrote:

Hi all, looking for a true RMS meter that can do roughly ~DC to ~100
kHz measurement down to 1 uV and up to say tens of volts. Hoping to
find something inexpensive and can be calibrated. Can someone suggest
anything? Either Tek or HP/A/K or something else. Bench or portable is
fine. Battery operation a big plus either way. I'd like 5 digits or
more, but fewer is fine around 1 uV. My budget is a few hundred $.

Thanks






Re: Looking for sensitive audio RMS meter

Liam Perkins
 

I have a couple of HP 3457s in good nick and looking for a home.

We just parted out a pair of 3456As for gold scrap (broke my heart) and
I can tell you the '57s are smaller and much lighter, same width give or
take but about 4" shallower front to back.

Liam

On Mon, May 10, 2021 at 3:18 PM cheater cheater <cheater00social@gmail.com>
wrote:

Hi all, looking for a true RMS meter that can do roughly ~DC to ~100
kHz measurement down to 1 uV and up to say tens of volts. Hoping to
find something inexpensive and can be calibrated. Can someone suggest
anything? Either Tek or HP/A/K or something else. Bench or portable is
fine. Battery operation a big plus either way. I'd like 5 digits or
more, but fewer is fine around 1 uV. My budget is a few hundred $.

Thanks






Re: Looking for sensitive audio RMS meter

 

Thanks. The specs are impressive. However, I am also hoping to find
something much more compact than that if that's possible - any
suggestions?

On Mon, May 10, 2021 at 11:20 PM Bob Albert via groups.io
<bob91343=yahoo.com@groups.io> wrote:

HP 3456A
On Monday, May 10, 2021, 02:18:40 PM PDT, cheater cheater <cheater00social@gmail.com> wrote:

Hi all, looking for a true RMS meter that can do roughly ~DC to ~100
kHz measurement down to 1 uV and up to say tens of volts. Hoping to
find something inexpensive and can be calibrated. Can someone suggest
anything? Either Tek or HP/A/K or something else. Bench or portable is
fine. Battery operation a big plus either way. I'd like 5 digits or
more, but fewer is fine around 1 uV. My budget is a few hundred $.

Thanks