[Tek2440] Java GUI feedback requested. #photo-notice

Jan de Jongh
 

Dear Tekkers,

Over the past two months or so, I've been working on a [Java] GUI for my Tek2440 accessed through a ProLogix GPIB-ETH . Since I still do not know all the Tek2440's features that well, I'm asking this group for feedback on the GUI I constructed so far. A screenshot is in Photos (don't know how to embed/refer; tips WELCOME!). In the figure, all gray rectangles pop up a dialog for controlling the particular group settings.

Main questions are:
- Should specific functions/settings now embedded in dialogs be available in the main application window?
- Should functions be moved into dialogs?

Needless to say, I cannot make all settings available on 900p/1080p, sp I have to make choices...

Thanks for your response; FYI, the code is on https://github.com/jandejongh/jinstrument although I'm planning on forking a Tek2440 specific project.

BR
jan - pa3gyf.


Re: Special Offer from Peter Keller to TekScopes Members: The Cathode Ray Tube, Technology, History, and Applications"

 

Is it too late to get on this lists?


Re: Difference between a 475 and a 475A

John Gord
 

Jeff,
My unresearched recollection is that the 475A was basically the same as the 475, but adjusted for higher bandwidth but perhaps slightly inferior pulse response. The 475A lacked the 2mV/div vertical setting.
--John Gord


Difference between a 475 and a 475A

 

I have a 475 that belonged to my father, manufactured in late 1974, that I have a sentimental attachment to and would like to keep running. To this end I have purchased several "for parts" scopes on eBay, both 475s and 475As, because many of the mechanical parts appear to be completely interchangeable. I had terrible luck with the first parts scope: while it is in terrible cosmetic condition (busted knobs, spattered with something that does not wash off with soap and water or with isopropyl alcohol), it appears to be operating flawlessly. Next I received a 475A which appears to be a much better donor candidate as it appears to be quite sick.

I have done thorough inspections of both scopes when I received them, and I noticed something odd: the sweep board of the first parts scope (475) does NOT look like the sweep board on my father's 475, but it DOES look almost identical to the sweep board on the 475A, which makes me wonder if it might not actually be a 475A with the faceplate of a 475.

I pulled down the schematics for both the 475 and the 475A and compared their vertical amplifier sections, on the basis that a higher bandwidth scope would require upgrades to the vertical amplifiers. I also made a thorough assay of the part numbers for the PCBs and custom ICs in all three scopes, hoping to find some definitive evidence for which scopes were 475s and which were 475As. Sadly, aside from the layout of the sweep board, I have not found anything conclusive.

The collected information on the three scopes follows:

475, vintage 1974
vertical amp board PN 670-224-01 or GH-2780-01
- switch IC PN 155-0091
- 1st pre-amp IC PN 155-0085-01
- 2nd pre-amp IC PN 434-0078-02 / 437-0078-02
main board PN 670-2239-04 or GE-2779-01
trigger board PN 670-2241-03 or GH-2781-01
- sweep IC PN 155-0059-01
- trigger ICs PN 155-0032-01
sweep board PN 670-2244 or GI-2784-05

475/475A? vintage 1976
vertical amp board PN 670-2240-08 or GH-2780-01
- switch IC PN 155-0091-00
- 1st pre-amp IC PN 155-0085-10
- 2nd pre-amp IC PN 019-0078-10 / 936-0078-10
main board PN 670-2239 or GE-2779-01
trigger board PN 670-2241-03 or GE-2781-02
- sweep IC PN 155-0049-02
- trigger ICs PN 155-0217-00
sweep board PN ? or GD-3980-02

475A, vintage 1978
vertical amp board PN 670-2240 or GH-2780-01
- switch IC PN 155-0091-00
- 1st pre-amp IC PN 155-0085-01
- 2nd pre-amp IC PN 813-0098-03
main board PN 670-2239 or GF-2779-01
trigger board PN 670-2241-02 or GA-2781-02
- sweep IC PN 155-0049-02
- trigger IC PN 155-0032-01
sweep board PN ? GD-3990-02

There are also differences between the parts in the 3rd pre-amp, but I'm not sure what to make of them because the physical circuit on the boards differs significantly from the schematics and parts lists.

Are the 475 and 475A really just the same scope with a marketing upgrade? I know that Tek over-engineered their scopes, and that a 475 may well be able to handle signals up to (and beyond) 250 MHz, but I expected to find noticeable upgrades to the vertical amplifiers. I suppose that it's possible that my father's scope was upgraded in the field (that is, by Tek field service personnel, not by my father), but other such modifications have been accompanied by option indicators, and there are not such indicators in my father's scope.

Have I been looking in the wrong place for the modifications that would distinguish a 475 from a 475A?

Are there, in fact, no such modifications, and the 475A was merely a marketing upgrade?

How would I be able to tell if a scope had had its front panel swapped with another scope? Is the scope serial number recorded anywhere other than on the plate at the bottom of the front panel?

Thanks,

Jeff Dutky


Re: Tektronix Technical Briefs (or whatever they're called in 2020)

toby@...
 

On 2020-11-19 4:41 p.m., Dennis Tillman W7pF wrote:
These are a small fraction of the total number of Application Notes that were available from Tek. I accumulated this list because they have something to do with the 7000 Series of Scopes. This list was assembled from many different sources because they were of interest to me.
Dennis Tillman W7pF

Is the content accessible anywhere?

--Toby


22W-11230-4 Telephone Access Network Measurements
25M1.0 TDR for Cable Testing
26W-7071 EMI Applications Using the Tektronix 2712 Spectrum Analyzer
35J1.0 Jitter-Free Oscilloscope Displays of Disc File Data
42AX-2686-1 Measuring Memory Core I/O Signals with Digital Accuracy
42AX-2774-1 Why Use Display-by-Events in Your Measurement Applications?
42AX-3085 Using Storage to Find Troublesome Logic Glitches
42AX-3198 Variable Persistence Storage Applications
42AX-3199 Bistable Storage Applications
42AX-3200 Fast and Multimode Storage Applications
42AX-3379-1 Pulse and Digital Timing Measurements - A Better Technique
42AX-3957 X-Y Displays with Interval Timing for Measuring SOA
42AX-4194 World's Fastest Oscilloscope Breaks the 1GHz Barrier (article)
42AX-4682 Introduction to 7854 Oscilloscope Measurements and Programming Techniques
42W-2659-3 Accurate Radar Pulse Measurements Using Digital Delay
42W-2680-3 Measuring Time Interval Between Non-Adjacent Digital Word Pulses
42W-2687-2 Measuring Disc Drive Access Time and Access Voltages
42W-3632-2 DAC Measurements: The Sampling Oscilloscope Approach
42W-3681-1 Pulse-Echo Measurements with Digital Accuracy
42W-4281-1 Measurement Variety. An Engineering Challenge Featuring the 7854
42W-4416-1 Using the 7854 in a GPIB Configuration
42W-4935 Extending Waveform Measurement Capability with Special Purpose Plug-ins
42W-5017-1 Increased Measurement Accuracy Using a 7D15 Universal Counter/Timer in any 7000 Series Oscilloscope
42W-5079-1 7D20 Programmable Digitizer: Digitizing Performance and Versatility in a Powerful Plug-in
42W-5085 The 7D20 Programmable Digitizer: Performing a Wide Range of Measurement Tasks Easier, Faster, and More Accurately
42W-5195 Sampling for High Speed Measurements
42W-5311 Human Pattern Recognition Speeds Automated Testing (article)
42W-5314 Microchannel-plate CRT Added to Oscilloscope Speeds Fault Finding (article)
42W-5315 Storage Scopes: A Variety of Techniques and Capabilities (article)
42W-5325 Measurement Techniques with Differential Amplifiers
42W-5334-1 Automated TDR Testing Made Easy with the 7854 Oscilloscope / 7S12 Sampler Plug-in
42W-5335-1 Applying Photographic Writing Rate to High Speed Signal Measurements
42W-5588 Advanced Triggering Techniques
42W-5629 Viewing Low Amplitude Pulses
42W-5630 Displaying Bus Contention
42W-5645 Using Analog Sampling and Digital Storage to Improve ECL Testing (article)
42W-5646-1 Use a Personal Computer and DFT to Extract Data from Noisy Signals (article)
42W-5700 Power Supply / Device Testing
42W-5802 Basic Software Programs for Communicating between the 7854 and IBM PC
42W-5903 Logic-Triggered Amplifier Upgrades Oscilloscopes for Digital Troubleshooting (article)
42W-5918 The Evolution of Oscilloscopes: Faster Than Any Other Type of Test Equipment (article)
42W-5926 7854 Programming Primer
42W-5947 Storage Scopes Solve Tough Signal Capture Problems (article)
42W-5968 7854 Measurement Primer
42W-5969 Sampling Primer
44L1.0 The Boxcar Integrator
45W-5280 7D20 Instrument Interfacing Guide
52AX-3221 7844-400MHz/5444-60MHz - Low Repetition-Rate pulse Pairs and the Dual Beam Oscilloscope
61AX-4803 Performance Analysis
61AX-4804 The Family Emulator
61AX-4806 Interrupt Analysis
61AX-4807 Memory Allocation
99AX-4607 Tektronix Codes and Formats for GPIB Instruments
A-2495 Measuring Distortions in the Television Signal
A-2496, TV Products App Note #4 Remote TV Transmitter Monitoring with tektronix Television Equipment
A-2509, TV Products App Note #7 Picture Monitor Color Temperature Adjustment Using the Tektronix J16
A-2588 7000 Series Oscilloscope Systems (catalog), October 1972
A-2618 3 New Probes and LED Adapter
A-2661, TV Products App Note #5 Measuring Chrominance-To-Luminance Gain and Delay
A-2663-1, TV Products App Note #6 Monitoring and Interpreting the Vertical Interval Reference Signal
A-2719, TV Products App Note #8 Using the Vertical Interval Reference Signal
A-2772 TM 500 Series Application Notes #2. FG 501 Swept Frequency Applications
A-2786, TV Products App Note #8A TV Transmitter Precorrection with the Tektronix 1440
A2912 Practical Lighting Measurements with the Tektronix J16
A-2926 Picture Monitor Color Temperature Adjustment Using the Tektronix J16
A-2935 Rapid Scanning Spectrometer Sales Guide (Company Confidential)
A-3107 The Tektronix Cookbook of Standard Audio Tests
A-3107 The Tektronix CookBook of Standard Audio Tests
A-3183 TM 500 Modular Test and Measurement Instruments (catalog)
A-3186 Suggested TM 500 Power Supply Circuits
A-3263 16 Channel Logic Analyzer
A-3269 Easier, Faster, More Accurate Oscilloscope Timing Measurements
A-3341 Tektronix Logic Analyzers (for the Digital Domain) (Color Glossy Brochure)
A-3357-3 7D01 Data Sheet
AN-3266 AM Broadcast Measurements Using the Spectrum Analyzer
AX-2930, TV Products App Note #10 Chrominance to Luminance Gain Correction and Delay Measurements
AX-2932, 47N1.0 R7912 Transient Digitizer… A Solution to Pulse Laser Measurement Problems
AX-2933-1 Sideband Analysis for TV
AX-2936, 45A1.0 Mechanical Measurements Using the DPO
AX-2937, 45A1.1 Engine Performance Measurements
AX-2983 Pulsed Laser Measurements Using the R7912 Transient Digitizer
AX-3074, 30U1.1 Measurement of Color Coordinates with the Tek 7J20 Rapid-Scan Spectrometer System
AX-3145, 52G1.2 Troubleshooting a Logic Circuit
AX-3170, 60M1.0 X-Ray Tube Current Measurements
AX-3187, 45F1.0 DPO Program Library Techniques
AX-3217, 75M1.0 Generating Complex Waveforms with TM 500 Instruments
AX-3218, 75M2.0 Integration Through V to F Conversion
AX-3259-1 Noise Measurements Using the Spectrum Analyzer, Part 2: Impulse Noise
AX-3260 Noise Measurements Using the Spectrum Analyzer, Part 1: Random Noise
AX-3266 AM Broadcast Measurements using the Spectrum Analyzer
AX-3281 The Tracking Generator/Spectrum Analyzer System
AX-3323 Television Operational Measurements; Video and RF for NTSC Systems
AX-3336, 47L.0 Windowing to Control FFT Leakage
AX-3349, 41G1.0 Using Delayed Sweep in Measuring Digital Word Trains
AX-3388, TV Products App Note #23 Multiburst Testing with the 1470
AX-3406 EMI Applications Using the Spectrum Analyzer
AX-3428, TV Products App Note #24 A Simple Color Background Generator
AX-3433 Baseband Measurements using the Spectrum Analyzer
AX-3492 Analyzing A/D Activity Through Mapping
AX-3524, 57K1.0 Troubleshooting a Microprocessor
AX-3535 Crystal Device Measurements Using the Spectrum Analyzer
AX-3582-1 FM Broadcast Measurements Using the Spectrum Analyzer
AX-3632, 42K1.0 DAC Measurements: The Sampling Oscilloscope Approach
AX-3682 The Spectrum Analyzer as a Frequency Selective Level Meter
AX-3810 Automating Swept RF Measurements
AX-3814 7D01 Logic Analyzer Laboratory Workbook
AX-3816 Tektronix Logic Analyzers Features Description and Glossary
AX-3903 Keeping Pace with Changing Needs in Optical Fiber Evaluation
AX-3999 Isolating Problems on GPIB
AX-4000 Detecting Data glitches Through Latching
AX-4006-1 An Introduction to Time and Frequency Domain Modulation and Waveform Analysis
AX-4012 Troubleshooting in the Synchronous & Asynchronous Modes
AX-4070 Simultaneous Display of Digital and Analog Test Data
AX-4098 Displaying I/O Address Activity Through Clock Qualification
AX-4117 Probing a CMOS Microcomputer Emulator
AX-4401 Capture Fast Waveforms Accurately with a 2-Channel Programmable Digitizer
AX-4416 GPIB Communication with the 7854
C. W. Rhodes, 1972 The 12.5T Modulated Sine-Squared pulse for NTSC (IEEE Trans. On Broadcasting Vol BC-18, #1, Mar. 1972)
C. W. Rhodes, 1972 Measuring Distortions in the television Signal
C. W. Rhodes, 1972 Test Signals and their Monitoring for Remote Control of Television Transmitters
Digital Family # 8 Logic Analysis Using Delay by Events Count
Internal Tek Application Note Application Note to Integrate the Area Under a Curve
TV Products App Note #1 Using the Tektronix 140, 144, or 146 as a VIT Signal Source
TV Products App Note #2 Measuring Luminance Cross Modulation with the Tektronix Types 140, 144, and 146
TV Products App Note #3 In-Service Random Noise Measurements Using the Tektronix 147
HANDSHAKE Application Library Catalog
The Logic Analyzer (Mini-Micro Systems, Sept. 1977)
What to Look for in Logic Timing Analyzers (magazine reprint)
Crack tough System problems With a Dual-Timebase Analyzer (EDN, Apr 28, 1983)
State Analyzers Move from Lab to Production Area (Electronic Design, May 13, 1982)
Clock Trigger Versatility Bolsters Logic Analysis (Electronic Design, Sept. 2, 1982)
Rapid Color Measurement (R/D magazine Nov. 1974
Rapid Scanning Spectrometry (American Laboratory, Nov. 1972)
Quality Control of the Color TV Signal (BM/E, Dec. 1969)
Get to Know Op Amps; Use a Curve Tracer (IEEE Spectrum, Vol 11, #9, Sept. 1978)
The Evolution of Scopes: Faster Than Any Other Type Of Test Equipment
Using the AA 501 and the SG 505 in Common Audio-Frequency Measurements
X-3585 Spectrum Analyzers / Swept Frequency Systems
Pulsed RF Spectrum Analysis
2EW-8190-0 The spectral connection.... Television RF measurements with the spectrum analyzer
063-0566-00 Television Measurements NTSC Systems
2EW-8380-0 Spectrum Analyzer Fundamentals

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Roy Thistle
Sent: Thursday, November 19, 2020 9:47 AM
To: TekScopes@groups.io
Subject: [TekScopes] Tektronix Technical Briefs (or whatever they're called in 2020)

For several decades, Tektronix has released these "Technical Briefs:" a.k.a. Technical Documents, Application Notes, and possibly other names.
Is there a single point(s) of curation for them?







LAST CALL for Peter Keller's Book and Current Orders I have

 

This is a last call for Peter Keller's fabulous book. I will accept orders
until Friday evening my time. At present I have orders for 75 books.
At that time I will publish the final list of people I have received an
order from and I will explain how to make payment.
Below are the people who have been placed on Peter's book order. Please
confirm your name is on this list.

If you have missed all of the excitement here is a summary
Pete Keller (from Tektronix) is prolific writer about cathode ray tubes and
many other kinds of display technologies.
He published a very detailed book on the subject in 1991 which I highly
recommend. The book title is:
"The Cathode Ray Tube, Technology, History, and Applications", Peter A.
Keller. 320 p.
ISBN 0-9631559-0-3;
TK7871.73.K46;
621.3815'42-DC20

I paid $70 for my copy years ago when I bought it from Peter. There is a
copy of his book currently on eBay listed for $650!!!
Peter's book is still available in hard cover and he has copies of it to
sell to us.
Because of recent interest in it from TekScopes members I went out on a limb
and asked Peter if a discount would be available for our members if we
bought enough books to make it worth his while and if I did all the work for
him.
He agreed and offered our members a huge discount. The cost will be $30 +
shipping.
I am hoping I can even get him to autograph each copy.
I promised Peter I would keep his effort to an absolute minimum by
collecting the names and addresses, payments which I would send to him in
one lump sum; then when he ships all the books to me I will mail each book
for him.

The cost for a book going anywhere in the US is $30 + $4 Media Rate Postage
+ $3 for supplies and my time. The total will be $37.00.
The cost for orders overseas will be $30 + approximately $35 for Flat Rate
International Priority Mail (this includes tracking).

If you want a copy of Peter's book send your mailing address to me OFF-LIST
at dennis at ridesoft dot com. INCLUDE YOUR ADDRESS.

These are the orders I have as of 2:15PM (UTC-08:00) Coordinated Universal
Time-08.
If your name is not on this list then I missed your order (my apologies) so
resend it with your address.

UNITED STATES ORDERS
Eric Spendel Fairfield, OH 45014
Dave Daniel Mims, FL 32754
Chuck Azzalina Perkasie, PA 18944
Jean Paul Novato, CA 94947
Kurt Rosenfeld Ossining, NY 10562
Vince Vielhaber K8ZW Oxford, MI 48371
Stan Perkins N6BYU San Diego, CA 92109-2348
Bruce Lane Kent, WA 98030-8803
Mark Huffstutter Seattle, WA 98115
Joe Rigdon Oviedo, FL 32765
Jeffrey S. Dutky Silver Spring, MD 20901
Larry Snyder Springboro, OH 45066-9761
Byron Hayes Jr. WA6ATN Toluca Lake, CA 91602-2914
Michael Drum Fanwood, NJ 07023-1008
Jack Reynolds Howell, MI, 48843
Phil Erickson Clinton, MA 01510
Edward Oscarson New Hartford, CT 06057
Steve Berg Casselberry, FL 32707
Charles Daves Aurora, CO 80015-1422
Chuck Harris Damascus, MD 20872
John Malec IV Northglenn, CO 80234
Marvin Moss Marietta, GA 30064
Tom Norman Escalon, CA 95320
Bill Lavick WA2SMF Edwards, NY 13635
Jeff Frantzen Olathe, KS 66062-3693
Glenn Little Goose Creek, SC 29445
Mark Vincent N. Chesterfield, VA 23236
Lance Lieberman Albertson, NY 11507-1022
Steven Horii Bryn Mawr, PA 19010-1226
Charles Nalley Burien, WA 98166
Carl Miles Rio Rancho, NM 87124
Larry Schneider Otis, ME 04605-7652
Dennis McCreery Bellevue, WA 98008
Tim Laing Lima, OH 45801-4644
Chris Wilkson Detroit, MI 48206
Timothy Koeth Brandywine, MD 20613
John Griessen Albuquerque, NM 87107
Greg Muir Great Falls, MT 59405-3144
Kurt Swanson Wilton, CA 95693-9765
Bruce Gentry KA2IVY Mattydale, NY 13211
Peter Brown Broomfield, CO 80023
Tim Pierce Cottage Grove, OR 97424
Bob Darlington N3XKB Los Alamos, NM 87544
Hugh Vartanian Littleton, MA 01460
Richard Brittingham W4MCD Edenton, NC 27932
Jim Rawlings AF6VF Livermore, CA 94550
Monte Meredith Reno, NV 89511
Chris Loggans Haymarket, VA 20169
Steve Bates Concord, MA 01742
John Glass, O'Brien Electric Dallas, OR 97338

INTERNATIONAL ORDERS
Sigurdur Asgeirsson CANADA
Shaun Merrigan CANADA
Bill Perkins, PEARL, Inc. CANADA
Dan Gajanovic CANADA
Andy Guelzow VA7NNM CANADA
Simon Jarman FRANCE
Christoph F. Bruggaier GERMANY
Heinz Breuer GERMANY
Jan Wuesten GERMANY
Red Dot Finder THE NETHERLANDS
Leo Potjewijd THE NETHERLANDS
Mario Giganti I1CWZ ITALY
Colin Herbert UNITED KINGDOM
Alan Ainslie UNITED KINGDOM
Robert Angell UNITED KINGDOM
Thomas S. Knutsen NORWAY
Jared Cabot JAPAN
Gangyi, Le SINGAPORE


Re: Tektronix Technical Briefs (or whatever they're called in 2020)

 

These are a small fraction of the total number of Application Notes that were available from Tek. I accumulated this list because they have something to do with the 7000 Series of Scopes. This list was assembled from many different sources because they were of interest to me.
Dennis Tillman W7pF

22W-11230-4 Telephone Access Network Measurements
25M1.0 TDR for Cable Testing
26W-7071 EMI Applications Using the Tektronix 2712 Spectrum Analyzer
35J1.0 Jitter-Free Oscilloscope Displays of Disc File Data
42AX-2686-1 Measuring Memory Core I/O Signals with Digital Accuracy
42AX-2774-1 Why Use Display-by-Events in Your Measurement Applications?
42AX-3085 Using Storage to Find Troublesome Logic Glitches
42AX-3198 Variable Persistence Storage Applications
42AX-3199 Bistable Storage Applications
42AX-3200 Fast and Multimode Storage Applications
42AX-3379-1 Pulse and Digital Timing Measurements - A Better Technique
42AX-3957 X-Y Displays with Interval Timing for Measuring SOA
42AX-4194 World's Fastest Oscilloscope Breaks the 1GHz Barrier (article)
42AX-4682 Introduction to 7854 Oscilloscope Measurements and Programming Techniques
42W-2659-3 Accurate Radar Pulse Measurements Using Digital Delay
42W-2680-3 Measuring Time Interval Between Non-Adjacent Digital Word Pulses
42W-2687-2 Measuring Disc Drive Access Time and Access Voltages
42W-3632-2 DAC Measurements: The Sampling Oscilloscope Approach
42W-3681-1 Pulse-Echo Measurements with Digital Accuracy
42W-4281-1 Measurement Variety. An Engineering Challenge Featuring the 7854
42W-4416-1 Using the 7854 in a GPIB Configuration
42W-4935 Extending Waveform Measurement Capability with Special Purpose Plug-ins
42W-5017-1 Increased Measurement Accuracy Using a 7D15 Universal Counter/Timer in any 7000 Series Oscilloscope
42W-5079-1 7D20 Programmable Digitizer: Digitizing Performance and Versatility in a Powerful Plug-in
42W-5085 The 7D20 Programmable Digitizer: Performing a Wide Range of Measurement Tasks Easier, Faster, and More Accurately
42W-5195 Sampling for High Speed Measurements
42W-5311 Human Pattern Recognition Speeds Automated Testing (article)
42W-5314 Microchannel-plate CRT Added to Oscilloscope Speeds Fault Finding (article)
42W-5315 Storage Scopes: A Variety of Techniques and Capabilities (article)
42W-5325 Measurement Techniques with Differential Amplifiers
42W-5334-1 Automated TDR Testing Made Easy with the 7854 Oscilloscope / 7S12 Sampler Plug-in
42W-5335-1 Applying Photographic Writing Rate to High Speed Signal Measurements
42W-5588 Advanced Triggering Techniques
42W-5629 Viewing Low Amplitude Pulses
42W-5630 Displaying Bus Contention
42W-5645 Using Analog Sampling and Digital Storage to Improve ECL Testing (article)
42W-5646-1 Use a Personal Computer and DFT to Extract Data from Noisy Signals (article)
42W-5700 Power Supply / Device Testing
42W-5802 Basic Software Programs for Communicating between the 7854 and IBM PC
42W-5903 Logic-Triggered Amplifier Upgrades Oscilloscopes for Digital Troubleshooting (article)
42W-5918 The Evolution of Oscilloscopes: Faster Than Any Other Type of Test Equipment (article)
42W-5926 7854 Programming Primer
42W-5947 Storage Scopes Solve Tough Signal Capture Problems (article)
42W-5968 7854 Measurement Primer
42W-5969 Sampling Primer
44L1.0 The Boxcar Integrator
45W-5280 7D20 Instrument Interfacing Guide
52AX-3221 7844-400MHz/5444-60MHz - Low Repetition-Rate pulse Pairs and the Dual Beam Oscilloscope
61AX-4803 Performance Analysis
61AX-4804 The Family Emulator
61AX-4806 Interrupt Analysis
61AX-4807 Memory Allocation
99AX-4607 Tektronix Codes and Formats for GPIB Instruments
A-2495 Measuring Distortions in the Television Signal
A-2496, TV Products App Note #4 Remote TV Transmitter Monitoring with tektronix Television Equipment
A-2509, TV Products App Note #7 Picture Monitor Color Temperature Adjustment Using the Tektronix J16
A-2588 7000 Series Oscilloscope Systems (catalog), October 1972
A-2618 3 New Probes and LED Adapter
A-2661, TV Products App Note #5 Measuring Chrominance-To-Luminance Gain and Delay
A-2663-1, TV Products App Note #6 Monitoring and Interpreting the Vertical Interval Reference Signal
A-2719, TV Products App Note #8 Using the Vertical Interval Reference Signal
A-2772 TM 500 Series Application Notes #2. FG 501 Swept Frequency Applications
A-2786, TV Products App Note #8A TV Transmitter Precorrection with the Tektronix 1440
A2912 Practical Lighting Measurements with the Tektronix J16
A-2926 Picture Monitor Color Temperature Adjustment Using the Tektronix J16
A-2935 Rapid Scanning Spectrometer Sales Guide (Company Confidential)
A-3107 The Tektronix Cookbook of Standard Audio Tests
A-3107 The Tektronix CookBook of Standard Audio Tests
A-3183 TM 500 Modular Test and Measurement Instruments (catalog)
A-3186 Suggested TM 500 Power Supply Circuits
A-3263 16 Channel Logic Analyzer
A-3269 Easier, Faster, More Accurate Oscilloscope Timing Measurements
A-3341 Tektronix Logic Analyzers (for the Digital Domain) (Color Glossy Brochure)
A-3357-3 7D01 Data Sheet
AN-3266 AM Broadcast Measurements Using the Spectrum Analyzer
AX-2930, TV Products App Note #10 Chrominance to Luminance Gain Correction and Delay Measurements
AX-2932, 47N1.0 R7912 Transient Digitizer… A Solution to Pulse Laser Measurement Problems
AX-2933-1 Sideband Analysis for TV
AX-2936, 45A1.0 Mechanical Measurements Using the DPO
AX-2937, 45A1.1 Engine Performance Measurements
AX-2983 Pulsed Laser Measurements Using the R7912 Transient Digitizer
AX-3074, 30U1.1 Measurement of Color Coordinates with the Tek 7J20 Rapid-Scan Spectrometer System
AX-3145, 52G1.2 Troubleshooting a Logic Circuit
AX-3170, 60M1.0 X-Ray Tube Current Measurements
AX-3187, 45F1.0 DPO Program Library Techniques
AX-3217, 75M1.0 Generating Complex Waveforms with TM 500 Instruments
AX-3218, 75M2.0 Integration Through V to F Conversion
AX-3259-1 Noise Measurements Using the Spectrum Analyzer, Part 2: Impulse Noise
AX-3260 Noise Measurements Using the Spectrum Analyzer, Part 1: Random Noise
AX-3266 AM Broadcast Measurements using the Spectrum Analyzer
AX-3281 The Tracking Generator/Spectrum Analyzer System
AX-3323 Television Operational Measurements; Video and RF for NTSC Systems
AX-3336, 47L.0 Windowing to Control FFT Leakage
AX-3349, 41G1.0 Using Delayed Sweep in Measuring Digital Word Trains
AX-3388, TV Products App Note #23 Multiburst Testing with the 1470
AX-3406 EMI Applications Using the Spectrum Analyzer
AX-3428, TV Products App Note #24 A Simple Color Background Generator
AX-3433 Baseband Measurements using the Spectrum Analyzer
AX-3492 Analyzing A/D Activity Through Mapping
AX-3524, 57K1.0 Troubleshooting a Microprocessor
AX-3535 Crystal Device Measurements Using the Spectrum Analyzer
AX-3582-1 FM Broadcast Measurements Using the Spectrum Analyzer
AX-3632, 42K1.0 DAC Measurements: The Sampling Oscilloscope Approach
AX-3682 The Spectrum Analyzer as a Frequency Selective Level Meter
AX-3810 Automating Swept RF Measurements
AX-3814 7D01 Logic Analyzer Laboratory Workbook
AX-3816 Tektronix Logic Analyzers Features Description and Glossary
AX-3903 Keeping Pace with Changing Needs in Optical Fiber Evaluation
AX-3999 Isolating Problems on GPIB
AX-4000 Detecting Data glitches Through Latching
AX-4006-1 An Introduction to Time and Frequency Domain Modulation and Waveform Analysis
AX-4012 Troubleshooting in the Synchronous & Asynchronous Modes
AX-4070 Simultaneous Display of Digital and Analog Test Data
AX-4098 Displaying I/O Address Activity Through Clock Qualification
AX-4117 Probing a CMOS Microcomputer Emulator
AX-4401 Capture Fast Waveforms Accurately with a 2-Channel Programmable Digitizer
AX-4416 GPIB Communication with the 7854
C. W. Rhodes, 1972 The 12.5T Modulated Sine-Squared pulse for NTSC (IEEE Trans. On Broadcasting Vol BC-18, #1, Mar. 1972)
C. W. Rhodes, 1972 Measuring Distortions in the television Signal
C. W. Rhodes, 1972 Test Signals and their Monitoring for Remote Control of Television Transmitters
Digital Family # 8 Logic Analysis Using Delay by Events Count
Internal Tek Application Note Application Note to Integrate the Area Under a Curve
TV Products App Note #1 Using the Tektronix 140, 144, or 146 as a VIT Signal Source
TV Products App Note #2 Measuring Luminance Cross Modulation with the Tektronix Types 140, 144, and 146
TV Products App Note #3 In-Service Random Noise Measurements Using the Tektronix 147
HANDSHAKE Application Library Catalog
The Logic Analyzer (Mini-Micro Systems, Sept. 1977)
What to Look for in Logic Timing Analyzers (magazine reprint)
Crack tough System problems With a Dual-Timebase Analyzer (EDN, Apr 28, 1983)
State Analyzers Move from Lab to Production Area (Electronic Design, May 13, 1982)
Clock Trigger Versatility Bolsters Logic Analysis (Electronic Design, Sept. 2, 1982)
Rapid Color Measurement (R/D magazine Nov. 1974
Rapid Scanning Spectrometry (American Laboratory, Nov. 1972)
Quality Control of the Color TV Signal (BM/E, Dec. 1969)
Get to Know Op Amps; Use a Curve Tracer (IEEE Spectrum, Vol 11, #9, Sept. 1978)
The Evolution of Scopes: Faster Than Any Other Type Of Test Equipment
Using the AA 501 and the SG 505 in Common Audio-Frequency Measurements
X-3585 Spectrum Analyzers / Swept Frequency Systems
Pulsed RF Spectrum Analysis
2EW-8190-0 The spectral connection.... Television RF measurements with the spectrum analyzer
063-0566-00 Television Measurements NTSC Systems
2EW-8380-0 Spectrum Analyzer Fundamentals

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Roy Thistle
Sent: Thursday, November 19, 2020 9:47 AM
To: TekScopes@groups.io
Subject: [TekScopes] Tektronix Technical Briefs (or whatever they're called in 2020)

For several decades, Tektronix has released these "Technical Briefs:" a.k.a. Technical Documents, Application Notes, and possibly other names.
Is there a single point(s) of curation for them?







--
Dennis Tillman W7pF
TekScopes Moderator


Re: Tektronix Technical Briefs (or whatever they're called in 2020)

 

Hi Roy,
In the 1970s and 1980s they were known generically as Application Notes. Tek changed their numbering system several times for various reasons.
* In some cases they were published by the group that created the instruments (Spectrum Analyzer articles came from that group for instance).
* In some cases they were created by the marketing person for a particular product line (like the TM500 instruments).
* In some cases they were created by a separate division of Tek (like the TV Instruments, or Laboratory Scopes divisions).
* In some cases they were articles written for trade publications which Tek had permission to reprint.
* Some excellent articles were written for Tek Periodicals such as Service Scope, Handshake, etc.
As if that wasn't confusing enough the same Application Notes were printed with different numbering systems when Tek tried to place all of these things under one group within the company.

So to answer your question: There is no rhyme or reason for how Tek did it.

Dennis Tillman W7pF

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Roy Thistle
Sent: Thursday, November 19, 2020 9:47 AM
To: TekScopes@groups.io
Subject: [TekScopes] Tektronix Technical Briefs (or whatever they're called in 2020)

For several decades, Tektronix has released these "Technical Briefs:" a.k.a. Technical Documents, Application Notes, and possibly other names.
Is there a single point(s) of curation for them?







--
Dennis Tillman W7pF
TekScopes Moderator


Tektronix Technical Briefs (or whatever they're called in 2020)

Roy Thistle
 

For several decades, Tektronix has released these "Technical Briefs:" a.k.a. Technical Documents, Application Notes, and possibly other names.
Is there a single point(s) of curation for them?


Re: “High Bandwidth Transient Capture,” Tektronix Publication 1991

Roy Thistle
 

High Bandwidth Transient Capture, Technical Brief No: 45W-8088, Tektronics circa (1991?)


Re: Broken Tektronix scopes free

Michael W. Lynch
 

I would like to hear more about the 465's

--
Michael Lynch
Dardanelle, AR


Re: Broken Tektronix scopes free

john dear
 

Hi Peter,
Where are you located?
Best wishes,
Emil

On 19-Nov-20 16:13, Peter Margolis via groups.io wrote:
I have two broken Tektronix 465s and one 2336 that I want to give away for the cost of shipping. Anyone interest in one or both?

Pete




Re: Broken Tektronix scopes free

 

Interested in the 2336
Chris n9llo@aol.com

On Nov 19, 2020, at 9:13 AM, Peter Margolis via groups.io <pmargolis=mac.com@groups.io> wrote:

I have two broken Tektronix 465s and one 2336 that I want to give away for the cost of shipping. Anyone interest in one or both?

Pete





Broken Tektronix scopes free

Peter Margolis
 

I have two broken Tektronix 465s and one 2336 that I want to give away for the cost of shipping. Anyone interest in one or both?

Pete


Re: Peter Keller Book Order INTERIM UPDATE

 

On Wed, Nov 18, 2020 at 02:30 PM, Raymond Domp Frank wrote:


Dennis,
Uithoorn and Amsterdam may be combined in one shipment to Uithoorn. The places
are close together and I (in Uithoorn) will take care of it, once I know the
name and address of the person in Amsterdam. BTW, my lab is in Uithoorn but I
live in Amsterdam...
Hi Dennis,
The person living in Amsterdam has contacted me, I've met him before. We have agreed that you may send both his (Amsterdam) and my (Uithoorn) book to me.
Payment arrangements will be made separately.

Raymond


TDS 784A / 754A WARNING-ERROR MESSAGE

Gif Sim
 

good morning to the whole group

I have a tds754a upgraded to tds784a by configuring the resistors and removing the capacitors in the acquisition card

the acquisition card is 671-3953-00
the microp board is 679-3488-00
the serial is b010976
the memory board is 1M

all electrolytic capacitors have been replaced and the oscilloscope works perfectly
the ds1486 was replaced with a working battery
It was originally mounted a DS1650y that has been replaced with a DS1250y-100 with a working battery

when I connect the tek FAS software with gpib and start the first calibration phases
in the error log I get some truncated warning messages:

"WARNING: 600 remote nv memory a ...."

the calibration then proceeds smoothly (but I haven't finished it yet)
what does this warning message mean and what can be the cause?
thanks to everyone who can be of help

A greeting

Simon


Re: Tektronix 7D13 and 7A42 wanted

 

Hi Ed and Dan,
I'll toss the battery and remove the jumper in mine also.
Dennis

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Ed Breya via groups.io
Sent: Wednesday, November 18, 2020 7:14 PM
To: TekScopes@groups.io
Subject: Re: [TekScopes] Tektronix 7D13 and 7A42 wanted

Hi Dennis,

I didn't realize earlier your question "What did you use for a battery?" was to me. There is no battery. For something like this, it's simple enough to set up as needed, for occasional use. Having it remember is a nice convenience, but not worth the trouble and leakage risk. All I did was check it out after enough burn-in, and I removed the BE jumper as Dan suggested. It's running right now for one last time before mothballing.

Ed







--
Dennis Tillman W7pF
TekScopes Moderator


Re: Tektronix 7D13 and 7A42 wanted

Ed Breya
 

Hi Dennis,

I didn't realize earlier your question "What did you use for a battery?" was to me. There is no battery. For something like this, it's simple enough to set up as needed, for occasional use. Having it remember is a nice convenience, but not worth the trouble and leakage risk. All I did was check it out after enough burn-in, and I removed the BE jumper as Dan suggested. It's running right now for one last time before mothballing.

Ed


Re: Q122/222 Jfet J300 subs for 2215A scope ?

Ed Breya
 

Oops - I meant to say Vds here, not Vgs:

The Vds should be fairly constant since the required output excursion is only about +/-100 mV.

Ed


Re: Q122/222 Jfet J300 subs for 2215A scope ?

Ed Breya
 

Yup, I see it. I finally looked at the schematic for the 2215A. Looks like the quiescent drain current should be around 11 mA, and the source at Vbe.I'm looking at the J310/J309 data sheet here:

https://www.mouser.com/datasheet/2/308/J309-D-109346.pdf

If I'm interpreting Fig. 1 properly, the lower curve set is for the J309, and the Id is about right for Vgs= -Vbe. The upper curves for the J310 are about 3 X higher - too far out there for the nominal circuit conditions, although Q15 could probably pull it hard enough to DC balance (but things would start getting weird). The JFET runs with Vds about 8V, so pretty close to the 10V used in the specs. The Vgs should be fairly constant since the required output excursion is only about +/-100 mV.

This circuit reminds me of the ones in all the other 2200 series, and may be essentially exactly the same. More info may be discerned by looking at those others to see what JFETs were used. The 2N5486 keeps sticking in my brain.

Ed