Re: Tek 465: CH1 No signal, CH2 signal reads a fraction of input, V Pos distorts signal
You need to get the service manual - it tells you how to remove the CRT.
toggle quoted messageShow quoted text
The problem is likely the silicone sealant where the EHT is connected has failed. You'll need to clean off the old and re-seal. Please don't use an acetic acid cure silicone - get a neutral cure type. David
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Greenpitu@gmail.com Sent: 20 August 2020 12:40 To: TekScopes@groups.io Subject: Re: [TekScopes] Tek 465: CH1 No signal, CH2 signal reads a fraction of input, V Pos distorts signal Hello Roger - thanks once again! Went into a dark room and spotted the spark straight away. It's coming from the top right corner of the CRT. I can see the spark lighting through a very tiny opening that is available in the chassis. Please see short video with the exact location: https://youtu.be/7CxqL82_3VI Now, I have no idea how to get into that spot... Any thoughts? I will get on with the other measurements and revert back. I think I can carefully and safely work my way through these checks. Best regards Luis
|
|
Re: Tek 465: CH1 No signal, CH2 signal reads a fraction of input, V Pos distorts signal
Greenpitu@...
Hello Roger - thanks once again!
Went into a dark room and spotted the spark straight away. It's coming from the top right corner of the CRT. I can see the spark lighting through a very tiny opening that is available in the chassis. Please see short video with the exact location: https://youtu.be/7CxqL82_3VI Now, I have no idea how to get into that spot... Any thoughts? I will get on with the other measurements and revert back. I think I can carefully and safely work my way through these checks. Best regards Luis
|
|
Re: Looking for front cover for 465 and/or 485
Tom Gardner
I've considered making custom covers from laser-cut birch ply.
toggle quoted messageShow quoted text
That should be trivial given the ability to generate finger joints, e.g. https://en.makercase.com/#/basicbox (no recommendation, simply the first I googled)
On 20/08/20 07:07, dabono@gmail.com wrote:
Here's an interesting one - custom-made from clear acrylic. IMO it's a bit expensive ($85), about what you'd pay for an original Tektronix front cover.
|
|
Re: Tek 465: CH1 No signal, CH2 signal reads a fraction of input, V Pos distorts signal
Roger Evans
A crackling sound from the front of the CRT is not good. If you take the scope into a dark room with the cover off you may see some flashes from the region where the high voltage lead attaches near the front of the CRT. In any case you need to clean that area with IPA, but wait for maybe 30 minutes after the scope is powered off to allow the EHT to discharge.
Can you do some measurements on TP322 and TP324? With the scope displaying Ch1, input selector to ground, and the trace on the vertical centre line, these two test points should measure close to zero volts. If not can you adjust the vertical position to get TP322 and TP324 to read the same value, and then where is the beam? Also the Ch1 vertical position control should be somewhere near the mid point of its travel. If these tests come out OK then the fault is most likely in the vertical amplifier not being biassed correctly. The manual I have doesn't give any voltage readings around U440 but you could measure the resistances of R446, R447 and R448 and the voltages on each end of R446 and R447. The bias on the output transistors is adjustable with R455 but don't move it yet as it affects the vertical calibration and frequency response. Measure the voltages on each end of R457 and R452 to make sure that the transistors are drawing some base current and the voltage across R454 to ensure that the variable resistor R455 has not gone open circuit. It is also worth removing U440, cleaning the pins and reinserting. If the mounting allows, I would apply some IPA around the pin/socket area and leave for an hour or so to penetrate before trying to remove the IC - they can be very stubborn. Check that there is not a nut on the other side of the PCB holding it in! At this point I am running well beyond my competence and if anyone else would like to add anything I would be very grateful. Regards, Roger
|
|
Re: 2246A Power supply failure turning on / off
This https://www.rc-cam.com/forum/index.php?/topic/4165-tektronix-2245a-oscilloscope-smps-repair-re-cap/ may be of some interest.
On a side note to everyone; is there a cannonical design or schematic for a dummy load suitable for 224x PSUs? Cheers, Brian.
|
|
Re: Looking for front cover for 465 and/or 485
dabono@...
Here's an interesting one - custom-made from clear acrylic. IMO it's a bit expensive ($85), about what you'd pay for an original Tektronix front cover.
https://www.ebay.com/itm/TEKTRONIX-465B-465-OSCILLOSCOPE-FRONT-COVER-CUSTOM-MADE-CLEAR-ACRYLIC/283951063134?hash=item421ccd605e:g:j5kAAOSwSiZfEbTg
|
|
Re: 2246A Power supply failure turning on / off
JorgeTop
By now I have removed almost all the electrolytic capacitors and have observed that they all have much higher capacitance than stated, for example 1000uF have 1400uf, 470uf go for 600-700uF. How troublesome can this be?
For the diodes I have removed one of the pins to perform the diode measurement with a multimeter in diode function and they seem normal. I have not found open or shorted diodes. Zener diodes, I have changed most, as well as some transistors. It remains for me to change the MC310 and the IC that I have not found locally. I'm going to review the 0.1Ohm resistor you mention. Thanks, I will update this week
|
|
Re: A possibly sick 7904A
On Wed, Aug 19, 2020 at 11:04 PM, Eric wrote:
Very few of us use veteran analog 'scopes because they are better suited to the job. That's only true in some situations, if compared with fast-refreshing digital 'scopes. Raymond
|
|
Re: FG501 repair
um-gs@...
Hi,
just want to continue with my lessons learned on the same subject. The device was operative, suplly voltages were adjustable to their nominal values, but no output. 1. No output a) open R251 and inject signal voltage: Voltage appears at output, so amplifier is working b) checking voltage at junction R32/R34 when turning the frequency dial: voltage varies c) checking voltage at output of U15 pin 6 : does not vary - replace U15 which was defect - oscillator is working again, output signal is generated 2. Sine wave output distorted and adjustment R150 has no effect a) replacing both R150 and R170 allows adjustment of distortion free sine wave output 3. Output signal not at symmetrical duty cycle a) adjustment of R45 (step7) has no effect b) output voltage at U40 pin6 does not vary when R45 is adjusted c) swiper of R45 constantly at zero d) opening up/replacing of C40 solved the problem (it was shorted) - output duty cycle can be adjusted Regards, Gordian
|
|
Re: A possibly sick 7904A
Eric
I will blaspheme here for a little if I will be indulged. Don't get me wrong I do love the 7904A but for the VERY slow and the very fast above 1Ghz I will take the screen of a digital scope. Seeing 1hZ sign waves on a CRO is not easy with out a camera. of course then again.. I might be a noob
toggle quoted messageShow quoted text
Eric
On 8/19/2020 5:00 PM, Raymond Domp Frank wrote:
On Wed, Aug 19, 2020 at 10:47 PM, Eric wrote:Due to the difficulty of being able to source a tube if I needed one.That's been my excuse for having two or, better even, three copies of all models for a few years now. But we're treading on very different ground here...
|
|
Re: A possibly sick 7904A
On Wed, Aug 19, 2020 at 10:47 PM, Eric wrote:
That's been my excuse for having two or, better even, three copies of all models for a few years now. But we're treading on very different ground here... Raymond
|
|
Re: Tek 465: CH1 No signal, CH2 signal reads a fraction of input, V Pos distorts signal
Greenpitu@...
Hello Roger,
I am back with some measurements. ** Vertical Switching ** Both measurements taken in ALT mode, no probe attached, and lowest TIME/DIV setting. TP364 (-2V and -0.102V) TP374 (-2V and -0.071V) ** Trig View ** With calibrator signal feeding Trig In, Trig Source set to Ext Trigger Level Centered: when the beam is at the centre of the screen, the beam width is just under 1x sub-division Trigger Level +++: when the beam is at the highest pos, that is +2.5 divisions above centre, the beam is just over 2 sub-divisions in width Trigger Level ---: when the beam is at the lowest pos, that is -1.0 divisions below centre, the beam is also just over 2 sub-divisions in width When at the center, there is a fuzzy glimpse of the square signal. However, when on +++ or --- positions then beam becomes entirely flat. ** Q322 and Q324 ** Using tester got the following values: Q322: NPN, Gain: 75, Vf: 707mV Q323: NPN, Gain: 71, Vf: 701mV I should say I got (I think) bad news. I started hearing concerning cracking sounds for the first time while I was running the sweep at its slowest speed. At first, I heard like 5-7 in rapid succession. I powered off the scope and waited a little while. The scope cover was removed at the time, so I didn't get too close to the circuitry just in case. These sounds weren't too friendly. I increased the sweep speed before I turned it back on. Unfortunately, the crackling sound still persist - it doesn't seem to be related to the speed at all. In fact, it seems to be coming from the front of the cathode ray tube, but I can't be certain. I made a short video where I captured 2 or 3 of the crackling sounds. My phone was capping out the percussive effect somewhat; the real sound is higher in pitch and volume. https://youtu.be/DgHwnqH9xJs Is this a concern?
|
|
Re: A possibly sick 7904A
On Wed, Aug 19, 2020 at 10:47 PM, Eric wrote:
Not having an MCP CRT hardly ever is (or rather: was) a problem. It's not like "You don't have a good 'analog 'scope unless it has an MCP CRT"... Raymond
|
|
Re: A possibly sick 7904A
On Wed, Aug 19, 2020 at 09:24 PM, Eric wrote:
I think that's generally true, though it shouldn't be exaggerated. Acceptable brightness levels are fine and the 'scope does have some protections. OTOH, some people like to see how bright the image of their 'scope will go. That's not what these 'scopes were designed for: A few years ago, I had a friend who proudly showed me the screen brightness of his newly acquired 7104, the only problem being that brightness couldn't be turned down. Obviously a DC restorer problem or the like. After repair there was very little permanent (local) loss of brightness and no burned spots were visible. Quite different from the HP 1727A analog storage 'scope that turned out to have burnt its target. The CRT in the 2467 appears to be far more resilient than the much older design in the 7104. It has better protection as well I guess. At low/normal brightness levels, life should be fine. I only use a 7104 when I especially feel like it. It's not the best choice as a regular 'scope. Its screen graticule size is a bit smaller than regular models' as well (8mm/div vs. 10mm/div). AFAIK, the 2467BHD, with its MCP CRT, was intended for regular HDTV-work. FWIW, the CRT of my 7104 is as new, as are the CRT's of my 2467B and 2467BHD. Raymond
|
|
Re: A possibly sick 7904A
Eric
The lack of an MCP is why I went with the 7904 as apposed to the 7104. Due to the difficulty of being able to source a tube if I needed one.
toggle quoted messageShow quoted text
On 8/19/2020 4:03 PM, Dave Daniel wrote:
AFAIK, the 7904A does *not* use and MCP CRT. The 7104 and 2467 do use MCP CRTs.
|
|
Re: Tek 465 Sweep Start Calibration
On Wed, Aug 19, 2020 at 09:59 PM, david wrote:
What is the DTM mechanical max CCW reading ? It should be 0.20 http://www.hakanh.com/dl/docs/hardtofind/dtmsettings.pdf /HÃ¥kan
|
|
Re: A possibly sick 7904A
On Wed, Aug 19, 2020 at 10:03 PM, Dave Daniel wrote:
That's correct or you wouldn't experience the writing rate being as "limited" as with all the other regular models. Raymond
|
|
Re: Last Call - Tek 222/224 Battery Charger Project
Just wanted to chime in and let people know that they're also available at a slightly lower cost on www.n0dy.com<http://www.n0dy.com>. The cost there is lower because Ebay doesn't take a cut as a commission.
Thanks for the kind words, Henry! Jeff ________________________________ From: TekScopes@groups.io <TekScopes@groups.io> on behalf of Henry Martinez <hrmartinez@gmail.com> Sent: Tuesday, August 18, 2020 6:58 PM To: TekScopes@groups.io <TekScopes@groups.io> Subject: Re: [TekScopes] Last Call - Tek 222/224 Battery Charger Project Sorry if I'm late to the party, Just wanted people to know the kits are still available on eBay and they work great. I just put one into my Tektronix 224 and it is performing exactly as advertised. Thanks Jeff for pulling this all together.
|
|
Re: A possibly sick 7904A
Dave Daniel
AFAIK, the 7904A does *not* use and MCP CRT. The 7104 and 2467 do use MCP CRTs.
toggle quoted messageShow quoted text
DaveD
On Aug 19, 2020, at 15:24, Eric <ericsp@gmail.com> wrote:
|
|
Tek 465 Sweep Start Calibration
david
Hi guys, I need your help again, I am trying to do horizontal calibration on my 465 but I cannot set step one of calibration, "Check/Adjust Sweep Start and A Sweep Calibration", With Sweep start adjustment R1115 fully counter clockwise, the Intensified sweep starts at 0.96 on Delay Time position dial, cannot get it to 1.0, Is out of range of R1115. Trimmer R1115 is in the middle of a voltage divider between R1113, R1112 and R 1117. Has anybody had this problem? In service manual, parts list says R1113 is "196 ohm (NOM VALUE) SEL". My 465 has a 145 ohm resistor for R1113, changing R1113 to 196 ohm would probably get enough range to fix Sweep Start. I don't know where else to look for the problem. Also have a newer version of service manual where parts list says R1113 is "196 ohm". Does not say nominal value select. The rest of the horizontal cal went OK.
David
|
|