Date   

Re: Stuck switches on 453

Stephen
 

If this is of any help, I had something similar on a 7B71, 7000 series time base plug-in.
The VAR didn’t want to move at all. It was completely stuck. I ended up de-soldering the pot and freeing it from the shaft extension a bit. It turns out it was pushed way too much into it and, as a result, wasn’t moving. That wasn’t easy, but now it’s completely free, and working as it should.

I don’t know how yours is mounted, but I hope this help...


Re: Trying to understand the 453 calibrator

Richard P
 

Ahhhh. That’s above my pay grade, I suppose. I just thought you used the little bar to compensate a basic probe.


Re: Trying to understand the 453 calibrator

Bob Albert
 

It's not delivering a square wave of voltage.  It's a current loop for calibrating a current probe.
Bob

On Monday, August 3, 2020, 10:12:11 AM PDT, Richard P via groups.io <wb5nen=yahoo.com@groups.io> wrote:

I’ve always been able to get a good 1 kc square wave from the front panel BNC calibrator output. Works ok for compensating a probe. But I never could get the 1 kc signal from the probe loop on the side panel. Today, I verified the signal is being delivered to one side of the loop through parallel resistors equaling 20 ohms, in agreement with the schematic. But since the other side of the loop is tied to ground, also in agreement with the schematic, it seems to me the square wave is going to ground, and hence, that is why I have never seen any square wave on the loop. Never, ever.  The loop is a ground. How is this supposed to work? Thanks.

Richard


Re: Why would you do this?

snapdiode
 

I have a CombBind machine and spines in any case, it's just the odd cropping of the pictures that made me suspicious.
Turns out the spine is pretty good and a very clever and handsome person bought the manual.
Oh and he smells good too.


Trying to understand the 453 calibrator

Richard P
 

I’ve always been able to get a good 1 kc square wave from the front panel BNC calibrator output. Works ok for compensating a probe. But I never could get the 1 kc signal from the probe loop on the side panel. Today, I verified the signal is being delivered to one side of the loop through parallel resistors equaling 20 ohms, in agreement with the schematic. But since the other side of the loop is tied to ground, also in agreement with the schematic, it seems to me the square wave is going to ground, and hence, that is why I have never seen any square wave on the loop. Never, ever. The loop is a ground. How is this supposed to work? Thanks.

Richard


Re: Stuck switches on 453

Richard P
 

I barely managed to free up (a little) one of the two stuck potentiometer/switching knobs on the variable time/division control. (I see in an earlier post or two that I mistakenly started started saying it was the volts/div controls that were stuck. Wrong. It is time/div that are stuck. Regardless, it was a similar problem. It was stuck. Apologies.) Anyway, the “Uncal” light is off and things are working acceptably. I thank everyone for help. But boy, these guys are stuck! I can live with this.

Richard


Re: TM500 DM511 Digital Multimeter Service Manual?

Jared Cabot
 

On Tue, Aug 4, 2020 at 12:33 AM, Ken Eckert wrote:


https://elektrotanya.com/tektronix_dm5110_511_4%2C5-digit_dvm_1989_sm.zip/download.html#dl
That's the one! Much thanks! :)


Re: TM500 DM511 Digital Multimeter Service Manual?

Ken Eckert
 


Re: TM500 DM511 Digital Multimeter Service Manual?

Stephen
 

On Mon, Aug 3, 2020 at 03:51 AM, Jared Cabot wrote:
Unfortunately that is just the user manual, there's no service info in there
like calibration procedures, troubleshooting, parts list or schematics etc.
Maybe someone here has an original one, and would be willing to scan it for you for a small fee.
Or, if it’s not urgent, you could always wait for one to pops up on eBay or another auction site. I don’t know how often they do, though.
Sorry I can’t be of anymore help.


Re: TM500 DM511 Digital Multimeter Service Manual?

Dave Daniel
 

https://www.ebay.com/itm/TEKTRONIX-Ops-Service-Manuals-for-DM511-DM5110-Programmable-Digitlal-Multimeter-/370469694062

There are pobably others out there.

I assume you’ve checked Artek manuals?

http://artekmanuals.com/

They are not free, but what you get is a very high-quality PDF.

DaveD

On Aug 3, 2020, at 10:51, Jared Cabot via groups.io <jaredcabot=protonmail.com@groups.io> wrote:

On Mon, Aug 3, 2020 at 11:38 PM, Stephen wrote:


http://w140.com/tek_dm5110_dm511_user.pdf

Unfortunately that is just the user manual, there's no service info in there like calibration procedures, troubleshooting, parts list or schematics etc.



Re: TM500 DM511 Digital Multimeter Service Manual?

Dave Daniel
 

That is true.

There are also the Davmar, KO4BB, Barrytech, kahrs.us and other sites (my apologies to the owners of those sites for my declining memory) which sometimes offer different information, including non-Tektronix instrument resources.

DaveD

On Aug 3, 2020, at 10:35, Colin Herbert via groups.io <colingherbert=blueyonder.co.uk@groups.io> wrote:

I was about to link to that! TekWiki is a resource everyone interested in Tektronix gear should be aware of. It won't protect you from Coronavirus or cure your ingrowing toenails, but it does almost everything else. Enjoy it.
Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Dave Daniel
Sent: 03 August 2020 15:26
To: TekScopes@groups.io
Subject: Re: [TekScopes] TM500 DM511 Digital Multimeter Service Manual?

http://w140.com/tekwiki/wiki/DM511

DaveD


On 8/3/2020 10:24 AM, Jared Cabot via groups.io wrote:
Hi,

I'm looking for a service manual for a TM500 module DM511 multimeter, does anyone know where I can download this from?

Thanks!



--
This email has been checked for viruses by Avast antivirus software.
https://www.avast.com/antivirus








Re: TM500 DM511 Digital Multimeter Service Manual?

Jared Cabot
 

On Mon, Aug 3, 2020 at 11:38 PM, Stephen wrote:


http://w140.com/tek_dm5110_dm511_user.pdf

Unfortunately that is just the user manual, there's no service info in there like calibration procedures, troubleshooting, parts list or schematics etc.


Re: TM500 DM511 Digital Multimeter Service Manual?

Stephen
 


Re: TM500 DM511 Digital Multimeter Service Manual?

Colin Herbert
 

I was about to link to that! TekWiki is a resource everyone interested in Tektronix gear should be aware of. It won't protect you from Coronavirus or cure your ingrowing toenails, but it does almost everything else. Enjoy it.
Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Dave Daniel
Sent: 03 August 2020 15:26
To: TekScopes@groups.io
Subject: Re: [TekScopes] TM500 DM511 Digital Multimeter Service Manual?

http://w140.com/tekwiki/wiki/DM511

DaveD


On 8/3/2020 10:24 AM, Jared Cabot via groups.io wrote:
Hi,

I'm looking for a service manual for a TM500 module DM511 multimeter, does anyone know where I can download this from?

Thanks!



--
This email has been checked for viruses by Avast antivirus software.
https://www.avast.com/antivirus


Re: TM500 DM511 Digital Multimeter Service Manual?

Dave Daniel
 

On 8/3/2020 10:24 AM, Jared Cabot via groups.io wrote:
Hi,

I'm looking for a service manual for a TM500 module DM511 multimeter, does anyone know where I can download this from?

Thanks!

--
This email has been checked for viruses by Avast antivirus software.
https://www.avast.com/antivirus


TM500 DM511 Digital Multimeter Service Manual?

Jared Cabot
 

Hi,

I'm looking for a service manual for a TM500 module DM511 multimeter, does anyone know where I can download this from?

Thanks!


Re: OT - Any interest in starting a Tektronix Logic Analyzer Group? (with added bonus question!)

jslcanuck
 

What, no snappy retort to put me in my place?

You talk about redundancy like it's a bad thing. Lemme tell you about redundancy:
It gives you error detection and it gives you fault tolerance. And a search function is
only as good as the nature and consistency of the data set you're searching within,
which in the case of 7000 not-particularly-organized contributors is always going to
be less than stellar.

So taxonomy and hierarchy are your best friends if you want to have a hope of finding
what you're looking for without extensive, repetitive, and increasingly-speculative
searching in a big flat swamp of files and directories. And if you think that's somehow
superior to having a directory called "465", then you're in possession of some new
magic which has not yet been brought to my attention.

Now, I don't claim to know everything about the capabilities of the groups.io system,
but what I do know is that they provide a blockheaded and ham-handed set of tools that
constantly presents new surprises in what it *can't* do, for example, the inability to
create a file hierarchy in the photos section or to cleanly embed a link in the files section.

I could go on, but I think I've made my point.

Jonathan


Re: Tek 4041 GPIB Controller

Monty McGraw
 

I just added eight service notes in a new folder for the Tektronix 4041 on my github repository: https://github.com/mmcgraw74/Tektronix-4041-GPIB-Controller/tree/master/4041_Service_Notes

One of the notes discusses the difference between v2.0 firmware and v2.1 firmware.

I extracted these notes from ServiceTekNotes I found at http://w140.com/tekwiki/wiki/Tektronix_Magazines


Re: Tektronix 2230

Saroj Pradhan
 

Hi satbeginner
Last photo is taken after confirmation of 1×probe setting
Thanks
Saroj


Re: Parting Out 7912AD

Dick Wilson at DiLette Sales <cmdrdick42@...>
 

I'd consider taking the whole set if it wasn't a huge expense.Where are you and what would you take for it?

On Sunday, August 2, 2020, 10:44:44 PM EDT, Dave Casey <polara413@...> wrote:

I have a complete 7912AD with a trashed scan converter tube (damaged in
shipping). Since it weighs a ton and does nothing useful, it's probably
best off as a donor for someone's project. Please send your requests for
any boards/assemblies offline.

Dave Casey