Re: TDS3000B USB floppy emulator
Holger Lübben
Hi Oz!
I've checked my mails and I have to correct myself. It was the manufacturer who answered my technical questions. You can reach them under sales@gotekemulator.com They had a minimum order quantity of 10 items so at the end I bought it somewhere on aliexpress. Holger
|
|
Re: TDS820 help
Dave Casey
Have you tried running Signal Path Compensation (SPC) to see if the scope
will null it out for you? Dave Casey On Wed, Feb 5, 2020 at 10:45 AM george gonzalez <grg2gonzalez@gmail.com> wrote: I bought an old TDS820 with a somewhat blown second channel.
|
|
Re: 2465B - Weak Readout Intensity
Mark Litwack
I replaced a CRT on a 2465 and have done exactly as Chuck suggested with only adjusting the various vertical and horizontal pots. It took about 10 minutes.
toggle quoted messageShow quoted text
The only suggestion I have is to use the cursors to do the alignment. Once for each axis, set them to show a value that is the same as the full graticule range, and then adjust the pots (gain and center) to make the cursors land on the edge graticule lines. I also adjusted the transient response, but that isn't difficult either if you have a pulse generator with a fast edge. -mark
On Tue, Feb 4, 2020 at 05:57 PM, Chuck Harris wrote:
|
|
Re: 2465B - Weak Readout Intensity
Lawrance A. Schneider
On Tue, Feb 4, 2020 at 04:58 PM, <flanneltuba@gmail.com> wrote:
Congratulations!
|
|
TDS820 help
george gonzalez <grg2gonzalez@...>
I bought an old TDS820 with a somewhat blown second channel.
It “works” and seems to be mighty speedy, somewhere near spec, anyway it’s faster than my Jim Williams rise time checker, however but has a large and intermittent negative offset of about 1 volt. Are there any known common fault modes for this that are easily repairable? Like hitting the input stage with a small hammer? I can almost null the offset out with the offset adjust. Thanks, George
|
|
Re: TDS3000B USB floppy emulator
Oz-in-DFW
Holger,
Which reseller did you work with? Oz
|
|
Re: Knob and help.
James R. Bartlett
Thanks Tom,.
toggle quoted messageShow quoted text
I would like to see the files if it is not too much trouble. Appreciate the reply and the offer. Be good to get a good knob or fix up the other one. Thanks again Regards Jim
On Wed, 5 Feb 2020 at 13:03, Tom Gardner <tggzzz@gmail.com> wrote:
When I had a 475 with missing V/div legend, I
|
|
Re: Knob and help.
James R. Bartlett
Thanks Colin.
Ok on the knobs, yes they are pretty scarce. QService are out of stock. THe probes do have the necessary pin, they are mainly TEK switched x1/x10 It appears to be a poor contact somewhere as they work sometimes. Thanks for the reply. regards Jim On Wed, 5 Feb 2020 at 13:17, Colin Herbert via Groups.Io <colingherbert= blueyonder.co.uk@groups.io> wrote: Hi,
|
|
Re: Knob and help.
Colin Herbert
Hi,
toggle quoted messageShow quoted text
If you are referring to the Volts/Div knobs, then it is possible to repair the lettering on them. I did it some time ago with a knob on a 468; those knobs are unobtainium because of the extra positions and black background. I think I used a gel-pen. With regard to the lamps behind the knob skirts which light up the appropriate settings for X1 and X10 probes, to begin with, the probe must have the sensor pin to contact the extra ring on BNC connector. It can be checked that the switching and bulbs are OK by connecting between the extra ring and ground with an 11 k ohm resistor. If this doesn't work, there is a possible broken connection inside, but space is limited in there and it isn't easy to access the input BNC connectors. I also recall that the lamps will not be lit unless the appropriate channel is selected to be active. Colin.
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of James R. Bartlett Sent: 05 February 2020 11:05 To: Tekscopes@groups.io Subject: [TekScopes] Knob and help. Greetings to the group, Regular reader but seldom poster needs help. I am working on a 475 at the moment. I am in need of a vertical sensitivity knob as the lettering on the Channel 2 one is very badly worn. Anyone have a spare one? Also have a problem with the voltage indication changing on the Channel knob. When using a x10 probe sometimes it will not switch to the x10 lamp. moving the probe around at the BNC connector sometimes it comes on, Probes are mainly Tek ones Some switchable others are fixed x10. any thoughts ? Many thanks Jim Ei2BB
|
|
Re: Knob and help.
Tom Gardner
When I had a 475 with missing V/div legend, I
toggle quoted messageShow quoted text
* created my own legend * printed it on transparent plastic * used UHU "rubber solution glue" to stick it behind the plastic on the knob Send me an email if you want the files I used for the legend
On 05/02/20 11:04, James R. Bartlett wrote:
Greetings to the group,
|
|
Re: TDS3000B USB floppy emulator
Holger Lübben
Hi Tom!
My 420A also has a 26 pin interface. I've talked to a reseller of the emulator and they gave me an overview about the different model numbers. The main differences were: - 26 pin or 32 pin interface - 720KB or 1.44MB - Accepting a generic USB Stick or accepting only preformatted sticks of the exact target size. I needed the 26pin interface with a target size of 1.44MB and I wanted to use a generic USB stick - so I ended up with the SFR1M44-DU26. The interface handshake of a floppy drive is a little bit different from vendor to vendor. And I had no information about how Tek implemented the interface in the 420A. That's why I had to test several jumper configurations. The selected emulator works perfectly in my 420A, so there was no need to touch the firmware. Holger
|
|
Re: Stan Griffiths
magnustoelle
I am very sorry to hear about Stan's passing. My condolances to his family and friends.
I like to imagine that he has found a place in heaven, sitting close to Deane Kidd, Jim Williams and Bob Pease... May he rest in peace. Magnus
|
|
Knob and help.
James R. Bartlett
Greetings to the group,
Regular reader but seldom poster needs help. I am working on a 475 at the moment. I am in need of a vertical sensitivity knob as the lettering on the Channel 2 one is very badly worn. Anyone have a spare one? Also have a problem with the voltage indication changing on the Channel knob. When using a x10 probe sometimes it will not switch to the x10 lamp. moving the probe around at the BNC connector sometimes it comes on, Probes are mainly Tek ones Some switchable others are fixed x10. any thoughts ? Many thanks Jim Ei2BB
|
|
Re: Switch retainers for 7000 series plug ins
Hi Martin,
toggle quoted messageShow quoted text
That sounds like the Source INT/LINE/EXT pushbutton switch. If I am correct it uses three modular 2-piece switches. Those are fragile and have several different failure mechanisms. There is a Pushbutton Switch repair guide on TekWiki. I believe it discusses how to fix them if you have the parts. That is a big IF. http://w140.com/tekwiki/images/2/24/Repairing_tek-made_board-mounted_push-button_switches.pdf I know from experience how to fix them because I have had many of them break. Each time I first have to find a donor plugin which gets harder and harder. Good luck, Dennis Tillman W7pF
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Martin Whybrow Sent: Tuesday, February 04, 2020 2:49 PM To: TekScopes@groups.io Subject: [TekScopes] Switch retainers for 7000 series plug ins I've just acquired a 7854 scope for a nice price; the 7B15 timebase module that was included has a pushed-in switch (S90) which I have found to be due to a broken plastic catch that should be holding the switch onto the PCB. Is there a source for these, other than from a scrap plug in? -- Dennis Tillman W7PF TekScopes Moderator
|
|
Re: Switch retainers for 7000 series plug ins
Dave Trainor
I have a couple of parts TM500 plugins, one a FG400 (witch was sorta
diagonally crushed in shipping) (very sad). It has pushbutton switches, does anyone know if the pushbuttons are the same? If so I may have 3 or 4 that are not damaged. But I don¹t know not having a 7000 series plugin if they are even close to the same. 73- Dave N8ZFM On 2/4/20, 11:16 PM, "TekScopes@groups.io on behalf of Clark Foley" <TekScopes@groups.io on behalf of clarkfoley@msn.com> wrote: I had a similar problem on an early 7k prototype mainframe Vert Mode
|
|
Re: Switch retainers for 7000 series plug ins
Clark Foley
I had a similar problem on an early 7k prototype mainframe Vert Mode Select switch. I could not find loose replacement parts.
*******WARNING********** Collectors and instrument restorers should not read further as I do not wish to offend! *******WARNING********** I drilled two small holes through the PCB and strung some no.24 (AWG) steel wire through the holes and looped the wire over the tail of the switch where the nylon clip grabbed the body. Twist the wire together on the back side of the PCB. Crude but effective. Of course you need to avoid traces. I would not have the heart to do this to an otherwise unsuspecting and pristine PCB. I warned you!
|
|
Re: 2465B - Weak Readout Intensity
Congratulations!!!
toggle quoted messageShow quoted text
On Tuesday, February 4, 2020, 03:58:52 PM CST, flanneltuba@gmail.com <flanneltuba@gmail.com> wrote:
The replacement CRT arrived yesterday afternoon, so I promptly sequestered myself out in my unheated shop (in a drafty barn) in 30 degree weather to swap in the new tube. Before I could even think of installing it, I had the very frustrating task of straightening all the bent deflector pins on the neck of the tube, which had been badly bent due to shabby packing for its long trip from Israel. (Who was it on this thread that said something like, every time I order parts from Israel, I've regretted it?) So I spent the first 35 chilly minutes very carefully bending back the bent neck pins. Some of these are fairly stiff and made me cringe as I applied fairly strong mechanical pressure to them to straighten them. I did wear safety glasses and gloves in the event of catastrophic glass failure, all the while praying I would not create a metal-glass leak. Luck was with me, and the only detectable real damage was minor cratering of the glass at the base of the heavier pins. So with that accomplished, I slid in the new (used replacement) CRT. Some tips on technique are due here which you won't find in the manual, which I had to learn the hard way: Because of the delicate pins on the neck of the tube, "sliding" the tube into the chassis must be done in a thoughtful and deliberate manner or you risk bending one or more of these pins if they contact the sleeve as it is being inserted. Two ways to mitigate this possibility are, 1) turn the scope onto its right side, so that as the tube travels into the sleeve, the inevitable pull of gravity on the neck of the tube will cause the neck to slide along its =side= where there are no pin connectors, or 2) set the chassis onto its back side and lower the tube into the sleeve from the top using your best Milton Bradly "Operation" game skills to avoid letting the neck pins hit the sides. It's helpful to position the chassis on the edge of the bench such that the back of the CRT sleeve is accessible from the the underside so you can support the tube in its final centimeter or so of descent, since your ability to grasp the front of the tube will be lost as it enters the chassis at the top. This second technique has the added advantage that the four corner support wedges sit nicely in place during the descent, which they aren't so happy to do in the first. Ok, cutting to the chase! With the new tube mounted and all connectors in place, (with more Milton Bradly maneuvers required for the deflector connections), I fired it up. The new CRT came to life with a beautifully bright and clear display that would happily allow the intensity to go all the way up to full brightness. SUCCESS! Chuck Harris nailed it on the first reply on this thread. Being that my hands were now quite frozen, (remember, I'm doing all this in a 36 degree barn) I decided to leave the rest of the alignment and intensity calibrations for another day. Speaking of calibrations, one poster noted that after a CRT replacement a full calibration would be necessary. That doesn't seem quite right to me, but this is where my knowledge turns to conjecture. Obviously I will need to conduct all the CRT related adjustments, but my question for the group is, what other post CRT replacement calibrations are truly and absolutely necessary? - Scott
|
|
Re: TDS3000B USB floppy emulator
Tom B
Hello Holger,
toggle quoted messageShow quoted text
From what I have learned, the Gotek emulators are all similar but have different firmware. The TDS3000 series use a 26 pin interface, so the cable is a bit different from the 420A. Did you have to change the firmware on the Gotek? I put a new floppy drive in my TDS3044B and it tells me it has invalid media. The same disks work fine in my TDS3012B. So, maybe I have another problem with the floppy interface. Tom Bryan N3AJA
On 2/4/2020 3:56 AM, Holger Lübben wrote:
Hi!
|
|
Re: 2465B - Weak Readout Intensity
Chuck Harris <cfharris@...>
Really, the answer is yes.
toggle quoted messageShow quoted text
The new CRT will certainly have different gain for its deflection plates, its control grid will have a different curve to match. The geometry plates will be a little different. Tilt, and Y axis rotation will certainly be different. And, the transient response is going to be different, because the vertical deflection plates are really a lumped transmission line. The manual says full calibration if you replace a major assembly... the CRT applies. That stuff said, you might be able to satisfy your own needs by doing the CRT calibration, and the rotations. The pots for vertical gain and centering, and Hx1, Hx10, and H-center could be done without additional test equipment. Worth a try... most won't notice the transient response... until they get to the high frequencies. -Chuck Harris flanneltuba@gmail.com wrote:
The replacement CRT arrived yesterday afternoon, so I promptly sequestered myself out in my unheated shop (in a drafty barn) in 30 degree weather to swap in the new tube. Before I could even think of installing it, I had the very frustrating task of straightening all the bent deflector pins on the neck of the tube, which had been badly bent due to shabby packing for its long trip from Israel. (Who was it on this thread that said something like, every time I order parts from Israel, I've regretted it?) So I spent the first 35 chilly minutes very carefully bending back the bent neck pins. Some of these are fairly stiff and made me cringe as I applied fairly strong mechanical pressure to them to straighten them. I did wear safety glasses and gloves in the event of catastrophic glass failure, all the while praying I would not create a metal-glass leak. Luck was with me, and the only detectable real damage was minor cratering of the glass at the base of the heavier pins.
|
|
Re: 7000 horizontal plugins cheap
Martin Whybrow
Just what I'm after, unfortunately about 3,000 miles away from me.
|
|