Re: Your generosity is STUNNING!
Hi Raymond,
toggle quoted messageShow quoted text
You are one of the few people who did correctly identify your contribution in my list. This is the information I have on your contribution: ### DENNIS' LOCAL TIME $$$ STATUS CONTRIBUTOR PAYPAL TRANSACTION 128 Wed 1/1/2020 2:50 PM $22.57 TO BE RETURNED Raymond Domp Frank 9SK435895B100851N Yes, Siggi is correct PayPal did this for everyone. They also obscure the time you made your contribution because they strip out your UTC offset (time zone) so I can’t tell the hour of the day you sent it. The minutes are close however. Thank you for your contribution. We received too much money to keep it all so we will be refunding anything above $10.00 as soon as I have had a chance to contact everyone who asked for confirmation that I received their contribution. Dennis
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Raymond Domp Frank Sent: Monday, January 20, 2020 9:45 AM To: TekScopes@groups.io Subject: Re: [TekScopes] Your generosity is STUNNING! Dennis, My transaction could well be 9SK43*******0851N $22.57 TO BE REFUNDED That's the only transaction in your list for the amount that I transferred. Siggi may be right in assuming that Paypal references change between sender and receiver (nice!) for international transactions. Raymond
|
|
FW: CRT database and Stan's Reprise database of 151 part numbers
This is the reply from Dave Brown at the vintageTEK Museum regarding the CRT
toggle quoted messageShow quoted text
catalog the museum was working on. It is available at https://vintagetek.org/crt-database/ Dave said he does not remember Stan's Reprise data base of 151 part numbers. Dave also said the reprise database is gone forever. I (Dennis) only remember Stan and Bill's CRT database. Dennis Tillman W7PF
-----Original Message-----
From: Dave Brown [mailto:davebr@modularsynthesis.com] Sent: Saturday, January 25, 2020 10:25 AM To: dennis@ridesoft.com Subject: RE: CRT database The CRT database https://vintagetek.org/crt-database/ has all the information. It is just not in the form of a database so the user has to spend maybe 5 minutes to find the data. The reprise database is gone forever. I don't remember the 151- database but it is also gone forever. However, all that information is in the common parts catalogs which we got a release for and are on Tekwiki. Again, someone has to find the 151- to manufacturer pages in the transistor/diode book on Tekwiki and look it up. I assume that is where the database actually came from. Attached but here is the link to the document. http://w140.com/tek_common_design_parts_catalog_transistors_diodes_and_misc_ may_1988.pdf Dave -----Original Message-----site. Also, are you familiar with Stan's Reprise database of 151 part numbers?apparently it is no longer available. -- Dennis Tillman W7PF TekScopes Moderator
|
|
Re: TDS3012B to TDS3052B upgrade
Dave Casey
It works because the hardware was designed for the 500MHz bandwidth from
toggle quoted messageShow quoted text
the beginning. The firmware limits the bandwidth based on model number. That said, there is some debate as to whether there was any "binning" of hardware based on performance (such that scopes which didn't meet spec at 500MHz but did at 300MHz became TDS3032s and 3034s instead of 3052s and 3054s). This doesn't mean that such a scope wouldn't be usable at higher frequencies if the bandwidth limitation were removed, just that it might not meet Tek's advertised specs. There's another camp that says they're all good to the full bandwidth, carte blanche. Either way, your 3012 was probably not calibrated above 100MHz, so if you do the upgrade you may not realize the full benefit without a calibration. Dave Casey
On Sat, Jan 25, 2020 at 9:55 PM Tom B <tbryan@nova.org> wrote:
Hello All,
|
|
TDS3012B to TDS3052B upgrade
Tom B
Hello All,
I have seen a lot of postings from people who have upgraded the bandwidth of their TDS3000 scopes with firmware. I have a TDS3012B that I was thinking about doing that with but it seems strange that you could go from 100MHz to 500MHz without any hardware changes. What I want to know is; how well does this work? Are there any issues to watch out for? Tom Bryan N3AJA
|
|
Re: Your generosity is STUNNING!
Paul McClay
Jay: you don't get a 'show quoted text' button w/Dennis' mesg?
toggle quoted messageShow quoted text
On Sat, Jan 25, 2020 at 06:18 AM, Jay Walling wrote:
Hi Dave,Say, didn't you just post about putting context in replies? ;^)
|
|
Re: 2465B - Weak Readout Intensity
On Fri, Jan 24, 2020 at 01:12 AM, satbeginner wrote:
The 2465x uses what I call a "fly by wire" control system.... The sample and hold circuits are build using a simple opamp circuit.Thank you for pointing this out! This illuminates an entire section of the DAC "fly by wire" circuit that I have heretofore overlooked. Certainly before I go to the monumental effort of performing a CRT swap, I will sniff my way carefully thought this MUX'd sample-and-hold circuitry. I have ordered a CRT, which hopefully will be of the 2465A/B flavor, but I'd be nothing but glad to just put it up on a shelf or resell it instead of having to install it! Also let me add my "+" to the massively informative, if not exasperatingly long and multilayered EEVBlog "2465B Teardown" thread. I spent hours pouring over that thread—hours well spent for anyone diving into a 2565B restoration. It was that thread that essentially illustrated the finer points of repairing the later model A5 board with its notorious leaky 'lytics. - Scott
|
|
Re: P6201 probe Q100 FET replacement
Gary Appel
I think that years ago I got some replacement FETs from Dean Kidd, and they had to be mounted upside down because the lead configuration was reversed.
toggle quoted messageShow quoted text
Gary Appel
On 1/25/2020 12:59 AM, Nenad Filipovic wrote:
Dear Group,
|
|
Re: 2400 series Test 4 Fail 3,13, 1
GerryR <totalautomation1@...>
Craig,
toggle quoted messageShow quoted text
My guess would be that you got an old battery from Mouser. I purchased mine from Mouser a few months back, and it came with no date code stamped on it, so who knows how long its been on the shelf. Mine is still working, however, and the one that I took out with a date code of 07-87 still has its full voltage on it. If my new battery fails anytime soon, Mouser is going to hear about it! The calibration process, as you know is quite a tedious process. Regards, GerryR KK4GER
----- Original Message -----
From: "Craig Cramb" <electronixtoolbox@gmail.com> To: <TekScopes@groups.io> Sent: Saturday, January 25, 2020 12:02 PM Subject: Re: [TekScopes] 2400 series Test 4 Fail 3,13, 1 Dealing with 3 scopes all having the Test4 Failures 2465A S/N B011512 Test4 Fail 3 (Fixed) 2465B S/N B010817 Test4 Fail 13 ( This Post) 2465A DM S/N B010824 Test 4 Fail 2 (Haven’t started) Thank to all for the info in this post. This has been long journey but glad to see an underlying condition could be existing. I’m looking at completing the repair of the 2465B instrument pre 50,000 S/N. Got started on it about 6 months ago troubleshooting with help from members. Did the repair by replacing the U2460 (original part # NEC D4364C-15LL, replaced with part # NEC D4364C-20L) and replaced the Dallas Keeper battery purchased from Mouser. That was June 2019. Other things took my interest so just now getting back to calibration. After doing the required calibration I kept getting diagnostic of test 04 fail 13, Test 04 fail 11, Test 06 fail 01, then back to test 04 fail 01. Couldn’t seem to get ahead of it. So did what I should have done before attempting the calibration and checked the battery voltage. Yea I found it at .6VDC. No wonder I’m dealing with these faults. So now onto why would the battery be drained in 6 months. The old battery I removed that was from the 1988 vintage was in the same condition so I just thought changing it 6 months ago would allow for another 10 yrs. But that wasn’t the case. So tracing thru the circuit there isn’t anything that I see that would allow for a quick drain of this battery. Feeding two IC, an amplifier, and two caps. Not sure how long it took to drain the battery. Craig
|
|
Re: 2400 series Test 4 Fail 3,13, 1
Dealing with 3 scopes all having the Test4 Failures
2465A S/N B011512 Test4 Fail 3 (Fixed) 2465B S/N B010817 Test4 Fail 13 ( This Post) 2465A DM S/N B010824 Test 4 Fail 2 (Haven’t started) Thank to all for the info in this post. This has been long journey but glad to see an underlying condition could be existing. I’m looking at completing the repair of the 2465B instrument pre 50,000 S/N. Got started on it about 6 months ago troubleshooting with help from members. Did the repair by replacing the U2460 (original part # NEC D4364C-15LL, replaced with part # NEC D4364C-20L) and replaced the Dallas Keeper battery purchased from Mouser. That was June 2019. Other things took my interest so just now getting back to calibration. After doing the required calibration I kept getting diagnostic of test 04 fail 13, Test 04 fail 11, Test 06 fail 01, then back to test 04 fail 01. Couldn’t seem to get ahead of it. So did what I should have done before attempting the calibration and checked the battery voltage. Yea I found it at .6VDC. No wonder I’m dealing with these faults. So now onto why would the battery be drained in 6 months. The old battery I removed that was from the 1988 vintage was in the same condition so I just thought changing it 6 months ago would allow for another 10 yrs. But that wasn’t the case. So tracing thru the circuit there isn’t anything that I see that would allow for a quick drain of this battery. Feeding two IC, an amplifier, and two caps. Not sure how long it took to drain the battery. Craig
|
|
Re: Tek 222 Probe ground lead
tekscopegroup@...
The probes on the 222 are permanently attached to the scope, so I would think it would be difficult to get just a whole probe, unless coming from a parts unit. I just looked at one of the detachable probe ground leads of my 222 and it does not look easy to find an alternative that fits nicely. But if the ground on both probes is common as it probably should be (not sure the fact that the inputs of this scope can be floated up to some serious voltage would change this) maybe you could leave the one ground cable on the ch1 probe and the if you need to use ch2 on the same circuit with ch1 already connected, chances are you would not need the second ground. I hope someone with a parts unit will be able to help you.
|
|
Re: 11801 questions - funny fan noise and light
Chuck Harris <cfharris@...>
Mine is normal until after the scope has been turned on for
toggle quoted messageShow quoted text
about 30 minutes, and then it starts going "Brummm..... Brummm..." I suspect a capacitor failure, as it is annoying, and would have caused notice if it were a "normal" condition. Annoying, though it may be, I can't see how it would cause any harm... unless it presages a total failure. What it isn't is a bearing failure, or lack of lubrication. My fan spins for what seems like half a minute after power is turned off. -Chuck Harris Reginald Beardsley via Groups.Io wrote:
I don't recall mine ever doing that. I'm very sensitive to noise as I investigate anything I can't identify, so I think I would have noticed it.
|
|
Re: Mounting and dismounting curve tracer 177 fixture
Chuck Harris <cfharris@...>
To paraphrase the 577 manual on gaining access to the 577's plugin
toggle quoted messageShow quoted text
connector for replacement: ... remove the 4 screws from the bottom cover of the scope. Pay note that two of the screws are shorter than the rest, and must be placed in the front two holes... It seems likely that some technician, at some point in the scope's life, removed the bottom cover plate from the scope, and scrambled the screws... thus locking the plugin into the scope. -Chuck Harris Gary Robert Bosworth wrote:
Thank you for your assistance. Sure enough, there were 2 screws on the
|
|
Re: Your generosity is STUNNING!
Hi Dave,Say, didn't you just post about putting context in replies? ;^) Jay
|
|
Re: 11801 questions - funny fan noise and light
Reginald Beardsley
I don't recall mine ever doing that. I'm very sensitive to noise as I investigate anything I can't identify, so I think I would have noticed it.
A fan on one of my HP Z400 workstations was doing that. I shut it down to reduce the heat in the room and when I tried to boot it the fan was dead. I've not had time to correct it yet. Good luck, Reg
|
|
Re: P6201 probe Q100 FET replacement
YES! I'm the OP for the cited post. My P6201 works perfectly with D and S switched using an MMBF4416
toggle quoted messageShow quoted text
David
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Nenad Filipovic Sent: 25 January 2020 11:00 To: TekScopes@groups.io Subject: [TekScopes] P6201 probe Q100 FET replacement Dear Group, I got a dead P6201 off eBay, and everything points to a defective input FET (probe serial B065531). So I obtained a handful of MMBF4416 (code 151-5025-00 in the service manual, factory default for later serial numbers). Tek also released a product modification document under 050-1413-01 code: http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Tek/050-1413-01.pdf same document: http://hakanh.com/dl/docs/kitinstructions/050-1413-01.pdf outlining this replacement. The problem I have now is that drain and source of the 151-5025-00 FET in the above document are reversed compared to MMBF4416. None of the datasheets for MMBF4416 say it's symmetrical, and I somehow doubt an RF JFET would be without optimizations on the drain side. This comment however states it is symmetrical: https://groups.io/g/TekScopes/message/131217 but I see no facts this claim is based on. Does anyone have direct experience with this particular replacement and D-S polarity? Disassembling the probe head is a pain, I'd like to get this right on the first go. Best Regards, Nenad
|
|
Re: P6201 probe Q100 FET replacement
Nenad Filipovic
Is the pinout correct if you install it upside down?Yes, and this is most likely what I will do. But anyway I wanted to hear from people who already went through this. I just find it strange that Tek document has it reversed without any comment or regard.
|
|
Re: P6201 probe Q100 FET replacement
Is the pinout correct if you install it upside down?
toggle quoted messageShow quoted text
On Sat, Jan 25, 2020 at 6:00 AM Nenad Filipovic <ilmuerte@gmail.com> wrote:
Dear Group,
|
|
P6201 probe Q100 FET replacement
Nenad Filipovic
Dear Group,
I got a dead P6201 off eBay, and everything points to a defective input FET (probe serial B065531). So I obtained a handful of MMBF4416 (code 151-5025-00 in the service manual, factory default for later serial numbers). Tek also released a product modification document under 050-1413-01 code: http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Tek/050-1413-01.pdf same document: http://hakanh.com/dl/docs/kitinstructions/050-1413-01.pdf outlining this replacement. The problem I have now is that drain and source of the 151-5025-00 FET in the above document are reversed compared to MMBF4416. None of the datasheets for MMBF4416 say it's symmetrical, and I somehow doubt an RF JFET would be without optimizations on the drain side. This comment however states it is symmetrical: https://groups.io/g/TekScopes/message/131217 but I see no facts this claim is based on. Does anyone have direct experience with this particular replacement and D-S polarity? Disassembling the probe head is a pain, I'd like to get this right on the first go. Best Regards, Nenad
|
|
Re: Your generosity is STUNNING!
Ken Wright
Hi Dennis. This has been tremendous response, my personal views are that this group should be secured for the future, after that I would leave it to your discretion.
toggle quoted messageShow quoted text
Ken Wright M0KHW
On Saturday, 25 January 2020, 02:21:20 GMT, Dennis Tillman W7PF <dennis@ridesoft.com> wrote:
Hi Ken, By now you have probably seen the reason your transaction number wasn't found. PayPal issues you one transaction number and issues me a totally different number. I had no idea this was the case until a few days ago. Your contribution came through just fine. This is the information I received: ### DENNIS' LOCAL TIME $$$ STATUS CONTRIBUTOR PAYPAL TRANSACTION 168 Fri 1/3/2020 3:45 AM $10.20 Ken says keep it LU27EUKen Wright 5UU92005CL6140032 Thank you for your contribution, Dennis -----Original Message----- From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Ken Wright via Groups.Io Sent: Sunday, January 19, 2020 2:54 AM To: TekScopes@groups.io Subject: Re: [TekScopes] Your generosity is STUNNING! Hello Dennis.I cannot see my contribution on your list ($10.20), PayPal number. 4CT072513H300731W.I am quite happy for you to retain it, I would like to know that you had received it. Best wishes.Ken Wright M0KHW On Saturday, 18 January 2020, 16:57:15 GMT, Dennis Tillman W7PF <dennis@ridesoft.com <mailto:dennis@ridesoft.com> > wrote: Wrapping up the TekScopes pledge drive, Michael Dunn and I are STUNNED by your generosity and all the positive comments we received. Your contributions have exceeded the $340 paid to Groups.io by almost 1,000%. You sent us 203 contributions totaling $3,341.77! We agree this is a problem that I will have to fix since it was my bright idea. I did not realize when I sent my request for help that people would be so generous. I wrote, "So I am asking for small contributions". What happened was anything but "small". Leo Bodnar, a friend, stunned me when he contributed the entire required amount two days later. This was the first sign I should have chosen a synonym such as "tiny" or "teeny" instead of small. It was too late at that point. I could see there was a tsunami headed my way and I had no way to stop it. I think the contributions have finally stopped. The last one (I hope) arrived three days ago. Michael and I have decided that this is far too much money for us to keep on behalf of TekScopes. The simplest solution is for us to tear up the checks and return all the PayPal contributions above $10.00 (actually all those above $10.05) over the next few days, with our sincerest thanks. This will result in exactly half of the contributions being kept and half being returned. PayPal allows me to do a total refund very quickly; we considered partial refunds, but the process would take much longer. If, after you receive your refund, you would still like to make a contribution to TekScopes, please don't send more than $10.00. If you do, we'll be forced to spend it on wine and DMMs. Thank You, Dennis Tillman W7PF Michael Dunn VA1XY As I promised, here is every contribution I received. The four checking account contributions are listed in numerical order by their bank account number. Following those, the 199 PayPal contributions are listed in numerical order by their transaction number. I double and triple checked the PayPal notices that came to me or to my wife, Marian, to be sure I didn't miss any. If I missed your PayPal contribution please let me know so I can find out what happened to it. The most common contribution was $10.00 and the average contribution was $15.94. CHECKING ACCOUNT 0000*255 $50.00 TO BE REFUNDED 0035*054 $20.00 TO BE REFUNDED 4650*2065 $50.00 TO BE REFUNDED 9179*1067 $50.00 TO BE REFUNDED PAYPAL TRANSACTION 02079*******9581B $10.00 03E74*******9033E $100.00 TO BE REFUNDED 05P54*******33704 $20.00 TO BE REFUNDED 0DV52*******9640N $20.00 TO BE REFUNDED 0FX24*******0225D $12.61 TO BE REFUNDED 0GH61*******5321H $10.00 0H164*******7324J $5.00 0HP31*******79205 $10.00 0TB16*******6233S $10.00 0V228*******9483Y $15.00 TO BE REFUNDED 0WE54*******4492F $20.00 TO BE REFUNDED 0XD49*******99831 $10.00 0XX33*******5571C $20.00 TO BE REFUNDED 10F89*******59157 $20.00 TO BE REFUNDED 11D81*******1581M $25.00 TO BE REFUNDED 13W62*******8082B $5.00 16621*******82104 $20.00 TO BE REFUNDED 16T81*******89037 $21.00 TO BE REFUNDED 16U30*******4884J $10.00 17T42*******7831E $10.00 1C179*******04615 $10.00 1C830*******8112P $10.00 1KB10*******35308 $10.00 1KC35*******1823E $10.00 1KK41*******81702 $40.00 TO BE REFUNDED 1MJ87*******7921S $20.00 TO BE REFUNDED 1PL71*******8194S $20.00 TO BE REFUNDED 1SF61*******75325 $6.28 1U469*******7321A $10.00 1UN19*******71331 $21.50 TO BE REFUNDED 1XL88*******0010V $10.00 21K47*******0942J $5.00 26M35*******2652B $20.00 TO BE REFUNDED 26W48*******18439 $10.00 27955*******6435G $2.00 28784*******31459 $10.00 2AJ63*******9742M $20.00 TO BE REFUNDED 2BB60*******7284T $10.05 2C041*******7435F $15.00 TO BE REFUNDED 2CF51*******55300 $10.00 2GN07*******6691P $21.36 TO BE REFUNDED 2HA37*******2763N $20.00 TO BE REFUNDED 2JP56*******9614M $6.00 2L669*******5001S $20.03 TO BE REFUNDED 2LL81*******2181A $10.00 2NX73*******6334E $25.00 TO BE REFUNDED 2PE83*******8434J $20.00 TO BE REFUNDED 2R866*******6484Y $25.00 TO BE REFUNDED 2S663*******60054 $10.00 2UD84*******92512 $20.00 TO BE REFUNDED 2WG55*******8174X $10.00 30988*******3613G $10.00 31532*******3440H $25.00 TO BE REFUNDED 32S52*******3263Y $25.00 TO BE REFUNDED 33N46*******8874F $20.00 TO BE REFUNDED 35J04*******2071X $5.00 37611*******3481X $30.28 TO BE REFUNDED 38D56*******6900H $40.00 TO BE REFUNDED 38H00*******74637 $10.00 39503*******91844 $10.00 3C222*******60057 $5.00 3DL81*******49307 $10.00 3DN03*******6580W $10.00 3HN95*******74316 $20.00 TO BE REFUNDED 3JY75*******0415H $10.00 3MK47*******8000V $20.00 TO BE REFUNDED 3ML12*******7670M $30.00 TO BE REFUNDED 3RW61*******7583B $110.00 TO BE REFUNDED 3SV52*******0093K $10.00 3U066*******6064R $10.00 3W759*******4273V $10.00 3WN84*******32644 $20.00 TO BE REFUNDED 3X630*******09454 $10.00 43V15*******1315P $30.00 TO BE REFUNDED 43W14*******2963X $10.00 4C132*******25402 $20.78 TO BE REFUNDED 4CP43*******1271K $15.00 TO BE REFUNDED 4E489*******92025 $20.00 TO BE REFUNDED 4GU02*******14949 $10.00 4H892*******5443S $10.00 4NK97*******39943 $30.00 TO BE REFUNDED 4NS27*******8414L $20.00 TO BE REFUNDED 4PM01*******5380X $10.00 4RA41*******16029 $20.00 TO BE REFUNDED 4V179*******50029 $25.00 TO BE REFUNDED 4VU25*******5244C $10.00 4WE10*******2420U $12.00 TO BE REFUNDED 4WG41*******2952L $20.00 TO BE REFUNDED 4XH55*******98246 $10.00 4XM94*******9951L $20.00 TO BE REFUNDED 4XU29*******2203A $5.00 52Y94*******9314H $10.00 53E69*******9450U $15.00 TO BE REFUNDED 53M48*******1162D $10.04 54264*******83607 $20.00 TO BE REFUNDED 54350*******3123J $10.00 56419*******9740B $10.68 TO BE REFUNDED 57958*******19248 $10.00 57M28*******3463Y $15.00 TO BE REFUNDED 58J72*******8751M $5.05 59K74*******6614F $10.00 59M25*******3721R $10.72 TO BE REFUNDED 5FC22*******9632C $10.00 5GA43*******4223M $5.00 5K650*******2512S $10.00 5L001*******9890K $30.00 TO BE REFUNDED 5L631*******0780F $5.00 5P795*******19647 $20.00 TO BE REFUNDED 5SE25*******18324 $20.00 TO BE REFUNDED 5UP40*******89642 $20.00 TO BE REFUNDED 5UU92*******40032 $10.20 61H86*******4671Y $15.00 TO BE REFUNDED 61W04*******0651P $5.00 62172*******4981K $10.03 62762*******2074B REFUNDED 64M95*******0450J $20.20 TO BE REFUNDED 69567*******7993L $10.00 69G97*******6681W $5.00 6AX91*******2221W $10.00 6DC49*******7192J $2.20 6DY20*******43532 $10.00 6FR77*******2981V $20.00 TO BE REFUNDED 6GH42*******5101K $25.00 TO BE REFUNDED 6JF13*******51538 $13.42 TO BE REFUNDED 6M239*******2634X $20.00 TO BE REFUNDED 6M818*******3010R $10.75 TO BE REFUNDED 6MW55*******8294T $4.00 6NN32*******4542V $10.00 6PN72*******3184X $20.00 TO BE REFUNDED 6SJ24*******7280W $10.00 6TX54*******6691J $10.00 6UC87*******60625 $5.00 6W050*******3122G $25.00 TO BE REFUNDED 6YC11*******1694V $10.00 6YR09*******24211 $10.00 70R50*******48816 $40.00 TO BE REFUNDED 71A57*******04246 $20.00 TO BE REFUNDED 71B15*******9210H $5.00 77749*******2345C $5.00 7AA50*******6582S $5.00 7BH77*******10502 $10.00 7C686*******0453C $20.00 TO BE REFUNDED 7C834*******0923E $10.00 7CK01*******1245K $5.00 7D080*******7730E $10.00 7HC88*******8542G $50.00 TO BE REFUNDED 7JC48*******6100G $50.00 TO BE REFUNDED 7LV25*******2193G $30.00 TO BE REFUNDED 7M285*******92740 $25.00 TO BE REFUNDED 7MK85*******1842C $10.00 7MS44*******5362B $15.00 TO BE REFUNDED 7NK34*******5412S $31.44 TO BE REFUNDED 7PE90*******5070C $15.00 TO BE REFUNDED 7PW51*******11256 $5.00 7SB72*******7192F $20.00 TO BE REFUNDED 7UJ45*******9004U $5.00 7V986*******7741K $10.00 7WL39*******4140E $10.00 80874*******3273N $20.00 TO BE REFUNDED 82A24*******71901 $20.00 TO BE REFUNDED 85T72*******59347 $10.00 86464*******0522V $25.00 TO BE REFUNDED 89L93*******2080C $20.00 TO BE REFUNDED 8AB06*******94251 $20.00 TO BE REFUNDED 8BM27*******0873M $20.00 TO BE REFUNDED 8BP86*******45402 $20.00 TO BE REFUNDED 8DF32*******6052V $10.00 8EJ68*******08541 $12.00 TO BE REFUNDED 8FJ75*******8872W $20.00 TO BE REFUNDED 8GC22*******34141 $25.00 TO BE REFUNDED 8GT73*******09326 $12.61 TO BE REFUNDED 8MC99*******59946 $11.00 TO BE REFUNDED 8N867*******6281R $20.00 TO BE REFUNDED 8P783*******6711A $23.00 TO BE REFUNDED 8PJ76*******6821F $10.00 8TL81*******09540 $20.00 TO BE REFUNDED 8X573*******06030 $10.00 8XN50*******4422L $20.00 TO BE REFUNDED 90T91*******35713 $10.00 91410*******3142B $10.00 91U38*******2105T $11.00 TO BE REFUNDED 92261*******0350X $10.00 94C22*******3914J $50.00 TO BE REFUNDED 97618*******3223U $20.00 TO BE REFUNDED 98Y68*******70508 $1.24 99954*******8720E $5.00 9A088*******0853S $20.00 TO BE REFUNDED 9AP46*******63352 $10.00 9B390*******8403F $10.00 9DH47*******6461J $10.00 9GJ81*******0164P $10.00 9HK96*******73147 $25.00 TO BE REFUNDED 9HS74*******5364F $20.00 TO BE REFUNDED 9HY36*******94500 $10.00 9KN17*******5330K $5.00 9N383*******7605L $5.36 9SK43*******0851N $22.57 TO BE REFUNDED 9SN87*******3420T $10.00 9UB93*******8990M $20.00 TO BE REFUNDED 9V074*******7422K $5.37 -- Dennis Tillman W7PF TekScopes Moderator
|
|
Re: 7L5 dot frequency reset bug - any clues or a fix?
Dave Casey
I can try one in a 7603 at some point in the not too distant future.
Dave Casey On Fri, Jan 24, 2020 at 10:55 PM Cliff Carrie <cliffcarrie@hotmail.com> wrote: Thanks for the data, Dave.
|
|