Date   

Re: Tek 533 -150 test point

N7QL
 

Thank you very much! I appreciate it.


Re: Tektronix 532 LV rail shorted

randolphbeebe@...
 

Rajesh,

I have not had time to look at it again yet. Hopefully this weekend. One spot of good news is that I just checked the fuse and it was blown. That would sure account for no voltage at C640!

One thing that is curious however, upon reviewing the schematics of the LV power supply I read my notes and recalled that when I measured the voltage at pin 16 on the transformer it measured +89V AC. I do not expect to see the voltage near the correct numerical value on the crippled circuit but POSITIVE voltage? Would that account for the symptoms as well since both C640 and C110 are installed positive to ground?

Can that positive voltage be coming from one of the other supply rails perhaps, the +350, +225 or the +100? Or a fault in the transformer. It gives me something to think about when I read about the circuit description.

Your time distribution method is excellent, I am trying to get away from the neophyte's instinct to just shotgun parts to "see if it works now".

Randy


Re: MEMBERS PLEASE READ: Our annual Group.io payment is due in 2 weeks.

Lawrance A. Schneider
 

Next year is AN OTHER YEAR.
We will come to the same place as far as needing contributions.
Some succeeded and some did not (me).

So, what do we do next year. I very much enjoy this group! I would be happy to contribute!

HOW????

larry


Re: 7A26 transient response 5 nS all atten

Albert Otten
 

Hello Jon,
My other 7A26s have this format SN, but this has no letter, 708058. I think its a late model, but see no IC date codes.
Readout PCB is 672-0051-13 Amplifier PCB is 670-2549-21 attenuators are PN200-1442-00
The Readout board is probably marked 670-2310-09. Then both board versions were introduced at B251090. You need the -01 version manual for that, Revised SEP 1985. I think the ICs have 3-digit date codes, the first indicating the year (5xx pointing to 1985). Of course the plugin itself can be newer than the ICs in it. My 713121 is from 1986.
BTW did you mean 7904 or 7104?

Albert


Re: MEMBERS PLEASE READ: Our annual Group.io payment is due in 2 weeks.

Paul Amaranth
 

Email addresses are of the form

<ID> @ <Organization>

If it isn't in that form it's not an email address. Period.
Color has no meaning for email.

Groups.io disguises email addresses so harvesting of email address from
the public website is more difficult. This is a good thing as it
prevents some amount of spam.

I generally do not use the web interface, but I think the assumption
here is you would go to the member's profile and either get the
correct email address from there or send them a private message.

If you got the list as an email instead of looking at the website,
email addresses would be visible since they are not redacted.

I think groups.io did a really nice job with that.

Paul

On Fri, Jan 17, 2020 at 06:42:49AM -0800, Lawrance A. Schneider wrote:
Same happens to me.
further, what it the difference between @Dennis_Tillman_W&PF and the same thing in blue??? I saw a missive saying to use the above in blue and the same thing in black. I again tried each at PayPal and got the same thing asking for an :an error saying please enter valid email, name or mobile phone m=number.
I know I'm getting old, but what is going on???????

larry
--
Paul Amaranth, GCIH | Manchester MI, USA
Aurora Group of Michigan, LLC | Security, Systems & Software
paul@AuroraGrp.Com | Unix & Windows


Re: 2445 EPROMs

Mark Litwack
 

Hi Chuck and Chris,

The external clock input on the 6802 is divided internally by four to create an internal two-phase clock (one phase is output as signal "E"). So, with a 5MHz input, the cycle time is actually 800ns, which well within range of the Intel D27128 with 250ns access time I mentioned before.

There is also the NOP exerciser which can be enabled with jumper P503. This is good for getting to the bottom of possible processor and/or bus buffer issues by looking at E, address lines, and other signals as the processor is looping. Its operation is described in the service manual.

I do recall one instance of a bad processor, but granted, they are rare.

-mark


Re: MEMBERS PLEASE READ: Our annual Group.io payment is due in 2 weeks.

Lawrance A. Schneider
 

Same happens to me.
further, what it the difference between @Dennis_Tillman_W&PF and the same thing in blue??? I saw a missive saying to use the above in blue and the same thing in black. I again tried each at PayPal and got the same thing asking for an :an error saying please enter valid email, name or mobile phone m=number.
I know I'm getting old, but what is going on???????

larry


Re: 7A26 transient response 5 nS all atten

Jean-Paul
 

Colin! Many thanks, My manuals cover only B serials. Changes in Guernsey version?

All: My first reaction was calibration; page 5-7 sec 8, High Freq compensation.
But there are 10 adjustments per channel listed!
What order of adjustment? Which one might be for just the first 5-10 nS?

Mille mercis,

Jon


Re: Remote CSA803 / HP1180x programming (yet again)

GerryR <totalautomation1@...>
 

It must have been "Recommended Standard," because I had (have somewhere?) a book that had 132 or so versions of the RS-232 "standard" interface. That's a lot of versions of a so-called standard!
GerryR
KK4GER

----- Original Message -----
From: "Leo Bodnar" <leo@leobodnar.com>
To: <TekScopes@groups.io>
Sent: Friday, January 17, 2020 8:18 AM
Subject: Re: [TekScopes] Remote CSA803 / HP1180x programming (yet again)


What I really like about RS-232 is that RS stands for "Recommended Standard"!
Leo
On Fri, Jan 17, 2020 at 10:59 AM, Roy Morgan wrote:
Among my coworkers were the folks who organized and chaired the international committees that wrote the standards. (One fellow had chaired the committee that developed RS-232.)


Re: "liquid crystal driver" sighting

Glenn Little
 

I am currently working on drawing a schematic. Slow going. Hope to have it done and posted sometime in February.

Glenn

On 1/17/2020 3:04 AM, p mc wrote:
Thank you, Mr. Little, for starting to fill in a blank spot on the map.

At TekWiki I cleaned out the obsolete stub text and poor front panel image.


--
-----------------------------------------------------------------------
Glenn Little ARRL Technical Specialist QCWA LM 28417
Amateur Callsign: WB4UIV wb4uiv@arrl.net AMSAT LM 2178
QTH: Goose Creek, SC USA (EM92xx) USSVI LM NRA LM SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license"


Re: Looking for cabinet top for 465M

Bert Haskins
 

On 1/16/2020 7:29 PM, KB6NAX wrote:
I'll buy a junked 465M if it will go for cheap just to get a top cover. Just let me know. Thanks. -Ard
Where are you located?


Re: Remote CSA803 / HP1180x programming (yet again)

Leo Bodnar
 

What I really like about RS-232 is that RS stands for "Recommended Standard"!
Leo

On Fri, Jan 17, 2020 at 10:59 AM, Roy Morgan wrote:
Among my coworkers were the folks who organized and chaired the international committees that wrote the standards. (One fellow had chaired the committee that developed RS-232.)


Re: 7A26 transient response 5 nS all atten

Colin Herbert
 

Hi,
I, too, have a 7A26 with no letter at the front of the serial-number (102925). The label with this serial-number on it also identifies the plug-in as having been manufactured in Tektronix, Guernsey, which is why there is no "B" or "H" in front of the serial-number.
My manual, though originally produced in Beaverton (as I think all may have been), has a page in the "Change Information" section at the back referring to changes to Guernsey-produced 7A26s from tentative S/N 100396 and 100777.
I don't know how much use this will be to you, but I also have a source of the ICs used in the 7A26 (unless they have all been sold). They don't usually fail, but swapping them between channels (they are plug-in but be careful of bending the legs) is a pretty simple way of verifying them.
There are also a couple of resistors, one in each channel, which are at about their wattage limit, get hot and sometimes change value because of this. They are R1327 and R1329 (8K2, 0.25W. Carbon comp.) and would benefit in being replaced with resistors of higher wattage.
Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Jean-Paul
Sent: 17 January 2020 12:20
To: TekScopes@groups.io
Subject: [TekScopes] 7A26 transient response 5 nS all atten

Hello all: My workhorse 7A26 needed rework, in 7804 mainframe.
Cleaned the AC/DC/GR contacts on attn with the iso alcohol-paper strips, and exercised some scratchy controls.
Poor fast transient response on CH2, same dip and peak ~ 5 nS at ALL V/div.

Setup: Leo Bodnar pulser 40 pS>Mini Ckt HAT-15>>Mini Ckt HAT 20>>precision 50 Ohm thru>>7A26 input
Same result if 7A26 is in L or R side 7804.
Shots are: Affected 7A26, CH2, CH1, Good 7A26 Ch2 Ch1
http://crypto-museum.org/TEK/7A26/7A26.pdf

Since the aberration is constant on every V/DIV, the attenuator and 2X amp seem to be exonerated.

Before I dare probe the CH 2 amp circuit, a few questions please:

1/ Test points mentioned on schematic are NOT in PCB photos, only the parts numbers. Need a board layout with test pts.
2/ Using normal P6137 probe. placement of ground lead ?
3/ Precautions in probing around the custom TEK hybrid ICs, eg U2350, U2450?
4/ Serial: Manual has sch versions for SN B179999 < and B180000>.
My other 7A26s have this format SN, but this has no letter, 708058. I think its a late model, but see no IC date codes.
Readout PCB is 672-0051-13 Amplifier PCB is 670-2549-21 attenuators are PN200-1442-00

Hope to receive some tips from our TEK veterans!

Jon in Paris


Re: Remote CSA803 / HP1180x programming (yet again)

Don Bitters
 

Ah yes, an impressive o-scope, that I too loved to hate.  With the right plugins it would do 20+ GHz.  My experience with them was to calibrate them - manually! for HP/Agilent.  IBM had a ton of them on their optical amp production lines.  Nobody at Tek or HP ever automated the cal procedure.  I have cal'd about 50 of the CS803's and 1180x's with the SD22 and SD24 plugins, the plugins have plugins!!!.  Through a lot of diligence and practice I could get the cal procedure down to 3.5 hrs on the mainframe and 5.5 hrs. on the plugins with full data taken.  I think the data stack for 12 ea. 11803's with plugins with 3 slots occupied with 2ea. SD24's/ plugin was over 10 in. tall. That was over a week of calibration alone at 10+ hr. days, definitely not fun.I have never heard of anyone automating the cal on these, probably too much programming effort.IBM did run them full automation on their production lines for their testing.Don Bitters


7A26 transient response 5 nS all atten

Jean-Paul
 

Hello all: My workhorse 7A26 needed rework, in 7804 mainframe.
Cleaned the AC/DC/GR contacts on attn with the iso alcohol-paper strips, and exercised some scratchy controls.
Poor fast transient response on CH2, same dip and peak ~ 5 nS at ALL V/div.

Setup: Leo Bodnar pulser 40 pS>Mini Ckt HAT-15>>Mini Ckt HAT 20>>precision 50 Ohm thru>>7A26 input
Same result if 7A26 is in L or R side 7804.
Shots are: Affected 7A26, CH2, CH1, Good 7A26 Ch2 Ch1
http://crypto-museum.org/TEK/7A26/7A26.pdf

Since the aberration is constant on every V/DIV, the attenuator and 2X amp seem to be exonerated.

Before I dare probe the CH 2 amp circuit, a few questions please:

1/ Test points mentioned on schematic are NOT in PCB photos, only the parts numbers. Need a board layout with test pts.
2/ Using normal P6137 probe. placement of ground lead ?
3/ Precautions in probing around the custom TEK hybrid ICs, eg U2350, U2450?
4/ Serial: Manual has sch versions for SN B179999 < and B180000>.
My other 7A26s have this format SN, but this has no letter, 708058. I think its a late model, but see no IC date codes.
Readout PCB is 672-0051-13 Amplifier PCB is 670-2549-21 attenuators are PN200-1442-00

Hope to receive some tips from our TEK veterans!

Jon in Paris


Re: Risetime calculator (in tekwiki)

Jean-Paul
 

I have the circular slide rule version, Tektronix Circuit Computer (TEK 003-023) circa 1961, risetime, resonance, impedance, etc.
https://vintagetek.org/tektronix-circuit-computer/

Any others have one?
Looking for another one!

Jon


Re: Risetime calculator (in tekwiki)

GerryR <totalautomation1@...>
 

Is there a Tek part number associated with this unit? How is it used to calculate rise time?
Thanks,
GerryR

----- Original Message -----
From: "Tim Phillips" <timexucl@gmail.com>
To: <TekScopes@groups.io>
Sent: Friday, January 17, 2020 2:32 AM
Subject: [TekScopes] Risetime calculator (in tekwiki)


from Tim P (UK)
What is this, please? It seems to be a simple analog computer, maybe with
Log pots.
In case the link doesn't work, the device is in tekwiki Other Instruments.

http://w140.com/tekwiki/wiki/Risetime_Calculator

Appears to give f = sqrt(a^2 + b^2 + c^2)

thanks
Tim P


Re: Remote CSA803 / HP1180x programming (yet again)

Roy Morgan
 

Leo,

I do remember Hays modem commands, though only dimly.

I wonder if "AT" stands for "Attention, Terminal ! ".

On a related bit of lore: while I worked at NIST, I used a bit of SQL and learned that the letters originally stood for Structured Query Language but that later the official name of the standard was changed to "Database Language SQL", with no meaning for the letters.

Among my coworkers were the folks who organized and chaired the international committees that wrote the standards. (One fellow had chaired the committee that developed RS-232.)

IBM used to mean "International Business Machines" didn't it?

And so we "Graybeards" mumble off into the mists of time. ... ...

Roy Morgan
K1LKY Western Mass

On Jan 17, 2020, at 4:31 AM, Leo Bodnar <leo@leobodnar.com> wrote:
...
If you are old enough - think modems Hayes AT commands.
...


Re: Risetime calculator (in tekwiki)

Leo Bodnar
 

It can be also done with linear pots and quadratic scale markings around the pot.
Or a slide rule with two sliders.
Leo


Remote CSA803 / HP1180x programming (yet again)

Leo Bodnar
 

Has anybody produced anything interesting recently for CSA803 / 1180x series that allows remote comms? Apart from screen shot capture and printing.
The talks about these instruments spring up now and then only to die again.

Here is what I have cooked up quickly to avoid facing its excruciating UI. As you can see - I don't know python (and not willing to learn it.)

http://www.leobodnar.com/pulser/

I have used Prologix for this so "++.." commands are GPIB adapter configuration. If you are old enough - think modems Hayes AT commands.

CSA803 is the instrument I love to hate. There is nothing right with it apart from its raw functionality and market abundance.
I'd like to write more code to get nicer looking charts but I have run out of patience.

Cheers
Leo

21461 - 21480 of 184780