Re: 2465 Fan Collet Thingy
Chuck Harris
Very cool, but not useful as the strength of these composite
toggle quoted messageShow quoted text
filaments is entirely in the plastic binder. -Chuck Harris Vince Vielhaber wrote:
Here ya go Chuck. https://www.micromark.com/search?keywords=filament
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
Ed Breya
Sampling gates and other front-end stuff have traditionally been biased with constant current sources (high resistances from fairly high supply voltages) and associated circuitry. There is nothing wrong with this, and that's why you'll see earlier scope circuits with transistors taking advantage of tube plate and other high voltages for constant current sources - especially during the evolution where Vs and Qs were used together extensively. Nowadays, solid state would not need such tricks, since plenty of current source circuits can be made directly at low voltage. ICs have lots of them built in.
Regarding the7S14 ("perfectly" floating battery bias) versus any other samplers (fed from various bias topologies, but not floating), you'd have to look further downstream in the preamp and memory gate circuits to see the difference.The preamp characteristics for the chosen design in the 7S14 may have required no common-mode currents to deal with, for instance, while another sampler design may be OK with that, in exchange for more complexity, or an adjustment. There are lots of ways, pieces, and tradeoffs, and you can be pretty sure they wouldn't have put in Hg biasing and its grief, if they could have just stuck in a few more parts here and there. The overall design must have needed this particular setup to meet the complexity, size, cost, etc requirements. You can redesign and modify things later, but it's hard to know the exact thought process and situation they had at the time - about fifty years ago, in this case. Ed
|
|
Re: Tek 576 noise and looping
Ed Breya
There is a looping compensation variable capacitor in the collector supply, that normally should help. However, the capacitor shaft may be seized up, and the driving shaft extension or knob may be too loose to actually rotate the cap or show that it's stuck. So, while the knob may turn, the cap may be stuck (for many years) in one spot, and have no apparent effect. Both of my 576s have this problem, and before I finally figured out what was going on years ago, I thought it just plain didn't work - a useless feature. Because it's such a PITA to work on, I let it slide, and adapted myself to the looping and ignore it.
If you've got it apart anyway, you may want to check for proper mechanical operation of the cap. As I recall, the extension shaft is plastic for electrical isolation, making it even harder to keep a tight grip. Don't worry about distortion of the collector sinewave - line voltage looks pretty ugly anyway, due to all the nonlinear loads on it, regardless of what the 576 is doing. Ed
|
|
Re: Type 184 questions
Glenn Little
Are all of the power supply bypass capacitors in good condition, low esr, correct value, low leakage?
toggle quoted messageShow quoted text
Glenn
On 11/10/2019 10:17 PM, bill K7WXW wrote:
I am refurbishing a type 184 time mark generator to go with the 109 and 191. The 0.5us marker has a little extra stub that I have to figure out but other than that the basic functions work and the 10HMz oscillator is well within specification and not drifting. But all is not well. There is 10MHz noise on the DC rails, about 200mV pp on 30VDC and 12VDC, 120mV pp on 125VDC. There is also noise on the trigger output that appears to be a combination of the 10MHz oscillator and the marker output, at least when the marker is set to 0.1us or faster. This noise varies between 60 and 100mV pp. I am pretty sure that's not how it is supposed to look. --
----------------------------------------------------------------------- Glenn Little ARRL Technical Specialist QCWA LM 28417 Amateur Callsign: WB4UIV wb4uiv@arrl.net AMSAT LM 2178 QTH: Goose Creek, SC USA (EM92xx) USSVI LM NRA LM SBE ARRL TAPR "It is not the class of license that the Amateur holds but the class of the Amateur that holds the license"
|
|
Type 184 questions
I am refurbishing a type 184 time mark generator to go with the 109 and 191. The 0.5us marker has a little extra stub that I have to figure out but other than that the basic functions work and the 10HMz oscillator is well within specification and not drifting. But all is not well. There is 10MHz noise on the DC rails, about 200mV pp on 30VDC and 12VDC, 120mV pp on 125VDC. There is also noise on the trigger output that appears to be a combination of the 10MHz oscillator and the marker output, at least when the marker is set to 0.1us or faster. This noise varies between 60 and 100mV pp. I am pretty sure that's not how it is supposed to look.
Other than the usual tracking it back from the BNC, any good places to start looking for noise injection points? There's no obvious physical problems with the unit, though it was clearly not stored carefully and the switches were in need of a good cleaning. Other than trying to track the noise down, any hints where to start? I bought this as a project, and I guess I got one. ;)
|
|
Re: Tek 576 noise and looping
Kevin Oconnor
Is the looping truly visible in every day transistor measurement?
Mine is a virgin. It has some looping under some conditions but others the sweep lines are fine. Knobs & switches are somewhat noisy but not worth putting unit into the ER. There is a line of thought that says sometimes you should leave well enough alone. Are you there or is major surgery called for? Kjo KO3Y
|
|
Re: 2465 Fan Collet Thingy
Shapeways will "print" brass as a lost wax casting. Their process is to print the wax on a 3D printer, then use lost wax casting to produce the part. They don't give specs on the precision of their process however and I have not tried it.
toggle quoted messageShow quoted text
On Sunday, November 10, 2019, 08:23:18 AM CST, Chuck Harris <cfharris@erols.com> wrote:
Hi Tom, Ahh! My fervent wish is to be able to 3D print brass, aluminum, or steel... But alas, that won't be for a while. What I see that says it is 3D printed brass, or aluminum, is in actuality a slurry of brass flakes, or aluminum flakes, and some organic (plastic) meltable binder. The brass or aluminum comes to the surface in a way that makes it appear to be brass or aluminum, but the strength is all plastic. I read an article of a M1911A2 being entirely printed in steel, but I never heard how they did that, only that they would happily sell you one of their printed guns for $30K.... To help fund research, of course. I suppose something like a mig welder controlled by a 3D mechanism would do something... but it would be brutally inefficient. -Chuck Harris Tom Gardner wrote: I've no doubt you are correct.
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Mon, Nov 11, 2019 at 12:29 AM, Miguel Work wrote:
Hi Miguel, - The 7S11 wasn't made specifically for the relatively slow S1 but supports discrete samplers that are 14 times as fast. - The 7S14's vertical input is much simpler and therefore of lower cost than a 7S11 plus sampler setup. See the documents referred to earlier. As I said in msg #161516, I think the 7S14 can "get away with" the simpler (battery) solution because of the relatively low BW. To describe my thoughts differently again: The 7S14's solution is cheaper than the 7S11 + Sx sampler solution, so it's less attractive for the 7S14 to be built as an integrated 7S11 + sampler. Rather the opposite from your thoughts that it would be desirable for the 7S14 to be designed like the 7S11 + sampler general solution. Raymond
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Mon, Nov 11, 2019 at 12:02 AM, Raymond Domp Frank wrote:
Specifically, msg #161516 contains a possible explanation. Also, for any particular functionality, several design solutions usually exist.Again, did you read previous replies to your question? Please let us know what you think of my suggestion and if maybe you disagree. In connection with that, msgs #161506 and #161510 may help. Raymond
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
Miguel Work
Yes, but I don´t find the answer. Why 7S14 can´t work as the S1?
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Sun, Nov 10, 2019 at 11:57 PM, Miguel Work wrote:
Again, did you read previous replies to your question? Have a look at msgs #161506 and #161516 Raymond
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
Miguel Work
On Sun, Nov 10, 2019 at 02:47 PM, Raymond Domp Frank wrote:
:), and now the question, why one works with cells and the other no.
|
|
Re: TEK 2430A Changing settings
frank carcia
Bob,My 7904 could see a 1GHz sine wave so really miss it. I turned on the 2430A today to record the faults and it came up. I looked at a square wave signal that was very rounded off. I cycled the power and next time it came up with faults. I recorded all the faults so I can check the manual. I almost bought a TDS a while back. I think I will check the battery and cycle all the boards and connectors. Maybe just dirty contacts somewhere. Frank WA1GFZ
toggle quoted messageShow quoted text
On Sunday, November 10, 2019, 04:03:42 PM EST, Bob Self via Groups.Io <bobmiscmail=yahoo.com@groups.io> wrote:
Hi Frank, Similar to your aquistion, my 2465B was a company scope that lost horiz drive so I paid a few bucks and took a chance on it. Fortunately I had a late friend who repaired HP & Tek products on the side, so with some parts from his junk pile I had a working scope. He also gifted me the 2430A which I don't use very much but it does come in handy once in a while. The 1990 date code Dallas NVRam modules lasted until now, not too bad for 10 yr guaranteed parts. Bob On Sunday, November 10, 2019, 10:09:48 AM PST, frank carcia <carcia@sbcglobal.net> wrote: Hi Bob,A couple years ago a VP at work signed a 2430A slated for the dumpster to me. It worked fine for a while and then started pulling faults.Reading these threats I learned about the battery and found a service manual on line. Then my beautiful 7904 died that appears to be a power supply issue. I'm debating unloading all of them and picking up a modern light weight scope. I have about 5 different 7000 series scopes and only one works. Not sure how much time I want to put into repairing these things. I have a couple NOS CRTs that I think go in a 7404 and maybe 7904. Frank WA1GFZ On Sunday, November 10, 2019, 10:52:02 AM EST, Bob Self via Groups.Io <bobmiscmail=yahoo.com@groups.io> wrote: I have a 2430A that I just replaced the NVRAM modules on and all is well relative to that, but I'm wondering if anyone has any thoughts on another problem that it has had for several years. The settings keep changing on the scope, the longer it has gone without use, the worse it is. If it has been on for hours, the problem becomes very infrequent and is tolerable. I worked on it months ago to no avail. I checked all the power supplies and re-seated all the connectors and also the eproms. I monitored the voltage on the bezel switches for a while and did not see any unexpected voltages. I am open to any ideas on where to look. I don't have another digital scope to catch occasional events, only a voltmeter and 2465 scope. Thanks, Bob WB6OFS
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Sun, Nov 10, 2019 at 11:26 PM, Raymond Domp Frank wrote:
Found that! The voltages on the sampler diodes in the S4 (e.g.) are derived from it. Effectively about +/- 2.3 V to the sampler diode circuits. Raymond
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Sun, Nov 10, 2019 at 07:15 PM, Miguel Work wrote:
Hi Miguel, Did you read and understand previous replies to your initial question? And where do you see +/- 50V going to the sampler in the 7S11? Raymond
|
|
Re: anyone have 310 or 360 parts machine?
oliver johnson
I have the same scope that i will repair some time in the future, can you send me a diagram of each interconnect cable , i have the ends but not sure is its a one to one connection.
toggle quoted messageShow quoted text
On Sun, Nov 10, 2019 at 11:00 AM, Dennis Tillman W7PF<dennis@ridesoft.com> wrote: Hi Phil, I MAY have one. I am on the road until Friday so you will have to wait until then. Next Saturday send me an email off list to dennis at ridesoft dot com to remind me and I will take a look. Dennis Tillman W7PF
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Phil Cirocco Sent: Saturday, November 09, 2019 9:40 PM To: TekScopes@groups.io Subject: [TekScopes] anyone have 310 or 360 parts machine? I am looking for a graticule lens for my Tektronix 360 indicator module. This is the old scope module for the 160 series with the round CRT.The 310 scope used the same part. Tektronix part #331-0027-00. I will gladly pay for it and shipping. Any help would be appreciated. Thanks Phil Cirocco -- Dennis Tillman W7PF TekScopes Moderator
|
|
Re: 2465 Fan Collet Thingy
Vince Vielhaber
Here ya go Chuck. https://www.micromark.com/search?keywords=filament
toggle quoted messageShow quoted text
The magnetic iron filament will actually rust! They have stainless, iron and copper. Vince.
On 11/10/2019 09:22 AM, Chuck Harris wrote:
Hi Tom,
|
|
Tek 576 noise and looping
peter bunge
The performance check/calibration that starts on page 5-4 was perfect up to step 16. Check Horizontal and Vertical Displayed Noise on page 5-15. The noise is twice what it should be for the horizontal and four times for the vertical. I 'scoped the power supplies and there is no ripple. Any suggestions?
I then switched to step 3. Locating Malfunctioning Circuits on page 4-5. Step D revealed that the rectified sine wave is flattened at the top and distorted on the right. This may be a cause of looping but not the noise problem. Any suggestions? I have a parts unit that I decided to repair so I could compare signals. I had to remove the Collector switch box to repair shaft couplings so got to find out where the collector bridge rectifiers are and how to set up the shafts of the switches. I was surprised to find that the floating 50 volt supply is only 5 volts checked on both units with a 'scope and DVM. There is no ripple. It is plainly marked 50 volts in several places. Any comments? I am going through the calibration on the second 576 but it is not looking good either. Too bad I cannot show the photos here but if anyone wants my notes to get involved please ask off line. PeterB
|
|
Re: TEK 2430A Changing settings
Bob Self <bobmiscmail@...>
Hi Frank,
toggle quoted messageShow quoted text
Similar to your aquistion, my 2465B was a company scope that lost horiz drive so I paid a few bucks and took a chance on it. Fortunately I had a late friend who repaired HP & Tek products on the side, so with some parts from his junk pile I had a working scope. He also gifted me the 2430A which I don't use very much but it does come in handy once in a while. The 1990 date code Dallas NVRam modules lasted until now, not too bad for 10 yr guaranteed parts. Bob
On Sunday, November 10, 2019, 10:09:48 AM PST, frank carcia <carcia@sbcglobal.net> wrote:
Hi Bob,A couple years ago a VP at work signed a 2430A slated for the dumpster to me. It worked fine for a while and then started pulling faults.Reading these threats I learned about the battery and found a service manual on line. Then my beautiful 7904 died that appears to be a power supply issue. I'm debating unloading all of them and picking up a modern light weight scope. I have about 5 different 7000 series scopes and only one works. Not sure how much time I want to put into repairing these things. I have a couple NOS CRTs that I think go in a 7404 and maybe 7904. Frank WA1GFZ On Sunday, November 10, 2019, 10:52:02 AM EST, Bob Self via Groups.Io <bobmiscmail=yahoo.com@groups.io> wrote: I have a 2430A that I just replaced the NVRAM modules on and all is well relative to that, but I'm wondering if anyone has any thoughts on another problem that it has had for several years. The settings keep changing on the scope, the longer it has gone without use, the worse it is. If it has been on for hours, the problem becomes very infrequent and is tolerable. I worked on it months ago to no avail. I checked all the power supplies and re-seated all the connectors and also the eproms. I monitored the voltage on the bezel switches for a while and did not see any unexpected voltages. I am open to any ideas on where to look. I don't have another digital scope to catch occasional events, only a voltmeter and 2465 scope. Thanks, Bob WB6OFS
|
|
Re: 2445 CH 3 and CH 4 vertical setting - volts/ div where ?
Chuck Harris
A free pdf file can be found on the BAMA site. You can find it by using
toggle quoted messageShow quoted text
your favorite search engine. Do you have real probes on those inputs? 10x probes have a resistor that will identify the attenuation factor to the scope, and the scope will change its scale readings to be 10x. There is one selection button for each channel. If pressing one button causes both channels to change their attenuation, you have a CPU board problem. -Chuck Harris vaclav_sal via Groups.Io wrote:
Thanks,I sort-of figured that out by myself late at night.I was using calibrator as a source and that of course is of fixed level.Basically these channels could be used to read 5V logic signals.
|
|