Re: 2440 nvram (was: NVRAM back-up battery change on a 2715..... avoiding disasters)
Szabolcs Szigeti
Hi,
toggle quoted messageShow quoted text
I have replaced the nvram in my 2440, without saving (it actually lost the content) did a recal according to the service manual and works just fine and meets the performance specs. For test I actually cleared the content by doing a cold start, did a self cal, ext cal and it was all OK. The service manual says that self cal + ext cal will bring the scope back to in spec. So I would not worry about saving the content, simply go ahead and replace it. I may be alone with this opinion, but I think this whole saving of the nvram for the 24xx series DSOs is treated with a bit of a mystery. This may stem from the 24xx analogue scopes, where one is screwed if the cal constants are lost. BTW, one of the nvram chips only store saved waveforms, etc, so you have to save the content of only one of them if you really want. Szabolcs Bob Albert via Groups.Io <bob91343=yahoo.com@groups.io> ezt írta (időpont: 2019. nov. 9., Szo, 22:58):
So is there a bottom line here? I have ordered replacement NVRAM modules
|
|
2465 Fan Collet Thingy
Are there any of the fan to motor collet thingies available anywhere?
When I removed the back panel, the fan fell out. I think someone tightened it up too much and broke it on the scope I'm rejuvinating.
|
|
Re: 2440 nvram (was: NVRAM back-up battery change on a 2715..... avoiding disasters)
Bob Albert
So is there a bottom line here? I have ordered replacement NVRAM modules but now am leery of installing them, as I may lose calibration, which seems to be good even though the batteries are down.
toggle quoted messageShow quoted text
Bob
On Saturday, November 9, 2019, 01:26:28 PM PST, Raymond Domp Frank <hewpatek@gmail.com> wrote:
On Sat, Nov 9, 2019 at 08:39 PM, Szabolcs Szigeti wrote: Quite likely. True for the operational settings, not true for the Adjustment (chapter 5) data that is stored in the "store..." menus during Adjustment. Here I disagree. If this were true, what purpose would the menu-supported adjustments (like setting numbers, turning external pots) and storing by pressing keys etc., see msg #161473, during manual calibration ("Adjustment", chapter 5) serve? They are quite different from the saving of operational settings, which happen all the time during use. In these respects, the NVRAM function is quite comparable to that in 'scopes like the 2445/65 series. Calibration constants are stored in NVRAM, as are the operational settings. The latter change all the time when working with the 'scope, whereas the former cannot be restored by just pressing a button: The scope needs to be Adjusted as per SM chapter 5. Many 'scopes also store user-initiated attenuator offset corrections in NVRAM. Not if the constants that were set during Adjustment (chapter 5) were lost because of exchanging the NVRAM , unless the 'scope does have another place to store them. It seems that we don't know where that would be in this model. Raymond
|
|
Re: 2440 nvram (was: NVRAM back-up battery change on a 2715..... avoiding disasters)
On Sat, Nov 9, 2019 at 08:39 PM, Szabolcs Szigeti wrote:
Quite likely. True for the operational settings, not true for the Adjustment (chapter 5) data that is stored in the "store..." menus during Adjustment. Here I disagree. If this were true, what purpose would the menu-supported adjustments (like setting numbers, turning external pots) and storing by pressing keys etc., see msg #161473, during manual calibration ("Adjustment", chapter 5) serve? They are quite different from the saving of operational settings, which happen all the time during use. In these respects, the NVRAM function is quite comparable to that in 'scopes like the 2445/65 series. Calibration constants are stored in NVRAM, as are the operational settings. The latter change all the time when working with the 'scope, whereas the former cannot be restored by just pressing a button: The scope needs to be Adjusted as per SM chapter 5. Many 'scopes also store user-initiated attenuator offset corrections in NVRAM. Not if the constants that were set during Adjustment (chapter 5) were lost because of exchanging the NVRAM , unless the 'scope does have another place to store them. It seems that we don't know where that would be in this model. Raymond
|
|
7854 power supply disassembly
Alberto, IZ2EWV
Hello friends,
The 7854 that I recently bought, after some minutes of operation started emitting a whistle, a burst, some smoke from the rear ad an awful burning smell... "as usual!" I thought, remembering the past experience with my 7904A, in which the problem was in the AC filter. But here the bad component is the RC snubber C5 in the A23 switching board. The problem is: how to disassemble the A23 board from the A12? I tried to unscrew the A12 from the chassis and the rear heatsink, hoping that it could come with the transformer, but some wires of the latter are soldered to the A23 board! Anybody could explain to me the easier way to separate the two boards? I would like to avoid to make disasters, considered that I have to replace only one component! And the wires of the transformer and the other coils are leaving me afraid. I'm not the kind of who cuts the bad part and solders the new one by the component side of the board, or simply bypasses it, so I need to know how at Tektronix they service this board. Thank you in advance, Alberto, IZ2EWV
|
|
Re: Trade P6137 for P6139A probes
Jean-Paul
Dear Manuel, I may be able to do it, but am traveling away from the lab for a while.
If you can wait till early or mid December remind me with private message and I will check my inventory. How many 6137 are available and how many 6139A do you want? Are you located in USA? Best Regards Jon
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Sat, Nov 9, 2019 at 06:23 PM, Miguel Work wrote:
I guess the following is relevant: The samplers as used in the 7S11 need much sharper/shorter strobe pulses and a special physical layout to achieve the much higher bandwidth and the battery circuits are incompatible with that. Remember that the 7S14 is made for a much lower bandwidth (1 GHz) than the 7S11, which goes up to 14 GHz with an S4 sampler. In the circuit diagram of the 7S11 the sampler is shown as a rather simple thing but what you see there is only a block diagram because samplers (the plugins) are each different. Looking at the diagrams of any of the sampler plugins, you'll see that they in fact are rather complicated beasts. By comparison, the 7S14 is much simpler, being "LF" by comparison. So, it probably more a matter of the 7S14 "getting away with" the much simpler battery biasing scheme. Raymond
|
|
Re: 2440 nvram (was: NVRAM back-up battery change on a 2715..... avoiding disasters)
Szabolcs Szigeti
Hi,
toggle quoted messageShow quoted text
No, there is no other writable store in the instrument what i know about, and there doesn't seem to be any on the schematics. I think what the SM says here is when you press the button, it recalculates some values and writes it to the nvram. But this needs to be done after every change in setting the pots. So for the original question of is there a real need to save the 2440 nvram when _it is in cal_ but the nvram needs to be changed, the answer is no, as the scope can be brought back to calibrated state by running the self cal and ext cal. This does not need any physical setting inside the scope (well except for the cal enable jumper). If the scope is out of spec (which an old one most likely will be), then the answer is no again, because you have to do the full adjustment procedure anyway. However, I'm no authority on this, I did do calibration on a number of 2440s and experimented with them too, but I don't know how the firmware actually works. Would be interesting to reverse engineer though, as there are a number of really cryptic features in the calibration menus, about which the Sm doesn't say a word. Szabolcs Raymond Domp Frank <hewpatek@gmail.com> ezt írta (időpont: 2019. nov. 8., P, 17:30):
On Fri, Nov 8, 2019 at 02:45 PM, Szabolcs Szigeti wrote:I had a quick look in the SM. The Adjustments chapter (5) mentions e.g. on
|
|
AWG 2021conversion from rackmount, looking for clues on parts.
Szabolcs Szigeti
Hi,
I scored a nice AWG 2021, in a rackmont setup. I had to do some metal bending, as it was dropped at one time, but it is working now nicely and has all the nice options of 2nd channel and the FFT editr. Since I don't have a rack, and the rack part was damaged anyway, I want to convert it back to a normal desktop instrument as it would fit under my 2440 with the Tekmate so it cando direct wafeorm transfer. I could figure out the obvious things, like different length screws on the back, and missing feet and hadle, but the floppy drive puzzles me a bit. Looks like there is a completely different frame holding the connector for the cable to use the floppy externally and also there is a different cover plate. I know that the floppy is of little use nowadays and it is located at a very inconvenient place on this instrument anyway, much better to use it is now, brought to the front on a cable, but it bugs my OCD :-) Would anyone have the instructions for the rack mount kit (Option 1R) so i could see what needs to be changed or maybe take some photos. or have the parts for sale :-) Thanks a lot in advance. Szabolcs
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
Miguel Work
Thanks for your answer Raymond,
I have studied 7S11 and 7S14 vertical blocs. Question is why 7S11 works without cells, and 7s14 need it. And if is possible to use 7s14 with the same scheme. 7s11 uses +-50 v to bias sampling bridge. Regards Miguel -----Mensaje original----- De: TekScopes@groups.io [mailto:TekScopes@groups.io] En nombre de Raymond Domp Frank Enviado el: sábado, 9 de noviembre de 2019 16:44 Para: TekScopes@groups.io Asunto: Re: [TekScopes] 7S14 7S11 SAMPLER bias QUESTION On Sat, Nov 9, 2019 at 04:00 PM, Peter H wrote: You have to read the doc of the 7S11 and the sampling plugins as well. Note that the 7S14 is a complete horizontal and vertical sampling unit and the 7S11 is just a vertical plugin for the 7000 series *without* its own sampler. To make use of the 7S11, you need a sampler like an S-4 or S-6. Logically, the floating batteries of the 7S14 are in the samplers but the S-x 'es live without a floating battery supply. Have a look here as well: http://w140.com/tekwiki/wiki/7S11 Raymond
|
|
Re: DPO7054 question about Edge button
Tam Hanna
Hello,
maybe this video helps you - soldering SMD leds, down to 0603 is doable. Just be careful not to "crush" the LED with the pincer when it is hot: https://www.youtube.com/watch?v=FAuKQgwyQuk -- With best regards Tam HANNA Enjoy electronics? Join 15k7 other followers by visiting the Crazy Electronics Lab at https://www.instagram.com/tam.hanna/
|
|
Re: 555 for sale in Maryland - $250 with cart and goodies
Jamie,
Same here! I was trying to figure out a way to pick this up or make arrangements. I just cannot make it happen. This is a treasure trove of great stuff! -- Michael Lynch Dardanelle, AR
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
On Sat, Nov 9, 2019 at 04:00 PM, Peter H wrote:
You have to read the doc of the 7S11 and the sampling plugins as well. Note that the 7S14 is a complete horizontal and vertical sampling unit and the 7S11 is just a vertical plugin for the 7000 series *without* its own sampler. To make use of the 7S11, you need a sampler like an S-4 or S-6. Logically, the floating batteries of the 7S14 are in the samplers but the S-x 'es live without a floating battery supply. Have a look here as well: http://w140.com/tekwiki/wiki/7S11 Raymond
|
|
555 for sale in Maryland - $250 with cart and goodies
Jamie Ostrowski
It makes me so sad I but I simply can't afford to travel and buy this beautiful gem right now. I thought someone here might be interested:
https://www.facebook.com/marketplace/item/472746093502260/
|
|
Re: A few pics from a new-to-me exhibit in the National Museum of Nuclear Science and History
Glydeck
Dan,
toggle quoted messageShow quoted text
You are correct. The 535 came from Stan Griffiths at the Vintage Tek museum in trade for a 570 they had. He said he tried to explain that Tektronix was a post war company and that any Tek scope would be an anachronism. The nuclear history museum said they understood but just needed something for the exhibit. BTW, the 570 is on display at the Vintage Tek museum. George
On Nov 7, 2019, at 6:29 AM, Daniel Koller via Groups.Io <kaboomdk=yahoo.com@groups.io> wrote:
|
|
Re: vintageTEK scanned transformer drawings for 120-0998-01 and up
quadzillatech@...
is the 453a-1 transfo the same,ie the one in the hv compartment?,as mine is stuffed!,cheers Paul m3-vuv.
|
|
Re: 7S14 7S11 SAMPLER bias QUESTION
Peter H
Take a look at http://w140.com/tekwiki/wiki/7S14/Repairs and the 7S14 circuit diagram.
|
|
Re: NVRAM back-up battery change on a 2715..... avoiding disasters
Chuck Harris <cfharris@...>
Congratulations!
toggle quoted messageShow quoted text
It will be the first of many. -Chuck Harris michele.trozzi@gmail.com wrote:
The cell voltage was 2.74 V, now it is 3.23 V.
|
|
Re: NVRAM back-up battery change on a 2715..... avoiding disasters
The cell voltage was 2.74 V, now it is 3.23 V.
I just successfully replaced. I followed your advice, using a second cell soldered to pins 4-8 of DS1210. My wife saw me during the work and told me: you put it in extracorporeal circulation. She is a nurse. I had several heart strokes (pin removed from the original cell, positive battery terminal near gnd and the like) but I did it! At the end I drank a large glass of red wine. Next replacement scheduled for November 9, 2029.
|
|
Re: NVRAM back-up battery change on a 2715..... avoiding disasters
Chuck Harris <cfharris@...>
What is the cell's voltage now?
toggle quoted messageShow quoted text
-Chuck Harris michele.trozzi@gmail.com wrote:
Reading the manual I have understood that after data loss in the NVRAM, you can do a simple re-calibration using the internal calibrator.
|
|