In search of Impeller Fan Fastener For Tek 2465 - Part # 243104001

Phil Wong
 

I got hold of a Tek 2465 (the one with the Siemens motor), but the inside
fan is cracked.

I could get epoxy and glue it back (suggestion appreciated), but this
fastener piece is in 4-5 small pieces. Could somebody here suggest a place
where I could order one of these?

2247A Readout issue

contact@...
 

‍Hello,
As shown on the picture attached  one of my Tekscope  2247A givestrouble shooting with no readout display at power on. Only a vertical read outline . Intensity of readout vertical line change with action of front panel pot .

Fortunately all the front panel commands are actives both pushbuttons functions orstep attennuators and voltdivisions OK.
Photo shows trace CH1 and CH2  OK .  Vertical Pos. command OKHorizontal Time base OK No fail in the trigger  led code table (Service Manual )Only the read out that is not displayed full screen .
Supplies voltage corrects within Service Manual  A16 processor board.
Readout position mixer U2415-2414  give correct vertical and horizontal waveform before amplifiers .
But if  Field & Mixer Control U2411  receive data at  R0-R7 inputs theMIX0-1-2-3 outputs  are locked at 0V with no waveforms so the followingReadout position mixers  don't receive waveforms characters to be mixedwith the Hor and Vert ramps .
Replaced U2411 with no effect. . This control latch seems to be activated witha MIX STB CLOCK  pin 11  ( that comes from the READOUT processor U2400) .But there is no wave form in the S. Manual for the kind of pulse to command thisU2411. Seems that it is locked
If anyone has suggestions because this scope is a first hand clean circuitry ?
Thank you for further infos . 


Andre GEORGES 

 

Re: Tektronix 555 oscilloscope indicator unit on eBay

John Williams
 

On the other hand, here is a pretty good listing. Lots of descriptive details. In addition to the scope there is a cart, lots of plugins and tons of components. All well documented at a pretty reasonable price. I don’t have any interest or knowledge of this sale, but would be interested in buying this if I wasn’t winding down in the hobby.

https://www.ebay.com/itm/Tektronix-585A-Working-Oscilloscope-accessories-electronic-components/283634038050?hash=item4209e7f522:g:jc0AAOSwSp5dNad8

Re: 2445 display problems

pete shepherd
 

Good morning Siggi.Mine is the 2445.Pete.

On Tuesday, 8 October 2019, 01:24:08 BST, Siggi <siggi@...> wrote:

Hey Pete,

is that a 2445, a 2445A or a 2445B? There are significant differences
between them, plus some 2445Bs have a surface-mount A5 board that has
well-known problems.

Regards,
Siggi

On Mon, Oct 7, 2019 at 4:16 PM pete shepherd via Groups.Io <pete.shepherd4=
btinternet.com@groups.io> wrote:

Hi folks. Just wondering if anyone can give me a few pointers. I have a
2445 with no display function. On switch on some of the front panel
indicators flash on and the swp trigd indicator is on. I have followed the
preliminary flow chart in the service manual up to checking the low volt
psu supplies which are all correct. Can anyone give me a few points to look
for from here? I have not checked the HV supplies as of yet. Also no front
panel leds flash when I operate any of the switches, and the cooling fan is
not turning.Any help offered.  Pete



Re: P6139a with broken tip

Stephen Hanselman
 

Sorry we didn’t have what you needed and thank you for the correction

Regards,

Stephen Hanselman
Datagate Systems, LLC

On Oct 7, 2019, at 16:29, Jan Kolek <jgkolek@...> wrote:

Steve,

I looked those over as well and didn't see a match. The 206-0378-00 is a 3.5 mm probe tip assembly with black on the probe tip end and clear/natural on the tail (threaded) end.

Regards,

Jan Kolek


Re: 2445 display problems

Chuck Harris
 

2445

Siggi wrote:

Hey Pete,

is that a 2445, a 2445A or a 2445B? There are significant differences
between them, plus some 2445Bs have a surface-mount A5 board that has
well-known problems.

Regards,
Siggi

On Mon, Oct 7, 2019 at 4:16 PM pete shepherd via Groups.Io <pete.shepherd4=
btinternet.com@groups.io> wrote:

Hi folks. Just wondering if anyone can give me a few pointers. I have a
2445 with no display function. On switch on some of the front panel
indicators flash on and the swp trigd indicator is on. I have followed the
preliminary flow chart in the service manual up to checking the low volt
psu supplies which are all correct. Can anyone give me a few points to look
for from here? I have not checked the HV supplies as of yet. Also no front
panel leds flash when I operate any of the switches, and the cooling fan is
not turning.Any help offered. Pete





Re: P6042 Current Probe Repair -- Need new probe cable

snapdiode
 

The most vexing thing about the P6042 is how fragile the probe is mechanically, a very short drop on a hard surface can permanently damage the transformer inside the probe, and huge DC offset is the result. It's not repairable.

I bought a perfect P6042 almost 10 years ago and after an accidental drop, *balfap*, it was useless.

Through sheer luck I found a complete transformer replacement kit for 23$. This has never happened again!

I also found that one NOS cable assembly, that also never happened again.

The prices for P6042s are all over the place from 100$ to over 1000$. It is unpredictable.

*"balfap" sound effect courtesy of Don Martin

Re: 2445 display problems

Siggi
 

Hey Pete,

is that a 2445, a 2445A or a 2445B? There are significant differences
between them, plus some 2445Bs have a surface-mount A5 board that has
well-known problems.

Regards,
Siggi

On Mon, Oct 7, 2019 at 4:16 PM pete shepherd via Groups.Io <pete.shepherd4=
btinternet.com@groups.io> wrote:

Hi folks. Just wondering if anyone can give me a few pointers. I have a
2445 with no display function. On switch on some of the front panel
indicators flash on and the swp trigd indicator is on. I have followed the
preliminary flow chart in the service manual up to checking the low volt
psu supplies which are all correct. Can anyone give me a few points to look
for from here? I have not checked the HV supplies as of yet. Also no front
panel leds flash when I operate any of the switches, and the cooling fan is
not turning.Any help offered. Pete



Re: P6139a with broken tip

Jan Kolek
 

Steve,

The probe tip assembly you have labeled as "Blue--Yellow - 206-0440-00 - 7ea" might actually be 206-0441-00. P/N 206-0440-00 is the part number of the connector/compensation box for the Tektronix P6139A oscilloscope probe.

Regards,

Jan Kolek

Re: P6139a with broken tip

Jan Kolek
 

Steve,

I looked those over as well and didn't see a match. The 206-0378-00 is a 3.5 mm probe tip assembly with black on the probe tip end and clear/natural on the tail (threaded) end.

Regards,

Jan Kolek

Re: P6139a with broken tip

Stephen Hanselman
 

Jan,
As you may have noted there are several w/o part numbers. What are the colors probe tip and the other end?

Regards,

Stephen Hanselman
Datagate Systems, LLC

On Oct 7, 2019, at 15:36, jgkolek@... wrote:

Dear Mr Hanselman,

Thank you for expending the effort to publish the list above with oscilloscope probe replacement tips. Unfortunately, I don't see any of the type that I need (206-0378-00). Hopefully some others can benefit from your inventory list.

Regards,

Jan Kolek


Re: Passive Scope Probe Tip Assembly Replacement

Jan Kolek
 

See the thread "Re: [TekScopes] P6139a with broken tip"

Regards,

JGK

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Jeff Woolsey
Sent: Monday, October 7, 2019 5:55 PM
To: TekScopes@groups.io
Subject: Re: [TekScopes] Passive Scope Probe Tip Assembly Replacement

On 10/7/19 6:16 AM, Lawrance A. Schneider wrote:
I would also be thankful for ONE (1).

Thanks larry

As would I.


On Oct 6, 2019, at 8:27 PM, jgkolek@... wrote:

Mr. Hanselman,

Thank you for your offer of assistance.

Just to be clear, I am in need of one (1) of Tektronix probe tip assembly P/N 206-0378-00, used in Tektronix oscilloscope probes P6137 and P5050.

I anxiously await your findings.

Regards,

Jan G Kolek


--
Jeff Woolsey {{woolsey,jlw}@jlw,first.last@{gmail,jlw}}.com
Nature abhors straight antennas, clean lenses, and empty storage.
"Delete! Delete! OK!" -Dr. Bronner on disk space management
Card-sorting, Joel. -Crow on solitaire

Re: Passive Scope Probe Tip Assembly Replacement

Jeff Woolsey
 

On 10/7/19 6:16 AM, Lawrance A. Schneider wrote:
I would also be thankful for ONE (1).

Thanks larry

As would I.


On Oct 6, 2019, at 8:27 PM, jgkolek@... wrote:

Mr. Hanselman,

Thank you for your offer of assistance.

Just to be clear, I am in need of one (1) of Tektronix probe tip assembly P/N 206-0378-00, used in Tektronix oscilloscope probes P6137 and P5050.

I anxiously await your findings.

Regards,

Jan G Kolek


--
Jeff Woolsey {{woolsey,jlw}@jlw,first.last@{gmail,jlw}}.com
Nature abhors straight antennas, clean lenses, and empty storage.
"Delete! Delete! OK!" -Dr. Bronner on disk space management
Card-sorting, Joel. -Crow on solitaire

Re: P6139a with broken tip

Jan Kolek
 

Dear Mr Hanselman,

Thank you for expending the effort to publish the list above with oscilloscope probe replacement tips. Unfortunately, I don't see any of the type that I need (206-0378-00). Hopefully some others can benefit from your inventory list.

Regards,

Jan Kolek

Re: 2445 display problems

Jan Kolek
 

You might want to be careful about running the O'scope with the cooling fan inoperative. Apparently cooling is critical to keeping some custom ICs from burning out. You can read a horror story about repairing a 2445 with a failed fan on EEVBLOG (from 2014).

Regards,

JGK

Re: 453a-1 no trace

quadzillatech@...
 

well im making sort of progress,when i got the scope q1034 on the z axis board was dead,,i replaced it and it brought up a trace,but something killed the replacement,ive replaced that too but its running well hot and still the r900 hv pot does nothing!,i pulled q1024 and q1034 still ran hot,then pulled q 1043,it then ran cool,so with either q 1024 or q 1043 in place q 1034 gets too hot,still the hv pot does nothing,any ideas where to look?,cheers Paul m3vuv.

2445 display problems

pete shepherd
 

Hi folks. Just wondering if anyone can give me a few pointers. I have a 2445 with no display function. On switch on some of the front panel indicators flash on and the swp trigd indicator is on. I have followed the preliminary flow chart in the service manual up to checking the low volt psu supplies which are all correct. Can anyone give me a few points to look for from here? I have not checked the HV supplies as of yet. Also no front panel leds flash when I operate any of the switches, and the cooling fan is not turning.Any help offered. Pete

Re: P6139a with broken tip

Stephen Hanselman
 

Good Morning,

I have the following probe tips
3.5mm
Tip-tail - Part Number - Qty on hand
Natural-Grey - 206-0389-01 - 6ea
Blue--Yellow - 206-0440-00 - 7ea
Black-red - no p/n - 8ea
Grey-Blue - 206-0392-00 - 13ea
Black-Blue - no p/n - 1ea
Brown(could be funky red)-natural - no p/n - 3ea If red then p/n 206-0416-00
2.5mm
Clear-Clear - 206-0265-13 - 5ea (NOS), 8ea(pull)
Clear-Blue - 206-0265-10 - 20ea (NOS), >20ea (pull)
Yellow-Blue - no p/n - 2ea
Clear-Clear - 206-0265-13 - 8ea (pull)
Brown-Red - no p/n - 10ea
Clear-Yellow - 206-0265-12 - 8ea

196-2963-00 - P6469 Lead set
174-0976-00 - P6101 probe cable 2ea NOS
6106 probe body
131-4997-01 - Probe tip
131-4353-00 - PCB to probe adapter - 20ea
"clip" style grounds - >20ea
"ring" style grounds >20ea
015-0540-00 - Cable for P6501 - 1Ea
175-9411-00 - cable - 1ea
174-0976-00 cable 1ea
174-0971-00 - 2ea
010-0361-00 P6046 10X Attenuator

Sorry for the bandwidth but I though the group of tek collectors might like the inventory. There are other hard to describe easily spares also

Contact off list if interested

Steve

-----Original Message-----
From: TekScopes@groups.io <TekScopes@groups.io> On Behalf Of jgkolek@...
Sent: Saturday, October 5, 2019 11:31 AM
To: TekScopes@groups.io
Subject: Re: [TekScopes] P6139a with broken tip

Hi,

I just posted a new query elsewhere on this board with similar thoughts: can I use a CT3667 (206-0441-00 equivalent) used in P6139A to replace a 206-0378-00 used in a P6137?

Regards,
JGK

Re: Passive Scope Probe Tip Assembly Replacement

Lawrance A. Schneider
 

I would also be thankful for ONE (1).

Thanks larry

On Oct 6, 2019, at 8:27 PM, jgkolek@... wrote:

Mr. Hanselman,

Thank you for your offer of assistance.

Just to be clear, I am in need of one (1) of Tektronix probe tip assembly P/N 206-0378-00, used in Tektronix oscilloscope probes P6137 and P5050.

I anxiously await your findings.

Regards,

Jan G Kolek


Re: P6042 Current Probe Repair -- Need new probe cable

 

On Mon, Oct 7, 2019 at 03:47 AM, @0culus wrote:


Thanks for the part number. I will keep a watch.
Just a thought ......
Comparing the schematics of A6302 and P6042 the cable looks very similar.
I have several scrapped A6302 so if you think it can be used as a substitute
I can send a cable for the cost of shipping. Contact me off list if you are interested.

/Håkan