Re: Tek 576 chatters
Chuck Harris
I don't know what shape your 576 is in, but I have found
toggle quoted messageShow quoted text
many with bad 5V filter capacitors. They suddenly go open circuit. They are the Mallory FP style, and the internal aluminum strip that connects to the solder lugs etches open circuit. The usual symptom is the 576 goes completely goofy. The relays chatter, the screen goes away, bat bleep crazy. You can safely parallel a modern capacitor to the terminals on the FP can and put the capacitors somewhere convenient while you wait for a better solution. Worth a look see. -Chuck Harris peter bunge wrote:
I have not had any replies on the TDS644A power supply problem. I have my repair log for the last repair if anyone is interested. It is not a simple power supply.
|
|
Tek 576 chatters
peter bunge
I have not had any replies on the TDS644A power supply problem. I have my repair log for the last repair if anyone is interested. It is not a simple power supply.
While troubleshooting I needed my Tek 576 curve tracer to check the TVS and HV transistors, especially the replacement BU508A (the last batch from China were only good to 400v). I had used MJE8501 (pencilled in on a TDS544A schematic) but have not been able to find a parts list to show the correct transistor. HELP! My curve tracer now does not show any curves. While troubleshooting it suddenly made a sound like a relay chattering and the display went blurry and jumped around. The +100v supply was at 60v. However turning it off and on it ran for a few minutes then did the same thing. I left it off for dinner and when I returned it ran for over an hour while I continued troubleshooting the step generator which does not seem to be getting to the test transistor. The manual says to check the DAC but does not say where it is. I am about to start again. Any suggestions? PeterB
|
|
Re: Why U800 is such a big problem in Tek Scopes
We don't offer the re-bake process on U800's any longer at Sphere (because Toru in Japan does not do it any longer). The actual root cause of the primary problem with U800 is as follows:
The modified DIP package has a very large heatsink pad, *bonding the chip to the pad is very difficult*, and it was a process that sometimes left the chip attached with essentially a cold solder joint, with high thermal resistance. This made the chip exhibit a range of problems as it warmed up. The re-bake process essentially re-seated the chip to the pad by re-flowing the attachment. *IF* the faults were related to this issue, the success rate was quite high. The chip also had other stress failures not related to this specific issue, they cannot be cured by re-flowing the die attachment. Heatsinking the chip seems to help improve life, and I did it to any 24xx scope that I worked on. Even a 2" square of plate aluminum attached to the package makes a very big difference in package and die temperature. This is not the only 24xx series ship to exhibit die attachment thermal problems, it also occurs in the 155-0236, and they have also been recovered by this process. While I like the 24xx series operationally, the product did not age as well as earlier designs, and repairs and calibration often prove quite difficult after 30 years, especially if U800 is involved. best regards, walter -- Walter Shawlee 2 Sphere Research Corp. 3394 Sunnyside Rd. West Kelowna, BC, V1Z 2V4 CANADA Phone: +1 (250-769-1834 -:- http://www.sphere.bc.ca +We're all in one boat, no matter how it looks to you. (WS2) +All you need is love. (John Lennon) +But, that doesn't mean other things don't come in handy. (WS2) +Nature is trying very hard to make us succeed, but nature does not depend on us. We are not the only experiment. (R. Buckminster Fuller)
|
|
Re: Will this capacitor fail?
George and DW
Definitely a 170uF Cap. I would check the ESR on that old cap very carefully, if it checks "good" for capacity, ESR and ripple, leave it alone. I just finished with a 577 and the filter caps all checked good for ESR and Ripple current. In my experience, the 577 was much more plagued with bad tantalum caps. I found 7 bad ones on mine, which completely disabled the thing. Just my opinion and experience, of course. Best of luck in your project. -- Michael Lynch Dardanelle, AR
|
|
Re: 2235A Vertical Trace Issue
Chuck Harris
Think of a CRT's electron gun as being like a machine gun. Just like
toggle quoted messageShow quoted text
a machine gun fires bullets at a fixed continuous rate, so does an electron gun with electrons. Let's imagine our electron gun fires off electrons at 10 electrons per second. If you sweep our electron gun across the width of the screen in 10 seconds, there will be about 100 electrons striking the screen, evenly spaced along a line. Similarly, the same thing will happen if the electron gun's beam of electrons travels from the top to the bottom of the screen in 10 seconds. Further suppose the beam goes from top to bottom in only 1 second, while it is also traveling from left to right in 10 seconds. Since the trace is only 10 seconds long, it will still contain a total of 100 electrons, but during the time it is traveling from top to bottom, there will only be 1 second's worth of electrons, or about 10 electrons. Suppose still further that the beam travels top to bottom in 1/10th of a second while it is taking 10 seconds to travel from left to right. The total trace will still contain only 100 electrons, but the section that goes from top to bottom will now contain only 1 or maybe 2 electrons! If we keep reducing the time it takes for our beam to go from top to bottom, while it is taking its 10 seconds to travel from left to right, eventually it will traverse from top to bottom of the screen without recording even a single electron striking the screen. The record of our electron beam's travels will look as though the beam instantly traveled from the top to the bottom of the screen. This is exactly what is happening on your scope. The number of electrons the electron gun shoots at the screen is at a rate that is a million times faster, but parts of the trace where the bean travels great distances in very little time will not have as many electrons hitting the screen as will parts where the beam is puttering along slowly... The beam will appear to have no brightness in those sections where it is traveling really fast.. -Chuck Harris Ed Pavlovic wrote:
I'm slowly going through my 2235A scope, replacing the Schaffner line filter and the RIFA capacitors on the Pre Regulator board. The RIFA caps were visibly crazed, so I'm glad I'm replacing them before they failed like the ones in an old computer PSU I have.
|
|
Re: 7B15, 7B10 questions
Roy Thistle
On Sun, Nov 19, 2017 at 10:12 AM, Dennis Tillman W7PF wrote:
Hi Dennis: We have 2 7104s... and occasionally I see one for sale, and the picture of it usually shows it operating with the amber "limited viewing time" indicator illuminated. I'd only pay the prices most are asking for a 7104, if I thought I could get a good spare MCP CRT, with some good life left in it. A cheap 7104 might be worth it, just for the parts though. I take it by your post... that even though the janitor tripped over a 7104 in the hallway, outside the research lab, and is now selling it... the CRT should be good, even though it may have operated in "limited viewing time" mode frequently? Best regards and wishes. Roy
|
|
Re: Will this capacitor fail?
Glydeck
You should be fine. 20% above 200 volts would be 240 volts. BTW, the schematic shows the value C744 as 170 micro farad.
toggle quoted messageShow quoted text
George KD6NEW
On Aug 20, 2019, at 9:16 PM, DW <wilson2115@outlook.com> wrote:
|
|
Re: Need Help Troubleshooting Tektronix PS280
daven9ooq <daven9ooq@...>
I would stick with what you have, maybe it's time to do a little service on it, sometimes you need a dual supply or even 3voltages it's rare though.
I should have done more resurch on this before I bought it. But it's the same era as my TM500 equipment so it should fit right in. Around 1981.Many thanks !Daven9ooq
|
|
Re: Thoughts on TDS744A
greenboxmaven
The condensers you mention were a worldwide scourge. Many millions of them were made when an angry employee stole the electrolyte formula from one company and took it to another. The problem was the formula was not complete. In many cases the electrolyte gasses off slowly and escapes without leaving a residue, but the condenser has no capacitance. Having restored many items using these bad condensers, I strongly recommend that you mark the locations, value, and polarity of all the condensers, remove them all, and then gently scrub the board with hot soapy water. Rinse it thouroughly and dry it in flowing warm air before soldering the new condensers on.
toggle quoted messageShow quoted text
Bruce Gentry, KA2IVY
On 8/21/19 5:22 AM, Jay Walling via Groups.Io wrote:
Hi Siggi,
|
|
Re: 2235A Vertical Trace Issue
Glenn Little
The dim traces on the vertical part of a square wave is normal.
toggle quoted messageShow quoted text
This is due to the speed which they are "drawn" on the CRT face. The faster the rise time of the trace, the dimmer the trace will be. The faster the trace speed, the dimmer the trace. It is like painting with a spray can. If you hold the nozzle down and paint a horizontal line then quickly paint a vertical line, the vertical line will have less paint on it per unit of length than the horizontal line. Glenn
On 8/21/2019 1:43 AM, Ed Pavlovic wrote:
I'm slowly going through my 2235A scope, replacing the Schaffner line filter and the RIFA capacitors on the Pre Regulator board. The RIFA caps were visibly crazed, so I'm glad I'm replacing them before they failed like the ones in an old computer PSU I have. --
----------------------------------------------------------------------- Glenn Little ARRL Technical Specialist QCWA LM 28417 Amateur Callsign: WB4UIV wb4uiv@arrl.net AMSAT LM 2178 QTH: Goose Creek, SC USA (EM92xx) USSVI LM NRA LM SBE ARRL TAPR "It is not the class of license that the Amateur holds but the class of the Amateur that holds the license"
|
|
Re: Thoughts on TDS744A
Hi Siggi,
While the models I mention do have aluminum SMT electrolytics on the front panel, I have not seen any that were leaking. I suspect that the leaking caps on the older models were a quality problem with the cap vendor they were using at the time. Jay
|
|
Re: General Radio 874 Connector, inner conductor "bendies"
Richard Knoppow
The General Radio part of the IEC site has many old GR catalogs. They will help identify the connectors used and those available as parts. GR used a couple of variations of binding posts, mostly combined with banana plugs or jacks. I have not checked but think the 274 type dual banana plugs and jacks had the same spacing for decades making them compatible. On many GR instruments with 874 connectors there is a combination binding post mounted under the coaxial connector. These are ground connections but a 274 dual banana plug with fit between the center part of the 874 and the binding post. This is sometimes handy where one is dealing with audio or low frequency RF although it does not have the effective shielding of the 874 connector.
toggle quoted messageShow quoted text
GR made adapters going to nearly anything from the 874 and both GR and others, like Pomona, made adapters of all sorts for the 274.
On 8/20/2019 11:16 PM, Roy Thistle wrote:
On Sat, Aug 17, 2019 at 08:56 AM, Dale H. Cook wrote:I found a 1001A that I though had a BNC output connector; but, it was an 874 to BNC adapter, inserted into the front panel 874. --
Richard Knoppow dickburk@ix.netcom.com WB6KBL
|
|
Re: 7000 series carrying handles - how strong are they?
Roy Thistle
On Sun, Aug 18, 2019 at 05:54 PM, evan foss wrote:
I've carried the 500s and 7000s, for long distances occasionally 1Km plus, and frequently carried them a couple of hundred meters.. and lots of times carried them 10 to 20 meters, from lab to lab, through the halls... or across the lab. (I know... there is a new invention called wheels; but, the scope carts are always jammed up with junk!) The only times a metal strap broke... and that was before I lifted the scope... the steel strap broke (just by pulling on it)... it was very obviously rotted and rusted through, . A couple of times, on the 500s, I braided a rope (1.5 mm rope/string) handle on, just so I could carry it... when I thought the rusty steel handle present might suddenly fail, under inconvenient circumstances. That said... the steel strap is only as "strong", as the fittings it engages, and the bolts (screws really) that hold the whole assembly to the body of the scope. Checking those are sound, and will continue to be sound, seems prudent. I get the willies too, wondering whether the handle is going to break, on these heavy scopes... but under the weight and strain, I quickly start to trust the steel, and I worry more about banging the front panel into something, and bending those little knobs! I have a 1200 series Tek logic analyzer, that has a bail type handle... but it has one corner of the bottom compressed, upwards... like it was dropped, straight down, from letting go of the handle... and then the thing landing on just one corner... so how does that happen?... the handle is fine. Roy
|
|
Re: 7000 series carrying handles - how strong are they?
Roy Thistle
On Sun, Aug 18, 2019 at 03:38 PM, Bruce Griffiths wrote:
unless it is cold worked (like stamping it?)... and then it develops a magnetic response. The strength of which [compared to stainless numbers that are magnetic]… I guess... depends on the amount of cold working. Roy
|
|
Re: General Radio 874 Connector, inner conductor "bendies"
Roy Thistle
On Sat, Aug 17, 2019 at 08:56 AM, Dale H. Cook wrote:
I found a 1001A that I though had a BNC output connector; but, it was an 874 to BNC adapter, inserted into the front panel 874. I have some GR units, mostly older than 1950, mostly audio, 600 ohm stuff, I guess... mostly in wooden cabinets, so old I guess too... like a 546-C, with binding posts. (sometimes seen referred to 546-0, because of the front panel stenciling?). Anyway, even though I have, and have run across GR stuff... my introduction to the 874 was on Tek stuff (like pulse gens, sample heads, and 7000 plugins)… which I was not interested in, at the time... so just made a note of that odd looking connector, on the front panel. I guess that was just the peculiarity of my situation... which led me to think it was an old Tek connector, when it was really GR! Best regards and wishes all. Roy
|
|
2235A Vertical Trace Issue
Ed Pavlovic
I'm slowly going through my 2235A scope, replacing the Schaffner line filter and the RIFA capacitors on the Pre Regulator board. The RIFA caps were visibly crazed, so I'm glad I'm replacing them before they failed like the ones in an old computer PSU I have.
Issue I'm having is the vertical traces are very dim compared to horizontal traces. Most noticeable on a square wave, but if you introduce a horizontal element, a sinewave or sawtooth pattern, the trace looks normal. I'm waiting on parts from Mouser to replace the failure prone inlet caps with new ones, so I'm unable to check voltages and ripple at the moment. Is it just an issue with the width of a vertical trace affecting the brightness on the screen, or is there maybe a deeper issue going on here? Thanks Ed Pavlovic KC9MMM
|
|
Will this capacitor fail?
DW
I have a Tektronix 577 curve tracer and I am recapping the power supply board that is located at the rear of the instrument. I am looking at replacing C744 which is a 250V 1700uF capacitor. I found a 250V 1800uF capacitor but the capacitor installed is a 275V capacitor though the manual states 250V. I am curious how much voltage typically runs across C744 so I don't destroy a capacitor, thanks
|
|
Re: General Radio 874 Connector, inner conductor "bendies"
Ken Eckert
While it probably isn't "news" IET Labs still sells GR874 connectors along
toggle quoted messageShow quoted text
with other GR equipment that they still manufacture. https://www.ietlabs.com/
On Sat, Aug 17, 2019 at 11:28 AM Roy Morgan <k1lky68@gmail.com> wrote:
The GR 1232A Tuned Amplifier and Null Detector has an 874 connector at the
|
|
Re: 494 and 2756 Spectrum Analyzer questions...
Ken Eckert
I have no knowledge of the particular component, I use a chemistry hot
toggle quoted messageShow quoted text
plate that I have roughly calibrated for pre heating, just dandy for certain surface boards where one leg of the component is tied to a big ground plane. 125 - 175 C is the recommendation usually, your mileage and experience may be different.....
On Tuesday, August 20, 2019, John Miles <john@miles.io> wrote:
Any suggestions about the best way to open it inI usually use a Dremel tool with a cutting wheel. The filter takes a lot
|
|
Re: TDS754A
Neil Gruending
My 754D once had a problem with random reboots. Turned out that a couple of the conductors in the ribbon cables between the boards had burned out. It was fine after replacing the cable, maybe your 754A is constructed the same way.
Cheers, Neil ________________________________ From: TekScopes@groups.io <TekScopes@groups.io> on behalf of Stephen Hanselman via Groups.Io <kc4sw.io=kc4sw.com@groups.io> Sent: August 20, 2019 8:02 AM To: TekScopes@groups.io <TekScopes@groups.io> Subject: Re: [TekScopes] TDS754A I would suggest you open it up remove the proc board and PS cover. Then blow out the dust. We have found any “new” unit to our facility was fairly dirty. The split wall between the PS and HVPS is perforated but with small holes which clog easily. Regards, Stephen Hanselman Datagate Systems, LLC 3107 North Deer Run Road #24 Carson City, Nevada, 89701 (775) 882-5117 office (775) 720-6020 mobile s.hanselman@datagatesystems.com www.datagatesystems.com<http://www.datagatesystems.com> a Service Disabled, Veteran Owned Small Business DISCLAIMER: This e-mail and any attachments are intended only for use by the addressee(s) named herein and may contain legally privileged and/or proprietary information. If you are not the intended recipient, any dissemination, distribution or copying of this e-mail and any attachments is strictly prohibited. If you have received this e-mail in error, please immediately notify me and permanently delete the original and all copies and printouts of this e-mail and any attachments. On Aug 20, 2019, at 07:37, Siggi <siggi@undo.com> wrote:
|
|