Date   

Re: still trying to track down 2216 manuals

Jim Olson <v_12eng@...>
 

Colin, thanks that was my thought also. Checked with the site nothing there. I did send him an email letting him know that Q service had a complete service manual with schematics for the 2212 on CD it is also on ebay.

Jim O

On March 6, 2019 at 3:48 AM "Colin Herbert via Groups.Io" <colingherbert=blueyonder.co.uk@groups.io mailto:colingherbert=blueyonder.co.uk@groups.io > wrote:


Jim,
My post was relating to Bill K7WXW's earlier post (see below). I don't own a 2216; I was just trying to help. I don't know if your suggestion has possibilities or not....
Colin.

-----Original Message-----
From: TekScopes@groups.io mailto:TekScopes@groups.io [mailto:TekScopes@groups.io mailto:TekScopes@groups.io ] On Behalf Of Jim Olson
Sent: 05 March 2019 20:18
To: TekScopes@groups.io mailto:TekScopes@groups.io ; Colin Herbert via Groups.Io
Subject: Re: [TekScopes] still trying to track down 2216 manuals

Colin, have you tried the Vintage TEK Museum in Beaverton Oregon? they have a good collection of original Tek manuals.

Jim O

> > On March 5, 2019 at 10:42 AM "Colin Herbert via Groups.Io" <colingherbert=blueyonder.co.uk@groups.io mailto:colingherbert=blueyonder.co.uk@groups.io mailto:colingherbert=blueyonder.co.uk@groups.io mailto:colingherbert=blueyonder.co.uk@groups.io > wrote:


I don't know whether this is at all valid, but after doing a bit of searching with Bing, I found this:

http://www.jetecnet.com/others.html

It claims to be Tektronix 2216, 4 Channel digital Storage Oscilloscope Service Manual for $20 and is 332 pages. Oddly, they state that the Operator's Manual is "Not Available" which we know is untrue. Since the User Manual is 328 pages, I'm inclined to think it is that.
Is it worth a punt?
Colin.

-----Original Message-----
From: TekScopes@groups.io mailto:TekScopes@groups.io mailto:TekScopes@groups.io mailto:TekScopes@groups.io [mailto:TekScopes@groups.io mailto:TekScopes@groups.io mailto:TekScopes@groups.io mailto:TekScopes@groups.io ] On Behalf Of bill K7WXW
Sent: 04 March 2019 20:03
To: TekScopes@groups.io mailto:TekScopes@groups.io mailto:TekScopes@groups.io mailto:TekScopes@groups.io
Subject: [TekScopes] still trying to track down 2216 manuals

I've search through the archives here and saw the 2014(?) discussion on the 2216 service manual. I have copies of the partial manuals for the 2216 and 2212. Alas, no schematics. Yet.

I am surmising that the manuals' unavailability is related to the fact that both the 2212 and the 2216 were designed in Heerenveen. I would guess that they were only manufactured in Holland and repaired at the Benelux Repair Center. Since Benelux was closed in 2000 as part of the Tek breakup, who knows what happened to the stock of manuals? I am hoping that they all didn't end up in the tip.

Unfortunately, I am now in "dog chasing bone" mode now wrt to finding documentation. Does anyone know if there's a European-based Tek fan or alumni club? I want to see if I can track down anyone that worked in Benelux. So as not to inflict my obsession on others, I can be contacted off list, my call sign at arrl dot net.






>




7834 with bad HV transformer free to good home

 

I've reached the end of my repair efforts on this 7834. I'd hoped to fix it
and sell it, but I'm not equipped to rewind the transformer and I need the
space it has occupied.

So if anyone in the UK wants it then they are welcome to collect it from
Kenilworth.

If no-one wants to take the whole thing, then I'm happy to remove boards or
other parts that people want but you'll need to collect them or pay postage
and packing costs.

I need to have it gone within a week or so - it goes to the local recycling
centre after that :(

I will be keeping one piece of the metal work from the PSU as I need that
for a 7854 ... unless one comes my way!

PS if anyone needs CRTs for 7633 or 2564 I have to clear those out as well.

David


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

bbortnick@...
 

Hi Tom,
Thanks for putting me on the right path. Indeed, there was no continuity between the pads of the x10 fingers. Redid the IPA thing, and now I have a signal at all settings! On to the timing switch - it's quite crusty, not sure if that is repairable. May post a question on that one here later.
Thanks again!


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

Colin Herbert
 

My guess also would be the X10 attenuator. These attenuators can develop a connection fault inside the module. It *can* be fixed with a small soldering-iron and a steady hand, once the top has been popped off, but it is tricky and could ruin the attenuator module. If you haven't already tried this, swap the X10 attenuator modules between the channels and see if the problem goes with the module or stays with the channel.

Cleaning of these cam-switches has been discussed in some detail on a recent thread titled " Help with 7A18 problem". Whatever you do, don't get anything other than IPA or any other cleaning product anywhere near the attenuator assembly circuit-boards - it is likely to ruin them.

Good Luck, Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Tom Gardner
Sent: 06 March 2019 17:38
To: TekScopes@groups.io
Subject: Re: [TekScopes] Help with 465 repair - no signal on 50mv/100mv/200mv ranges

On 06/03/19 17:12, bbortnick@2bcool.ca wrote:
The fingers are gold plated and looked very clean before I IPA'd them. Is there anything else in the circuit that could be causing the problem? Here is the relevant page from the service manual:
https://groups.io/g/TekScopes/album?id=86337
If the 5V/div range is also faulty, my guess would be the *10 attenuator.

Have you tried applying a signal to the faulty channel's input, and using the
/other/ channel to follow the signal forward until it disappears? That could
indicate whether the attenuator or the switch was faulty.

If you then suspect the attenuator, remove it and /cautiously/ replace it with a
wire jumper. If the signal then "gets through" (albeit at the wrong amplitude
and possibly distorted due to the wire and poor connections), then the
attenuator is faulty.

Alternatively, you can use an ohmmeter to measure the contact resistance of the
switch fingers.


Re: TEST MESAGE. "Fw: You have been removed from TekScopes@groups.io"

Brendan
 

This happened to me a week or so ago. Like you I accidentally marked a message as spam in yahoo mail.

On Wed, Mar 6, 2019 at 11:08 AM, Daniel Koller wrote:


Hi Folks,
    This is a test for me to make sure I am re-subscribed, and a warning to
others to be careful.
    I use Yahoo mail to read my tekscopes group messages (perhaps I should
not use Yahoo Mail).  This morning I *accidentally* clicked on a message and
reported it as spam, then immediately restored it to my inbox realizing it was
the wrong one.  It seems groups.io is on a hair trigger because it sent me
the message below.  This must be a new setting because...
...I have accidentally reported messages from Tekscopes as spam before, and
this has not happened, and some Tekscopes messages regularly end up in my spam
folder anyway, and I have to check there from time to time.   So, wonder if
Yahoo is now reporting when you move a message to spam, or groups.io is now
recording when yahoo sends such a report.  
   Either way, heads up to everyone that the system is sensitive!
  Dennis, is there a way to set up the groups.io system so that it takes 3 or
so reports of spam before groups.io unsubscribes a user?  Thanks,
  Dan

   
----- Forwarded Message ----- From: TekScopes@groups.io
<TekScopes+owner@groups.io>To: "kaboomdk@yahoo.com" <kaboomdk@yahoo.com>Sent:
Wednesday, March 6, 2019, 10:23:58 AM ESTSubject: You have been removed from
TekScopes@groups.io
Hello,
Because your Email Service Provider reported to us that one or more messages
sent to you from TekScopes@groups.io has been marked as spam you have been
automatically removed from the group. We suggest that you check your spam box
as soon as possible to see if your Email Service Provider has diverted
legitimate messages into it.

You will receive no more emails from that group. If this was a mistake, you
can resume your subscription within the next 7 days by clicking the following
link:

Resume Subscription

Groups.io groups offer several options for controlling how much and what types
of emails to receive. See this Help Center article for more information.

If this happens repeatedly please check with your email service to learn about
how they decide to send feedback reports to list services (such as Groups.io
groups).

Cheers,
The Groups.io Team


TEST MESAGE. "Fw: You have been removed from TekScopes@groups.io"

Daniel Koller
 

Hi Folks,
    This is a test for me to make sure I am re-subscribed, and a warning to others to be careful.
    I use Yahoo mail to read my tekscopes group messages (perhaps I should not use Yahoo Mail).  This morning I *accidentally* clicked on a message and reported it as spam, then immediately restored it to my inbox realizing it was the wrong one.  It seems groups.io is on a hair trigger because it sent me the message below.  This must be a new setting because...
...I have accidentally reported messages from Tekscopes as spam before, and this has not happened, and some Tekscopes messages regularly end up in my spam folder anyway, and I have to check there from time to time.   So, wonder if Yahoo is now reporting when you move a message to spam, or groups.io is now recording when yahoo sends such a report.  
   Either way, heads up to everyone that the system is sensitive!
  Dennis, is there a way to set up the groups.io system so that it takes 3 or so reports of spam before groups.io unsubscribes a user?  Thanks,
  Dan

   
----- Forwarded Message ----- From: TekScopes@groups.io <TekScopes+owner@groups.io>To: "kaboomdk@yahoo.com" <kaboomdk@yahoo.com>Sent: Wednesday, March 6, 2019, 10:23:58 AM ESTSubject: You have been removed from TekScopes@groups.io
Hello,
Because your Email Service Provider reported to us that one or more messages sent to you from TekScopes@groups.io has been marked as spam you have been automatically removed from the group. We suggest that you check your spam box as soon as possible to see if your Email Service Provider has diverted legitimate messages into it.

You will receive no more emails from that group. If this was a mistake, you can resume your subscription within the next 7 days by clicking the following link:

Resume Subscription

Groups.io groups offer several options for controlling how much and what types of emails to receive. See this Help Center article for more information.

If this happens repeatedly please check with your email service to learn about how they decide to send feedback reports to list services (such as Groups.io groups).

Cheers,
The Groups.io Team


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

Tom Gardner
 

On 06/03/19 17:12, bbortnick@2bcool.ca wrote:
The fingers are gold plated and looked very clean before I IPA'd them. Is there anything else in the circuit that could be causing the problem? Here is the relevant page from the service manual:
https://groups.io/g/TekScopes/album?id=86337
If the 5V/div range is also faulty, my guess would be the *10 attenuator.

Have you tried applying a signal to the faulty channel's input, and using the /other/ channel to follow the signal forward until it disappears? That could indicate whether the attenuator or the switch was faulty.

If you then suspect the attenuator, remove it and /cautiously/ replace it with a wire jumper. If the signal then "gets through" (albeit at the wrong amplitude and possibly distorted due to the wire and poor connections), then the attenuator is faulty.

Alternatively, you can use an ohmmeter to measure the contact resistance of the switch fingers.


Re: Uncooperative 7834

 

When the EHT transformer is connected to the 25kHz HV drive, then pin 13 on the control IC hits 0V pretty much immediately after the inverter starts causing a shutdown after 10mS. Anything below +50mV on this pin indicates a fault condition (inverter current > 5A).

I tried disconnecting the HV Multiplier, but that wasn't the problem. My suspicion is shorted turns in the EHT transformer.

With the EHT transformer disconnected the supply also ticks, but not because of over-current, but because Bal-Sense (pin 2) is just over +200mV when the 10mS timer expires. From what others have said this is pretty much normal behaviour.

Ho-Hum

Who needs some 7834 spares (including a working PSU)?

Dave


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

bbortnick@...
 

The fingers are gold plated and looked very clean before I IPA'd them. Is there anything else in the circuit that could be causing the problem? Here is the relevant page from the service manual:
https://groups.io/g/TekScopes/album?id=86337


Re: still trying to track down 2216 manuals

bill K7WXW
 

Thanks for the help, all. I have a lead on the 2216 service manual. The 2216 operators manual is at the vintage tek museum website. It will be interesting to see how much the 2212 and the 2216 have in common. Cheers, bill


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

Don A
 

Fair enough -if the contacts are gold then there won't be corrosion and IPA
is just fine.
I've wrestled with lots of silver plated tube sockets and connectors and
there Deoxit and NevrDull are really helpful
Don

On Wed, Mar 6, 2019 at 10:58 AM n4buq <n4buq@knology.net> wrote:

I may be wrong, but if this is in regards to the attenuator fingers, I
think they're gold plated so no corrosion should be present and only
cleaning with IPA is necessary/recommended.

Thanks,
Barry - N4BUQ

----- Original Message -----
From: "Don A" <don.asquin@gmail.com>
To: TekScopes@groups.io
Sent: Wednesday, March 6, 2019 9:34:10 AM
Subject: Re: [TekScopes] Help with 465 repair - no signal on
50mv/100mv/200mv ranges

If you need to clean switches I don't find Isopropyl Alcohol (IPA)does a
great job. I use the Deoxit products from CAIG. https://caig.com/
They have products for cleaning switches Deoxit D series and products for
cleaning potentiometers the Deoxit Fader series.

To me, the problem with IPA is that it is a solvent that cleans grease
but
won't remove tarnish. Deoxit D series will. For very tarnished surfaces
(say
tube sockets or connector pins) I use a product called Nevr Dull
http://www.nevrdull.com/ This a very fine polish that works great.

Yes, these product are expensive but the last a very long time for many
projects.

My thoughts
Don






Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

n4buq
 

I may be wrong, but if this is in regards to the attenuator fingers, I think they're gold plated so no corrosion should be present and only cleaning with IPA is necessary/recommended.

Thanks,
Barry - N4BUQ

----- Original Message -----
From: "Don A" <don.asquin@gmail.com>
To: TekScopes@groups.io
Sent: Wednesday, March 6, 2019 9:34:10 AM
Subject: Re: [TekScopes] Help with 465 repair - no signal on 50mv/100mv/200mv ranges

If you need to clean switches I don't find Isopropyl Alcohol (IPA)does a
great job. I use the Deoxit products from CAIG. https://caig.com/
They have products for cleaning switches Deoxit D series and products for
cleaning potentiometers the Deoxit Fader series.

To me, the problem with IPA is that it is a solvent that cleans grease but
won't remove tarnish. Deoxit D series will. For very tarnished surfaces (say
tube sockets or connector pins) I use a product called Nevr Dull
http://www.nevrdull.com/ This a very fine polish that works great.

Yes, these product are expensive but the last a very long time for many
projects.

My thoughts
Don




Re: Wanted 7B15

Fred S.
 

Let us all know how that turns out
that one for $55 looked very clean at first glance

Hi, when I got the 7B15, it looked very clean and had no shorts, but no function as well.
After a "quart" of IPA, I got some life out of it, but no full size traces. I removed the hybrid and cleaned the socket, pcb, and the the hybrid pads and, bingo,
the whole thing worked.
So, I would describe the experience as positive, we can not expect to get 100% working unit for $59. I believe it worked when it was decommissioned, but corroded while being stored.
Now my 7104 is complete: 2x7A29, 7B15 and 7B10, I am very happy!!!

--
Cheers,

Fred S.


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

Fred S.
 

Hi, just clean the contacs of the input attenuator with IPA. There is a video on youtube showing how to clean them on a Tek 485, but that procedure is absolutely the same.

--
Best regards,

Fred S.


Re: Help with 465 repair - no signal on 50mv/100mv/200mv ranges

Don A
 

If you need to clean switches I don't find Isopropyl Alcohol (IPA)does a great job. I use the Deoxit products from CAIG. https://caig.com/
They have products for cleaning switches Deoxit D series and products for cleaning potentiometers the Deoxit Fader series.

To me, the problem with IPA is that it is a solvent that cleans grease but won't remove tarnish. Deoxit D series will. For very tarnished surfaces (say tube sockets or connector pins) I use a product called Nevr Dull http://www.nevrdull.com/ This a very fine polish that works great.

Yes, these product are expensive but the last a very long time for many projects.

My thoughts
Don


Re: still trying to track down 2216 manuals

Siggi
 

Hey Bill,

you might try to ping the Vintage Tek museum folks and see whether they
have the service manual on fiche. There's a tentative index of the
instruments covered in this posting:
https://vintagetek.org/microfiche-scans/, which does have the 2212, but it
doesn't say whether this includes the service manual.

Siggi

On Mon, Mar 4, 2019 at 3:02 PM bill K7WXW <K7WXW@qrpdx.org> wrote:

I've search through the archives here and saw the 2014(?) discussion on
the 2216 service manual. I have copies of the partial manuals for the 2216
and 2212. Alas, no schematics. Yet.

I am surmising that the manuals' unavailability is related to the fact
that both the 2212 and the 2216 were designed in Heerenveen. I would guess
that they were only manufactured in Holland and repaired at the Benelux
Repair Center. Since Benelux was closed in 2000 as part of the Tek breakup,
who knows what happened to the stock of manuals? I am hoping that they all
didn't end up in the tip.

Unfortunately, I am now in "dog chasing bone" mode now wrt to finding
documentation. Does anyone know if there's a European-based Tek fan or
alumni club? I want to see if I can track down anyone that worked in
Benelux. So as not to inflict my obsession on others, I can be contacted
off list, my call sign at arrl dot net.




Re: 91% Isopropyl Rubbing Alcohol to clean potentiometers?

Don A
 

Caig fader lube had a name change to deoxit fader lube a few years back.
Terrific product designed specifically for this task.
There also a product for cleaning switch contacts deoxit D5. Also fantastic.

Yup you will pay a few bucks for a can but they last forever through
countless projects.

My thoughts
Don


On Wed., Mar. 6, 2019, 12:20 a.m. Frank DuVal via Groups.Io, <corvairduval=
netscape.net@groups.io> wrote:

I hope you mean brake cleaner, as brake fluid comes in two types,
neither of which seem good for inside of potentiometers.... one Glycol
based and the other a synthetic base both with a viscosity greater than
water.

Yes, brake cleaner is very aggressive, just like carburetor cleaner. I
would not use it around plastic until I tried it on a sample disposable
part.

Cleaning tuners and volume controls was standard service back 30 years
ago. Many spray chemicals for the job. Using that little red straw got
the cleaner into the pot well.

But, using that spray on fader controls rendered them bad instantly! Not
a great lesson to be learned by trying.

Another vote for DeOxit (Caig) Fader cleaner.

Frank DuVal

On 3/5/2019 9:02 PM, Bob Albert via Groups.Io wrote:
I think Brake fluid is potent as well and I wouldn't use it on
anything without the utmost care.







Re: still trying to track down 2216 manuals

Colin Herbert
 

Jim,
My post was relating to Bill K7WXW's earlier post (see below). I don't own a 2216; I was just trying to help. I don't know if your suggestion has possibilities or not....
Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Jim Olson
Sent: 05 March 2019 20:18
To: TekScopes@groups.io; Colin Herbert via Groups.Io
Subject: Re: [TekScopes] still trying to track down 2216 manuals

Colin, have you tried the Vintage TEK Museum in Beaverton Oregon? they have a good collection of original Tek manuals.

Jim O

On March 5, 2019 at 10:42 AM "Colin Herbert via Groups.Io" <colingherbert=blueyonder.co.uk@groups.io mailto:colingherbert=blueyonder.co.uk@groups.io > wrote:


I don't know whether this is at all valid, but after doing a bit of searching with Bing, I found this:

http://www.jetecnet.com/others.html

It claims to be Tektronix 2216, 4 Channel digital Storage Oscilloscope Service Manual for $20 and is 332 pages. Oddly, they state that the Operator's Manual is "Not Available" which we know is untrue. Since the User Manual is 328 pages, I'm inclined to think it is that.
Is it worth a punt?
Colin.

-----Original Message-----
From: TekScopes@groups.io mailto:TekScopes@groups.io [mailto:TekScopes@groups.io mailto:TekScopes@groups.io ] On Behalf Of bill K7WXW
Sent: 04 March 2019 20:03
To: TekScopes@groups.io mailto:TekScopes@groups.io
Subject: [TekScopes] still trying to track down 2216 manuals

I've search through the archives here and saw the 2014(?) discussion on the 2216 service manual. I have copies of the partial manuals for the 2216 and 2212. Alas, no schematics. Yet.

I am surmising that the manuals' unavailability is related to the fact that both the 2212 and the 2216 were designed in Heerenveen. I would guess that they were only manufactured in Holland and repaired at the Benelux Repair Center. Since Benelux was closed in 2000 as part of the Tek breakup, who knows what happened to the stock of manuals? I am hoping that they all didn't end up in the tip.

Unfortunately, I am now in "dog chasing bone" mode now wrt to finding documentation. Does anyone know if there's a European-based Tek fan or alumni club? I want to see if I can track down anyone that worked in Benelux. So as not to inflict my obsession on others, I can be contacted off list, my call sign at arrl dot net.







Re: Tektronix TDS400/500/600/TLS216 displays and drivers

Tam Hanna
 

Hello Ingo,
while I do not know to 100 percent, I think they should be compatible. Afaik, the color Palette of the DPO / InstaVue actually is quite limited - Danaher accomplished the trick via Palette Stunts similar to the Palm IIIC.

Tam
---
With best regards
Tam HANNA (emailing on a BlackBerry PRIV)

Enjoy electronics? Join 14k other followers by visiting the Crazy Electronics Lab at https://www.instagram.com/tam.hanna/


Help with 465 repair - no signal on 50mv/100mv/200mv ranges

bbortnick@...
 

Hi all, I am trying to get my 465 scope repaired. CH2 has a problem where I get a flat-line (with some noise - signal is the calibration square wave) when set to the 50mv/100mv/200mv ranges. This is obviously related to the #6/#10 volts/div rotors and related circuitry. I have taken the scope apart to clean the fingers (using the suggested paper soaked in IPA,) and I also re-soldered some suspect joints. However, putting it all back together did not fix the problem. I also swapped the 10x capacitor modules between channels, but that did not change anything either.
Suggestions? Thanks for the help!

31341 - 31360 of 186345