Re: 465 No Go
Dm Armstrong
I have the 465 manual the one with the Blue Cover (PDF) and a Operator Owner Manual. I have Captured a Close Up of all the Power Supplies. I have tried to follow the flow charts and will do it again before I undo more. It seems like it told me to check and repair the power supplies in an order with the -8 last.
This is what I have done so far .Replaced the 3000 uf Electrolytic and a 33 uf Tantalum in the -8 volt power section. Checked all Transistors an Resistors in this section including the Large Transistor heat sinked to the Frame. I traded places with the two Identical IC Amplifiers . Chanced two Tantalum Cap under the High Voltage cover All other power supplies are Spot On, Right On to what they are supposed to be. The Negative 8 is the only one not right. I think I said I had .79 volts dc , and maybe .008 Ac. I did check the -8 Test Point pin to ground and had 53.6 Ohms I have a bright green Dot on the screen that I can focus and change the brightness. I can't move it up or down , left or right. The whole unit is pulling .61 amps
|
|
Re: 7104 Power Supply Dummy Load
John Griessen
On 1/8/19 3:39 PM, Dennis Tillman W7PF wrote:
The dummy load doesn't seem toSo, will you risk trying the dummy load on a good 7104?
|
|
Re: 465 No Go
Paul
These guys are doing a great job of helping you. I've been following along, but I did not see anything in the thread to indicate whether you have the Tek 465 manual? It has a really good troubleshooting section for the power supply with the schematic showing what voltages you should see at various points. I have a PDF of the entire 465 manual (it is 30mb), but I just made a smaller PDF of just the low voltage power supply portion (1.15mb). If you don't have it, email me at paulinsunriver2@netscape.net and I'd be happy to send you that power supply section. I'm not sure my browser will send a 30mb file attachment. If I am remembering correctly I may have found the link to it on the Tektronix 465 section of Wikipedia. Might have been somewhere else. Also, did I miss your answer about whether you have a second scope, or if you could borrow one? That is so helpful, as you know from tube radios, and true for power supplies, because you can literally see what is (is hot) happening. Those traces show you very clearly what is hard or impossible to measure with a meter. You can look at what is going into, and coming off that bridge rectifier, with and without the loads (the manual has isolation procedures to unload the different power supply sections. Anyway, no intention to step on the good coaching you are getting, but just some thoughts to try and help.
|
|
Re: 465 No Go
tands13462@...
WHOA....Hold on!
After repairing hundreds of Tek 465's, I've learned that the first thing to do is determine if a voltage isn't there or it's there and being pulled down by something downstream. If you have a service manual, there are trouble shooting flow charts in the back of the manual. Next thing is to determine which of the above conditions exist. With the instrument off, check the resistance from the -8 volt test point to ground. You should see some resistance. If you read no resistance (dead short), that's your problem....a shorted component further downstream. Follow the trouble shooting chart to localize the short. HTH...Tad
|
|
7104 Power Supply Dummy Load
I have a couple of 7104 switcher power
supplies that are not working. They do not tick. Instead the make an erratic faint buzzing sound. It appears the switching multivibrator is being "kicked" to get it started but it stops within microseconds before it can run long enough to make the usual "Tick" that indicates the supply is overloaded. In order to troubleshoot this I made a dummy load which I connected to the 10 pin male connector labeled P82 on the PC board. (It is called P1482 on the schematics). But so far I haven't gotten the dummy load to work. Can anyone see what I'm doing wrong with my dummy load. The loads were wired to the following P82 pins: 1: Not used 2: +5V, 12Ohm/5Watt 3: -15V, 30Ohm/10Watt 4: -50V, 330Ohm/10Watt 5: +50V, 330Ohm,10Watt 6: +15V, 30Ohm/10Watt 7: Ground 8: Ground 9: Lamp Common 10: +5V Lamp The power supply monoitors each of these voltages via a male 8 pin connector labeled P83 on the PC Board (P1483 on the schematic). These are the Voltage Sense (VS) inputs of the individual supplies. P82 was wired to P83 as follows 1: Not used 2: Not used 3: Ground 4: +5VS, from P82 pin 2 5: -50VS, from P82 pin 4 6: +50VS, from P82 pin 5 7: -15VS, from P82 pin 3 8: +15VS, from P82 pin 6 The dummy load doesn't seem to work. That could be because I haven't fixed the supply yet.
|
|
Re: 465 No Go
Definitely cut it apart with side cutters and remove the leads individually. You'll probably be able to solder in the replacement from above w/o removing the board (may need to tip over after fitting).
toggle quoted messageShow quoted text
May be advisable to uprate the bridge rectifier current rating too. Dave
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Dm Armstrong Sent: 08 January 2019 16:59 To: TekScopes@groups.io Subject: Re: [TekScopes] 465 No Go .008 volts AC at the -8 Volt Test Point so I am guessing that this Bridge Rectifier is open. It is soldered so tight to the board I can't even see Daylight under it .The only way I can test it is to Remove the whole board . I had thoughts of cutting it out and destroying it to get to the old pins and hoping there was enough to solder in a new one from the top without removing the board
|
|
Re: 465 No Go
John Griessen
On 1/8/19 10:59 AM, Dm Armstrong wrote:
.008 volts AC at the -8 Volt Test Point so I am guessing that this BridgeI'd interpret that as very little ripple -- and the bridge rectifier could be open. Not a lot of info yet. Do you measure AC volts at bridge input pins? (I always consider looking for open wires/connectors/switches as well as components) What reading is -8 Volt Test Point on DC? "destroying it to get to the old pins" can work well. If you later need to replace some caps and have to pull the pcb out you can tidy it up then.
|
|
Re: 465 No Go
Harvey White
On Tue, 8 Jan 2019 10:14:35 -0500, you wrote:
I am not some kind of Wizzard or Egg Head with Electronics. I hope I knowok, if you have another scope, look at the output of the bridge rectifier, it should be pulsating DC (camel humps) with the 3000 UF disconnected. If the capacitor is in, then you should get at least -10 volts (more like 11 or 12, check the schematic for voltage level). Bridges can fail in various ways, they can open (no output), one diode can fail open (missing humps), fail short (in which case the whole thing pretty much is shorted and fuses blow). This sounds open. Regulators (like the one here), take a higher voltage, typically at least 3 volts more than the desired output, and use the big transistor (pass transistor), as a variable resistor (think voltage divider), and give you the desired output voltage. All those extra parts are in there to control that transistor, and handle situations where too much current tries to be drawn. If you replace the bridge (and yes, most of them are nasty to replace), try finding one with higher current ratings and higher voltage ratings. You really would like a scope to diagnose this scope (which is typical). Harvey
|
|
Re: 465 No Go
Dm Armstrong
.008 volts AC at the -8 Volt Test Point so I am guessing that this Bridge
toggle quoted messageShow quoted text
Rectifier is open. It is soldered so tight to the board I can't even see Daylight under it .The only way I can test it is to Remove the whole board . I had thoughts of cutting it out and destroying it to get to the old pins and hoping there was enough to solder in a new one from the top without removing the board
On Tue, Jan 8, 2019, 11:33 AM Siggi <siggi@undo.com wrote:
Power supply problems can cause any kind of failure, so no trace is quite
|
|
Re: 465 No Go
Siggi
Power supply problems can cause any kind of failure, so no trace is quite
plausibly due to this failure. You might have other trouble, but there's no point in speculating until you've brought the supplies up and into spec. If your DMM has a diode mode, you can check the diodes in the bridge individually with that. Each diode should measure ~0.7V in the forward direction, and should measure open in the reverse direction. You can also try and measure the voltage on the -8V rail with the meter in AC mode. If you have a shorted diode, you'll see a gross AC ripple on the "DC" rail. On Tue, Jan 8, 2019 at 11:19 AM Dm Armstrong <dm.armstrong1@gmail.com> wrote: This Old Vacuum Tube Guy is assuming that like Tubes , Transistors need a
|
|
Re: 465 No Go
Dm Armstrong
This Old Vacuum Tube Guy is assuming that like Tubes , Transistors need a
toggle quoted messageShow quoted text
Negative Bias or feedback to get things going. Is this -8 problem enough to keep this Scope from Tracing or might I have other peoblems.? All I have is a New Multimeter and a Cheep conponet tester to work with. I wish I had an Oscoloscope to fix my Oscoloscope.
On Tue, Jan 8, 2019, 10:38 AM Chuck Harris <cfharris@erols.com wrote:
The bridge rectifier's diodes can be either: good, shorted, or open.
|
|
Re: 465 No Go
Chuck Harris
The bridge rectifier's diodes can be either: good, shorted, or open.
toggle quoted messageShow quoted text
If they are good, no problems, the voltage on its filter capacitor will be 1.4 x the Vrms it is being fed. There will be significant ripple under load, but it will be at 2x the line frequency. If any are shorted, the capacitor will be taking on the AC power line in a fight for its life. It will get very hot very quickly. As you might imagine, the scope will be drawing unusually high current from the power line.... given time, fuses will blow, capacitor guts and smoke will get released. If any are open, the capacitor will either be fed 1/2 wave rectified AC, or nothing. If it is fed 1/2 wave rectified AC, the voltage on the capacitor will be at best 0.8 x Vrms. You will see a very large 60Hz ripple component on the capacitor. If it is getting nothing, you will see, well, nothing. -Chuck Harris Dm Armstrong wrote:
I am not some kind of Wizzard or Egg Head with Electronics. I hope I know
|
|
Re: 465 No Go
george edmonds
Hi
toggle quoted messageShow quoted text
Just check that it looks like four diodes with a DMM set on diode test. 73 George G6HIG
On Tuesday, January 8, 2019 3:14 PM, Dm Armstrong <dm.armstrong1@gmail.com> wrote:
I am not some kind of Wizzard or Egg Head with Electronics. I hope I know enough to keep myself from being Electrocuted. The whole unit is drawing .61 amps . The Bridge Rectifier doesn't cost that much but looks like a big PAIN to replace properly. I am hoping someone can help me think this out before I do much more to it. And if I talk like an old Tube Guy that's because that's my first love . On Tue, Jan 8, 2019, 9:08 AM Dm Armstrong via Groups.Io <dm.armstrong1= gmail.com@groups.io wrote: The voltage is .79 positive. When I had the Big 3000 uf Electrolytic out I
|
|
Re: 465 No Go
tom jobe <tomjobe@...>
If you look in the Tekscopes message archive, there should be lots of discussion about how to change the 4XX bridge rectifiers.
toggle quoted messageShow quoted text
The two methods I've used were... either crush the rectifier piece by piece by nibbling it away a little piece at a time, or wiggle the complete rectifier gently side to side to fatigue the four leads until they break and then pull each lead out of the PCB separately. In either case, put in the highest rated rectifiers you can find and squeeze in there, with the leads bent over at an angle so you can easily reach under and solder the new rectifier to the board. Rectifiers are common failures in the these kinds of 4XX oscilloscopes. tom jobe...
On 1/8/2019 7:14 AM, Dm Armstrong wrote:
I am not some kind of Wizzard or Egg Head with Electronics. I hope I know
|
|
Re: 465 No Go
Dm Armstrong
I am not some kind of Wizzard or Egg Head with Electronics. I hope I know
enough to keep myself from being Electrocuted. The whole unit is drawing .61 amps . The Bridge Rectifier doesn't cost that much but looks like a big PAIN to replace properly. I am hoping someone can help me think this out before I do much more to it. And if I talk like an old Tube Guy that's because that's my first love . On Tue, Jan 8, 2019, 9:08 AM Dm Armstrong via Groups.Io <dm.armstrong1= gmail.com@groups.io wrote: The voltage is .79 positive. When I had the Big 3000 uf Electrolytic out I
|
|
Re: 465 No Go
Dm Armstrong
The voltage is .79 positive. When I had the Big 3000 uf Electrolytic out I
think I measured +5 on the positive post were the Cap was. After I got it out it still was OK. The next Question, New Old Stock or NEW from maybe Mouser? Does anyone have some idea what I need ? Can the Bridge Rectifier be replaced without removing the board? The way I have it figured is 1000 parts, 5 replaced only 995 to go untill it works. On Tue, Jan 8, 2019 at 8:38 AM Craig Sawyers <c.sawyers@tech-enterprise.com> wrote: The bridge rectifier is certainly a possibility. Tek changed the specs on
|
|
Re: 465 No Go
Craig Sawyers <c.sawyers@...>
The bridge rectifier is certainly a possibility. Tek changed the specs on the bridges several times because of a known weakness in earlier choices.
toggle quoted messageShow quoted text
What does the -8V read? Is it just not there at all, or is there excessive ripple? Craig
-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of dm.armstrong1@gmail.com Sent: 08 January 2019 00:58 To: TekScopes@groups.io Subject: [TekScopes] 465 No Go I bought a very used 465 for not much money. When I got it home I found the 1.5 amount fuse on the inside was bad . The Main fuse was OK. I replaced two Tantalim Caps under the high voltage cover , replaced the fuse. No more blowen fuses. What I have is a very bright green dot. I can focus it and adjust the brightness , but nothing else. I can't move it up or down or side to side. All my low voltages are right on except the -8 . I have replaced the big 3000 uf Electrolytic capacitor and a 33uf Tantalim in the -8 volt curcut. The IC is the same as another , swap places with no change. The Transistors are Transistoring , and the Resistors are Resisting . All that's left is The Bridge Rectifier, it is soldered so tight to the board I don't know how to check it. What the chance it is bad? Any help will be greatly appreciated. Thank You Dave
|
|
465 No Go
Dm Armstrong
I bought a very used 465 for not much money. When I got it home I found the 1.5 amount fuse on the inside was bad . The Main fuse was OK. I replaced two Tantalim Caps under the high voltage cover , replaced the fuse. No more blowen fuses.
What I have is a very bright green dot. I can focus it and adjust the brightness , but nothing else. I can't move it up or down or side to side. All my low voltages are right on except the -8 . I have replaced the big 3000 uf Electrolytic capacitor and a 33uf Tantalim in the -8 volt curcut. The IC is the same as another , swap places with no change. The Transistors are Transistoring , and the Resistors are Resisting . All that's left is The Bridge Rectifier, it is soldered so tight to the board I don't know how to check it. What the chance it is bad? Any help will be greatly appreciated. Thank You Dave
|
|
Re: Looking for U100 (ADG361 or MM9576-VIG) chip of TDS3032
Vince Vielhaber
I just had a return with some cables I got that were wrong. Ebay sent me a shipping label to send the cables back. Once received, I got the refund within a couple of hours.
toggle quoted messageShow quoted text
Vince.
On 01/08/2019 12:05 AM, Chuck Harris wrote:
Sorry, no. If it is not as described, ebay takes the funds
|
|
Re: Looking for U100 (ADG361 or MM9576-VIG) chip of TDS3032
Chuck Harris
Sorry, no. If it is not as described, ebay takes the funds
toggle quoted messageShow quoted text
from the seller, refunds, and generally you keep the item. If the seller wants it back, they have to pay to get it back. -Chuck Harris toby@telegraphics.com.au wrote:
On 2019-01-07 4:42 PM, David C. Partridge wrote:Customer does *not* pay for return if the item is "not as described", and a fake is most definitely "not as described".In practice, though, if the return address is in China, customer may
|
|