Re: tektronics 2465b 400mhz nvram battery info
Chuck Harris <cfharris@...>
By "the rest" I mean the rest of the 256 words
toggle quoted messageShow quoted text
in the block. -Chuck Harris Chuck Harris wrote:
If your copy program uses any ram, it will put
|
|
Re: tektronics 2465b 400mhz nvram battery info
Chuck Harris <cfharris@...>
If your copy program uses any ram, it will put
toggle quoted messageShow quoted text
its very own data in the NVRAM. The NVRAM is mostly used as the data space for the CPU, ... data, heap, stack..., and as such is refreshed every time the scope is powered up. There is only a very small portion (170 14bit words) that is calibration constants. I would suggest that the rest is the front panel state, and the timers, and a checksum. -Chuck Harris ADas via Groups.Io wrote:
Hi SImone,
|
|
Re: 7B53A broken switch (mixed mode switch)
Hello Max and Phil,
I've been recently through a very similar case, to the exception that on my 7B53A, the upper part of the switch, as well as the rolling golden contact pin, and its pressure spring, were all still present... on;y that the switch was already open on one side. I posted pictures back than that you can still see here: https://groups.io/g/TekScopes/album?id=35474 To Phil, if you're to disassemble your parts 7B53A to get a switch to Max, I endorse what Max said; it's simple to detach the "A6-Sweep Circuit Board" from the back of the "A1 Interface Circuit Board", to get access to the solder pads of the switch and further endorse that, putting it back together is the hardest part, because you will need to align some 20 pins on different spots of the two boards, in order to be able to pull them together again... But if it's a parts unit, maybe you won't want to have all the trouble... just keep the boards apart, or bend the pins. To Max, my advice to you is, while crimping the tabs of the upper metal cover back in place, around the switch's bottom part to hold them both together, be careful and don't apply too much pressure on the plier. The bottom part looks like plastic, but it's bakelite, and as you know, it's hard and strong, but it breaks easily. I managed to break mine and I had to be creative to glue the two broken pieces into one again, in a way that would stand being crimped and yet, survive the day-to-day stress... I think I was lucky and it ended up well, but you better be cautious while crimping the metal tabs back, around the bakelite bottom piece. all the best, Fabio
|
|
Re: tektronics 2465b 400mhz nvram battery info
Gif Sim
thank you
toggle quoted messageShow quoted text
Das the most complicated part for my skills is the combination ,with hex editor , of the two halves obtained with your program I'm waiting for all the components and then I'll try you did a great job I think it's the simplest and easiest method of all thank you very much for the help you give on the group !! Simone
|
|
Re: 7904 Mainframe damaged
JJ
Right Dave. There is enough room to scope the regulator board on top while
the PSU is hanging out the back of the mainframe. On Tue, Apr 10, 2018 at 4:01 PM, David DiGiacomo <daviddigiacomo@...> wrote: I forgot to mention that you don't actually need to build a load board
|
|
Re: 7904 Mainframe damaged
JJ
Bob, I found that the diodes seem OK when I disconnected one lead and
toggle quoted messageShow quoted text
measured both resistance and diode forward and reverse voltage. I found that the short across those two diodes was due to a shorted capacitor on the rectifier board - I was actually measuring the winding tap resistance through a weird path. I reconnected the PSU back into the mainframe after changing the cap. There was no glory - a very low tick coming from the supply. I'll need to continue debug by following the procedure in the that document. I measured the resistances in z-axis board as the procedure indicates while the PSU was out of the mainframe - they are pretty much in line - a couple of resistances were much higher than the table - none were lower . I'll measure them now while the power supply is in the mainframe, Also, I get different results using a VOM and a DVM. Maybe there's something wrong with my VOM - it's pretty old. Is a VOM required to get the proper measurements as indicated in the table? .Best, John
On Tue, Apr 10, 2018 at 2:41 PM, Robert Hay <bobh@...> wrote:
They say in the article that a low resistance indicates a problem in the
|
|
Re: 7904 Mainframe damaged
David DiGiacomo
I forgot to mention that you don't actually need to build a load board
to work on a 7904 power supply. The cables are long enough that you can let the PS module trail out the back of the scope onto the bench, and get decent access to most of it.
|
|
Re: 7904 Mainframe damaged
David DiGiacomo
On Tue, Apr 10, 2018 at 5:15 AM, JJ <jajustin@...> wrote:
Dave, I went through hundreds of files in the files section of the archiveThere was nothing uploaded, it was a short and sweet message: I used 4 ohms on both 5V, 20 on both 15V, and 500 on both 50V. It is not critical, you could probably get by with only 2 ohms on 5v. Jerry Massengale
|
|
Re: tektronics 2465b 400mhz nvram battery info
ADas
Hi SImone,
What you are describing is right way to use the copy program. It doesn't matter which program "ramhigh.dat" or "ramlow.dat" is run first since they copy independently the upper half and lower half of the old nvram memory locations. . Remember once one of the programs is run, which is on the fresh nvram, take the fresh nvram out, read it with the programmer to obtain data from locations 200hex to 11ff hex inclusive. The data from these locations obtained using "ramhigh.dat" contain copy the upper 4000 decimal bytes of the old nvram and "ramlow.dat" will contain the lower 4000 decimal bytes. Once the copied data (on the fresh nvram) is copied out using the programmer, program the "ramhigh.dat" (assuming ramlow.dat used first), into the fresh nvram and repeat the procedure described to obtain the second half of the old nvram contents. cheers Das
|
|
Re: tektronics 2465b 400mhz nvram battery info
Gif Sim
ciao mirco
toggle quoted messageShow quoted text
what programmer do you use for nvram? I'm not very experienced with the programming I'm waiting for all the components to arrive then I will try with the method tested by adas which I think is the easiest method saluti Simone Minelli
|
|
Re: tektronics 2465b 400mhz nvram battery info
zoldanmirco@...
Ciao Das
Thanks for your help now I get everything I need and I go with the procedure that you have tested, in the next few days I communicate the results. Ciao Mirco
|
|
Re: 7904 Mainframe damaged
bobh@joba.com
They say in the article that a low resistance indicates a problem in the mainframe which is true if you are looking at something shorting to ground. But, in your case I would not discount that high current caused a open circuit while trying to get back to the source. I suppose you already looked for any signs of burning or smoked components, leads, traces, etc.
toggle quoted messageShow quoted text
Bob.
On 4/10/2018 7:44 AM, JJ wrote:
Yes the resistances were very high. According to that doc, the issue should
|
|
Re: Trying to fix up my 2213A
Hello Lukus,
See some answers next to your questions... On Sun, Apr 8, 2018 at 05:57 pm, Lukus wrote: On April 3rd, I picked up a 2213A. While testing it, I noticed that the traceIndeed it's a weird waveform. Did you try to change the sweep speeds (Time/Div), and or the input attenuator (Volts/Div) settings? Does that change the way that "wave" shows on the screen, this may tell you/us something about where this "signal" is sneaking into the circuits. The other day, I opened upUsually the Tek manuals are very comprehensive and you should get instructions on how to open it properly. However, even with the gain turned to the max, the trace does not fill theI wouldn't have jumped directly into changing the scope's calibration, without first checking the basics, and I`m afraid that even doing that, you didn't really ruled out anything. There are two things that you will see on this forum, which are general and mandatory advice before diving into anything more involving (like fiddling with the calibration adjustments): 1. Check the "so called" low voltage power supplies...(LVPS is everything that is not the C.R.T. High voltages) not necessarily a "low" voltage. Among the low voltages there are usually voltages in the 50~60 volts ballpark, and in the 100~140V ballpark. 1a.for correct voltages (you will need a VoltMeter, at least, and preferably one that you can trust the measurement). 1b. For ripple, and since this is the first oscilloscope you have, I suppose you can't check the ripple and/or power supplies waveforms, so it must suffice you measure the AC voltage of the DC power supplies. Those ripple voltages are usually in the millivolts arena, at most, in the 100ths of volt, so you will need a relatively sensitive AC voltmeter (again that you can trust). 2. Looking for signs of false-contacts or conditions that are usually associated to false-contacts, like signs of corrosion, excess of dust or dirty innards, signs of humidity... and preemptively, trying to clean the easily accessible switches with I.P.A. or contact cleaner. If the latter, make sure you use a contact cleaner that's safe to plastics... The manual is your friend here. Don't apply any cleaner without first checking in the manual if that switch in particular has special recommendations for cleaning. In doubt... don't. Exercising the switches and potentiometers usually help to clean them, and also reveal if they are showing signs of dirtiness / false contacts. It's common that some lubricants dry out and leave an insulating film that sometimes render the switch or potentiometer totally dead or opened, and often this is not an irreversible condition. One example to this topic is the "Beam Finder" switch, false contact on it is known to cause this kind of problem (of compressed gain, on either X, or Y axis). The wave disappears if I ground the shroud surrounding the CRT. I will try toThe strange waveform may be nothing, it may just be noise pickup that's happening because the cover is partially opened (so that it's natural connection to ground is broken, as it remains grounded while closed) but at the same time, it's in close range to sensitive electronics that will pickup noise from the nearby big piece of metal... or the cover, while ungrounded, may be picking up strong EMI interference from the Power supply, and transmitting it (by capacitive coupling) to the sensitive parts that it may be nearby. Can someone help me work through calibrating this scope and getting it to workYou came to the right place. I don't know the 2213 in particular, but surely other folks will jump in and help. If you want more information and/or details, please let me know; I would beLet us know about the Low Voltage power supply voltages... and if the manual helped you with opening the scope (completely). By the way, how do I properly reply to a reply on a topic? I tried posting aDo you use an e-mail client? or you read / reply directly from Tekscopes Groups.io web interface? Rgrds, Fabio
|
|
Re: 2235 stuck cover
Ed Breya
Sometimes the cans can be a tight fit on the chassis. This can be aggravated by any dents, dings, grit, or corrosion that can make the Al surfaces gall and bind up. Once you are sure all the fasteners are removed, you can apply gentle brute force by placing the scope upright with its back end resting on a piece of carpeting or towels on the floor, and pushing the can downward. Hold the handle knobs on the sides to apply the force. Some rocking and twisting should help to slightly deform the whole thing, breaking any surface bonds.
Ed
|
|
Re: tektronics 2465b 400mhz nvram battery info
Mark Litwack
The calibration data can be dumped via GPIB, at least on the 2445/2465 and 2445A/2465A. It's not been confirmed it can be done on the 2465B, but I suspect it would work on that too. I'm not aware of an automatic way to dump the whole NVRAM, but the hidden GPIB commands to do it piecemeal have been posted here:
http://www.eevblog.com/forum/testgear/tektronix-2465b-oscilloscope-teardown/msg1449647/#msg1449647 If you (or anyone) try it on the 2465B, let us know if it works or not. As per that post, the data can also be written. So, if you have a GPIB interface it makes calibration backup and restore much simpler.
|
|
Re: 7B53A broken switch (mixed mode switch)
Hi Max,
toggle quoted messageShow quoted text
I was looking at the cross-connecting pins from the side and not from the bottom, and misinterpreted the connection. I have been working on replacing or rebuilding the fan on my 453, and have been displeased with how the fan is hardwired to the inside of the scope, rather than having a plug! It seems crazy to me, made all the more crazy when I pulled one of the wires out (accidentally). I guess I am (now) expecting the worst from Tek, where, before the fan, was always looking for the excellent engineering and finding it everywhere in every device... Of course, you are right. I will see what I can do with this tear down and will take your advice about looking at my working 7B53A for that switch "floating" and get it tightened. I'm wondering if I can't fashion a tool for that job? Thanks, and I will keep you advised. Phil
On 4/9/2018 4:14 AM, unclebanjoman wrote:
they are connected via several connectors
|
|
Re: 2235 stuck cover
tom jobe <tomjobe@...>
You don’t mention the screw on the side, but you do mention two screws being removed after the back cover was off.
toggle quoted messageShow quoted text
I think there are only those two ,the side and bottom screws but I’m not at home to have a look at any 22xx’s
On Apr 10, 2018, at 7:27 AM, Bert Haskins <bhaskins@...> wrote:
|
|
Re: 7904 Mainframe damaged
JJ
All the resistances with the PSU disconnected from the mainframe are way
higher than those indicated in the "Servicing the 7904 High-efficiency PS" doc. When I fix the issues that I have found and the resistance measurements line up with that doc, then I should be good to reconnect the PSU to the mainframe, On Tue, Apr 10, 2018 at 10:22 AM, Velik Kazakov via Groups.Io < velik_kazakov@...> wrote: On Tue, Apr 10, 2018 at 04:16 am, JJ wrote:archive
|
|
Re: 7904 Mainframe damaged
JJ
Yes the resistances were very high. According to that doc, the issue should
toggle quoted messageShow quoted text
be on the PSU. I'm going to put the PSU back into the mainframe, connect the cables, and make those voltage measurements on the LV regulator board. I didn't check for voltages before taking the PSU out because I didn't know at that time that the PSU needs to be under minimum load. I found a shorted cap C1360 on the rectifier board (+54v filter cap on output of Pi filter). I'm going to lift one lead to ensure it's the problem. I'm also going to validate that the two power diodes that I found are shorted - by disconnecting the wires going to those diodes and measuring. If the diodes are bad, I'll replace the bad 10A diodes with two 5A diodes in parallel temporarily - hoping that's OK, I'll then hook up the mainframe's cables to the PS and check the voltages. I think the 3 cables are long enough for the PSU to hang out the back. If not, I plan on making extender cables. Best, John
On Tue, Apr 10, 2018 at 10:16 AM, Robert Hay <bobh@...> wrote:
Have you checked resistances as suggested in the Service Scope article?
|
|
Re: tektronics 2465b 400mhz nvram battery info
Gif Sim
hi adas
toggle quoted messageShow quoted text
thank you very much for your information I think this method of copying is the fastest and most brilliant of all the methods described I bought the material from digikey and I'm waiting for delivery as a programmer I bought the GQ-4x4 that they tell me is compatible with dallas and fram a little info I ask you: when I insert the ram with the "ramlow.dat" program, I turn on the oscilloscope I short circuit pin 1-2 of the cpu, move the jumper P503 I turn off the oscilloscope remove the short-circuit pin 1-2 and move the jumper p503 then I repeat the whole sequence with "ramhight.dat" is the procedure correct? thanks adas also by all users who follow the post regards Simone
|
|