Re: Type CA Channel B Gain Adj Problem


Glydeck
 


I just looked at my CA manual and 535A manual. The DC value of 66 V is called out in the CA and 67 V in the 535A. -- George

On Apr 30, 2013, at 4:21 PM, larrys@... wrote:

 

The TU-7 manual (on bama) mentions it as .4V P-P is 4 div.
-ls-

David Wise <david_wise@...> wrote:
> Sorry, can't remember.
>
> Dave Wise
>
> -----Original Message-----
> From: TekScopes@... [mailto:TekScopes@...] On Behalf Of David
> Sent: Tuesday, April 30, 2013 4:00 PM
> To: TekScopes@...
> Subject: Re: [TekScopes] Re: Type CA Channel B Gain Adj Problem
>
> Did Tektronix write that down somewhere? That sounds like it comes
> from a service manual I should print up.
>
> On Tue, 30 Apr 2013 15:49:58 -0700, David Wise
> <david_wise@...> wrote:
>
> >67.5VDC common-mode, 0.1V/div differential.
> >
> >Dave Wise
> >
> >-----Original Message-----
> >From: TekScopes@... [mailto:TekScopes@...] On Behalf Of David
> >Sent: Tuesday, April 30, 2013 3:47 PM
> >To: TekScopes@...
> >Subject: Re: [TekScopes] Re: Type CA Channel B Gain Adj Problem
> >
> >I could not find any specifications for the 500 series vertical
> >plug-in interface voltage levels. If you know those, then you can
>
>
> ------------------------------------
>
> Yahoo! Groups Links
>
>
>