Wouldn't there still be a problem if the internal windings have already shorted? Bob
toggle quoted messageShow quoted text
--- In TekScopes@yahoogroups.com, "Bene's Mails" <bene.dienst@...> wrote: The solution comes from my dad, he’s an electro-engine builder (excuse me if the translation is bad but I don’t really know whats the name of “Elektro-Maschinenbauer� in english). He said too, that it ist a problem of material degeneration, exactly that the epoxy becomes brittle with the time (a problem that a lot of transformers had in earlier days, he said). So it get microfissures, because of AC caused vibrations, where the moisture becomes encased. The water decreases the insulation ratio of the impregnition and so the leakage current increase. This current heats up the transformer and cause flux degeneration in the ferrite core, so the transformers efectivity decrease and the secondary Voltage fell off. Heating the transformer for a long time under dry enviromental contitions, dehumidifys the epoxy, so he said. The new epoxy seals the old bad epoxy against the air humidity, so it can’t absorb new moisture.
This is how my father described the problem to me. For my part I can say that it has worked well.
greetings an a nice weekend ben
From: Chuck Harris Sent: Friday, April 12, 2013 12:07 AM To: TekScopes@yahoogroups.com Subject: Re: [TekScopes] Any advice on 547 power supply repair?
I cannot imagine that any solution that involves putting epoxy on a bad 547 HV transformer will work... heating it or whatever. I spent several years researching the issue, and found that the failure is a progressive and irreversible degradation of the epoxy material.
In my experience, the epoxy has changed from its original composition to something that is lossy in general, and will absorb water to become even more lossy.
The only solution I have found to the transformer problem which lasts long term is to remove the winding, and replace it with new.
I have been selling replacements for the 547 EHT for some years now.
-Chuck Harris
Bene's Mails wrote:
Cool! I hope the rest of the instrument works fine. Pay attention on the HV-Supply, I don’t know if you already hear it, but the transformers impregnation is infamos for absorbing moisture, so that the insulation breaks down and with it, the HV. Sympthoms for this are a fuzzy trace and inoperative Focus, Astigmatism and Intensity controls after a period of ca. 20 minutes. A solution which works in most cases is to dismount the Transformer out of the HV compartment and put it into the oven at 70-80°C (160-180°F) not higher and leave it for around five hours in it. Then you mix some epoxy-compound glue and put it on the windings and the insulation tape. Be careful with the glue, it becomes liquid by heating it up, the best way is to put it on the upper windings, wait until it’s dry (time for diffrent temperatures is printed on the pacage), then turn it and glue the other side. Next, lay it on the side, and glue the first half of the tape, wait again and then glue the other side. Wehn you finished it, the problem shoud never appear again.
ben
From: Dave Sent: Wednesday, April 10, 2013 6:30 PM To: mailto:TekScopes%40yahoogroups.com Cc: Bene's Mails ; mailto:davidwhess%40gmail.com ; mailto:stefan_trethan%40gmx.at Subject: Re: [TekScopes] Any advice on 547 power supply repair?
I replaced the 6BY8 in the 100V power supply and that fixed the problem. Thanks to all who responded ...
Dave
On 3/29/2013 11:18 AM, Bene's Mails wrote:
good luck! Write here again, when you’ve fixed it!
ben
From: Dave Sent: Friday, March 29, 2013 2:58 PM To: mailto:TekScopes%40yahoogroups.com Subject: Re: [TekScopes] Any advice on 547 power supply repair?
OK, I understand now. Thanks very much, Ben. Now, it's time to work on that 'scope...
Dave
On 3/29/2013 9:22 AM, Bene's Mails wrote:
Yes, but all other circuits where the +100V are used. I know it’s a lot of work, but when the supply is allright, then you must find the problem in the rest of the instrument.
ben
|