Re: 2465B DC Balance/Parametric measurement problem
Bob Koller <testtech@...>
Jim, You need to run CAL 2 and 3 first. The different Trig hybrid will need different cal constants, plus the bad one probably caused some too-far-afield constants to be generated. If the CAL's don't pass or correct the problem, then, maybe the Control board has a problem. Is the serial number of your instrument above B05xxx?
From: jmacwilliams To: TekScopes@... Sent: Monday, June 4, 2012 8:46 AM Subject: [TekScopes] Re: 2465B DC Balance/Parametric measurement problem With nothing connected to Ch1, Volts/Div at 5V, and input set to GND (it actually makes no difference what it's set to), I can set the reference line to 1 division from the top using the Position pot. Assume the bottom line is numbered 0, then there are 9 horizontal lines, numbered 0 to 8, 8 being the top line. Reference line is on line 7. Here's how it changes as I rotate Volts/Div thru its settings:
5v -- line 7 2v -- line 5.5 1v -- line 3.4 500mv -- line 7 200mv -- line 5.5 100mv -- line 3.4 50mv -- line 7 20mv -- line 5.5 10mv -- line 3.4 5mv -- just below the bottom of the display (line minus something) So there is a pattern: 5v = 500mv = 50mv and so on. This repetition repeats for ch2 when the same deltas. Hope that makes sense. I'm about to try changing the A5 board with a known good one. Jim --- In TekScopes@..., "Patrick Wong" wrote: > > Hi Jim, > > Back to your OP, how much does the trace move vertically, in major divisions? > > Patrick Wong AK6C > > --- In TekScopes@..., "jmacwilliams" wrote: > > > > Power on self test works with no problems reported. Switching Volts/Div for both Chn 1 and 2 causes the signal to move vertically on the display. I tried running the DC balance procedure and the scope executes the procedure, but the problem doesn't change. > > > > Parametric measurements do not work either. The scope tries to make a measurement, but ends up not finding the signal, and the Volts/Div ends up on the highest scale. > > > > The Trigger voltage reading is also incorrect. Connecting Chn 1 to the Cal signal, the scope will trigger on the signal when the Trigger voltage reading is somewhere -0.5 to -0.1 volts. The Cal signal is truly between 0.0 and 0.4 volts. > > > > I suspect all of these problems have a common cause. I've checked all the test voltages on J119 and all read correctly. > > > > Anyone have any ideas? > > >
|
|