Re: 2235 with 360 kHz 100mV P-P ringing on +100V PSU rail
tom jobe <tomjobe@...>
Hi Vauga,
toggle quoted messageShow quoted text
I was studying the service manual for the 2235 military version, which is what it looks like you have. The military 2235 seems to be a slightly higher quality 22xx scope in my experience, even though all of what I think of as the "Beaverton design" 22xx's are quite nice and quite similar. The +100 volt supply is used on several circuits that go to the CRT such as the Horizontal Deflection Plates, the Grid Bias, the Z axis ,and the Geometry circuits. The +100 volt supply is used very directly to these circuits with a very small capacitor right at the point of each use. Look at the far right side of schematic <6> and notice how many places the +100 volt supply is used just before those horizontal leads that go up to the CRT. It is no surprise that this seemingly small ripple might be causing so much grief. Since this problem seems so common in the 22xx scopes, I'm thinking that you are not looking for a defective component, I think we are looking for a modification of some kind. I wonder what kind of a modification it would take to eliminate the 100 mV ripple from a +100 volt supply? Maybe we will luck out and the brighter and more experienced folks on Tekscopes will offer some suggestions on this matter. "Thank You!" in advance to anyone who might have any thoughts on this problem or on my speculations. tom jobe...
----- Original Message -----
From: "vaugha69" <vaugha_brewchuk@yahoo.ca> To: <TekScopes@yahoogroups.com> Sent: Monday, May 23, 2011 10:37 AM Subject: [TekScopes] Re: 2235 with 360 kHz 100mV P-P ringing on +100V PSU rail Hello Tom, Thank you for your quick response! Yes, I replaced C940, but the closest I could source was a 1200uF Nichicon PW, which in the worst case scenario could fall outside of the maximum specified tolerance of 1000uF + 30%. Mind you, I had to do pretty much the same for most of the other capacitors, except all the other replacements fall within the original tolerance bands: C954 47uF vs. 33uF; C956 390uF vs. 270uF; C960 etc. 1200uF vs. 840uF; C942 & C943 15uF vs. 10uF. I also have at the ready a United Chemi-Con 100uF, 400V (could not source 450V) KXG series capacitor, so I can easily replace C906. I got a bit of a cold feet syndrome since the KXG is 1/4 the size of the Sprague cap, and I also read elsewhere that the Sprague telephone grade caps last "forever". I found that last bit hard to believe mind you, since it is an electrolytic after all. I could not find the ripple current rating of the Sprague capacitor, but I'm guessing that the KXG will work just fine. Maybe I should jus go ahead and replace C906? I was wondering whether replacing the original 0.01uF 3000V caps with polypropylene could have disrupted the power supply somehow? Or perhaps it could be the overall collection of slightly higher than nominal capacitance on all the output filter caps? I realize that I am shotgunning the whole issue, but I really did not mind replacing all the electrolytics. I have been doing this for most of my other equipment at around 15 to 20 years of use as preventative maintenance. Unfortunately I do not have access to an isolation transformer. Thank you again, Vaugha. --- In TekScopes@yahoogroups.com, "Tom Miller" <tmiller@...> wrote: transformer. shotgunning for the main problem. rail [1 Attachment] am also not an electronics technician, so please be gentle with me :-). However, when using the scope I noticed occasional small ripples in the traces. I opened the scope and tested the power supplies and determined that the ripple on all the rails was 2 to 3 times larger than the limit quoted in the service manual. Consequently I replaced all electrolytic capacitors in the power supply, except the C906 input filter capacitor, with Nichicon PW series capacitors. I used the exact same capacity components where the originals were +/-20% and the next larger capacity where the originals were -10% to +30, +50 or +100%. I also replaced the four 0.01uF 3000V capacitors with CDE 940C30S1K-F polypropylene capacitors. after the overhaul even though the traces seem to be a bit sharper. The ripples on all the power supply rails have significantly reduced, with the exception of the +100V rail. Here the actual low frequency ripple seems to be in check at approximately 30mV P-P, but every second ripple peak exhibits damped high frequency ringing at approximately 350 - 360kHz with 100mV P-P amplitude. I attached an image showing the ripple and the ringing on the +100V rail.
|
|